Replacement caps for Tektronix 434 Power supply board.

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • John McGivern
    Senior Member
    • Dec 2013
    • 195
    • uk

    #81
    Re: Replacement caps for Tektronix 434 Power supply board.

    I used 2x 200meg in series then 100k fitted inside Dmm plastic probe. Soldered 200meg to probe and gnd and neutral to end of 100k. Then red wire to junction of 2nd 200meg and 100k. Ptfe around all solder joints then shrink wrap then put all inside plastic tube then all inside another plastic tube about 10" long.
    Resistors rated for 10kv measuring -2000v.
    Last edited by John McGivern; 02-23-2017, 01:46 AM.

    Comment

    • John McGivern
      Senior Member
      • Dec 2013
      • 195
      • uk

      #82
      Re: Replacement caps for Tektronix 434 Power supply board.

      Red wire what is aquadag my multimeter is 8.4meg.
      Measured old 9v battery
      Dmm = 6.78v
      HV probe setup = 1.6mV
      Ratio = 4237.5:1
      Double checked same result.

      Comment

      • Longbow
        Badcaps Veteran
        • Jun 2011
        • 623
        • USA

        #83
        Re: Replacement caps for Tektronix 434 Power supply board.

        Interesting way of figuring out the division ratio. I calculate 4000:1, so on this tube measurement, 4KV would read 1 volt. Be sure the ground is connected solidly to the chassis and to your meter (-) terminal.

        Aquadag is a term used for the grey conductive coating near the front inside end of the tube. Essentially that coating is the "plate" element. TV sets put the +30KV on the aquadag w/r to ground. More coating was sprayed on the outside of the tube to produce a swell HV filter capacitor. This scope has the aquadag at ground, and the cathode at -4KV, so you don't have to go probing around in the usual place. There should be a HV test point near the tube socket. Naturally you will want to try out your probe AFTER replacing the faulty diodes.

        p.s. I'd like to see your finished probe!
        Last edited by Longbow; 02-23-2017, 09:35 AM.
        Is it plugged in?

        Comment

        • srhofmann
          Senior Member
          • Aug 2016
          • 135
          • usa

          #84
          Re: Replacement caps for Tektronix 434 Power supply board.

          Just a quick comment. If you're looking to build your own HV probe look for an old resistor from Beman. There nicely built, my Polaris 651 probe has an 800Mohm in it. Found this really quick using google.

          http://www.electronicsurplus.com/bem...ic-1090meg-ohm

          P.S. Great thread, it's reminding me of everything I've forgotten.....

          Comment

          • John McGivern
            Senior Member
            • Dec 2013
            • 195
            • uk

            #85
            Re: Replacement caps for Tektronix 434 Power supply board.

            Srhofmann glad your going down memory lane as you get older you realise you forgot more than you know. I have to try to learn just some of what you forgot ha ha.
            Longbow will post photo of finished result.

            Comment

            • John McGivern
              Senior Member
              • Dec 2013
              • 195
              • uk

              #86
              Re: Replacement caps for Tektronix 434 Power supply board.

              Tried HV probe today on my Hameg which has a fault doh. Anyway -1900v read -1892
              -1877 read -1868 and -1962 read -1960 seems ok.

              Comment

              • Longbow
                Badcaps Veteran
                • Jun 2011
                • 623
                • USA

                #87
                Re: Replacement caps for Tektronix 434 Power supply board.

                On a non-storage tube, 2KV is typical. Now quit playing with that probe and fix your 434.
                Last edited by Longbow; 02-24-2017, 08:41 AM.
                Is it plugged in?

                Comment

                • John McGivern
                  Senior Member
                  • Dec 2013
                  • 195
                  • uk

                  #88
                  Re: Replacement caps for Tektronix 434 Power supply board.

                  I am still waiting for the components ESR said there was a delay with pay pal but my account shows payment sent last Saturday very annoying phoned today again told in post said the same on Wednesday when I phoned.

                  Comment

                  • John McGivern
                    Senior Member
                    • Dec 2013
                    • 195
                    • uk

                    #89
                    Re: Replacement caps for Tektronix 434 Power supply board.

                    Hi Gents and Mr Longbow, parts arrived today fitted them tonight.
                    See photo's. Zoom in on photo's.
                    The trace does not cover the 10x graticules almost 1/2 a graticule short.
                    What is the bright green dot at the start of the trace never noticed it before.
                    The signal shown is with a 1x probe its the scope calibration signal 0.6v @ 1Khz.
                    See settings on scope.
                    Attached Files

                    Comment

                    • redwire
                      Badcaps Legend
                      • Dec 2010
                      • 3900
                      • Canada

                      #90
                      Re: Replacement caps for Tektronix 434 Power supply board.

                      I see insufficient horizontal sweep in those pics. It looks like it should be 11 graticules wide and you are getting 10.
                      The bright spot at the beginning of the trace is normal per service manual 3.L,M.

                      Can you comment on the action of the horizontal position control <->
                      I notice it's a bit shifted. Does it move the trace further to each side?
                      It is probably the H-sweep a bit out of cal but it could be the H-oscillator, at this point.
                      Can you check the timebase, if it is accurate or compressed 10/11's.

                      The H-amplifier uses +250VDC and the +/-15VDC rails (pg. 199)
                      I don't know the history of your adjustments but I would make sure the rails are correct and then consider doing a CAL on the H-sweep.
                      Adjusting the +15V rail affects the entire scope calibration! It has to be bang on per A.1 and changing it "the oscilloscope will require complete calibration". I'd leave it unless you affected that...

                      If there is no problem seen, I'd try calibrate the H-sweep.
                      Center the H-pos control and adjust R703 until things are in the middle, per D.1.f
                      If it is low sweep amplitude, adjust R817 until you get 11 graticules width per D.3.f

                      Comment

                      • John McGivern
                        Senior Member
                        • Dec 2013
                        • 195
                        • uk

                        #91
                        Re: Replacement caps for Tektronix 434 Power supply board.

                        Red wire are you looking at the correct serial number BELOW 500000.
                        I don't see any voltages on p199 but I will check them from my PDF.
                        Many thanks.

                        Comment

                        • srhofmann
                          Senior Member
                          • Aug 2016
                          • 135
                          • usa

                          #92
                          Re: Replacement caps for Tektronix 434 Power supply board.

                          Redwire is right it does look like horizontal sweep is compressed. Did you make sure the the var knob on the time base selector is locked? It looks like the signal is only 840-850 uSec. Should be a lot closer to 1,000 uS, 5 full divisions wide. The built in calibrators are usually pretty damn close 1Khz+-33. Might check it with a frequency counter if you've got one or use a known external source. Step down transformer with a diode/bridge to measure the ripple frequency.

                          If you've got a "find the trigger" button you might give it a couple of pushes, the one on my Tek scope occasionally acts up.

                          Comment

                          • John McGivern
                            Senior Member
                            • Dec 2013
                            • 195
                            • uk

                            #93
                            Re: Replacement caps for Tektronix 434 Power supply board.

                            Redwire see photo's after alterations you recommended.
                            Longbow see photo diy HV probe (worked out ok)
                            Many thanks to everyone any other advice is welcome.
                            Special thanks to Longbow for staying the distance and helping me your a gem mate cheers.
                            Attached Files

                            Comment

                            • John McGivern
                              Senior Member
                              • Dec 2013
                              • 195
                              • uk

                              #94
                              Re: Replacement caps for Tektronix 434 Power supply board.

                              Can someone check frequency from these photo's
                              10x probe see settings on scope.
                              0.6v 1khz calibration signal.

                              Many thanks.
                              Attached Files

                              Comment

                              • srhofmann
                                Senior Member
                                • Aug 2016
                                • 135
                                • usa

                                #95
                                Re: Replacement caps for Tektronix 434 Power supply board.

                                It looks close to 1 KHz. Take a look at something that you're pretty sure is pretty accurate. Got an old analog TV take a look at the colorburst crystal. It should be around 3.58 MHz for NTSC. Got an old router pop the top and look at the clock in it.

                                Don't forget 90+ percent of the time you're just looking for a signal with a scope. Is it there? Does it look like it should? Is it about the right amplitude and frequency?

                                Good job getting the scope working again.
                                Last edited by srhofmann; 03-02-2017, 09:26 AM.

                                Comment

                                • John McGivern
                                  Senior Member
                                  • Dec 2013
                                  • 195
                                  • uk

                                  #96
                                  Re: Replacement caps for Tektronix 434 Power supply board.

                                  Srhofmann thanks for replying I have a HP 3325B injected 1khz signal very accurate so scope should show 1khz but it looks wrong to me I thought there should be ten signals for each gradient or am I wrong.

                                  Comment

                                  • srhofmann
                                    Senior Member
                                    • Aug 2016
                                    • 135
                                    • usa

                                    #97
                                    Re: Replacement caps for Tektronix 434 Power supply board.

                                    Originally posted by John McGivern
                                    Srhofmann thanks for replying I have a HP 3325B injected 1khz signal very accurate so scope should show 1khz but it looks wrong to me I thought there should be ten signals for each gradient or am I wrong.
                                    Good catch, just proves I shouldn't do basic math before I've had an entire pot of coffee. Approximately 2 divisions for a 1Khz signal at .5 milliseconds per division. At 50uS or .05 milliseconds per division you shouldn't see a pulse train only a transition edge. It's like you're off by a factor of 10.

                                    1. Does the output look the same if you use the internal calibrator?
                                    2. The time/div knob can be pushed in or out and it controls a zoom feature. I don't remember exactly how this works you might check this feature out.

                                    Here's a picture of the internal calibrator of my scope at .5 mSec per division
                                    Attached Files

                                    Comment

                                    • John McGivern
                                      Senior Member
                                      • Dec 2013
                                      • 195
                                      • uk

                                      #98
                                      Re: Replacement caps for Tektronix 434 Power supply board.

                                      Yes it's the same with the internal calibrator.
                                      I think you are right about the factor of 10 but what would cause this is it possible for the Timebase dial to slip out of cal by a factor of 10 if yes how do you correct this???

                                      Comment

                                      • srhofmann
                                        Senior Member
                                        • Aug 2016
                                        • 135
                                        • usa

                                        #99
                                        Re: Replacement caps for Tektronix 434 Power supply board.

                                        If you've got the Timebase switch in the normal operating mode my guess is that something mechanical has failed in the rotary switch itself. Look in the last few pages of the schematic and you can see that the timebase switch actually has around 20 cams that open and close individual switches to change the timebase by switching different components in and out of the circuit. I've never opened one up and never had to troubleshoot one so I have no idea where to even begin.

                                        Have you tried looking at different frequencies from the generator. Maybe it's just a failure in this particular mode. Just occurred to me, is it possible that the knob isn't aligned correctly and it's rotated from where it really should be? Or could you just rotate the knob so it's correct for what's on the screen?
                                        Last edited by srhofmann; 03-02-2017, 06:17 PM.

                                        Comment

                                        • John McGivern
                                          Senior Member
                                          • Dec 2013
                                          • 195
                                          • uk

                                          #100
                                          Re: Replacement caps for Tektronix 434 Power supply board.

                                          Same on other frequencies knob hasn't moved may try opening up.

                                          Comment

                                          Related Topics

                                          Collapse

                                          • Tynan Dill
                                            Vizio e601i-A3 - Has Sound and Display, But No Backlight - Bad Power Supply Board or Bad LED Bulbs ?
                                            by Tynan Dill
                                            I was given this TV from my great uncle. He said it just wouldn't turn on one day out of nowhere, replaced the TV, and gave it to me to possibly fix and use for myself.

                                            Upon bringing it home and plugging it up, it showed a standby light.

                                            I powered it on and without a flashlight, the display showed the "V" but the lighting is very dim, but visible.

                                            The screen seems to blackout and stay black, but with a flashlight I can see the display.

                                            With my Playstation 4 connected via HDMI, and running a game I can hear sound.

                                            Assuming...
                                            11-22-2024, 01:46 PM
                                          • sam_sam_sam
                                            Panasonic switching power supply daughter board has two 27uf @ 50 volt capacitors which I do not keep in stock
                                            by sam_sam_sam
                                            I did not have any 27uf @ 50 volt capacitors but one nice thing is that it has enough room for two capacitors in parallel a 22uf and a 4.7uf @ 50 volts 22 plus 4.7 equals 26.7uf which so very close to 27uf I will show some pictures of it sometime tomorrow

                                            This is one of the shit est boards I seen in a very long time and it was the daughter board the traces just lifting off the board I had to use the capacitor leads to repair the the traces and the main board is not any better because I had to repair several traces for the daughter board to the main board

                                            After all of...
                                            07-13-2024, 08:14 PM
                                          • Soulzink
                                            Samsug washing machine. Power supply DC41-00189a board issue
                                            by Soulzink
                                            Hi everyone,

                                            I am trying to fix a samsung washinmachine power supply main board.
                                            I already replaced it with another one to fix the machine but i want to try understand the issue with the old board.
                                            Therefore i hope to be able to get some help to explain what to check in order to troubleshoot.
                                            The power supply board reference is the following: DC41-00189a
                                            I have followed one thread (https://www.badcaps.net/forum/troubl...ingmachine-rel...
                                            04-19-2025, 06:26 AM
                                          • captain150
                                            Help with switching power supply caps
                                            by captain150
                                            I'm trying to repair two old VCRs, they both have bad caps. One has leaky ones, the other would barely run until I subbed in some caps from another power supply I had laying around (though they are the wrong values). This vcr works for an hour or two, but then the power supply starts whining and the picture gets lines in it. I didn't replace all the secondary caps, so another voltage might still be problematic, or the values I used are too far off.
                                            I've been on mouser and digikey but the options are a bit overwhelming. I just need some new ones that will work. They don't need to be top quality,...
                                            03-16-2025, 07:34 PM
                                          • momaka
                                            HannStar Hanns-G HC194d LCD monitor repair
                                            by momaka
                                            Normally, I never post repairs this quick after I do them, because… I am usually very slow. But today, I’m making an exception here. Why? No idea. Perhaps only because the repair details are still “fresh” in my head… which is ironic, given this is a 16 year old monitor that hardly anyone will care about today. It is new to me, though. I picked it up last November from someone on my local Craigslist. It wasn’t very close to where I live, but was close to a family friend that I had to go visit anyways. So after watching the posting on Craigslist for a few weeks and seeing it getting...
                                            03-15-2023, 10:17 PM
                                          • Loading...
                                          • No more items.
                                          Working...