Damaged system of the VU indicator in the JVC KD-A11 tape recorder Request for repair
  • UNITRAX
    • Jan 2023
    • 64
    • POLAND
    JVC KD-A11 Cassette Deck
    Both channels play cassettes. The problem is that the left VU indicator does not work.
    I desoldered and checked the resistor R143.
    Capacitors
    C 141, C142, C143, C144, C902,
    C124, C126.
    Diodes
    DI51, DI52.
    Transistors
    X63,X64,X65,X66
    X101, X102, X103, X104.
    The tester shows that they are functional
    IC901 UPC 4557C
    Measured with an ohmmeter
    between the pins
    4 - 8 charges to 840 Ohms
    2 -3 charges to 180 kOhm
    5 - 6 charges to 2.4 MOhm
    4 - 1 19.2 KOhm
    4 - 7 19.2 KOhm
    When turning on the device, the indicators should stand up for a moment. The right one does it correctly. The left one gets up, but much less than the right one.
    The indicators themselves work fine. The ohmmeter shows the same result in both gauges.
    I don't know what to do next.
  • jiroy
    • Jun 2016
    • 2416
    • Lebanon
    Re: Damaged system of the VU indicator in the JVC KD-A11 tape recorder Request for re
    Capacitors
    C124, C126.
    Diodes
    DI51, DI52.
    Transistors
    X63,X64,X65,X66
    IC901 UPC 4557C
    between the pins
    4 - 1 19.2 KOhm
    4 - 7 19.2 KOhm
    If I were you , I would check the internal resistance of the VU meter itself ,then try to resolder its pins if on board.Try that first and if it didn't work ,repost .
    • UNITRAX
      • Jan 2023
      • 64
      • POLAND
      I did it at the beginning, both VU meters are ok, the resistance result is identical and the tips stand up under the ohmmeter. Note that the resistance at the first output of the upc 4557C opamp is several mega ohms. Does this indicate its damage?
      • CapLeaker
        Leaking Member
        • 8132
        the circuit between the VU meter and the UPC4557 is pretty simple. Since you got two channels (left and right), compare them to each other with an oscilloscope...
        Did you check VR151 if it's open? Check if the audio coming out of both channels are equal in volume. You got to figure out is it only the VU meter (circuit) that isn't working correctly or is one channel low in volume causing the VU meter to read low. Read page 11 and 12.
        • redwire
          • 3906
          See what signal you get out of the headphone jack because that is also used to drive the VU meters. It is the op-amp output.
          I had one of these JVC decks, built so good all metal buttons.
          • budwich
            • Jul 2015
            • 3097
            what does this mean? your previous posted impedance measurements of the outputs showed "similar" readings for either output. Hence, the question. Please explain.
            How did you check the output capacitors from the upc4557... maybe the capacitors are having a high / higher esr than is acceptable.
            • UNITRAX
              • Jan 2023
              • 64
              • POLAND
              I checked VR151, the left VU meter does not respond to adjustments.
              The Vu meter is OK, the fault is somewhere in the system. Tomorrow I will borrow an oscilloscope and then I can take measurements.
              The sound level is equal on both channels, headphone and output. Therefore, I say that the damage is in the circuit.
              I agree with the quality of the materials used to build this tape recorder, it is armored.
              Very similar to Technics RS-M10 and Panasonic 616
              When I measure Vu meters not connected to the circuit. The left and right have the same resistance, the vu meters readings are the same, so I say the gauges themselves are working perfect.
              I checked the following with a tester of electronic components:
              Capacitors
              C141, C142, C143, C144, C902,
              C124, C126.
              Diodes
              DI51, DI52.
              Transistors
              X63,X64,X65,X66
              The tester shows that they are fine.
              The UPC4557 resistance on the left and right outputs is not the same
              legs 2 - 3 180 kOhms
              legs 5 - 6 2.4 Mega Ohms!!!
              • budwich
                • Jul 2015
                • 3097
                quote: "The UPC4557 resistance on the left and right outputs is not the same
                legs 5 - 6 2.4 Mega Ohms!!!"
                Those are NOT output pins. They are signal input pins to the op amps. The main reason for the difference... likely is that if you look at the schematic, it appears that only one of the op amps is used in the circuit.... as there is likely a upc4557 for each side, if I understand the posted manual correctly.
                Anyway, when you said that you tested the caps with a tester, what did the ESR reading say for the caps in question?
                • UNITRAX
                  • Jan 2023
                  • 64
                  • POLAND
                  Thanks for the correction, I'm still new to electronics and I make mistakes everywhere.
                  I don't remember the exact esr of each capacitor, but I had new ones ready for replacement. The desoldered ones had better esr and resistance than the new ones.
                  • CapLeaker
                    Leaking Member
                    • 8132
                    I just checked the schematic and the board. The UPC4557 uses one side for one channel and the other side for the other. There are no two separate op amp IC's for each channel. I would look hard at caps in series (DC blocker) like C143. When these go high in ESR in a circuit like this, you loose audio volume. Other than that you could take UPC4557 out, solder in some wires to the board and wire up UPC4557 channels up in a manner that left is right and right is left. If the problem indeed is the one side of the op amp, the problem should be now on the other VU meter.
                    • budwich
                      • Jul 2015
                      • 3097
                      ok.... as you planned, a scope of the area with a nice 1khz input signal will likely quickly point to where the signal disappears (ie. after which component).
                      Hopefully, your "better" means lower values although I am somewhat "concerned" at your description of "better esr and resistance"... hmmmm.
                      Having said that, what capleaker just posted, only one upc4557 (ie. basically no right and left board), then your impedance test of the two sets of inputs DOES appear to point at an issue associated with the input "area" of the apc4557 either with connections coming towards it or within the chip itself. If you take it out of the circuit and check those pin pairs, that will tell you more about where the problem lies.
                      • UNITRAX
                        • Jan 2023
                        • 64
                        • POLAND
                        In the first photo, the value of the capacitor desoldered from the circuit.
                        The second photo shows a new capacitor.
                        The new ones had higher ESR and Ω
                        Tomorrow I can change them for new ones, or order capacitors with better parameters?
                        Thank you all for your involvement. I will also need tips on how to use the oscilloscope properly.
                        • budwich
                          • Jul 2015
                          • 3097
                          Those seem ok. NOTE: in your impedance check of the inputs to the op amps, your 2/3 measure is the one of concern because that is the input associated with the left channel which you say is not working correctly.
                          • redwire
                            • 3906
                            [...] The sound level is equal on both channels, headphone and output. Therefore, I say that the damage is in the circuit. [...]
                            If there is good audio (not distorted, good level) at the headphone jack, then it would be the diodes or 4.7uF capacitor C143 or 47uF capacitor C144. Why are you messing with the 22uF parts, there's none in the meter circuit.
                            The VU meter rectifier diodes D151, D152 are OA90 germanium diodes. So they have a low Vf around 0.34V@1mA on a multimeter.
                            Notice it is a bit strange the meter/cap are reversed polarity, the (+) goes to GND.
                            The usual problem with cassette decks is a dirty REC-PLAY switch. This long switch I move back and forth 100 times, and use pot/switch cleaner with PPE additive. Don't get any spray on the belts!
                            • budwich
                              • Jul 2015
                              • 3097
                              excellent pickup. That cap is likely bad IF that is what he took out of the system. It appears that the OP was just going by the fact that he got a reading with the meter as opposed to looking at the schematic for what the values actually are... maybe.
                              • UNITRAX
                                • Jan 2023
                                • 64
                                • POLAND
                                "... then it would be the diodes or 4.7uF capacitor C143 or 47uF capacitor C144. Why are you messing with the 22uF parts, there's none in the meter circuit.
                                The usual problem with cassette decks is a dirty REC-PLAY switch. This long switch I move back and forth 100 times, and use pot/switch cleaner with PPE additive. Don't get any spray on the belts!"
                                C 143 4.7uF/50V and C144 47uF/25V I changed to new ones.
                                Their polarity is marked on the PCB, both top and bottom. According to this, the old and new capacitors were soldered.
                                Measured diodes with a multimeter:
                                D151= 0.310V
                                D251= 0.324V
                                D152= 0.342V
                                D252= 0.326V
                                I know the issue of the recording switch, I checked its conduction, seems ok. I switched it many times to be sure. I know the tape mechanism is a different story. I ordered new belts, the old one is already weak. After the replacement, I will adjust the mechanism.
                                I saw the polarity reversed in the diagram after you mention it. PCB has descriptions of components and solder pads. I hope the old ones were installed correctly, because that's how I put the new ones in.
                                The LINE IN inputs a 1kHz Sine signal
                                The oscilloscope connected to the output, the results of Lch and Rch can be seen in the pictures.
                                I only have basic information from his service. I don't have an isolation transformer and the oscilloscope is connected to the same wall as the tape recorder.
                                Volts/Div = 0.2V
                                Time/Div= 0.1ms
                                Probe= X1
                                • budwich
                                  • Jul 2015
                                  • 3097
                                  good work... just to confirm, are you referring to the outputs of the 4557? You need to check the output just after the caps in question (just after the 4557 outs) to see what is being delivered to the meters.
                                  • petehall347
                                    • Jan 2015
                                    • 4426
                                    • United Kingdom
                                    have you tried recording yet ?
                                    • UNITRAX
                                      • Jan 2023
                                      • 64
                                      • POLAND
                                      Only when recording is turned on, I can hear the 1kHz sound from the generator on the output
                                      The right vu meter connected to the oscilloscope when I turn on 1kHz,
                                      The voltage drops by about 120 - 140mV
                                      Left vu meter to the oscilloscope. The voltage drops by only about 5mV
                                      The oscilloscope connected between -, + C143, shows nothing.
                                      same with the C243
                                      Oscilloscope between Vr151- gnd and C143 - negative = 0.5V
                                      Oscilloscope between Vr251- gnd and C243 - negative = 20mV
                                      • budwich
                                        • Jul 2015
                                        • 3097
                                        sorry I don't understand what you have posted.
                                        are there wave forms or just DC voltages? Not sure that you are using the scope correctly.... "randomly" connecting the ground of the probe to a circuit can be damaging as this is inserting a ground / short into the circuit (comment related to " connected between the -,+ c143").
                                        • howardc64
                                          by howardc64
                                          Problem
                                          08-04-2024, 10:36 PM
                                        • UNITRAX
                                          Radio tape recorder Hitachi TRK 8010k - How to install aux in place of am band?
                                          by UNITRAX
                                          Hitachi TRK 8010k
                                          Is it possible to have an aux input in place of the radio on the am band range? When connecting to the tape recorder, I have additional tape noise. There is an integrated circuit HA11251 in the schematic. Which pin is the am radio output on?
                                          Radio tape recorder diagrams and HA11251 datasheet included...
                                          02-18-2023, 07:23 AM
                                        • RetroComputingGrotto
                                          IBM LTO 7 Tape Drive - Dust Sensors??
                                          by RetroComputingGrotto
                                          I have a Quantum SuperLoader robotic tape library that uses an IBM LTO 7 tape drive which is constantly complaining that it needs cleaning. If you insert a cleaning tape, it performs the clean and resets the cleaning counter to 0 hours. It should only require cleaning every 50 or so working hours but after as little as just 1 hours use it requests cleaning. We've purchased a new cleaning tape along with new LTO backup tapes but that has not resolved the problem.
                                          I have taken the library apart and can see no noticeable signs of dust inside. I have removed the tape drive from the...
                                          09-19-2024, 01:54 AM
                                        • triplefour
                                          LG 75UN8570AUD panel issue fixed with tape.
                                          by triplefour
                                          I've been seeing a lot of the new TVs (samsung and LG only i think) maybe 1-2 years old suffering from sudden panel death and have been over 50% successful bringing them at least most of the way back to functionality if not all the way, with the tape masking of the LVDS pins method.
                                          add this one to the list. the issue was on the left side of the panel and i ended up blocking just 3 pins on the connector that joins the two panel boards that make up the left side of the panel (it has 4 boards total) i didnt get the picture back to 100% perfect. i spent hours fiddling with which pins to block...
                                          07-17-2023, 10:22 PM
                                          DVR recorder Switching power supply Failure
                                          by sam_sam_sam
                                          When I originally started with my DVR recorder it had 4 individual switching switching power supply’s that were sh*t so I bought a 7 amp 12 volt switching power supply and modified the output plug and put a four output power supply cable to match the input power supply cable inputs
                                          Which worked for over 14 years until a few days ago when the cameras went dark no picture anymore well it turns out now the switching power supply has stopped working altogether now so I have to troubleshoot what has happened to this power supply more later on today what I found
                                          This...
                                          05-27-2023, 03:03 AM
