Fluke 187 corroded battery terminals - how to clean?

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • retiredcaps
    Badcaps Legend
    • Apr 2010
    • 9271

    #1

    Fluke 187 corroded battery terminals - how to clean?

    Also posted on eevblog where most of the multimeter people hang out, but I want to get more opinions and suggestions from people here.

    I bought this used Fluke 187 off ebay. Opening up the battery compartment reveals some past battery leakage/corrosion. The supplied batteries are not leaking and are Duracell Procell with expiry date 2017.

    I have read about using baking soda and vinegar to neutralize the corrosion as per

    http://www.eevblog.com/forum/beginne...ked-batteries/

    So I applied the mix using a q-tip, but as you can see, there is still corrosion (the two middle cell holders)? Is it just a matter of applying more of the mix over and over again? Ideally, I would like to "soak" the entire back case overnight, but I'm afraid of damaging or staining the case.

    I don't *think* the main pcb was affected by the corrosion due to the design of the back plastic casing. I can't detect any corrison on the main board. All the rotary functions seem to work correctly.

    I want to make sure there are no long term problems as I expect this multimeter to last a long time. I will be using nimh batteries for this Fluke 187.

    Suggestions are welcome.
    Attached Files

    if you find these attachements useful please consider making a small donation to the site

    --- begin sig file ---

    If you are new to this forum, we can help a lot more if you please post clear focused pictures (max resolution 2000x2000 and 2MB) of your boards using the manage attachments button so they are hosted here. Information and picture clarity compositions should look like this post.

    We respectfully ask that you make some time and effort to read some of the guides available for basic troubleshooting. After you have read through them, then ask clarification questions or report your findings.

    Please do not post inline and offsite as they slow down the loading of pages.

    --- end sig file ---
  • ratdude747
    Black Sheep
    • Nov 2008
    • 17136
    • USA

    #2
    Re: Fluke 187 corroded battery terminals - how to clean?

    The meter should be ok, electrolyte is only conductive as a liquid (not a crystallized salt). As for the batteries, who knows... I'd try the meter with some fresh alakalines before I risked some NiMHs on it.
    sigpic

    (Insert witty quote here)

    Comment

    • t.j.
      Badcaps Veteran
      • Apr 2010
      • 383
      • canada

      #3
      Re: Fluke 187 corroded battery terminals - how to clean?

      emery cloth an a small screw driver, the baking soda vinegar neutralized it, can clean it with the emery an a small screw diver.

      Comment

      • ratdude747
        Black Sheep
        • Nov 2008
        • 17136
        • USA

        #4
        Re: Fluke 187 corroded battery terminals - how to clean?

        Baking soda is for acid batteries. Duracell Procells are alkaline batteries. Baking soda is not of use.

        if it's dried already, it shouldn't be of electrical concern, see my post above.
        sigpic

        (Insert witty quote here)

        Comment

        • 999999999
          Badcaps Veteran
          • Sep 2006
          • 774
          • USA

          #5
          Re: Fluke 187 corroded battery terminals - how to clean?

          Both of those battery contact strips are removable. There may be a little tab that needs depressed or rebent to provide tension again when you reinstall them. Pull them off, agitate the plastic casing in hot detergent solution and briefly scrub with a toothbrush, rinse, shake remaining water out, and air dry.

          Probably was alkaline batteries that leaked, you could try applying a mild acid to the contact strips then polish the metal with a toothbrush and metal polish or use fine sandpaper. As a last step if you happen to have any, you could very lightly tin the contacts with silver solder, though if you apply too much they won't fit back in the casing so in the worst case just heat the contacts and fling excess solder off while molten. Retinning the contacts isn't really necessary but it may be a longer lasting fix that way as the bare metal can oxidize again.

          Depending on the contact size and perhaps your ability to whittle, you might be able to get replacement contacts to fit. Either contact Fluke to see if the parts are available or places like Digikey have a few random styles though I don't know which are most compatible.

          http://www.digikey.ca/product-detail...209K-ND/299941
          http://www.digikey.ca/product-detail...210K-ND/151586
          http://www.digikey.ca/product-detail...209K-ND/151583
          http://www.digikey.ca/product-detail...90K-ND/2137886

          Comment

          • budm
            Badcaps Legend
            • Feb 2010
            • 40746
            • USA

            #6
            Re: Fluke 187 corroded battery terminals - how to clean?

            I use the car batteries terminal corrosion protection that you spray on the car batteries terminal, I just spray it on q-tip and rub on the batteries area, so far so good on most of my equipment that are damaged by leaking batteries.
            Never stop learning
            Basic LCD TV and Monitor troubleshooting guides.
            http://www.badcaps.net/forum/showthr...956#post305956

            Voltage Regulator (LDO) testing:
            http://www.badcaps.net/forum/showthr...999#post300999

            Inverter testing using old CFL:
            http://www.badcaps.net/forum/showthr...er+testing+cfl

            Tear down pictures : Hit the ">" Show Albums and stories" on the left side
            http://s807.photobucket.com/user/budm/library/

            TV Factory reset codes listing:
            http://www.badcaps.net/forum/showthread.php?t=24809

            Comment

            • retiredcaps
              Badcaps Legend
              • Apr 2010
              • 9271

              #7
              Re: Fluke 187 corroded battery terminals - how to clean?

              @ratdude747
              As always, thanks for the comments and hints.

              @t.j.

              Thanks for the suggestions regarding the emery and screwdriver. PS. Where is wonk, AB?

              @999999999

              I didn't know replacement parts existed, but Excavatoree over at eevblog mentioned they were available.

              However, they are $9.56 EACH (ripoff).

              http://www.ebay.com/itm/370609154201

              The seller also doesn't ship outside USA.

              So I will look through digikey for a possible replacement.

              @budm

              Check your Fluke 187.

              Previously, I purchased a Fluke 87IV which is essentially the same as a 187, but the seller completely didn't bother to test it and listed it as working. When I got it, the most basic functions didn't work. I opened it and found that the battery compartment has signs of leakage/corrosion and it got onto the main pcb.

              So I don't know if the 187/189/87IV are prone to corrision or not? Or am I just unlucky? I have purchased other used Fluke multimeters before and never had battery corrosion problems.
              --- begin sig file ---

              If you are new to this forum, we can help a lot more if you please post clear focused pictures (max resolution 2000x2000 and 2MB) of your boards using the manage attachments button so they are hosted here. Information and picture clarity compositions should look like this post.

              We respectfully ask that you make some time and effort to read some of the guides available for basic troubleshooting. After you have read through them, then ask clarification questions or report your findings.

              Please do not post inline and offsite as they slow down the loading of pages.

              --- end sig file ---

              Comment

              • retiredcaps
                Badcaps Legend
                • Apr 2010
                • 9271

                #8
                Re: Fluke 187 corroded battery terminals - how to clean?

                So thanks to all the helpful suggestions and comments, I ended being able to get the contacts out thanks to a-fluke's ebay auction photos. I couldn't see the "hidden" tabs late last night and it has been cloudy all day.

                I was able to get the worst corroded contact out without any damage and let it soak in vinegar for about 1 hour. Then I neutralized it by adding baking soda and let it sit for about 30 minutes. Then I gave it a wash with tap water (I don't have any distilled at home).

                Here is the result. Do you think it is the best I can do given the tools (vinegar, baking soda, toothbrush)?

                I wasn't able to remove the other not so badly corroded contact for fear of damaging it so I left it alone and tried my best to use a toothbrush/q-tip.

                PS. I have a reminder now to check my batteries in all my multimeters every 6 months.
                Attached Files

                if you find these attachements useful please consider making a small donation to the site

                --- begin sig file ---

                If you are new to this forum, we can help a lot more if you please post clear focused pictures (max resolution 2000x2000 and 2MB) of your boards using the manage attachments button so they are hosted here. Information and picture clarity compositions should look like this post.

                We respectfully ask that you make some time and effort to read some of the guides available for basic troubleshooting. After you have read through them, then ask clarification questions or report your findings.

                Please do not post inline and offsite as they slow down the loading of pages.

                --- end sig file ---

                Comment

                • Evil Lurker
                  Warranty Voider
                  • Feb 2011
                  • 454

                  #9
                  Re: Fluke 187 corroded battery terminals - how to clean?

                  I would definitely hit them with some emery cloth or fine grit sandpaper or something to brighten them up a bit.

                  Comment

                  • smason
                    Badcaps Legend
                    • Feb 2010
                    • 1652
                    • Canada

                    #10
                    Re: Fluke 187 corroded battery terminals - how to clean?

                    Personally I always use mechanical means (scrape with a blade, then fine-grit sandpaper) rather than chemicals.
                    36 Monitors, 3 TVs, 4 Laptops, 1 motherboard, 1 Printer, 1 iMac, 2 hard drive docks and one IP Phone repaired so far....

                    Comment

                    • budm
                      Badcaps Legend
                      • Feb 2010
                      • 40746
                      • USA

                      #11
                      Re: Fluke 187 corroded battery terminals - how to clean?

                      That is pretty good clean up jobs, may be you can coat it with Silicone grease.
                      Never stop learning
                      Basic LCD TV and Monitor troubleshooting guides.
                      http://www.badcaps.net/forum/showthr...956#post305956

                      Voltage Regulator (LDO) testing:
                      http://www.badcaps.net/forum/showthr...999#post300999

                      Inverter testing using old CFL:
                      http://www.badcaps.net/forum/showthr...er+testing+cfl

                      Tear down pictures : Hit the ">" Show Albums and stories" on the left side
                      http://s807.photobucket.com/user/budm/library/

                      TV Factory reset codes listing:
                      http://www.badcaps.net/forum/showthread.php?t=24809

                      Comment

                      • 999999999
                        Badcaps Veteran
                        • Sep 2006
                        • 774
                        • USA

                        #12
                        Re: Fluke 187 corroded battery terminals - how to clean?

                        They are clean enough except that I would use mechanical means to further clean the area of contact with the batteries like very (VERY) fine grit sandpaper then a metal polish. Same goes for the contact(s) still in the meter housing.

                        Comment

                        • Kiriakos GR
                          Banned
                          • May 2012
                          • 940
                          • Greece

                          #13
                          Re: Fluke 187 corroded battery terminals - how to clean?

                          Originally posted by ratdude747
                          As for the batteries, who knows... I'd try the meter with some fresh alkaline before I risked some NiMHs on it.
                          Recently I have updated my research over NiMH, due my experiences with many different brands and models of NiMH, I love digital photography and so I am big spender about AA NiMH.

                          Half of my story is here. http://www.ittsb.eu/mini_articles.html#SanyoEneloop

                          What I consider as good news, and it would also be one direct suggestion to retiredcaps, would be to try out those latest Sanyo Eneloop HR-3UTG-4BP,
                          that are capable to keep their 75% of charge even for three years.

                          Personally I have just receive 8xAA & 8xAAA so to use them exclusively in expensive multimeters.

                          With those batteries, there is no chance to have a corroded battery contact ever.
                          This new generation of eneloop it will kill the market share of alkaline in just few months from now.

                          This is my wild guess.

                          Comment

                          • budm
                            Badcaps Legend
                            • Feb 2010
                            • 40746
                            • USA

                            #14
                            Re: Fluke 187 corroded battery terminals - how to clean?

                            I also use Eneloop in my Digital camera which is great, but when use them in my 187, since 4 of them is = 4.8V full charge, the batteries warning keep coming up.
                            Never stop learning
                            Basic LCD TV and Monitor troubleshooting guides.
                            http://www.badcaps.net/forum/showthr...956#post305956

                            Voltage Regulator (LDO) testing:
                            http://www.badcaps.net/forum/showthr...999#post300999

                            Inverter testing using old CFL:
                            http://www.badcaps.net/forum/showthr...er+testing+cfl

                            Tear down pictures : Hit the ">" Show Albums and stories" on the left side
                            http://s807.photobucket.com/user/budm/library/

                            TV Factory reset codes listing:
                            http://www.badcaps.net/forum/showthread.php?t=24809

                            Comment

                            • Krankshaft
                              Badcaps Legend
                              • Jan 2007
                              • 2328
                              • USA

                              #15
                              Re: Fluke 187 corroded battery terminals - how to clean?

                              I have a bunch of Eneloops they're great batteries and have a real low self discharge rate.

                              Unlike the other brands that will self discharge in a few weeks if you don't constantly recharge them.

                              They have a 6 hour charge time but the slower charge rate is gentler on the batteries. As a result they can go through more charge / discharge cycles than the quick charge batteries. Stay away from the 15 minute charge batteries the rapid charging destroys them pretty quickly.

                              If you need a noisy fan to keep the cells cool during charging they're not going to last very long .
                              Last edited by Krankshaft; 06-14-2012, 03:29 PM.
                              Elements of the past and the future combining to make something not quite as good as either.

                              Comment

                              • retiredcaps
                                Badcaps Legend
                                • Apr 2010
                                • 9271

                                #16
                                Re: Fluke 187 corroded battery terminals - how to clean?

                                Originally posted by budm
                                I also use Eneloop in my Digital camera which is great, but when use them in my 187, since 4 of them is = 4.8V full charge, the batteries warning keep coming up.
                                I just put in some eneloops (taken from my Canon A80) into the 187 and they work fine with no battery warning. All 4 eneloops measured 1.2655V DC. When they are fully charged the eneloops can read as high as 1.501 fresh off the charger.
                                --- begin sig file ---

                                If you are new to this forum, we can help a lot more if you please post clear focused pictures (max resolution 2000x2000 and 2MB) of your boards using the manage attachments button so they are hosted here. Information and picture clarity compositions should look like this post.

                                We respectfully ask that you make some time and effort to read some of the guides available for basic troubleshooting. After you have read through them, then ask clarification questions or report your findings.

                                Please do not post inline and offsite as they slow down the loading of pages.

                                --- end sig file ---

                                Comment

                                • retiredcaps
                                  Badcaps Legend
                                  • Apr 2010
                                  • 9271

                                  #17
                                  Re: Fluke 187 corroded battery terminals - how to clean?

                                  Originally posted by Kiriakos GR
                                  Yes, I saw that when your site came alive. Congrats on the new site.

                                  I have been a very long time nimh battery user.
                                  --- begin sig file ---

                                  If you are new to this forum, we can help a lot more if you please post clear focused pictures (max resolution 2000x2000 and 2MB) of your boards using the manage attachments button so they are hosted here. Information and picture clarity compositions should look like this post.

                                  We respectfully ask that you make some time and effort to read some of the guides available for basic troubleshooting. After you have read through them, then ask clarification questions or report your findings.

                                  Please do not post inline and offsite as they slow down the loading of pages.

                                  --- end sig file ---

                                  Comment

                                  • budm
                                    Badcaps Legend
                                    • Feb 2010
                                    • 40746
                                    • USA

                                    #18
                                    Re: Fluke 187 corroded battery terminals - how to clean?

                                    I wonder what is the set point for the batteries warning point is, and how accurate the meter will be with low batteries.
                                    Never stop learning
                                    Basic LCD TV and Monitor troubleshooting guides.
                                    http://www.badcaps.net/forum/showthr...956#post305956

                                    Voltage Regulator (LDO) testing:
                                    http://www.badcaps.net/forum/showthr...999#post300999

                                    Inverter testing using old CFL:
                                    http://www.badcaps.net/forum/showthr...er+testing+cfl

                                    Tear down pictures : Hit the ">" Show Albums and stories" on the left side
                                    http://s807.photobucket.com/user/budm/library/

                                    TV Factory reset codes listing:
                                    http://www.badcaps.net/forum/showthread.php?t=24809

                                    Comment

                                    • retiredcaps
                                      Badcaps Legend
                                      • Apr 2010
                                      • 9271

                                      #19
                                      Re: Fluke 187 corroded battery terminals - how to clean?

                                      Dave Jones, Kiriakos GR, and Martin Lorton do a lot of multimeter testing and one of their "standard" tests is to connect a variable DC power supply to the multimeter. The multimeter is then set to measure some precision component (resistor or some DC voltage). They turn down the power supply until the low battery warning comes on and they see if the multimeter is still within spec or not.

                                      I'm fairly sure all the modern Flukes will display "batt" and won't let you take a reading well before they are out of spec.

                                      I just got a Fluke 27 and it came with a weak battery. It displayed the low battery indicator and I did a simple shorting out of the probes. The readings wouldn't settle and fluctuated. I thought I got a dud Fluke 27.

                                      However, I put in a fresh battery and got 0.2 ohms when shorting out the probes. So for the Fluke 27, the multimeter allows you to take readings with a low battery indicator with possible erroneous results.

                                      I think the "batt" is definitely the way to go. You don't want your multimeter showing 0V DC only to find out the big filter capacitor is still charged at 400V DC.

                                      A lot of people don't appreciate this small nuance.

                                      Martin's multimeter reviews
                                      http://www.youtube.com/playlist?list...6&feature=plcp

                                      Dave Jones
                                      http://www.eevblog.com/episodes/

                                      Kiriakos
                                      http://www.ittsb.eu/index.html

                                      PS. I didn't realize how BIG the Fluke 27 is. I have always seen pictures and knew it was made for the US government and military, but until I got it in my hands, I had no idea how hefty it is.
                                      --- begin sig file ---

                                      If you are new to this forum, we can help a lot more if you please post clear focused pictures (max resolution 2000x2000 and 2MB) of your boards using the manage attachments button so they are hosted here. Information and picture clarity compositions should look like this post.

                                      We respectfully ask that you make some time and effort to read some of the guides available for basic troubleshooting. After you have read through them, then ask clarification questions or report your findings.

                                      Please do not post inline and offsite as they slow down the loading of pages.

                                      --- end sig file ---

                                      Comment

                                      • retiredcaps
                                        Badcaps Legend
                                        • Apr 2010
                                        • 9271

                                        #20
                                        Re: Fluke 187 corroded battery terminals - how to clean?

                                        Originally posted by budm
                                        I wonder what is the set point for the batteries warning point is, and how accurate the meter will be with low batteries.
                                        If you have time, hookup the Fluke 187 to a variable DC power supply.
                                        --- begin sig file ---

                                        If you are new to this forum, we can help a lot more if you please post clear focused pictures (max resolution 2000x2000 and 2MB) of your boards using the manage attachments button so they are hosted here. Information and picture clarity compositions should look like this post.

                                        We respectfully ask that you make some time and effort to read some of the guides available for basic troubleshooting. After you have read through them, then ask clarification questions or report your findings.

                                        Please do not post inline and offsite as they slow down the loading of pages.

                                        --- end sig file ---

                                        Comment

                                        Related Topics

                                        Collapse

                                        • Bowline
                                          Dell E6400 PT434 type Battery Charging Stuck at 1%. Battery Pinouts?
                                          by Bowline
                                          My question pertains to understanding the control circuity inside a Dell PT434 type battery used in Dell E6400, E6410 and other Dell notebooks?

                                          Does anyone have a circuit diagram for the inside of a Dell PT434 battery and/or a description of how the control lines work to control this battery?

                                          My reason for asking is I have a battery which I cannot get to charge beyond 1% in a Dell E6410 or E6400.
                                          The LED bar graph charge indicator on the battery shows one LED lighted when the info button is pressed.
                                          I have tried multiple different Dell power adapters. My Dell...
                                          06-11-2025, 10:13 AM
                                        • mkdj
                                          ASUS X571 no power : battery issue?
                                          by mkdj
                                          Hello everyone,

                                          Mark and model: ASUS X571
                                          Symptoms: not turning on. There is only the LED battery indicator working. The owner of the laptop told me this problem arrived one week after having changed the battery. He then managed to power on the laptop only once, and then it never powered on again.

                                          Tests already made:
                                          - Battery is charging when laptop is plugged in. The LED battery indicator turns white after a couple of hours of charge, indicating that battery is fully charged. However, laptop won't turn on anyway, with or without AC plug.
                                          - I have removed...
                                          07-14-2025, 09:16 AM
                                        • techfreak9356
                                          Vostro 3400 (LA-K033P) won't charge battery, everything else ok
                                          by techfreak9356
                                          I bought a Vostro 3400 motherboard (LA-K033P) to upgrade over my damaged Inspiron 5593 (LA-J091P) board. Everything else is ok except for the battery. This battery I am using now is already a replacement one since I disposed of the bloated original one. Back in the Inspiron 5593 motherboard, the battery charges fine and is identified just fine. However, with the upgraded motherboard, suddenly the replacement battery is labeled as "Not a dell battery". When I borrow a similar battery from my colleague (which did not have replacements yet), it is identified and charges just fine.
                                          ...
                                          09-14-2023, 11:10 PM
                                        • Document Archive
                                          ISL88739AHRZ ISL88739 Laptop Battery Charger Datasheet
                                          by Document Archive
                                          The ISL88739 & ISL88739AHRZ are a highly versatile combo battery charger
                                          configurable for operating as either a Hybrid Power Boost (HPB)
                                          charger or a Narrow VDC (NVDC) charger, supporting 2-, 3-, or
                                          4-cell batteries. Both configurations allow the battery to work with
                                          the adapter together to supply the system load when it exceeds
                                          the adapter capability, referred to as system Turbo mode. HPB
                                          charger configuration reverse-boosts battery energy to the system
                                          bus to help the adapter provide the system power in Turbo mode.
                                          NVDC charger configuration...
                                          11-05-2024, 03:28 PM
                                        • ferah831208
                                          Strange battery present behaviour in Lenovo Legion 5-15ARH05 mb model nm-d041
                                          by ferah831208
                                          Hello guys, i need help with this problem, I have a NM-D041 mobo, it came with low impedance in 3v linear regulator rail, the cause was a damaged EC controller ITE8227e-192. I ordered a new one and replaced it. This laptop have a different chip for keyboard management and it8227e pins for programming ITE chips are not accesible through keyboard interface, I flashed it using i2c interface pins with vertyanov programmer because laptop was dead and SPI flash didn't have power (only 3v EC) so it couldn't access firmware stored in SPI memory. I read SPI and extracted it8227 firmware from original BIOS...
                                          03-18-2025, 01:18 AM