    Re: Meters and Microamps

    well the bussman/fluke fuse costs about the same as my 61e did!!!
    i think they are just fucking people because they know the cost of the equipment!

    Comment

    • eccerr0r
      Solder Sloth
      • Nov 2012
      • 8698
      • USA

      #62
      Re: Meters and Microamps

      We with Flukes are getting fleeced!

      Interesting:
      Fluke (hand): 15A HRC 10x38, I think 600mA HRC for 300mA mode.
      Tenma (hand): 10A ceramic 6x25(!!!), 500mA 5x20 for 200mA mode
      BKP (bench): 4A HRC 10x35, does not seem to be fuse for 20A mode? Doesn't sound right, perhaps someone put in a 4A HRC because they couldn't find a 20A HRC...
      HP (bench): 2A 6x32(?)
      Viz (bench): 2A glass heatshrink 6x32 Soldered in circuit
      Micronta Analog (hand): No fuse (3A max measured current), ?? for 300mA mode.
      Eico Analog (hand...): No fuse (10A max measured current)
      HFT Freebie (hand) : No fuse (10A), don't remember for 200mA mode.
      Tektronix (bench): 2A glass 6x32
      Mastech (hand): 10A noname ceramic 6x32
      Last edited by eccerr0r; 04-01-2021, 06:42 PM.

      Comment

      • stj
        Great Sage 齊天大聖
        • Dec 2009
        • 30985
        • Albion

        #63
        Re: Meters and Microamps

        out of curiosity,
        i wonder how much blood it would cost to buy a fuse from keysight!!!

        Comment

        • eccerr0r
          Solder Sloth
          • Nov 2012
          • 8698
          • USA

          #64
          Re: Meters and Microamps

          I suspect that common parts the expectation is people buy just like for Fluke from third party vendors... else https://www.keysight.com/find/parts ?

          BTW the old HP DMM is so old (pre-Agilent!) that it doesn't even have shrouded jacks, plus the fact it was meant to be a low current (2A) benchtop so no HRC fuses in that thing...
          Last edited by eccerr0r; 04-02-2021, 08:41 AM.

          Comment

          • stj
            Great Sage 齊天大聖
            • Dec 2009
            • 30985
            • Albion

            #65
            Re: Meters and Microamps

            but now they push HRC for the 440mA range!
            bastards.

            Comment

            • eccerr0r
              Solder Sloth
              • Nov 2012
              • 8698
              • USA

              #66
              Re: Meters and Microamps

              My Eico VOM 10A or even 500mA range has no fuse at all... Pin attach probes, no shrouding. Who needs fuses?

              "Back in my day we had to walk to school in the snow up hill, both ways!"

              Comment

              • stj
                Great Sage 齊天大聖
                • Dec 2009
                • 30985
                • Albion

                #67
                Re: Meters and Microamps

                did the old AVO-8 have fuses?

                Comment

                • eccerr0r
                  Solder Sloth
                  • Nov 2012
                  • 8698
                  • USA

                  #68
                  Re: Meters and Microamps

                  Looks like at least the AVO-8 Mark 7 version had fuses... The M.7 looks sort of like 1980s design perhaps. Older versions of the AVO-8, no idea.

                  I suspect that most 1990s multimeter designs started using HRC fuses?

                  Comment

                  • stj
                    Great Sage 齊天大聖
                    • Dec 2009
                    • 30985
                    • Albion

                    #69
                    Re: Meters and Microamps

                    most early 90's ones had 20mm glass fuses.
                    so HRC - define "High"

                    Comment

                    • eccerr0r
                      Solder Sloth
                      • Nov 2012
                      • 8698
                      • USA

                      #70
                      Re: Meters and Microamps

                      Hmm...yeah I suppose the early 90s still used 80s like designs, glass...
                      but later on in the decade the ceramics came in...

                      Then again if there's no regulations, there's no need for these, and all those eevblog naysayers giving crap to cheap chinese crap designs...never needed to replace the fuse they never needed...

                      Comment

                      • redwire
                        Badcaps Legend
                        • Dec 2010
                        • 3906
                        • Canada

                        #71
                        Re: Meters and Microamps

                        I remember at Radio Shack guy demanding a refund for his analog multimeter. He wanted to check if a mains outlet was 15A so he put his meter across it. The rotary switch basically exploded
                        All my old 80's multimeters have no fuse on the 20A range, it's trial by fire. But engineers in the lab blow fuses in multimeters and I'm the guy buying new ones, the boss thinks I popped them, I just wanna measure current.

                        A long time ago Fluke pushed for safety, patented many aspects of multimeter input protection, bigger fuses etc. So they pioneered 1010 Safety Standard which became 61010 the "multimeter standard". I think there were too many incidents of people using the current function and hitting high power systems.

                        A fuse's voltage rating is dictated by it's length, we would all like 5x20mm fuses good for 1,000V but that's too short, so we get 38mm piggies.

                        I can't pay $50 for a multimeter fuse nowadays. I could hoard them and make a ton of money off their appreciation. Not as much as Bitcoin, but maybe FuseCoin... lol

                        Comment

                        • stj
                          Great Sage 齊天大聖
                          • Dec 2009
                          • 30985
                          • Albion

                          #72
                          Re: Meters and Microamps

                          i wonder if that's why fluke used to sell meter probes with a fuse in the cable?

                          as for switches exploding,
                          i got a free meter from a guy that had the protection diode split and the switch contact and tracks for current completly burned away.

                          i dont know what he did, but i dont care.
                          i just blocked-up the socket for current, replaced the diode and gave it to a friend who just uses it for voltage and continuity.

                          Comment

                          • eccerr0r
                            Solder Sloth
                            • Nov 2012
                            • 8698
                            • USA

                            #73
                            Re: Meters and Microamps

                            When I got the Eico VOM (likewise was free) it had a burned resistor in it...

                            "Gee I wonder how it burned!"

                            Then again I wonder if fuses are in line with the non-current carrying modes especially with DMMs. Analog VOMs can actually pass significant current in the Rx1 mode...

                            (That Eico meter was pretty broken when I got it. I had to replace resistors, AC rectifier, and the current shunt (which was incorrect for the application, not because it blew... Imagine blowing the current shunt... ) )

                            Comment

                            • stj
                              Great Sage 齊天大聖
                              • Dec 2009
                              • 30985
                              • Albion

                              #74
                              Re: Meters and Microamps

                              lol
                              i'v seen a current-shunt that desoldered itself before.
                              people dont read the manuals - there is a duty-cycle for high current readings.
                              something like 15seconds use, 2min's cooldown.

                              till you get some fool running 8A through the meter for an hour while he monitors some running kit!

                              Comment

                              • redwire
                                Badcaps Legend
                                • Dec 2010
                                • 3906
                                • Canada

                                #75
                                Re: Meters and Microamps

                                Fuse wire - I've seen many different alloys, tinned copper seems fine but knockoffs are higher resistance likely aluminum, as they are in automotive fuses. You don't want a (high resistance) fuse that runs really hot melting things.
                                I want to get some fuse wire, that's the main ingredient in the $50 caviar.

                                The fuses protect only the current shunts (and the user) but not the voltage/ohms functions.
                                There are pics of a totally roasted Brymen - I figure the guy connected it to MOT and burned the MOV's and resistors on both channels, they arc'd to each other which is not supposed to happen. No fuse there.

                                Comment

                                • eccerr0r
                                  Solder Sloth
                                  • Nov 2012
                                  • 8698
                                  • USA

                                  #76
                                  Re: Meters and Microamps

                                  That Eico's shunt wire was exactly that - a wire - and apparently someone substituted a copper wire when assembling it for whatever reason. So even at 10A it had no voltage across it, probably had more voltage across the probe wires.

                                  I didn't have any resistance wire to repair it, but I did have a pile of 0.2Ω 2W resistors. It should have been 0.025Ω so I piled eight of the resistors in parallel. At 10A, this pile of 2W resistors should survive just fine I think, and now the Eico can once again measure 10A current.

                                  I wonder if the fuse wire is Bussmann/Littelfuse/Siba/...'s secret sauce...and why the caviar is $50.

                                  Comment

                                  • stj
                                    Great Sage 齊天大聖
                                    • Dec 2009
                                    • 30985
                                    • Albion

                                    #77
                                    Re: Meters and Microamps

                                    shit happens.
                                    there's a thread here somewhere, where a guy tried to test a fence charger with a fluke and a uni-trend pocket meter.
                                    fucked the fluke, but the uni-t actually survived and read something like 3KV even though it was only 600v rated!!!!
                                    i guess the little pocket ones have no mov's in them.

                                    Comment

                                    • stj
                                      Great Sage 齊天大聖
                                      • Dec 2009
                                      • 30985
                                      • Albion

                                      #78
                                      Re: Meters and Microamps

                                      siba lists the resistance of their fuses in the datasheets.
                                      Attached Files

                                      Comment

                                      • eccerr0r
                                        Solder Sloth
                                        • Nov 2012
                                        • 8698
                                        • USA

                                        #79
                                        Re: Meters and Microamps

                                        I'd imagine all reputable fuse companies should report DCR of their fuses, seems Siba is one like Bussmann and Littelfuse.

                                        ---

                                        If I knew I had to probe 3KV I'd go with the Eico that actually has a 5KV range - well past all the DMM's I have

                                        (I do have a 66MΩ and a 100MΩ HV resistors that I could use if push comes to shove though will need to do math to get the actual voltage...)
                                        Last edited by eccerr0r; 04-03-2021, 03:34 PM.

                                        Comment

                                        • redwire
                                          Badcaps Legend
                                          • Dec 2010
                                          • 3906
                                          • Canada

                                          #80
                                          Re: Meters and Microamps

                                          Looking at copper fuse wire, Onderdonk's equations from the days of the telegraph:
                                          20AWG (0.812mm) matches the DMM-11A curve e.g blows 10 seconds 43A. 22AWG is 10 seconds 30A.
                                          35AWG (0.413mm) matches the DMM-44/100 curve e.g blows 10 seconds 1.5A and 36AWG is 10 seconds 1.2A
                                          I think the sand filling should make the results higher due to the cooling, so 22AWG and 36AWG caviar should taste good

                                          Ali seems to offer only aluminum fuse wire, half the melting point of copper. Can't solder it and CCA probably sucks for a good connection.

                                          Comment

                                          Related Topics

                                          Collapse

                                          • eccerr0r
                                            lowest current galvanometer/analog meter (passively powered!)
                                            by eccerr0r
                                            Most analog meters are actually current driven - a current is needed to deflect the pointer full scale. They can be lots of different values. There are ones that can take several amperes to swing the pointer as they have internal shunts.

                                            On the other hand there are many that are rated in microamps for the intent that they draw as little power as possible from the devices under test. Some are 1mA, 500µA, 200µA, 100µA, and the ubiquitous 50µA frequently used in analog VOMs. There are even 20, 25, 30 µA and even down to 10µA. Been collecting a bunch of different ratings and...
                                            04-16-2025, 12:41 AM
                                          • chth96
                                            Is it possible to test small ceramic disc capacitor with analogue meter?
                                            by chth96
                                            I read EBOOK of Testing Electronic Component and found that

                                            "In order to test the small blue resin coated ceramic disk capacitor which has 104 50v specification,it is possible to test this small ceramic capacitor by making use of analogue meter.
                                            Set to times 10Kohm that has 12v output from the probe to check disorder of ceramic capacitor.
                                            It is not uncommon that testing result of digital capacitance meter is okay,But It show shorted reading when you check it with analogue meter."

                                            Is it really possible to test small ceramic disc capacitor with analogue...
                                            03-09-2024, 10:03 PM
                                          • jorpec
                                            Changing the scale of an analog meter
                                            by jorpec
                                            Hi

                                            I want to change the scale of an analog meter, to monitor the charge of an li-ion single cell, so the needle at the end would point 4.2 volts and the 0 on the meter would point 2.5 volts or so, i tried to add diodes or a zener in series with the meter but the changes of the ambient temperature also changes the voltage value on the meter
                                            I wonder how i can accomplish this so the meter would be precise at any temperature ?

                                            Thnaks
                                            01-23-2023, 10:37 AM
                                          • sam_sam_sam
                                            How an incandescent light and a ESR meter ** friend**have a machine blowing fuses
                                            by sam_sam_sam
                                            About a month ago we had a transformer jumper wire that burned in two different places between two transformers caused over $200 thousand dollars worth of damage to VFD drives at least 10 of them and a memory card and quite a few power supply as well

                                            Now we have most of the equipment that got damaged back up and running except a feeder that has issues and we are waiting for some parts for

                                            Now we also have another piece of equipment that was blowing fuses on the 110 volt controller switch circuit with the “E” stops and one switching power (24 volt) supply and a transformer...
                                            12-14-2023, 07:45 PM
                                          • sam_sam_sam
                                            Modification of a Fluke multi meter that uses three AA batteries to use a 14500 lithium ion battery
                                            by sam_sam_sam
                                            I have had with regular AA batteries that leak this meter is expensive new I already replaced the battery battery clips once and the same thing happened again and I am tried of so it is getting modified to accommodate lithium ion batteries that are the same size excluding the positive terminal and removing the battery clips I will have enough room for the BMS protection board

                                            I going to put 3 lithium ion battery in parallel with the BMS protection board and a way to charge the battery externally and heat shrink the battery pack and use a JST connector so I easily remove the battery...