Oscilloscope
    Re: Oscilloscope
    I have two oscilloscopes.
    One 100 MHz digital storage scope used for most of my non repair work as it's only 250V maximum input rated and poor single shot capability. (HP 54501A, cost me £150.)
    And one 20 MHz analog scope (cost £60), which I have used successfully to repair two plasma TVs; I chose it for 400V inputs and despite what people have said, with care a 20 MHz scope is sufficient for debugging a 720p plasma (and resolved an okay waveform on a 1080p Panasonic) although you are limited in the higher frequency ranges and you have to turn the intensity right up (and lose a bit of focus control). The most common adjustment is the Vsetup and Vsetdn (or Yfr/Yrr) which can be done if you can read 1/5th division.
    • PCBONEZ
      Grumpy Old Fart
      • Aug 2005
      • 10661
      Re: Oscilloscope
      Originally posted by TheLaw
      I just snagged a 100Mhz Tek 465M for $70 shipped off eBay.
      It's sweet.
      I used LOTS of those.
      Military version of the 465B.
      Very good scope but it's analog.
      Did you get the hard-shell front cover with it?
      Mann-Made Global Warming.
      - Dr Seuss
      • Peter9DO
        • Dec 2011
        • 277
        #43
        Re: Oscilloscope
        I got the hard shell cover for the front. The original manuals, the manual holder and it's in VERY good condition! I will post a picture of it here in a little while.
        • retiredcaps
          • Apr 2010
          • 9271
          #44
          Re: Oscilloscope
          From your description in post #36, I'm sure the guy selling takes care of his stuff and probably a little sad to see it go.
          --- begin sig file ---
          --- end sig file ---
          • Peter9DO
            • Dec 2011
            • 277
            #45
            Re: Oscilloscope
            Originally posted by PCBONEZ
            Yes.
            Here are the specs for the 'stock' probes for that.
            Need to get probes with a compensating range that covers the input capacitance.
            Need to get probes with the correct input resistance.
            Both the above are different between 1x and 10x probes.
            You probably want at least one of each of 1x and 10x or one with a 1x-10x switch.
            [If you connect to 50v with a 10x probe the scope only sees 5v.]
            When you get the probes connect to the test signal and adjust the compensation for a nice square square-wave on the screen.
            [That takes out error due to the cable size and length and the probe.]
            The adjuster is usually a very small screw. Location varies.
            Could be near the connection to the scope or on the probe itself.
            By the way, thank you for finding this PDF for me! Here are the images of my purchase.
            • PCBONEZ
              Grumpy Old Fart
              • Aug 2005
              • 10661
              Re: Oscilloscope
              Looks like a runner.
              See where it says "Probe Adjust, 500 mV P-P, 1kHz".
              That's the port to adjust compensation on the probes I mentioned earlier.
              - Dr Seuss
              • Peter9DO
                • Dec 2011
                • 277
                #47
                Re: Oscilloscope
                Yea, I was probing that hole... lol! You know I looked all over the internet today for about an hour and I was not able to come across a digital format of the service manual, however I did find a hard copy someone was trying to sell for $100 (that's a little crazy). Oh well it, is what it is, maybe one day I will be able to come across one. Anyways! To everyone that posted on this thread (especially PCBONEZ & RetiredCAPS) Thank you very very very very very very very very MUCH! For all the help on giving me some information on what to look for when purchasing an oscilloscope! You guys are awesome! I have borrowed some probes from my cousin (until the ones I ordered from eBay come in) and have been playing with this thing for a couple of hours now. Again, thank you for all the help!
                Okay, I have one more question, this is on the use of the oscilloscope. When you are testing the circuit of a power supply for a computer should you have the power supply plugged in to the DC outlet or do you keep all flowing current off during testing? Just a little confused? I'm getting a reading when there is no power being put threw it and I'm scared I'm going to mess the oscilloscope up if the power supply is plugged in, so I have not tested it while it has been plugged in.
                -Pete
                • retiredcaps
                  • Apr 2010
                  • 9271
                  #48
                  Re: Oscilloscope
                  Y I was not able to come across a digital format of the service manual,
                  Ask on eevblog.com. Tekfan might have one. He restores old Tek scopes.
                  • Agent24
                    I see dead caps
                    • Oct 2007
                    • 5097
                    • New Zealand
                    #49
                    Re: Oscilloscope
                    -Pete
                    Measuring the PSU while unloaded won't tell you much, you need to load the PSU if you want to know what kind of ripple\noise is going to be coming out when it's under normal conditions.
                    You won't damage anything by having it powering something or not, but I think if it's connected to a motherboard you might get extra noise coming FROM the motherboard etc and that can affect your reading.
                    -David VanHorn
                    • Th3_uN1Qu3
                      Believe in
                      • Jul 2010
                      • 6031
                      • Romania
                      #50
                      Re: Oscilloscope
                      Um, of course you plug it in. If there's no power to the thing there will be no electrical signals to view obviously. You also use dummy loads to load up the PSU and simulate a real computer, because hooking up a motherboard will affect the readings just like Agent24 said, and you are interested in what's coming from the PSU alone.
                      The power supply needs to be plugged in AND connected via an isolation transformer to the mains. If you do not have an isolation transformer then DO NOT attempt to probe anything in the PRIMARY section, otherwise you'll blow up the power supply, the scope, or yourself. An alternative to an isolation transformer is to use an inverter or an UPS on battery power to feed the PSU under test. Some people have those lying around.
                      A working TV? How boring!
                      • Peter9DO
                        • Dec 2011
                        • 277
                        #51
                        Re: Oscilloscope
                        I felt kind of stupid for asking that, I didn't understand why I was getting a signal off of it when it wasn't plugged in. The only thing I can think of is that the caps have not been fully discharged and I'm getting something off of them. Please keep in mind, I currently have no clue how to use this and I'm just starting my learning process now that I have one to learn on. I know I'm not going to learn it over night, but I just need somewhere to start, I had a broken power supply sitting there so I just took it apart and started messing with it. If you guys have an idea of something that would be more on the beginner level, I have tons of stuff here. computer parts, a bunch of laptops, Plasma TVs, LCD monitors and the list goes on.
                        • Peter9DO
                          • Dec 2011
                          • 277
                          #52
                          Re: Oscilloscope
                          Thank you! I posted on the forum! I will check it later to see if I got any response!
                          Thanks again,
                          Peter
                          • PCBONEZ
                            Grumpy Old Fart
                            • Aug 2005
                            • 10661
                            Re: Oscilloscope
                            For an experience builder use a good PSU and measure the Ripple on each of the outputs.
                            Don't even need to pull the PSU apart for that.
                            Will help you get the hang of what the knobs on the scope do.
                            - Dr Seuss
                            • godin
                              • Dec 2011
                              • 34
                              #54
                              Re: Oscilloscope
                              "If you do not have an isolation transformer then DO NOT attempt to probe anything in the PRIMARY section, otherwise you'll blow up the power supply, the scope, or yourself"
                              Is this true???
                              • PCBONEZ
                                Grumpy Old Fart
                                • Aug 2005
                                • 10661
                                Re: Oscilloscope
                                #11, #16 and some others.
                                http://www.allaboutcircuits.com/worksheets/scope1.html
                                - Dr Seuss
                                • Peter9DO
                                  • Dec 2011
                                  • 277
                                  #56
                                  Re: Oscilloscope
                                  Thanks, I will take a look at it now!
                                  • godin
                                    • Dec 2011
                                    • 34
                                    #57
                                    Re: Oscilloscope
                                    Great link, thanks.
                                    I think I am correct in my understanding that the ground clip if it is connected to ground is safe and the best way to measure in the examples given would be with 2 channels both grounded by their clips?
                                    • PCBONEZ
                                      Grumpy Old Fart
                                      • Aug 2005
                                      • 10661
                                      Re: Oscilloscope
                                      Originally posted by godin
                                      Great link, thanks.
                                      I think I am correct in my understanding that the ground clip if it is connected to ground is safe ...
                                      No, that's the possible problem.
                                      Depends on what you are calling 'ground' too.
                                      The clip on the probe is connected to the scope's chassis which is connected to the ground in the scope's power plug.
                                      If the 'ground' you connect it to isn't at the same potential as the scope's chassis -and- that chassis is connected to line ground then you just made a ground loop current path through site wiring, the chassis under test and the scope.
                                      I'm a bit beat and I'm not explaining it very well.
                                      Maybe the Boy Wonder from Romania has a good drawing to make it more clear?
                                      - Dr Seuss
                                      • godin
                                        • Dec 2011
                                        • 34
                                        #59
                                        Re: Oscilloscope
                                        Sooo, I think that makes a lot of sense. Therefore the 2 probe 2 channel with no earth clips connected is the correct approach. This would eliminate any chance of potential difference.
                                        • PCBONEZ
                                          Grumpy Old Fart
                                          • Aug 2005
                                          • 10661
                                          Re: Oscilloscope
                                          Maybe not.
                                          - I'm not real clear on what you are doing and I don't want to give you bad advice.
                                          To get a signal to the scope there has to be a return path [aka reference] for whatever the probe is connected to.
                                          Where is/are your return paths?
                                          - Dr Seuss
                                          • skwarszczow
                                            Macbook Pro A2141 020-01700-A GPU Artifacts then no S5 rails
                                            by skwarszczow
                                            This is my first post on the forum
                                            I recently purchased a MacBook A2141 from a colleague at work, which had issues with the dedicated GPU. The system occasionally displayed artifacts on the screen and sometimes experienced kernel panics related to the GPU. Interestingly, applying pressure to the VRAM module would temporarily resolve the artifact issues, which made me optimistic about the ease of repair.
                                            Board: 020-01700, Config: Intel i7, 512GB SSD, 16GB Ram.
                                            Before completing the purchase, we decided to erase...
                                            03-23-2025, 06:04 PM
                                          • prodigy83
                                            MacBook Pro M1 Pro Logic Board 820-02100 Mosfet where to purchase?
                                            by prodigy83
                                            My reference only: TT10141.
                                            Problem: I need to purchase a mosfet chip from board 820-02100 M1 Pro, that is supposed to be Q5820 but according to Digikey, the schematic diagram I sent them, neither are relevant to the actual mosfet specification I sent them. I just basically need to find this mosfet to purchase, I cannot find it anywhere and am hoping someone could point me in the right direction without have to purchase this entire allusive board. Already checked with Digikey, Mouser, Aliexpress, eBay. Any suggestions?
                                            Mosfet specs are as follows:
                                            Chip markings:...
                                            03-12-2024, 01:38 PM
                                          • Evets
                                            Dynamco analogue oscilloscope
                                            by Evets
                                            I have recently acquired a Dynamco D7100 analogue oscilloscope without a user manual. As this is the first time I have owned or used an oscilloscope I would like to try and find an instruction book. The manufacturers no longer hold any information and a Google search only found one site but when I downloaded a 'service manual' all I got was gobbledegook. Any help would be gratefully appreciated or even pointers to a generic manual where I could learn more.
                                            08-30-2024, 06:54 AM
                                          • SolomonMan
                                            Vintage BK Precision 1403 Oscilloscope....
                                            by SolomonMan
                                            All,
                                            Some of you have helped me in the past and I am very appreciative of all the help! (I have fixed TVs/Stereos/ and other small appliances with your efforts...Thanks so much!!!
                                            Today I went to a local HAM Radio Festival. It was one of the smaller ones in our area. My stepfather is the HAM operator. I tinker in electronics but am a Software Programmer (27 years now).
                                            I was looking for a few odds and ends for some electronics work I have to do for the household/farm. Typically these type of shows have the grab bags or a electronics vendor...
                                            12-03-2023, 10:39 PM
                                          • chth96
                                            what capacitor should I purchase for LCD display PCB?
                                            by chth96
                                            I want to fix Video door phone which works its doorbell but it is not able to display any screen.
                                            So I disassembled its plastic case and I detached FPC film cable from LCD display in order to inspect LCD display's PCB
                                            I measured ESR resistance of all smd capacitor on this PCB and I noticed that there is some SMD capacitor which has high ESR value(48ohm 50V 0.47uf,15 and 28 ohm 50v 3.3uf,10 and 8 ohm 35v 10uf)
                                            I desoldered this 5 suspicious SMD capacitor like below screenshot.
                                            images2.imgbox.com/51/b8/8WVZwDpV_o.jpg
                                            But I cannot identify specification of desoldered...
                                            05-24-2025, 09:25 PM
