Test Primary of SMPS with scope
  • stepgr
    • Nov 2008
    • 34
    • Greece
    can someone explain in a simple (if possible ) way , how to measure the primary (hot) side of an power supply without burning the scope or the PSU ?
    From reading various posts in here I came to the following procedure :
    1. Use an isolation transformer on the PSU and then connect the probes ground
    on the hot ground on the PSU and the probe wherever I want in the primary side.
    OR
    2. Use a dual channel scope in add or subtract mode without the need of an isolation transformer, connecting the two probes ground together and probe with the two probe tips the hot side.
    I have an old dual channel 25MHz scope who says max input 300V (DC +AC peak) and I want to do this mostly to learn .
    thanks in advance for any tips or recommendations.
  • Kiriakos GR
    Banned
    • May 2012
    • 940
    • Greece
    Re: Test Primary of SMPS with scope
    The isolation transformer is a good step but costly at 500W or more.
    One special probe at X100 it is what will solve the input voltage limit of your scope, but it is pointless getting one if your oscilloscope have settings just for X1 & X10 probes.
    • Uniballer
      • Jul 2013
      • 334
      The safety folks will tell you that it is dangerous to defeat the safety grounding of your oscilloscope and/or the device under test. They say you need fancy differential probes to do it right. And if you have them it works great and is quite safe. So that is what I officially recommend.
      I have not personally had great luck using the "two probes in add mode with one inverted" for a differential measurement. I'm not really sure why. Maybe I just didn't work at it hard enough.
      My older analog scope (Tek 475A) has a transformer-isolated linear power supply. I can float this scope simply by eliminating the safety ground with one of those 2 to 3 prong adapters. Note that this means if you connect the ground lead of the probe to something 400V above ground then all of the metal parts on the exterior of the scope will be at that potential relative to ground. And the grounds of the two probes are tied together as well. So there is certainly a very real safety concern in floating the scope.
      I also have an isolation transformer that will probably work with any scope (linear or SMPS power supply) if you use one of those 2 to 3 prong adapters to lift the ground lead. But the same safety concern applies.
      I see no problem using an x100 probe on a scope that doesn't understand it as long as you can remember that the real voltages are 100 times what the scope displays.
        I always put the UUT into the isolation transformer, the transformer is then plugged into GFI outlet.
        • Uniballer
          • Jul 2013
          • 334
          While this does not completely eliminate the safety problems, at least the oscilloscope front panel will not shock you if you touch it and ground at the same time. So this is probably a safer way to use an isolation transformer. My isolation transformer is rated for about 920 VA, and cost about $65 (with shipping within the continental US) on ebay a few years ago. It looks a lot like this one.
          • stepgr
            • Nov 2008
            • 34
            • Greece
            Can you give an estimate since we both seem to be in the same country ?
            I don't think I'll need more than 500W, my main items to test are PC PSU's
            and in general IT PSU's no grater than 400W (from lcd monitors etc).
            Also the x100 probe is a must from what I understand in all your answers.
            Again thank you all guys for providing useful information (and safety) .
              I was able to purchase a 500VA continuous, 750VA intermittent isolating transformer for less than £25 here, from a local test equipment shop (Stewart of Reading.) Check local area for any shops selling such.
              • Kiriakos GR
                Banned
                • May 2012
                • 940
                • Greece
                I can not offer any true estimate, copper price rising like crazy.
                Your best bet is to order one 220/220 transformer and to make one box for it.
                There is a workshop that makes transformers 2Km from my shop.
                I would suggest you to find such a source in your area too and to avoid any middle man (electronic parts shop).
                If you are not aware of were to search? pay a visit in to any large supply shop for electricians, they will have some one to suggest to you.
                Yes the 100:1 probe is a must have, if your scope support it.
                I got my own yesterday.
                http://www.ittsb.eu/forum/index.php?topic=657.0
                • stepgr
                  • Nov 2008
                  • 34
                  • Greece
                  I also saw this little item here Pintek DP-25 Differential Probe which seems to be affordable for amateur use (all other diff probes I found go in more than 400 hundred range for some mysterious reason... ). What do you guys think ?
                  I will look into transformer workshop as you suggested (I already asked in electronic shops and found 50 to 60 euros for 500 VA transformers).
                  • Uniballer
                    • Jul 2013
                    • 334
                    ... Pintek DP-25 Differential Probe ... What do you guys think ?
                    No personal experience. There is a Pintek DP-25 teardown on EEVblog.
                    • stepgr
                      • Nov 2008
                      • 34
                      • Greece
                      Yes I was reading that also yesterday, seems decent enough for it's price
                      • Kiriakos GR
                        Banned
                        • May 2012
                        • 940
                        • Greece
                        @stepgr
                        Basically you have not mention yet which oscilloscope you do own?
                        And how high voltage you need to measure.
                        There is many tools out there for very specific applications.
                        In some circumstances there is no solution for a free ride, but only the complete ticket at full price.
                        Therefore be more specific.
                        • stepgr
                          • Nov 2008
                          • 34
                          • Greece
                          You mean what brand it its ? take a look (I said old didn't I?)
                          • Kiriakos GR
                            Banned
                            • May 2012
                            • 940
                            • Greece
                            @stepgr
                            You should turn your calendar 50 years ahead and buy a proper oscilloscope.
                            Because non of the accessories of 2000-2013 are capable to work with this Acropolis shaped as Oscilloscope.
                            And this is an advice for all.
                            • stepgr
                              • Nov 2008
                              • 34
                              • Greece
                              @stepgr
                              when it's time for a new, I promise I get one OK?
                              • Kiriakos GR
                                Banned
                                • May 2012
                                • 940
                                • Greece
                                The breakthrough of the affordable DSO in the past five years, is that the Oscilloscope become capable to display measurements in numbers with an accuracy close to 1% regarding volts that is good enough.
                                The much older Oscilloscopes they do not display numbers and they have a far worse accuracy of 5% or more that is bad.
                                The analog projection on this screen it can be out of adjustment and so you can not be certain of what you see in them, if the Oscilloscope is not tested against a high quality digital function generator.
                                This Pintek DP-25 Differential Probe is actually an active step down transformer, with an output of few mV which the digital scope according the attenuation adjustment on it, it can multiply and display as to were real voltage.
                                In conclusion when you will manage to get a proper fresh scope, all the joy will be yours and not my.
                                From the other hand if these words of truth sounds unpleasantly,
                                this is not my fault.
                                • stepgr
                                  • Nov 2008
                                  • 34
                                  • Greece
                                  You're right , I already told you. It's better to spend some money to get a decent new scope.
                                  • JimBanville
                                    MOSFET mystery
                                    by JimBanville
                                    N channel mosfets testing as P type?
                                    Working on a smps.
                                    Using Fluke 115. Red probe in red socket. Black probe in black/common socket. Meter set to diode mode.
                                    Removed a pair of mosfets marked 1L02AB FDP
                                    26N40 from the circuit.
                                    The pair of Mosfets are still screwed to the single heatsink. Mosfet printing facing me.
                                    I did the test where I short the left and center pins. I put black probe on right pin and red probe on center pin. I get OL. While keeping black probe on right pin, I put red probe on left pin and get OL. After a few seconds I move red...
                                    09-14-2023, 04:22 PM
                                  • digitalDNA
                                    MOSFET maintains voltage drop across G and S only for 3 seconds
                                    by digitalDNA
                                    While trying to fix an SMPS, I hear a ticking sound.
                                    I've checked all capacitors, but none seem to be shorted.
                                    I then suspected an issue with the flyback transformer which has 3 leads on the primary (used in a push-pull configuration ?) and one of them, which is always connected to GND has a resistance reading in the order of Mega ohms with any of the other 2 leads on the primary.
                                    But I thought let's check the MOSFETs first::
                                    The circuit also contains a PFC circuit, controlled by a UCC3818D (D stands for the SMD version),...
                                    10-11-2022, 11:00 AM
                                  • bauto601
                                    KDM 480W PSU upgrade succes! (opinions?)
                                    by bauto601
                                    Back in 2020 i bought a (2nd hand) compact ATX psu to replace my current one. My PC case only fits very short ATX power supplies so i didn't really have the choice of buying a decent 80Plus unit. The current unit is a modded YoungYear unit that i made a thread about a while ago:
                                    A "nice" YoungYear unit? - Badcaps
                                    The "new" unit is a KDM-M6480 480W psu, the 480W number is a typical KDM bullshit claim of course, but the 24A rating on the 12V rail seemed reasonable and the "Active PFC" claim gave me a bit of hope that this was going to be decent-ish....
                                    12-26-2023, 03:05 PM
                                  • dkneyle
                                    Onkyo DX-V500 Laser Disc Player - weird SMPS fault
                                    by dkneyle
                                    This relic from a technology path that came to a dead end (but still has historical interest) was brought in with a dead SMPS, which is based around a MK1210 device. I got it going after a battle. It seems to have a strange design - if it's not getting feedback that the secondary is OK it cause the big FET to go into full conduction and overloads resulting in a blown fuse at best or something worse! I'm using a Dim Bulb Tester (70W/230V) which thankfully has saved the day a number of times whilst repairing.
                                    Now, it loads a CD OK. The motor voltage output (which is built onto the SMPS...
                                    02-01-2022, 02:08 AM
                                  • shamsudeen
                                    Hi, Can we use SMPS Bench Power Supply for Laptop troubleshooting?
                                    by shamsudeen
                                    Hi all, I just need to know that, can i use SMPS type power supply to test and short circuit detection on a laptop motherboard?
                                    I saw 95% of repair guys are using only Linear bench power supply. linear bench power supply is pricy, thats why i planned to get one SMPS supply.
                                    i know that in linear power supply the noise will be very low comparing to SMPS supply....
                                    07-19-2025, 07:14 AM
