LG W2234S Repair Help...
  • docksteaderluke
    • Jul 2011
    • 39
    I have an LG W2234S that isn't working. The screen turns on and shows the LG logo and then about 2 seconds later turns off (black screen, no power LED). Then it turns on again and shows the logo (and power LED is on)... Repeat.
    I've pulled it apart and none of the caps are bulging or leaking. The big filter cap (150uf 450V) only charges to about 160V though... I feel like this could be a problem? I've heard also that the u108 chip (25L1005) on the logic card can get corrupted. I'm not sure this is the issue since it does actually display the LG logo for a few seconds but I could be wrong.
    Any help would be greatly appreciated!
  • PlainBill
    • Feb 2009
    • 7034
    Re: LG W2234S Repair Help...
    The symptoms could indicate bad caps, or possibly another problem.
    1. Please provide sharp, well lighted pictures of top and bottom views of the power supply, and a top view of the signal card. Take the pictures from as close to directly over the cards as possible. A resolution as close to 2000 x 2000 is also helpful.
    2. For best results, the test you described above should be done with the monitor hooked to a computer displaying a stable image - a the desktop, for example.
    3. Please confirm that the backlights and power light go off at the same time, and come back on without any action on your part.
    PlainBill
    • docksteaderluke
      • Jul 2011
      • 39
      Thanks for your help! In response to your questions:
      1. See the attached photos.
      2. Unfortunately I can never get to the point where I have a stable image since the monitor is turning on and off every 2 seconds (literally). I did perform the same measurement anyways. When the screen is powered on it measure 148Vdc across the big cap and when it's off it's 157Vdc across that same cap.
      3. The backlight and power LED do infact turn both on and off at the exact same time. The controls are unresponsive during the time that the monitor is on.
      • seanc
        • Nov 2008
        • 1319
        If the power light is staying on, but the *backlights* are turning on and off every 2 seconds then I would go after the CCFL cables - one of them is likely loose/disconnected from the bulb. Had the same issue with a W2043S
        You can confirm that by hooking up known good bulbs to the inverter.
        • docksteaderluke
          • Jul 2011
          • 39
          Nope... the power LED is turning off too... Both the power LED AND the backlight are turning on and off at the same time...
          • PlainBill
            • Feb 2009
            • 7034
            Thanks for confirming that. The 'smoking gun' in this case is the voltage across the large cap. If Charlottetown, PE is tied into the Canadian power grid, you have given a very solid clue as to the problem. The voltage across the large cap should be 1.4 x your AC line voltage, and quite stable whether the monitor is on or off. With the variation you describe, either you have SOMETHING drawing way too much current, a thermistor that is not working, or the main filter cap is bad. Su'scon does not have a good reputation.
            There are a number of ways to test this theory. Choose one or more of the following methods.
            1. Disconnect the signal card from the power supply. Measure the voltage across the large cap. It should be VERY close to the 1.4 x line voltage figure (for 120 VAC the figure is 167 volts), and stable. If it is much lower, CAREFULLY run your hands near the board, looking for a source of heat. Do NOT touch any heat sinks!!!
            2. With everything hooked together and the monitor plugged in and flashing on and off, short across the thermistor. That is the green device just below the fuse. Since I'm asking you to do this while power is on, I'm also asking you to be very careful. If you elect not to try this, I understand.
            3. If you have access to a capacitor rated at at least 200 volts, and 100uF, wire it in parallel with the existing cap. If the flashing stops, that virtually proves the theory.
            4. Buy or borrow an automotive tail light. My favorite is the 1157 dual filament bulb. With the power supply / inverter disconnected from the signal board, temporarily hook it across C819. Check the voltage across the main filter cap.
            PlainBill
            • docksteaderluke
              • Jul 2011
              • 39
              Hi PlainBill,
              Here's what I've got:
              1. With the signal card disconnected from the power supply I get 160.4-160.7V. Seems low.
              2. Shorting the thermistor didn't do anything from what I could tell. Not sure if this was the desired effect or not...
              3. 100uF, 63V is the biggest cap I have in my collection at the moment... so I didn't try this step.
              4. I don't have an automotive tail light handy so I didn't try this step either.
              • alexanna
                • Mar 2010
                • 1346
                Wile you are wating for some more sugestions.Try reflowing solder in the red and blue circled areas.
                • docksteaderluke
                  • Jul 2011
                  • 39
                  Forgot to mention. In #1 I couldn't detect any noticeable heat sources...
                  • PlainBill
                    • Feb 2009
                    • 7034
                    Hi PlainBill,
                    Measure the voltage at the outlet (or the end of the power cord). Multiply it by 1.4. Subtract 1 from that number. 160 VDC corresponds almost exactly to a line voltage of 115 volts, well within the normal range.
                    One other test: Remove L802. This will disable the inverter. Test the monitor. Obviously you won't have backlights, but if the power LED behaves normally that will provide a small bit of information.
                    Unfortunately, if alexanna's suggestions don't help, I don't see many alternatives to either providing an artificial load, measuring the capacitance of the large cap, or replacing the large capacitor.
                    PlainBill
                    • docksteaderluke
                      • Jul 2011
                      • 39
                      Thanks for the help guys... here's what I've got:
                      165.6V is the calculation I get.
                      When I resoldered those points the voltage went up to 161.5V(max).
                      Removing L802 didn't change anything. Still blinks.
                      I found a cap on another monitor that I have (whose backlight or inverter is gone) that is 100uF 450V. I'm not 100% sure if the cap is in perfect working order but I'm gonna try wiring it in parallel with the big cap after I re-solder L802 back on.
                      • docksteaderluke
                        • Jul 2011
                        • 39
                        Decided to wait on soldering L802 back on. Soldered the second cap in parallel with the first and no difference (on and off in 2-3 second intervals). Still measuring about 160.5V across them.
                        With the second cap wired in I could hear a slight charging then tick (discharge) of the caps.
                        I'm trying to think of something else I can use as a load across c819. What kind of resistance are we talking about here (for the load)? I've got tons of resistors lying around that might do the trick.
                        • PlainBill
                          • Feb 2009
                          • 7034
                          One idea. Obviously you have the signal card connected. Do you also have the cable from the signal card to the LCD panel connected? SOMETHING is overloading the power supply or a power rail. I've been assuming (I know, I KNOW!!) it's the inverter, or a bad filter cap. If you haven't already done so, try this with the LCD panel disconnected.
                          Do you have a pair of resistors in roughly a 3:2 ratio? For example, a 3.3K and a 2.2K would be ideal.
                          PlainBill
                          • docksteaderluke
                            • Jul 2011
                            • 39
                            When you say to try it with the LCD panel disconnected do you mean the VGA cable or the cables going from the inverter side of the power supply to the lcd?
                            Yes I have a 3.3k and a 2.2k right here.
                            • alexanna
                              • Mar 2010
                              • 1346
                              Disconnect the LDVS cable from the logic board to the LCD panel
                              • PlainBill
                                • Feb 2009
                                • 7034
                                Yeah, what Alexanna said.
                                PlainBill
                                • docksteaderluke
                                  • Jul 2011
                                  • 39
                                  Ok... gotcha. Did it and still no change.
                                  • PlainBill
                                    • Feb 2009
                                    • 7034
                                    I'm getting very close to recommending a total recap, but let's try this first.
                                    Work from the back of the power supply / inverter.
                                    1. Take the two resistors and twist one lead of each together, but staggered so one lead extends beyond the other.
                                    2. On the power supply / inverter lay a strip of tape - electrical, duct, or surgical between the pins of CN801 - the output connector - and the edge of the board.. You want to avoid accidental shorts between the leads of the resistors and the ground bus.
                                    3. Solder the two twisted leads of the resistor together.
                                    4. Solder the longer lead of the twist to pin 5 of CN801.
                                    5. Solder the free end of the 3.3K resistor to pin 3 or 4.
                                    6. Solder the free end of the 2.2 K resistor to pin 1 or 2.
                                    7. Check your work; make sure you don't have any shorts.
                                    8. Lay the power supply / inverter on an insulating surface - a piece of dry cardboard is fine.
                                    9. Plug the 4 CCFL leads into their sockets.
                                    10. Take a deep breath, cross your fingers, and plug the power cord into the power supply.
                                    At this point the CCFLs should come on. If they are flashing at about the same rate as when the signal card was plugged in, the problem is the power supply / inverter card. If they are on steadily, the problem is the signal card.
                                    Disconnect the power cord, unsolder the resistor network, and report the results.
                                    PlainBill
                                    • docksteaderluke
                                      • Jul 2011
                                      • 39
                                      Should the signal card be plugged back in for this?
                                      I plugged it back in. Soldered up the resistor network and plugged both a VGA cable in and the power cable in.
                                      What I got was a white screen for about a second, then the LG logo for a second or so and then off for a second or two. Repeat.
                                      If I didn't have the signal card connected I couldn't see anything on the screen... it was just black and with no backlight (or power LED) either.
                                      • PlainBill
                                        • Feb 2009
                                        • 7034
                                        No, I didn't want the signal board hooked up. The resistor network was to turn the inverter on. Perhaps it would be a good idea if I also had you feed a voltage into the brightness input - pin 6 of CN801. The idea was to see if the inverter would work properly if it were fed appropriate voltages without the influence of the signal card.
                                        Make up a voltage divider similar to the one driving pin 5. But this one should have a 3.9K resistor going to 5V and a 1K going to GND. The common point would go to pin 6 of CN801. And while power is applied, measure the voltages at pins 5 and 6, if you would.
                                        PlainBill
                                        • taslan
                                          Surface 7 Pro wont boot, shows Logo and immediately turns off
                                          by taslan
                                          as stated in the title, my surface 7 pro wont boot anymore.
                                          When i press the power button it shows the square windows logo for a second and immediately turtns off again.
                                          I dont even have a chance to get into bios or boot from usb.
                                          While trying to find help online it was suggested to look for help/repair here on badcaps and it might be a ME Region corruption in bios.
                                          I dont know what that is exactly, but i never tinkered with the bios or windows in terms of hacking/patching etc.
                                          Bought it directly from microsoft and used it now and then and only...
                                          01-16-2025, 01:25 AM
                                        • Super360
                                          Vizio E701i-A3 Vizio logo flashing then turns off
                                          by Super360
                                          Ok so here's the story. I was watching tv one day when all of a sudden I heard a pop from the back of the tv set and rushed to unplug the set. By this time the tv already had smoke coming out the back. It was not storming or inclement weather that day to cause a power surge to the best of my abilities. When I opened the tv up, I found that the three capacitors on the psu (highlighted in red) were all bulging and looked they had exploded. I replaced all of them right away and when I went to start up the tv after the "fix" it started just showing the flashing vizio logo as soon as power...
                                          07-25-2024, 07:06 PM
                                        • Teban1306
                                          [REQUEST] ThinkPad T14 Gen 1 (20S1) - Custom Boot Logo Modification
                                          by Teban1306
                                          Hello everyone!
                                          I hope this is the right section for this request. I'm looking for help to modify the boot logo on my ThinkPad T14 Gen 1.
                                          System Information:
                                          • Model: Lenovo ThinkPad T14 Gen 1
                                          • Machine Type: 20S1S4E600 (Intel i7 variant)
                                          • Current BIOS: N2XET42W (1.33) - Latest version
                                          • BIOS File: n2xuj25w.exe (12.98 MB) - Downloaded from Lenovo support
                                          What I'm trying to achieve:
                                          Replace the default "LENOVO" boot logo with a custom image during POST.
                                          🔧 What I've tried:
                                          09-14-2025, 11:49 AM
                                        • USmechS
                                          Hisense 65H6570F Logo shows then turns off
                                          by USmechS
                                          Hello everybody,
                                          I had this Hisense TV since November 2019 (Bestbuy Blackfriday special). Couple of nights ago my dad turned it off normally. In the morning it does turn on, shows the Hisense logo, then the LED light flashes and it turns off. It remains in the same state.
                                          Things I have done:
                                          A. Plug it off and on.
                                          B. Plug it off, pushing the power button for up to 2 minutes, plug it in.
                                          C. Contacted the customer service, They sent me a firmware update, with the file name:
                                          upgrade_loader_no_tvcertificate.pkg
                                          The zip file has this...
                                          04-25-2022, 03:06 PM
                                        • Stephen
                                          MacBook Pro 14" M1 Pro Turns on functions then turns off when I plug in USB C cable
                                          by Stephen
                                          MacBook Pro 14" M1 Pro Turns on functions then turns off when I plug in USB C cable to reset the computer. I disconnect the battery fully and then reconnect it and it won't turn back on. After letting it sit for a day I plug the battery back up, and the MagSafe 3 cable and it powers on and stays ons, once I plug in a USB C cable into the port closes to the MagSafe port it turns off and the Amber light turns off and won't turn on. No water damage, nothing, I want to do a DFU restore to wipe the computer since the customer stated they did an update and ever since it just decided to not work...
                                          03-15-2024, 09:37 AM
