Bad DC Offset TL072 Op-amp
  • Syn-fi
    • May 2013
    • 29
    • United Kingdom
    hi,
    I'm new this forum but hope to add my insights and contribute in respect of Samplers and Loewe TVs
    My present problem is related to a TL072 OP-Amp on an Ensoniq DP4+ FX rack. I am getting a 'Bad DC offset' message on boot up and indeed the DC is not zero across the op-amp. This is causing feedback across channel 1 and noise.
    I have recapped the Alu-Cap and also swapped out two SMD 0.1uf caps, but neither of those components seems to be the issue and I wonder if it is a bad Op-Amp.
    I attach a photograph of the circuit board and also the schematic.
    Any pointers would be gratefully received.
    Re: Bad DC Offset TL072 Op-amp
    Since the circuit is direct couple circuit, any DC present at the input of the opamp, it will be multiply by the gain setting of the op amp. How high is the DC offset at the output of the opamp? Is it positive or negative offset voltage?
    Learn about offset voltage:
    https://cdn.badcaps-static.com/pdfs/...edf24d297a.pdf
    http://www.analog.com/en/precision-o...cts/index.html
    • Syn-fi
      • May 2013
      • 29
      • United Kingdom
      Hi, thanks for responding so quickly.
      however, I am struggling with the pin outs of the SOIC package tl072 . I assume where the small circle, is pin 1 and possibly output 1 (ch1). I can only find datasheets for the dip version of the tl072
      separately i have 15v across the op-amp which seems right.
      • Khron
        • Sep 2006
        • 1350
        • Finland
        The pinouts of SOIC's (and MSOP and whatever other 8-pin opamp packages are used) are quite standardized and "mirrored" between each other The silicon die's the same anyway.
        Point being, pin 1 of the DIP is the same as the pin 1 of the SOIC
        Second google result: https://cdn.badcaps-static.com/pdfs/...4612b37d90.pdf
        Page 4 - D package = SOIC, P/N package = PDIP (as you can read in pages 2-3)
        Khron's Cave - Electronics - Audio - Teardowns - Mods - Repairs - Projects - Music - Rants - Shenanigans
        • Syn-fi
          • May 2013
          • 29
          • United Kingdom
          Ok got it, over and out, will be testing this morning.
          One thing I have discovered; is if swap the tl072 over to ch3/4 I get DC on ch3 but also on ch1. Swap them back and it goes back to dc on ch1. That seems to suggests that the opamp could be bad and there could be a problem elsewhere. Unfortunately I cannot test the op amps on a breadboard.
          • Syn-fi
            • May 2013
            • 29
            • United Kingdom
            Ch1 on tl072 seems to be 0.9v
            Ch2 is zero
            • redwire
              • 3912
              In the pic I don't see the TL072's pins 5-8 soldered well, can you touch that up. It's a $0.72 part so instead of swapping them, I would just replace it before the PCB tracks get damaged.
              Did you notice the trimpot labelled offset? Sometimes things drift with age and you just have to recalibrate. It looks like it runs to pins 5-6.
              The DC could be upstream of the op-amp, but I'm not sure what the cct looks like.
              • Syn-fi
                • May 2013
                • 29
                • United Kingdom
                Hi, I have already re-seated the tl072 and also cleaned the pads underneath. There is a trimpot for the offset but it does not make any difference. The troublesome tl072 is also running very hot compared to the others, too hot to touch. Prior to the op-amp there are inverted audio streams (balanced audio) and there some alu caps on the way but i have replaced those.
                There is some further history to this board. It looks like someone, stateside tried to swap out a converter on the output and the blew off a transistor and caused irregularities on the voltage regulators. There was track damage near where the output converter had been removed, i have fixed all of that and I can now get audio from ch3/4 but there is noise on ch1.
                • Khron
                  • Sep 2006
                  • 1350
                  • Finland
                  Running hot? Well, if it's not shorted internally, it might be oscillating. Do you happen to have / can you borrow an oscilloscope?
                  • Syn-fi
                    • May 2013
                    • 29
                    • United Kingdom
                    no oscilloscope at the moment but i'm on the hunt for one. I will buy some TL072 replacements anyway to narrow this down.
                    The fact that the TL072 is running hot suggests the problem is either with the TL072 or before rather than being anything to do with the offset control which occurs later.
                    • Khron
                      • Sep 2006
                      • 1350
                      • Finland
                      Is there some way you can attach / link the schematic where this opamp "fits into"? The snippet in the first post "barely" shows the opamp itself, with not much clue where its inputs are coming from (earlier than those switches).
                      • Syn-fi
                        • May 2013
                        • 29
                        • United Kingdom
                        here you go.
                        Just be reading an article regarding oscillation and this does some to be the likely problem. But i can't see that the 0.1uf ceramic caps are at fault. Could it be the film caps?
                        • Syn-fi
                          • May 2013
                          • 29
                          • United Kingdom
                          the entire schematic is here:
                          • redwire
                            • 3912
                            Thanks for posting the schematic, that makes it much easier.
                            Note the dual PSU voltages involved: +/-5V and +/-15V. Can you measure those, especially after seeing someone else was in there.
                            *If the TL072 outputs (offset) ever goes over +/-5.7V, it will dump current into the 5V rails and get hot (because it runs off +/-15V). I'm saying a good op-amp would heat up then. It might move the 5V rail too.
                            Possible reasons for too much offset would be:
                            Damaged op-amp TL072
                            CMOS switch IC 74HC4053 is no good
                            ADC IC current output1 has too much offset
                            The TL072 (in-) pin2/6 should be 0V, it is a summing point for many signals. If it's even +/- 0.005V the op-amp will have many volts of offset on its output.
                            I hope some of this helps. I'm not sure how far you can troubleshoot with a multimeter. The ADC normally runs warm.
                            • Syn-fi
                              • May 2013
                              • 29
                              • United Kingdom
                              Thanks for your detailed analysis. I have swapped over the hc4053 switches and it doesn't appear that they are at fault. Equally the TL072 seems fine too, when swapped over with the other channels. Re voltages I am going to struggle to test these across the board with a multimeter. But surely if the voltages were off, that would effect the other channels and introduce DC on ch3/4.
                              Could the ADC be sending too much current that isn't compensated for later ? Is it worthwhile flipping out the ADC and swapping it for ch3/4
                              • redwire
                                • 3912
                                The 5V powered stuff (A/D, 4053) can get damaged by the 15V powered stuff (TL072). So you have to check voltages first. If you just swap things, you can make things worse. I would check the five voltage regulators to see that they are in spec. and let us know what you have on all pins of the op-amp.
                                To see if I am following what you've tried: U18 (TL072) runs hot, U22 (4053) was also swapped but distorted audio on Ch.3 is still a problem (I would replace parts instead of swapping them...)
                                That leaves U19 and the A/D in the signal path. I would compare voltages on CIN1 and CIN2 as they should be close. The A/D has a current output to U18 to keep it zeroed, but careful the voltage on the TL072 pin 2 should never go over +/-5.7V or it will damage the A/D. So if you swap the A/D it could get damaged. Always check voltages first.
                                • Syn-fi
                                  • May 2013
                                  • 29
                                  • United Kingdom
                                  VR2 looks to be out
                                  VR7 should be + 15v
                                  In - 29.49v
                                  out 14.82v
                                  VR1
                                  In - 29.49v
                                  out 4.44v
                                  VR2 -should be -15v
                                  In 29.29v
                                  Out -24.30v
                                  VR8
                                  In 29.28v
                                  Out 13.63v
                                  VR6
                                  in - 14.36v
                                  out 9.34v
                                  • redwire
                                    • 3912
                                    I expect different readings from the PSU, I think you used the wrong ground point with your multimeter, or there is a bad ground to the PSU:
                                    Input you measure -29.49V:
                                    VR1 (7805) should be +5.0V; you measure +4.44V (low, 4.85V min.)
                                    VR7 (7815) should be +15V; you measure +14.82V; ok
                                    Input you measure +29.28V:
                                    VR2 (7905) should be -5.0V; you measure -24.3V (oh boy, way too much)
                                    VR8 (7915) should be -15V; you measure +13.63V; ???
                                    Input you measure -14.36V:
                                    VR6 (7905) should be -5.0V; you measure +9.34V; ???
                                    At the very least VR2 is bad, and VR8 might be feeding into VR6. Kinda scary.
                                    • Syn-fi
                                      • May 2013
                                      • 29
                                      • United Kingdom
                                      I'll report back when i have checked the regulators properly. They seem to have two ground points because the heatsink also has a grounding
                                        78xx series Center pin and the Tab are tied together, they are Ground.
                                        79xx center pin and tab is the input, left pin is ground.
                                        • jjne
                                          Bad DC offset on Onkyo TX NR414 - doesn't appear to be in the analogue section
                                          by jjne
                                          I have a TX NR414, well I have two actually which are both exhibiting the same issue.
                                          The one I've had since 2012 developed DC offset of around a volt under load on the front right speaker output. Not enough to trip the protection but enough to be noticeable in terms of sound quality. I checked the preamp board and couldnt see anything obvious. The offset is too high really to be a drift issue right?
                                          I put the amplifier away for a while. I managed to pick up another example with a dead HDMI board very cheaply (common on these - mine was replaced by Onkyo several years...
                                          07-28-2025, 07:51 AM
                                        • dragometin
                                          Hitachi 40HYC42
                                          by dragometin
                                          TV Model: Hitachi 40HYC42
                                          Power Supply: 17IPS71
                                          MainBoard: 17MB95M
                                          Panel: VES400UNDS-2D-NO3
                                          When the set came in it was dead. Fuse was open primary side and the power mosfet was shorted. So I ordered a new power supply. Now it turns on, only displays "Hitachi" and then power off red and blue led flashing.
                                          I had to modify the 0ohm resistors in the led backlight of the new PS like the original power supply.
                                          Is there any variants on 17IPS71 power supply's?
                                          UART output pins 10 and 12 on scart connector:
                                          Code:
                                          12:24:04.595
                                          08-08-2024, 05:53 AM
                                        • mferna14
                                          BIOS for Hp 24-xa0019, offset required
                                          by mferna14
                                          Hi Friends., working on Hp All-in-one Hp 24-xa0019. I got BID from the Original BIOS. & downloaded BIN from HP 084EE. I am bit confused about the serial # & Win Keys offsets. Can some help find the correct block for both. For Win Keys I found 691000 ~ 691030 (Correct me if this is wrong). But for the serial # kindly guide me. Original BIOS uploaded (NO DISPLAY, but beeps when Memory removed). Thank dear
                                          Model # 24-xa0019
                                          Serial # 8CC84208V4
                                          Product # 4NM99AA#ABL
                                          Motherboard # DA0N76MB6E0 Rev: E​...
                                          12-03-2024, 05:46 PM
                                        • mossieurclo
                                          Marantz PM7200 Can't adjust Bias or DC Offset
                                          by mossieurclo
                                          Thanks for reading me:
                                          After i repaired the speaker protection board of that amplifier, all was ok.
                                          So I then proceeded to adjust iddle current, got it, then DC offset. Oh, bboy.
                                          At that stage, on the left channel, after first seeing normal readings while moving the trimpot to get to the correct setting, suddenly readings were all over the place and did not respond to my turning the trimpot.
                                          I played music to earphones, to find that there was no sound on left channel. Only distorted crackles on peaks.
                                          I took out the board...
                                          09-30-2020, 08:11 PM
                                        • PanicMechanic
                                          Extech EX330 multimeter current range offset fault
                                          by PanicMechanic
                                          My multimeter has developed this fault, a few months ago. I tried to fix it, but had no success.
                                          The fault is that on the current ranges it never shows 0. Does not matter if the probes are open or shorted.
                                          As you can see from the pics, the offser is the same (just multiplied for other rages):
                                          -4.9 uA
                                          -0.49 mA
                                          -0.49 A
                                          I don't know what exactlly caused the fault.
                                          The meter had a battery leak before I noticed the fault, but I am not sure if the fault was present before that happened. I just did not noticed. The battery leaked...
                                          05-11-2021, 11:13 PM
