Oscilloscope
  • Th3_uN1Qu3
    Believe in
    • Jul 2010
    • 6031
    • Romania
    #81
    Re: Oscilloscope
    Because in the primary the "ground" reference is the negative side of the filter cap(s), which is rectified mains. If you connect the ground of the scope to the negative of the filter caps, you make a direct short across the mains (hot to earth) and you blow the power supply's fuse.
    A working TV? How boring!
    • godin
      • Dec 2011
      • 34
      #82
      Re: Oscilloscope
      Thanks, I guess the safe thing to do is always check to see if there is a voltage between what you are about to clip to and the ground clip!
      • PCBONEZ
        Grumpy Old Fart
        • Aug 2005
        • 10661
        Re: Oscilloscope
        Without isolation you CANNOT probe in the primary of a SMPS with an earthed chassis scope - only with one powered by batteries and without a metallic frame.
        That a regurgitation of the same old 'ancient knowledge' -what- that's been repeated since at least the '50's.
        [You back when much equipment and even most buildings didn't have grounds...]
        As I already expressed I can't think of a single -why- for that -what- when it comes to modern grounded equipment.
        .. And it seems neither can anyone else. - Including Tektronix..
        Please explain your -why-.
        Mann-Made Global Warming.
        - Dr Seuss
        • godin
          • Dec 2011
          • 34
          #84
          Re: Oscilloscope
          I must admit I am accepting it at the moment but not truly understanding. I will have to think more. Having looked at an smps schematic, the bridge rectifier is before the transformer which is not something I have encountered before.
          • PCBONEZ
            Grumpy Old Fart
            • Aug 2005
            • 10661
            Re: Oscilloscope
            Basic SMPS block diagram flow path.
            Input filter
            Rectifier AC-DC [High Volts]
            Switching Section [makes HV AC at Higher Frequency]
            Step-Down Transformer
            Rectifiers AC-DC [OP Volts]
            Output Filters
            In modern gear there is a ground in just about every stage.
            - Dr Seuss
            • PCBONEZ
              Grumpy Old Fart
              • Aug 2005
              • 10661
              Re: Oscilloscope
              And adding an additional transformer on the AC side of the rectifier changes that - how?
              - Dr Seuss
              • Th3_uN1Qu3
                Believe in
                • Jul 2010
                • 6031
                • Romania
                #87
                Re: Oscilloscope
                Adding a transformer means there is no more direct path to the mains hot and neutral, the outputs of the transformer are floating and you can connect whatever side you want to earth ground. Wire the negative side of the primary filter caps to earth ground without an isolation transformer and tell me what happens.
                • godin
                  • Dec 2011
                  • 34
                  #88
                  Re: Oscilloscope
                  Do you mean an isolation transformer that floats the ground too and just leaves a floating AC with no ground connection to the neutral?
                  • PCBONEZ
                    Grumpy Old Fart
                    • Aug 2005
                    • 10661
                    Re: Oscilloscope
                    Thanks for making me think but I'm still not buying the story.
                    - If I have a GCE somewhere in my thinking then point it out.
                    Here goes...
                    Wire either side of the Iso-Xfmr secondary to ground and tell me what happens.
                    You seem to be trying to tell me it magically isn't going to pass AC current to ground.
                    I'm thinkin' it will.
                    Your whole theory is based on two misconceptions.
                    #1 - One side of a rectifier isn't -always- 'at ground'.
                    #2 - A negative voltage -IS- a 'HOT'. - If it will shock you, it's a HOT.
                    Looks to me like:
                    "Wire the negative side of the primary filter caps to earth ground"
                    is going be the exactly same in a modern SMPS regardless of an Iso-Xfmr.
                    There is a chassis/earth ground between CY1 and CY2 before the rectifier so adding a transformer is not going to change anything at all.
                    All the Iso-Xfmr seems to be doing is forcing the neutral before the rectifier to be below ground potential. If anything that's MORE dangerous.
                    Either way [Iso-Xfmr or not] the DC negative side of that rectifier is below earth ground potential. [A negative voltage to ground is STILL a 'HOT'.]
                    - The CY caps arrangement sets the ground reference and ensures that.
                    Either way [Iso-Xfmr or not] connecting ground to either side of the rectifier it will cause a short because BOTH sides of the rectifier are HOT relative to earth ground.
                    [BOTH sides of the rectifier are HOT. - WHY would anyone connect a ground straight to a HOT?]
                    Avoiding the problem you think is there is easy.
                    Always connect the scope ground leads to earth ground.
                    If you need to check a + HOT to a - HOT [or any two HOTs + or - ] then use two leads and use 'add'. [Invert as necessary.]
                    It's not that hard.
                    Since it's DC in this case using the right tool [a DMM] would be even better.
                    That's how I was taught and probably why I couldn't see where you thought there was a problem.
                    I never connect a scope ground to a HOT [negative or positive] in the first place so the problem is a non-issue.
                    - Dr Seuss
                    • Th3_uN1Qu3
                      Believe in
                      • Jul 2010
                      • 6031
                      • Romania
                      #90
                      Re: Oscilloscope
                      Yes.
                      It's not that hard.
                      I didn't have a dual channel scope when i needed to check the primary. And i'm talking about seeing the switching waveform to the primary transistors - anywhere you put the ground clip in the primary, it's still "hot" re earth without an isolation transformer. Anyway, i do not own an isolation transformer so i just used an inverter and battery. Worked fine. As far as the safety aspect goes you have to be careful messing around in the primary anyway, but as long as the scope remains earthed there is no danger in touching the case of the scope... you'd have no business touching anything in the primary of the SMPS anyway.
                      • Peter9DO
                        • Dec 2011
                        • 277
                        #91
                        Re: Oscilloscope
                        I'm not going to lie the last few post between Th3 uN1Qu3 and PCBONEZ is great reading and entertaining but I'm so very very confused. I'm so confused I'm not even going to ask for an explanation in softer terms. But can you answer this for me please? Are we using the grounding clip or not and are we 86ing the isolation transformer or are we going to say "it's a good investment"?
                        • joshnz
                          • Feb 2011
                          • 969
                          • New Zealand
                          #92
                          Re: Oscilloscope
                          I too am confused.
                          I am going to give a situation please provide answer.
                          1. Have only a single channel scope and need to measure waveform on primary side.
                          My pc
                          CPU : AMD PHENOM II x4 @ 3.5Ghz
                          MB : ASUS M4A89TD PRO USB3
                          RAM : Kingston ValueRAM 16gb DDR3
                          PSU : Cooler Master 850W Silent Pro
                          GPU : ATI Radeon HD 6850
                          • PCBONEZ
                            Grumpy Old Fart
                            • Aug 2005
                            • 10661
                            Re: Oscilloscope
                            With your 2 Channel scope you should never need one.
                            If you ever have to measure a HOT to a HOT use two probes both with grounds to earth ground.
                            Both + ADD with no Invert on CH2 will display the total.
                            Both + ADD with Invert on CH2 will display the difference.
                            - Dr Seuss
                            • PCBONEZ
                              Grumpy Old Fart
                              • Aug 2005
                              • 10661
                              Re: Oscilloscope
                              Originally posted by joshnz
                              I too am confused.
                              Depends on which waveform you are looking at.
                              The PWM probably references to earth ground.
                              The OP of the switchers could probably be measured on the secondary side just as well - and do the math for the xfmr.
                              Beyond that you'll have to look at what you're working on in greater detail because they aren't all as 'ungrounded' as people seem to think they -all- are.
                              They can all be different. [See Pic].
                              The one in the Pic shows an earth ground right where Th3_uN1Qu3 says connecting a ground would blow it up.
                              That doesn't mean he's wrong for all PSUs though. - Or even most..
                              There's just too many different designs to generalize something like that one way or the other.
                              Worst case is you measure both sides to ground and do the math.
                              - Dr Seuss
                              • godin
                                • Dec 2011
                                • 34
                                #95
                                Re: Oscilloscope
                                Me too, confused as hell. I can't see the problem with the smps but bow to greater knowledge.
                                How come if it is earth and not possible to clip to is it not a shock hazard on the casing? Thinking specifically of a PC power supply.
                                If you connect your scope clip to ground and not anything above ground what can be the problem. I am missing something!
                                If a filter cap is passing AC to ground then ground must be ground.... mustn't it?
                                If you float with an isolation transformer and float the earth then I can understand that there would be a voltage present that was not being returned and would use the scope probe earth as a route but not a short.
                                In PCBONEZ schematic, if the earth were removed from the emi filter between the 2 caps by floating/removing the earth connection as suggested, then the casing and ground of the equipment would become live as the caps become a voltage divider and if you connected a clip to the casing it would be above ground. The whole problem is caused by removing the earth connection.
                                Hope to have my confusion fixed by the knowledgeable
                                • joshnz
                                  • Feb 2011
                                  • 969
                                  • New Zealand
                                  #96
                                  Re: Oscilloscope
                                  PCBONEZ I don't know how that psu even works if those grounds are in fact tied together half of the cycle is shorted.
                                  I don't understand how this psu works.
                                  unless L1 is the key.
                                  My pc
                                  • PCBONEZ
                                    Grumpy Old Fart
                                    • Aug 2005
                                    • 10661
                                    Re: Oscilloscope
                                    It's DC at that point.
                                    I don't get that one either.
                                    The point was they aren't all the same and you have to look to see what's going on.
                                    - Dr Seuss
                                    • godin
                                      • Dec 2011
                                      • 34
                                      #98
                                      Re: Oscilloscope
                                      DC doesn't cycle.
                                      .....and you should be able to connect the ground clip to the negative grounded DC leg of that rectifier output. Shouldn't you??
                                      • Th3_uN1Qu3
                                        Believe in
                                        • Jul 2010
                                        • 6031
                                        • Romania
                                        #99
                                        Re: Oscilloscope
                                        No, you can't. I've already shown above why.
                                        Anyway to sum it up - if you're serious about troubleshooting SMPS you need a two channel oscilloscope.
                                        • PCBONEZ
                                          Grumpy Old Fart
                                          • Aug 2005
                                          • 10661
                                          Re: Oscilloscope
                                          Some SMPS float the section between the first rectifier and the power transformer.
                                          Before that section is grounded and after that section is grounded but that section itself is not.
                                          What I'm saying is using an Iso-Xfmr to float the sections before and after -too- doesn't help anything or even make sense to me.
                                          - Dr Seuss
                                          • skwarszczow
                                            Macbook Pro A2141 020-01700-A GPU Artifacts then no S5 rails
                                            by skwarszczow
                                            This is my first post on the forum
                                            I recently purchased a MacBook A2141 from a colleague at work, which had issues with the dedicated GPU. The system occasionally displayed artifacts on the screen and sometimes experienced kernel panics related to the GPU. Interestingly, applying pressure to the VRAM module would temporarily resolve the artifact issues, which made me optimistic about the ease of repair.
                                            Board: 020-01700, Config: Intel i7, 512GB SSD, 16GB Ram.
                                            Before completing the purchase, we decided to erase...
                                            03-23-2025, 06:04 PM
                                          • prodigy83
                                            MacBook Pro M1 Pro Logic Board 820-02100 Mosfet where to purchase?
                                            by prodigy83
                                            My reference only: TT10141.
                                            Problem: I need to purchase a mosfet chip from board 820-02100 M1 Pro, that is supposed to be Q5820 but according to Digikey, the schematic diagram I sent them, neither are relevant to the actual mosfet specification I sent them. I just basically need to find this mosfet to purchase, I cannot find it anywhere and am hoping someone could point me in the right direction without have to purchase this entire allusive board. Already checked with Digikey, Mouser, Aliexpress, eBay. Any suggestions?
                                            Mosfet specs are as follows:
                                            Chip markings:...
                                            03-12-2024, 01:38 PM
                                          • Evets
                                            Dynamco analogue oscilloscope
                                            by Evets
                                            I have recently acquired a Dynamco D7100 analogue oscilloscope without a user manual. As this is the first time I have owned or used an oscilloscope I would like to try and find an instruction book. The manufacturers no longer hold any information and a Google search only found one site but when I downloaded a 'service manual' all I got was gobbledegook. Any help would be gratefully appreciated or even pointers to a generic manual where I could learn more.
                                            08-30-2024, 06:54 AM
                                          • SolomonMan
                                            Vintage BK Precision 1403 Oscilloscope....
                                            by SolomonMan
                                            All,
                                            Some of you have helped me in the past and I am very appreciative of all the help! (I have fixed TVs/Stereos/ and other small appliances with your efforts...Thanks so much!!!
                                            Today I went to a local HAM Radio Festival. It was one of the smaller ones in our area. My stepfather is the HAM operator. I tinker in electronics but am a Software Programmer (27 years now).
                                            I was looking for a few odds and ends for some electronics work I have to do for the household/farm. Typically these type of shows have the grab bags or a electronics vendor...
                                            12-03-2023, 10:39 PM
                                          • chth96
                                            what capacitor should I purchase for LCD display PCB?
                                            by chth96
                                            I want to fix Video door phone which works its doorbell but it is not able to display any screen.
                                            So I disassembled its plastic case and I detached FPC film cable from LCD display in order to inspect LCD display's PCB
                                            I measured ESR resistance of all smd capacitor on this PCB and I noticed that there is some SMD capacitor which has high ESR value(48ohm 50V 0.47uf,15 and 28 ohm 50v 3.3uf,10 and 8 ohm 35v 10uf)
                                            I desoldered this 5 suspicious SMD capacitor like below screenshot.
                                            images2.imgbox.com/51/b8/8WVZwDpV_o.jpg
                                            But I cannot identify specification of desoldered...
                                            05-24-2025, 09:25 PM
