    Fluke 87 going south?!

    Hello All

    Lately my old Fluke 87 have been behaving erratically , display fading, eating batteries in matter of days , measures going up/down constantly even without leads even connected and the like.

    Since this is a very old meter , i was thinking wheter it was worth or not to have it repaired/serviced/calibrated , a call to a representative yielded a quote of about 100$ plus S/H.

    So what would you guys suggest? a new tester? wich one would be better suited for light use in laptop/lcd troubleshouting wich i've started to get into recently?, or should i just repair the old one?.

    Ive considered getting the Blue ESR meter , so in case you suggest a good one what specs should i look for that would be a good complement for the esr meter?

    Your insights are very appreciated , thanks in advance
    Aiki
    Last edited by Aikiotto; 02-10-2012, 10:27 AM.
  • Krankshaft
    Badcaps Legend
    • Jan 2007
    • 2328
    • USA

    #2
    Re: Fluke 87 going south?!

    The display fading issue can be solved by replacing the zebra strip on the back of the display. Fluke has a kit available with a new one.

    Ignore the instructions to wipe the old one with a pencil eraser the effects are short lived and the display will begin to fade again.

    Is it possible the segments of the display are dropping out due to the bad zebra strip and you are misinterpreting that as a fluctuating reading?

    The Fluke 87 series is a good meter and based on what it cost new it would be worth a repair. If your issue is only a fading display you can do it yourself easily.

    If you wanted to upgrade to the newer 87-5 it cost me around $280 and that was the lowest price I could find. So you have to weigh whether the added features and the higher cost for a new unit are worth it.

    One things for sure there's no way you're going to get a better meter for the $100 repair fee.
    Last edited by Krankshaft; 02-10-2012, 11:00 AM.
    Elements of the past and the future combining to make something not quite as good as either.

    Comment

    • Aikiotto
      Member
      • Oct 2011
      • 30

      #3
      Re: Fluke 87 going south?!

      Hey Krankshaft

      I did read about the kits for the lcd from ebay (some 28$) and if it only were that issue i would not have doubts about solving , but im not 100% sure if thats the problem , since with new batteries the screen works perfectly....if only for a few days before it indicates low batt again.

      But even with a new batt I just switch to mDC measure for instance , it just start showing values out of nothing going up and down with no leads plugged , it also kind of turns on and shut off by itself for half a second if i hold to it tight and things like that.

      I checked the 87-5 its going for some 320$ at amazon wichs a bit steep considering how seldom im using it atm , so i tought about hearing opinions on others DMMs like the Fluke 115 (129$ at amazon) or any other brand/models you know about.

      Comment

      • retiredcaps
        Badcaps Legend
        • Apr 2010
        • 9271

        #4
        Re: Fluke 87 going south?!

        Originally posted by Aikiotto
        but im not 100% sure if thats the problem , since with new batteries the screen works perfectly....if only for a few days before it indicates low batt again.
        1) I haven't taken apart a Fluke 87 (I don't have one), but I do recall do some research on multimeters and ran into a couple of posts that might help?

        2) First is carbon contacts causing smears that drain the battery.

        http://www.eevblog.com/2010/02/08/ee.../#comment-5956

        3) Second is cleaning the contacts

        http://www.audiokarma.org/forums/sho...d.php?t=123354

        4) I guess you can also verify a parasitic drain by hooking up another multimeter and checking the amps draw while the Fluke 87 is turned off. Sort of like a car battery parasitic drain test.

        5) If none of this helps and you are considering buying a new one, PM me regarding your old Fluke 87.

        6) Don't forget about ebay. You can get some brand new Flukes for less money. I have seen the 117 go for about $100. Sometimes a 177 as well for $100.

        I believe the 11x series are all made in China. I think the 17x is made in the USA. Dave Jones at eevblog reviewed the 117 and found the made in China just as good as made in USA. See

        www.youtube.com/watch?v=3k9nVg03Wi4
        Last edited by retiredcaps; 02-10-2012, 04:35 PM.
        --- begin sig file ---

        If you are new to this forum, we can help a lot more if you please post clear focused pictures (max resolution 2000x2000 and 2MB) of your boards using the manage attachments button so they are hosted here. Information and picture clarity compositions should look like this post.

        We respectfully ask that you make some time and effort to read some of the guides available for basic troubleshooting. After you have read through them, then ask clarification questions or report your findings.

        Please do not post inline and offsite as they slow down the loading of pages.

        --- end sig file ---

        Comment

        • Aikiotto
          Member
          • Oct 2011
          • 30

          #5
          Re: Fluke 87 going south?!

          Thanks for the links Retired

          Whats described in the EEV blog from your link seems like exactly the problem with my unit , wich is the rotary switch wonder if thats servicable by a newbie.

          Do you think the meter would lose calibration somehow if i were to try and clean those carbon contacts? and whats the safest product to use for such a task?.

          Thanks again

          Comment

          • retiredcaps
            Badcaps Legend
            • Apr 2010
            • 9271

            #6
            Re: Fluke 87 going south?!

            Originally posted by Aikiotto
            Whats described in the EEV blog from your link seems like exactly the problem with my unit , wich is the rotary switch wonder if thats servicable by a newbie.
            Service manual is at



            See chapter 3. It has disassembly instructions and how to clean the rotary switch (section 3-12).

            Do you think the meter would lose calibration somehow if i were to try and clean those carbon contacts? and whats the safest product to use for such a task?.
            As long as you don't touch R21, R34, C2, C3, and C37 (section 3-29), you shouldn't affect any calibration settings. Don't even breathe in that direction!

            Section 3-12 has suggested cleaning tips and materials.

            I would like to ask a favour. Since I don't have a Fluke 87, can you please take pictures of the step by step process so I can "virtually" repair this? I have a huge interest in multimeters.

            Please use the manage attachment feature to upload your pics and please don't post inline for my sake.
            --- begin sig file ---

            If you are new to this forum, we can help a lot more if you please post clear focused pictures (max resolution 2000x2000 and 2MB) of your boards using the manage attachments button so they are hosted here. Information and picture clarity compositions should look like this post.

            We respectfully ask that you make some time and effort to read some of the guides available for basic troubleshooting. After you have read through them, then ask clarification questions or report your findings.

            Please do not post inline and offsite as they slow down the loading of pages.

            --- end sig file ---

            Comment

            • Krankshaft
              Badcaps Legend
              • Jan 2007
              • 2328
              • USA

              #7
              Re: Fluke 87 going south?!

              If you want to see inside an 87-5 Dave at EEVBlog has a tear down video of it.
              Elements of the past and the future combining to make something not quite as good as either.

              Comment

              • smason
                Badcaps Legend
                • Feb 2010
                • 1652
                • Canada

                #8
                Re: Fluke 87 going south?!

                It's absolutely worth fixing, especially since you can do it yourself.
                I'd definietly replace the elastomer on the display, and clean the switch, bet that's all it needs.
                36 Monitors, 3 TVs, 4 Laptops, 1 motherboard, 1 Printer, 1 iMac, 2 hard drive docks and one IP Phone repaired so far....

                Comment

                • cadiman
                  Badcaps Veteran
                  • Mar 2010
                  • 397

                  #9
                  Re: Fluke 87 going south?!

                  These meters are very easy to take apart.

                  I have an old 87 and every once in a while I take it apart and clean it and do some re- soldering. I have installed circuit breaker in place of the fuses.
                  (tired of replacing expensive fuses)

                  I drop it alot and subject it to cold and hot conditions all the time.

                  This is the best meter I have ever had and I bought it used for 120.00

                  It is worth fixing.

                  Start with a good quality battery and a good cleaning.

                  Make sure you are using good test leads too.
                  Last edited by cadiman; 02-24-2012, 09:02 PM.

                  Comment

                  • retiredcaps
                    Badcaps Legend
                    • Apr 2010
                    • 9271

                    #10
                    Re: Fluke 87 going south?!

                    Originally posted by cadiman
                    I have an old 87 and every once in a while I take it apart and clean it and do some re- soldering.
                    May I ask what parts are you resoldering? I'm thinking about getting a used Fluke 87 and just want to know if I have to be doing the same.

                    How has your calibration held up over the years?

                    I have installed circuit breaker in place of the fuses.
                    (tired of replacing expensive fuses)
                    I have purchased 2 used Flukes and both times the seller did not disclose that the milliamp fuse was blown.

                    Now I'm baffled why these milliamp fuses are blown for the following reason. If someone can spend the money to buy a Fluke, then I'm assuming that person is an experienced multimeter user which implies always testing the circuit on the highest Amp rating first (usually 10A). If the amp reading is comfortably lower than the rated multimeter milliamp scale, then and only then, move the probe to the milliamp scale to get better resolution.

                    I can see inexperienced users blowing out the fuse, if there is one, on the cheap 830 multimeter ebay variations, but not someone spending $400 on a Fluke 87-V.

                    Similary, if you even think the circuit is more than 10A, then I would expect an experienced user to get out their amp clamp and use that.
                    Last edited by retiredcaps; 02-24-2012, 10:02 PM.
                    --- begin sig file ---

                    If you are new to this forum, we can help a lot more if you please post clear focused pictures (max resolution 2000x2000 and 2MB) of your boards using the manage attachments button so they are hosted here. Information and picture clarity compositions should look like this post.

                    We respectfully ask that you make some time and effort to read some of the guides available for basic troubleshooting. After you have read through them, then ask clarification questions or report your findings.

                    Please do not post inline and offsite as they slow down the loading of pages.

                    --- end sig file ---

                    Comment

                    • smason
                      Badcaps Legend
                      • Feb 2010
                      • 1652
                      • Canada

                      #11
                      Re: Fluke 87 going south?!

                      I can't imagine a Fluke needing any soldering unless you toasted a compnent that needs replacing. Build quality (and soldering) have always been really good on any FLuke I've had apart.
                      36 Monitors, 3 TVs, 4 Laptops, 1 motherboard, 1 Printer, 1 iMac, 2 hard drive docks and one IP Phone repaired so far....

                      Comment

                      • KeriJane
                        Mac Enthusiast
                        • Sep 2008
                        • 681
                        • USA

                        #12
                        Re: Fluke 87 going south?!

                        +1 vote for careful disassembly, inspection and cleaning.

                        Be careful with replacing those expensive fuses with any kind of substitute.

                        If I recall, those expensive Fluke fuses are designed to prevent high-voltage arcing which can happen with a regular fuse or circuit breaker.

                        If, for example, one left the meter on an Amps setting and accidentally probed a 440v line the circuit breaker or "plain" fuse would immediately blow and prevent damage, right?

                        Wrong!

                        The fuse will blow but the high voltage can continue arcing through the space opened by the melted link and would most likely cause severe damage to the meter and user.

                        Ow.

                        Those expensive fuses are designed to minimize that risk.

                        If your meter uses the cheaper fuses or has been modded in any way it is best to label it clearly "LOW VOLTAGE ONLY" and then make sure that absolutely nobody else borrows it for anything, ever.

                        Or better, just get "those darn expensive fuses" and learn to stop making them pop.
                        The More You Learn The Less You Know!

                        Comment

                        • Agent24
                          I see dead caps
                          • Oct 2007
                          • 4958
                          • New Zealand

                          #13
                          Re: Fluke 87 going south?!

                          Are those "zebra stripe" LCD connectors a universal thing that can be obtained easily?

                          I've got a couple of handheld games with a fading display problem. Both use those strip connectors.
                          "Tantalum for the brave, Solid Aluminium for the wise, Wet Electrolytic for the adventurous"
                          -David VanHorn

                          Comment

                          • budm
                            Badcaps Legend
                            • Feb 2010
                            • 40746
                            • USA

                            #14
                            Re: Fluke 87 going south?!

                            @retiredcaps, I have about 6 of these 87 at works, what happen is that when you want to measure the current, you have to move the red probe over one of the two fused current inputs, one is for 400mA max, another one is for 10A max, if the person forget to move the red probe back to the V/Ohm input and use it to measure voltage, boom!, it blows the fuse.
                            I also order spare 4-input jack assemblies, what happen is that the contacts inside just worn out, or sometime crack solder joints on the jack to the PCB. It is great meter, I use it at home also.
                            Regular fuse only has 250v break down, not recommend if you are going to hand this meter out to your employees without having proper type of fuse installed.
                            Last edited by budm; 03-04-2012, 11:03 PM.
                            Never stop learning
                            Basic LCD TV and Monitor troubleshooting guides.
                            http://www.badcaps.net/forum/showthr...956#post305956

                            Voltage Regulator (LDO) testing:
                            http://www.badcaps.net/forum/showthr...999#post300999

                            Inverter testing using old CFL:
                            http://www.badcaps.net/forum/showthr...er+testing+cfl

                            Tear down pictures : Hit the ">" Show Albums and stories" on the left side
                            http://s807.photobucket.com/user/budm/library/

                            TV Factory reset codes listing:
                            http://www.badcaps.net/forum/showthread.php?t=24809

                            Comment

                            • retiredcaps
                              Badcaps Legend
                              • Apr 2010
                              • 9271

                              #15
                              Re: Fluke 87 going south?!

                              Originally posted by budm
                              @retiredcaps, I have about 6 of these 87 at works, what happen is that when you want to measure the current, you have to move the red probe over one of the two fused current inputs, one is for 400mA max, another one is for 10A max, if the person forget to move the red probe back to the V/Ohm input and use it to measure voltage, boom!, it blows the fuse.
                              One of the features of the Fluke 87 is the input alert. According to page 1 (one) of the manual ...

                              "An Input Alert (tm) that causes the beeper to sound if the test leads are plugged into the wrong input terminals for the function being performed."

                              So unless people aren't paying attention, the beeping should alert them to something wrong and they should double check their input jacks.

                              Some newer multimeters now have a shutter to close off the input jacks depending on the selected rotary setting. Others now prominently flash "LEAD" on the display.

                              It wouldn't be so bad, but Fluke fuses are $6 to $8 each. At that price, you guys should get the Harbour Freight $4 multimeter specials and use them exclusively for current measurement.
                              --- begin sig file ---

                              If you are new to this forum, we can help a lot more if you please post clear focused pictures (max resolution 2000x2000 and 2MB) of your boards using the manage attachments button so they are hosted here. Information and picture clarity compositions should look like this post.

                              We respectfully ask that you make some time and effort to read some of the guides available for basic troubleshooting. After you have read through them, then ask clarification questions or report your findings.

                              Please do not post inline and offsite as they slow down the loading of pages.

                              --- end sig file ---

                              Comment

                              • budm
                                Badcaps Legend
                                • Feb 2010
                                • 40746
                                • USA

                                #16
                                Re: Fluke 87 going south?!

                                Sometime the areas are very noisy and the users do not pay attention so it happens.
                                My work place required yearly calibration certificates for all of our test equipment since we are UL and CSA certified labs, cheap meters will not make it.
                                Never stop learning
                                Basic LCD TV and Monitor troubleshooting guides.
                                http://www.badcaps.net/forum/showthr...956#post305956

                                Voltage Regulator (LDO) testing:
                                http://www.badcaps.net/forum/showthr...999#post300999

                                Inverter testing using old CFL:
                                http://www.badcaps.net/forum/showthr...er+testing+cfl

                                Tear down pictures : Hit the ">" Show Albums and stories" on the left side
                                http://s807.photobucket.com/user/budm/library/

                                TV Factory reset codes listing:
                                http://www.badcaps.net/forum/showthread.php?t=24809

                                Comment

                                • retiredcaps
                                  Badcaps Legend
                                  • Apr 2010
                                  • 9271

                                  #17
                                  Re: Fluke 87 going south?!

                                  Originally posted by budm
                                  My work place required yearly calibration certificates for all of our test equipment since we are UL and CSA certified labs, cheap meters will not make it.
                                  Well stock up on the fuses then. I have purchased 3 used multimeters and each one has the milliamp fuse blown, but the sellers never disclosed that little fact.

                                  I don't mind paying for one Fluke 440ma replacement fuse because I'm confident that I will highly unlikely to blow it, but the shipping from digikey.ca is a minimum of $8 and ebay Fluke fuse sellers want $15 for shipping and handling for a fuse!!!
                                  --- begin sig file ---

                                  If you are new to this forum, we can help a lot more if you please post clear focused pictures (max resolution 2000x2000 and 2MB) of your boards using the manage attachments button so they are hosted here. Information and picture clarity compositions should look like this post.

                                  We respectfully ask that you make some time and effort to read some of the guides available for basic troubleshooting. After you have read through them, then ask clarification questions or report your findings.

                                  Please do not post inline and offsite as they slow down the loading of pages.

                                  --- end sig file ---

                                  Comment

                                  • budm
                                    Badcaps Legend
                                    • Feb 2010
                                    • 40746
                                    • USA

                                    #18
                                    Re: Fluke 87 going south?!

                                    We do stock them up, part of doing business. I do not want to tell you how much it costs us each year to have our equipment calibrated.
                                    We just bought new scope for $80K, probes are OPTIONS!
                                    UL, CSA, TuV for our products compliance testing are even worst.
                                    Never stop learning
                                    Basic LCD TV and Monitor troubleshooting guides.
                                    http://www.badcaps.net/forum/showthr...956#post305956

                                    Voltage Regulator (LDO) testing:
                                    http://www.badcaps.net/forum/showthr...999#post300999

                                    Inverter testing using old CFL:
                                    http://www.badcaps.net/forum/showthr...er+testing+cfl

                                    Tear down pictures : Hit the ">" Show Albums and stories" on the left side
                                    http://s807.photobucket.com/user/budm/library/

                                    TV Factory reset codes listing:
                                    http://www.badcaps.net/forum/showthread.php?t=24809

                                    Comment

                                    • retiredcaps
                                      Badcaps Legend
                                      • Apr 2010
                                      • 9271

                                      #19
                                      Re: Fluke 87 going south?!

                                      Originally posted by budm
                                      We do stock them up, part of doing business. I do not want to tell you how much it costs us each year to have our equipment calibrated.
                                      Part of me really wants, but does not need, a Fluke 87V. The only problem with buying an used Fluke 87V is I'm worried about its calibration (or lack of it). As I understand it, you need special software and calibrator to get it done since the 87V is closed case calibration.

                                      The original Fluke 87 and 87III had potentiometers so you could DIY if you had an accurate source.

                                      I recently bought an older pocket multimeter off ebay. The AC and DC readings were well off compared to my Fluke. I opened it up and found two pots. I figured out which was DC and I tuned it so now it reads similar to the Fluke.

                                      I also got an old Micronta multimeter for free and its accuracy was well off too. I had to tweak the pots.

                                      Let's hope my Fluke is accurate!

                                      I understand that software driven closed case calibration is probably more reliable and accurate than turning a screwdriver on a pot even at a calibration lab.

                                      Obviously, this is all for my home/personal use which is "good enough".
                                      Last edited by retiredcaps; 03-05-2012, 01:20 AM.
                                      --- begin sig file ---

                                      If you are new to this forum, we can help a lot more if you please post clear focused pictures (max resolution 2000x2000 and 2MB) of your boards using the manage attachments button so they are hosted here. Information and picture clarity compositions should look like this post.

                                      We respectfully ask that you make some time and effort to read some of the guides available for basic troubleshooting. After you have read through them, then ask clarification questions or report your findings.

                                      Please do not post inline and offsite as they slow down the loading of pages.

                                      --- end sig file ---

                                      Comment

                                      • Aikiotto
                                        Member
                                        • Oct 2011
                                        • 30

                                        #20
                                        Re: Fluke 87 going south?!

                                        Hello and sorry for the late update , buts been a crazy couple of months!!

                                        I finally got an used 87-V off Amazon (on the mail atm from the US to where im in Venezuela) with an industrial probes set , carrying case and whatnot for some 250$ , as for my old 87 i managed to clean the contacts on the rotary switch with pure alcohol and didnt exhibit the issue for a couple days , then its back to misbehaving.

                                        Once the new baby arrives i hope you guys can help me check its calibration , an issue retiredcaps just got me thinking about lol.

                                        Thanks for the help!

                                        Comment

                                        Related Topics

                                        Collapse

                                        • sam_sam_sam
                                          Modification of a Fluke multi meter that uses three AA batteries to use a 14500 lithium ion battery
                                          by sam_sam_sam
                                          I have had with regular AA batteries that leak this meter is expensive new I already replaced the battery battery clips once and the same thing happened again and I am tried of so it is getting modified to accommodate lithium ion batteries that are the same size excluding the positive terminal and removing the battery clips I will have enough room for the BMS protection board

                                          I going to put 3 lithium ion battery in parallel with the BMS protection board and a way to charge the battery externally and heat shrink the battery pack and use a JST connector so I easily remove the battery...
                                          12-31-2023, 06:04 AM
                                        • Baldbentley
                                          repairing a Fluke 335 clamp meter
                                          by Baldbentley
                                          I am repairing a Fluke 335 clamp meter. After contacting fluke it appears there is no service manual or circuit board/ component list available. I have determined that there us a faulty PNP bipolar junction transistors shown in the pictures. I am having trouble sourcing this transistor. I pulled a good one off of the board and tested it with my component analyzer but I cannot find anything that matches the specifications. Any help locating this part or a replacement would be greatly appreciated.

                                          Thanks...
                                          10-18-2023, 02:05 AM
                                        • chth96
                                          Is it possible to test small ceramic disc capacitor with analogue meter?
                                          by chth96
                                          I read EBOOK of Testing Electronic Component and found that

                                          "In order to test the small blue resin coated ceramic disk capacitor which has 104 50v specification,it is possible to test this small ceramic capacitor by making use of analogue meter.
                                          Set to times 10Kohm that has 12v output from the probe to check disorder of ceramic capacitor.
                                          It is not uncommon that testing result of digital capacitance meter is okay,But It show shorted reading when you check it with analogue meter."

                                          Is it really possible to test small ceramic disc capacitor with analogue...
                                          03-09-2024, 10:03 PM
                                        • Ross Kohler
                                          Diagnosing South Bridge (FH82H370) on Asus H370-Pro Gaming Motherboard
                                          by Ross Kohler
                                          Hi,

                                          I'm learning about motherboard repair and I believe I've traced the issue down for the South Bridge on the particular motherboard I'm working on. My path of concluding that:
                                          - There's no sign of any attempt to power on when pushing the power button.
                                          - No PSON signal is reaching the 24-pin ATX connector.
                                          - The Super IO chip is receiving the PS_SWITCH signal from the power button and is relaying this signal back to the South Bridge.
                                          - The Super IO is not receiving the S_SLPS#_R signal back from the South Bridge
                                          - If I pull the Super IO S_SLPS#_R pin...
                                          04-16-2024, 09:36 AM
                                        • jorpec
                                          Changing the scale of an analog meter
                                          by jorpec
                                          Hi

                                          I want to change the scale of an analog meter, to monitor the charge of an li-ion single cell, so the needle at the end would point 4.2 volts and the 0 on the meter would point 2.5 volts or so, i tried to add diodes or a zener in series with the meter but the changes of the ambient temperature also changes the voltage value on the meter
                                          I wonder how i can accomplish this so the meter would be precise at any temperature ?

                                          Thnaks
                                          01-23-2023, 10:37 AM
                                        • Loading...
                                        • No more items.
                                        Working...