    My new (used) Fluke 87

    I recently picked up another meter from a local pawn shop, this one is a Fluke 87. When I saw it I was immediately attracted to it because it had lots of features in one meter that I previously only had over several meters. I also liked the display but I later realized that it washes out really bad when viewed from the top. When viewed straight on and below it is bright and crisp. Although I knew there have been several iterations of the 87 model I was not aware of how old this meter really was, it was actually manufactured in the early nineties.

    It had a few other problems like the low current fuse was bad, the "kick stand" was missing and it was dirty. The probes it came with are Flukes but they were (as Dave says) very crusty. I think it cleaned up rather well in the end but you can judge for yourself as I have included some before and after pictures below. I have already made a crude wire kick stand for it but I intend to make something more permanent when time permits.

    The one thing I could kick myself for was that in my enthusiasm to rip it apart after getting it home I broke one of the clips that attaches the screen to the PCB. So far it has not affected the meters operation and I have an idea on how to bodge a repair when I have the time to do it.
    Attached Files
  • retiredcaps
    Badcaps Legend
    • Apr 2010
    • 9271

    #2
    Re: My new (used) Fluke 87

    Originally posted by LDSisHere
    I also liked the display but I later realized that it washes out really bad when viewed from the top.
    Some of the older original Fluke 80 series suffer from the faded digits due to contamination of the elastometer. Modemhead has documented the cleaning procedure at

    http://mrmodemhead.com/blog/fluke-8x-faded-lcd/

    I have cleaned two myself with good results.

    I think it cleaned up rather well in the end but you can judge for yourself as I have included some before and after pictures below.
    Pumice hand cleaner works really well to clean up the holster. It almost looks new and all the markings are still on it!!

    The fusible resistor (circled in red) seems a bit toasty? Or is that a photo trick? I would measure it to make sure it is around 1 kohm.

    Even though it was last calibrated in 1995, it is likely to hold and be within specifications.
    Attached Files
    --- begin sig file ---

    If you are new to this forum, we can help a lot more if you please post clear focused pictures (max resolution 2000x2000 and 2MB) of your boards using the manage attachments button so they are hosted here. Information and picture clarity compositions should look like this post.

    We respectfully ask that you make some time and effort to read some of the guides available for basic troubleshooting. After you have read through them, then ask clarification questions or report your findings.

    Please do not post inline and offsite as they slow down the loading of pages.

    --- end sig file ---

    Comment

    • LDSisHere
      Badcaps Veteran
      • May 2012
      • 727
      • U.S.A.

      #3
      Re: My new (used) Fluke 87

      Originally posted by retiredcaps
      Pumice hand cleaner works really well to clean up the holster. It almost looks new and all the markings are still on it!!
      I have learned my lesson about the markings and I now use extreme care when I am cleaning around them.

      The fusible resistor (circled in red) seems a bit toasty? Or is that a photo trick? I would measure it to make sure it is around 1 kohm.
      I will check that out as soon as I get a chance, but it did not raise any flags when I was looking it over. (Not that means much.)

      Even though it was last calibrated in 1995, it is likely to hold and be within specifications.
      I compared it with my Fluke 175 using two AA batteries (~3V) and they both gave me the same reading. I have not yet done any other comparisons but I need to do so.

      I will also check out Modemheads blog, I have probably already read that article but it will have a lot more meaning too me now.

      Comment

      • capkid
        Badcaps Legend
        • Oct 2010
        • 1339
        • United States

        #4
        Re: My new (used) Fluke 87

        I have a Fluke 87 III - good meter.
        LG Plasma Mal-Discharge Correction Service

        Comment

        • modemhead
          Lurking
          • Jun 2012
          • 52
          • USA

          #5
          Re: My new (used) Fluke 87

          Originally posted by LDSisHere
          I recently picked up another meter from a local pawn shop, this one is a Fluke 87.
          Nice find! Looks to be in excellent condition, especially with the clean-up job.

          The viewing angle thing is just a property of that LCD. It has the best contrast at a slight downward angle. Somewhere in the fog of my brain I remember reading something about that being a design trade-off, bias voltage vs. minimum operating temperature and other things I don't understand. The III and V generation displays are improved a bit in this respect.

          Being new to using an 87, here's something to be aware of. It LOVES to take itself out of auto-ranging mode. For example, if you select Ohms and then enable continuity (beeper) mode, and then go back to Ohms, it will be in 400 ohm range and auto-ranging will be off. It has to be re-enabled by holding the range button for a couple seconds or flipping the dial away and back to Ohms. This happens for a lot of the functions, like Min-Max, Hz, etc. It can lead to some head-scratchers if you forget. Like measuring a 1K resistor with it stuck in 400 ohm range. "Why did this resistor go open-circuit?"

          Sorry I don't have a fix for the broken clip. Duct tape?

          Comment

          • Kiriakos GR
            Banned
            • May 2012
            • 940
            • Greece

            #6
            Re: My new (used) Fluke 87

            Originally posted by modemhead
            It can lead to some head-scratchers if you forget.
            Few more things to worry about other than if it holds it calibration. LOL

            Comment

            • LDSisHere
              Badcaps Veteran
              • May 2012
              • 727
              • U.S.A.

              #7
              Re: My new (used) Fluke 87

              Originally posted by modemhead
              The viewing angle thing is just a property of that LCD. It has the best contrast at a slight downward angle. Somewhere in the fog of my brain I remember reading something about that being a design trade-off, bias voltage vs. minimum operating temperature and other things I don't understand.
              I figured it was something like that, good thing I do not use my meters upside down very often.

              Being new to using an 87, here's something to be aware of. It LOVES to take itself out of auto-ranging mode. For example, if you select Ohms and then enable continuity (beeper) mode, and then go back to Ohms, it will be in 400 ohm range and auto-ranging will be off. It has to be re-enabled by holding the range button for a couple seconds or flipping the dial away and back to Ohms. This happens for a lot of the functions, like Min-Max, Hz, etc. It can lead to some head-scratchers if you forget. Like measuring a 1K resistor with it stuck in 400 ohm range. "Why did this resistor go open-circuit?"
              Thanks for the heads up, this type of operation would make me think my meter is defective.

              Sorry I don't have a fix for the broken clip. Duct tape?
              I have a repair idea that I am going to try on the clip. If it is successful I will post the result. If it is not successful I should be no worse off than I am now. I just wish I had remembered your warning about these clips from your blog post. When I read it a while back ago it did not apply to me so I did not bother to commit it to memory. They break way too easily IMO.
              Last edited by LDSisHere; 11-02-2013, 12:55 PM. Reason: Typo

              Comment

              • PeteS in CA
                Badcaps Legend
                • Aug 2005
                • 3576
                • USA, Unsure of Planet

                #8
                Re: My new (used) Fluke 87

                One of the ICs has what looks to be a 1994 date on it; as does what appears to be a factory cal sticker (which probably cost the purchaser extra). So whatever version of the Fluke 87 was made then would be what you have. All in all, a very decent handheld DMM.
                PeteS in CA

                Power Supplies should be boring: No loud noises, no bright flashes, and no bad smells.
                ****************************
                To kill personal responsibility, initiative or success, punish it by taxing it. To encourage irresponsibility, improvidence, dependence and failure, reward it by subsidizing it.
                ****************************

                Comment

                • LDSisHere
                  Badcaps Veteran
                  • May 2012
                  • 727
                  • U.S.A.

                  #9
                  Re: My new (used) Fluke 87

                  Originally posted by retiredcaps
                  The fusible resistor (circled in red) seems a bit toasty? Or is that a photo trick? I would measure it to make sure it is around 1 kohm.
                  I think the color in the picture is due to the lighting. It actually looks really nice in person. (I know that sounds strange, but it is one of the best looking resistors I have ever seen.) It measured 1.001K ohms on my Fluke 175.

                  Originally posted by PeteS in CA
                  One of the ICs has what looks to be a 1994 date on it; as does what appears to be a factory cal sticker (which probably cost the purchaser extra). So whatever version of the Fluke 87 was made then would be what you have. All in all, a very decent handheld DMM.
                  Good catch on the date code, I did not notice it on the chip until you mentioned it. Since Fluke introduced the 80 series in 1988, it is not near as old as it could be.

                  Comment

                  • LDSisHere
                    Badcaps Veteran
                    • May 2012
                    • 727
                    • U.S.A.

                    #10
                    Re: My new (used) Fluke 87

                    Stupid plastic clips.

                    I took this meter back apart to attempt to repair the clip I broke the first time and I ended up breaking a second one. At least this time it did not snap completely off but it was close. These clips are way too fragile as I was very careful when I was removing the frame. So I had to repair two clips instead of just one.

                    The attached pictures show the repair job. I used the trimmings from PCB's that I previously etched for the bracing and CA gel adhesive as a binding agent. I waited between 12 to 24 hours between each step to make sure the glue was cured. I had to make sure to leave enough gap between the bracing and top of the clip so that the meter's board would fit between them. So far the repair is good, I just hope the glue does not let loose.
                    Attached Files
                    Last edited by LDSisHere; 11-04-2013, 04:23 PM. Reason: The pictures did not upload.

                    Comment

                    • retiredcaps
                      Badcaps Legend
                      • Apr 2010
                      • 9271

                      #11
                      Re: My new (used) Fluke 87

                      Originally posted by LDSisHere
                      Stupid plastic clips.
                      You can buy an entire new assembly on ebay, but the asking price is around $30 USD.

                      http://www.ebay.com/itm/370869182615

                      I believe the 87V now uses screws instead of plastic clips.

                      These clips are way too fragile as I was very careful when I was removing the frame.
                      I suggest leaving the holster on 100% of the time to cushion any falls to protect your repair. When you change batteries or fuses be careful not to drop it as that is when it is most vulnerable.

                      The attached pictures show the repair job. So far the repair is good, I just hope the glue does not let loose.
                      Maybe modemhead will include this repair in his user submitted section?
                      --- begin sig file ---

                      If you are new to this forum, we can help a lot more if you please post clear focused pictures (max resolution 2000x2000 and 2MB) of your boards using the manage attachments button so they are hosted here. Information and picture clarity compositions should look like this post.

                      We respectfully ask that you make some time and effort to read some of the guides available for basic troubleshooting. After you have read through them, then ask clarification questions or report your findings.

                      Please do not post inline and offsite as they slow down the loading of pages.

                      --- end sig file ---

                      Comment

                      • LDSisHere
                        Badcaps Veteran
                        • May 2012
                        • 727
                        • U.S.A.

                        #12
                        Re: My new (used) Fluke 87

                        Originally posted by retiredcaps
                        You can buy an entire new assembly on ebay, but the asking price is around $30 USD.

                        http://www.ebay.com/itm/370869182615

                        I believe the 87V now uses screws instead of plastic clips.
                        My cheap (compared to the 87) Fluke 21 from the same era even uses screws. I wonder if the accountants had a hand in the design process? $30 for a replacement is more than I want to pay but if it is you only option to keep the meter working then it is reasonable. This is why I came up with my own option.

                        I suggest leaving the holster on 100% of the time to cushion any falls to protect your repair. When you change batteries or fuses be careful not to drop it as that is when it is most vulnerable.
                        You bet this meter will be handled gently. I was considering using this meter as my main one, now I may just use it when I need multiple meters or features my 175 does not have.


                        Maybe modemhead will include this repair in his user submitted section?
                        He is more than welcome to use this on his site, I will send him the original shots if he is interested.

                        Comment

                        • modemhead
                          Lurking
                          • Jun 2012
                          • 52
                          • USA

                          #13
                          Re: My new (used) Fluke 87

                          With glue-type repairs, you never know how it's going to hold up, but yours looks good. The reinforcing material is a great idea.

                          Originally posted by LDSisHere
                          My cheap (compared to the 87) Fluke 21 from the same era even uses screws. I wonder if the accountants had a hand in the design process?
                          We used to call the value engineers "screw shorteners". In this case I think they went overboard. But Fluke did indeed go back to screws (4 of them) in the 87-V generation.

                          Originally posted by LDSisHere
                          He is more than welcome to use this on his site, I will send him the original shots if he is interested.
                          Always interested in good repairs! You have my email address. No hurry.

                          Comment

                          • Nitro468
                            Senior Member
                            • Oct 2013
                            • 65
                            • USA

                            #14
                            Re: My new (used) Fluke 87

                            I broke a clip on my Fluke 88... Packaging tape works well. I also sneaked in a paper shim to tighten up the elastometer connections. Now I need to get an input jack for it now.

                            Comment

                            • retiredcaps
                              Badcaps Legend
                              • Apr 2010
                              • 9271

                              #15
                              Re: My new (used) Fluke 87

                              Originally posted by Nitro468
                              Now I need to get an input jack for it now.
                              What do you mean above? Do you need the entire assembly? If yes, there are available on ebay at a hefty $30+ USD.

                              What is wrong the with current input jack?
                              --- begin sig file ---

                              If you are new to this forum, we can help a lot more if you please post clear focused pictures (max resolution 2000x2000 and 2MB) of your boards using the manage attachments button so they are hosted here. Information and picture clarity compositions should look like this post.

                              We respectfully ask that you make some time and effort to read some of the guides available for basic troubleshooting. After you have read through them, then ask clarification questions or report your findings.

                              Please do not post inline and offsite as they slow down the loading of pages.

                              --- end sig file ---

                              Comment

                              • PhaseCoder
                                Senior Member
                                • Nov 2013
                                • 62
                                • Uk

                                #16
                                Re: My new (used) Fluke 87

                                I picked one up from ebay a few year back and still using to this day.

                                Comment

                                • Nitro468
                                  Senior Member
                                  • Oct 2013
                                  • 65
                                  • USA

                                  #17
                                  Re: My new (used) Fluke 87

                                  The current input jack is kaput. Nothing left of the terminals. That is what I get for lending it out.

                                  Comment

                                  • retiredcaps
                                    Badcaps Legend
                                    • Apr 2010
                                    • 9271

                                    #18
                                    Re: My new (used) Fluke 87

                                    Originally posted by Nitro468
                                    The current input jack is kaput. Nothing left of the terminals. That is what I get for lending it out.
                                    Can you post a picture of the input jack?

                                    Some of us are interested in helping others keep their old multimeters working by coming with with safe, but creative repairs.

                                    PS. If I had to lend out meters, I lend out the ones with no fuses or ones that don't measure current.
                                    --- begin sig file ---

                                    If you are new to this forum, we can help a lot more if you please post clear focused pictures (max resolution 2000x2000 and 2MB) of your boards using the manage attachments button so they are hosted here. Information and picture clarity compositions should look like this post.

                                    We respectfully ask that you make some time and effort to read some of the guides available for basic troubleshooting. After you have read through them, then ask clarification questions or report your findings.

                                    Please do not post inline and offsite as they slow down the loading of pages.

                                    --- end sig file ---

                                    Comment

                                    • retiredcaps
                                      Badcaps Legend
                                      • Apr 2010
                                      • 9271

                                      #19
                                      Re: My new (used) Fluke 87

                                      Originally posted by LDSisHere
                                      The attached pictures show the repair job.
                                      I saw this repair job on ebay for the lcd broken clips.

                                      http://www.ebay.com/itm/231114471645

                                      Description from seller.

                                      "two of the plastic clips that hold the lcd screen to the circuit board snapped off while I tried to diagnose the problem, I've replaced them with some screws and it seems to be stronger than before. this is shown in the second picture."

                                      PS. At $40 + 5 s/h, it is definitely over-valued. There are probably numerous problems.
                                      Attached Files
                                      Last edited by retiredcaps; 12-11-2013, 06:23 PM.
                                      --- begin sig file ---

                                      If you are new to this forum, we can help a lot more if you please post clear focused pictures (max resolution 2000x2000 and 2MB) of your boards using the manage attachments button so they are hosted here. Information and picture clarity compositions should look like this post.

                                      We respectfully ask that you make some time and effort to read some of the guides available for basic troubleshooting. After you have read through them, then ask clarification questions or report your findings.

                                      Please do not post inline and offsite as they slow down the loading of pages.

                                      --- end sig file ---

                                      Comment

                                      • LDSisHere
                                        Badcaps Veteran
                                        • May 2012
                                        • 727
                                        • U.S.A.

                                        #20
                                        Re: My new (used) Fluke 87

                                        Did that guy really just drill through the PCB where he guessed it would be OK? I hope he had X-Ray vision or something.

                                        Comment

                                        Related Topics

                                        Collapse

                                        • sam_sam_sam
                                          Modification of a Fluke multi meter that uses three AA batteries to use a 14500 lithium ion battery
                                          by sam_sam_sam
                                          I have had with regular AA batteries that leak this meter is expensive new I already replaced the battery battery clips once and the same thing happened again and I am tried of so it is getting modified to accommodate lithium ion batteries that are the same size excluding the positive terminal and removing the battery clips I will have enough room for the BMS protection board

                                          I going to put 3 lithium ion battery in parallel with the BMS protection board and a way to charge the battery externally and heat shrink the battery pack and use a JST connector so I easily remove the battery...
                                          12-31-2023, 06:04 AM
                                        • Baldbentley
                                          repairing a Fluke 335 clamp meter
                                          by Baldbentley
                                          I am repairing a Fluke 335 clamp meter. After contacting fluke it appears there is no service manual or circuit board/ component list available. I have determined that there us a faulty PNP bipolar junction transistors shown in the pictures. I am having trouble sourcing this transistor. I pulled a good one off of the board and tested it with my component analyzer but I cannot find anything that matches the specifications. Any help locating this part or a replacement would be greatly appreciated.

                                          Thanks...
                                          10-18-2023, 02:05 AM
                                        • chth96
                                          Is it possible to test small ceramic disc capacitor with analogue meter?
                                          by chth96
                                          I read EBOOK of Testing Electronic Component and found that

                                          "In order to test the small blue resin coated ceramic disk capacitor which has 104 50v specification,it is possible to test this small ceramic capacitor by making use of analogue meter.
                                          Set to times 10Kohm that has 12v output from the probe to check disorder of ceramic capacitor.
                                          It is not uncommon that testing result of digital capacitance meter is okay,But It show shorted reading when you check it with analogue meter."

                                          Is it really possible to test small ceramic disc capacitor with analogue...
                                          03-09-2024, 10:03 PM
                                        • jorpec
                                          Changing the scale of an analog meter
                                          by jorpec
                                          Hi

                                          I want to change the scale of an analog meter, to monitor the charge of an li-ion single cell, so the needle at the end would point 4.2 volts and the 0 on the meter would point 2.5 volts or so, i tried to add diodes or a zener in series with the meter but the changes of the ambient temperature also changes the voltage value on the meter
                                          I wonder how i can accomplish this so the meter would be precise at any temperature ?

                                          Thnaks
                                          01-23-2023, 10:37 AM
                                        • sam_sam_sam
                                          How an incandescent light and a ESR meter ** friend**have a machine blowing fuses
                                          by sam_sam_sam
                                          About a month ago we had a transformer jumper wire that burned in two different places between two transformers caused over $200 thousand dollars worth of damage to VFD drives at least 10 of them and a memory card and quite a few power supply as well

                                          Now we have most of the equipment that got damaged back up and running except a feeder that has issues and we are waiting for some parts for

                                          Now we also have another piece of equipment that was blowing fuses on the 110 volt controller switch circuit with the ā€œEā€ stops and one switching power (24 volt) supply and a transformer...
                                          12-14-2023, 07:45 PM
                                        • Loading...
                                        • No more items.
                                        Working...