    Siemens D1052 oscilloscope (Gould DSO 4096) repair

    Hi to the best forum on electronics I know of...

    I need help on final step in repairing+restoring oscilloscope mentioned in the title
    (for which i think that nobody can find service manual)
    From start to finish I have done:
    cleaned a 2.4v NICD battery leakage across board with Hydrochloric acid, baking soda for neutralization, and finished with ethyl alcohol. Board does look like new now. Washed with water in between steps.
    Replaced what I have thought all Rifa capacitors with x and y class pp.
    Replaced output capacitors 2x1000uf35v + (1x470uF 25v - this one was manufactured with around double the capacity so I have replaced it with 1000uf one).
    Replaced both buk455-200a with equivalents *one was shorting out because of leaking capacitors
    Replaced BIOS battery with NIMH one.
    Replaced psu fuse.
    Had changed one small resistor as it seemed like it was acting up 0.5w 270k.
    Replaced 3 more capacitors in psu.
    Big blue Philips caps are left in psu as the esr meter showed good values on them.
    Magic smoke came again with Rifa capacitors inside mains ac filter plug.
    Bought a new Arcotronics filter. Feather light but expensive where I live, still wondering what quality is inside...

    Now comes the "good" stuff...

    PSU should give 3 main voltages +5v +16v -5v
    I measured voltages when unit is under normal load and there is a diference in outputs, +16.7v , +5.3v and -4.7v...Now that was strange for me so I checked for resistance between -5v and 0v on unit, and I think that I have found a problem IT IS 10.7 ohms.
    To me that looks like a short somewhere in the unit so I removed one by one plug in cards.
    One when removed raised resistance by huge amount so i found on her what I think is shorted -5v rail to ground or should I say 0v with 10.7 ohms.

    Well now mine question is this,

    1.May I apply higher current on this slot card than 1A with nominal positive voltage with my regulated power supply (which does not have negative output) may I detect some heat on shorted component with my thermal camera or I should not do that until I acquire negative output power supply.

    2. what are mine options with finishing this unit except to through it in garbage.
    Attached Files
    Last edited by Tamo Neko; 12-10-2020, 01:35 PM.
  • Per Hansson
    Super Moderator
    • Jul 2005
    • 5895
    • Sweden

    #2
    Re: Siemens D1052 oscilloscope (Gould DSO 4096) repair

    10.7Ω at -5v = 0.47A load, that seems normal to me for a board like that.
    You did not explain what the problem with the unit is?
    "The one who says it cannot be done should never interrupt the one who is doing it."

    Comment

    • Tamo Neko
      Member
      • May 2019
      • 47
      • Serbia

      #3
      Re: Siemens D1052 oscilloscope (Gould DSO 4096) repair

      Only voltage differences seems out, which I have to guess with the lack of schematics with nominal voltage tolerances. Other lookalike boards on this particular unit have 10 times higher resistance.

      Voltages imprinted on boards are not the same as measured...
      I have no idea of nominal psu voltages on theese 90s scopes.

      And Unit menages to auto calibrate itself this way.Seems to work. This is my first oscilloscope I acquired for learning how to work with them in general.
      Last edited by Tamo Neko; 12-10-2020, 03:30 PM.

      Comment

      • RJARRRPCGP
        Badcaps Legend
        • Jul 2004
        • 6301
        • USA

        #4
        Re: Siemens D1052 oscilloscope (Gould DSO 4096) repair

        Originally posted by Tamo Neko
        Feather light but expensive where I live, still wondering what quality is inside...
        Reminds me of overpriced LED light bulbs in some hardware stores, LOL.

        Pick some of them up, and it feels like zero heatsinking!
        ASRock B550 PG Velocita

        Ryzen 9 "Vermeer" 5900X

        32 GB G.Skill RipJaws V F4-3200C16D-32GVR

        Arc A770 16 GB

        eVGA Supernova G3 750W

        Western Digital Black SN850 1TB NVMe SSD

        Alienware AW3423DWF OLED




        "¡Me encanta "Me Encanta o Enlistarlo con Hilary Farr!" -Mí mismo

        "There's nothing more unattractive than a chick smoking a cigarette" -Topcat

        "Today's lesson in pissivity comes in the form of a ziplock baggie full of GPU extension brackets & hardware that for the last ~3 years have been on my bench, always in my way, getting moved around constantly....and yesterday I found myself in need of them....and the bastards are now nowhere to be found! Motherfracker!!" -Topcat

        "did I see a chair fly? I think I did! Time for popcorn!" -ratdude747

        Comment

        • Tamo Neko
          Member
          • May 2019
          • 47
          • Serbia

          #5
          Re: Siemens D1052 oscilloscope (Gould DSO 4096) repair

          I will post updates in a few days, as the shops for electronic components are not working because of covid situation....
          I have made a list to buy and replace all electrostatics on these plugin boards.

          Comment

          • Tamo Neko
            Member
            • May 2019
            • 47
            • Serbia

            #6
            Re: Siemens D1052 oscilloscope (Gould DSO 4096) repair

            Per Hanson is quite right. That board is working fine...Thanks Per Hanson for reply...
            I have just finished replacing many axial capacitors that had high vloss just in case.
            The unit manages to calibrate itself fine but I do have one major concern.

            -5v line is still low.*so i have to adjust main pot line for -5v to reach its nominal values...
            That gives me 18 volts on 16v line
            Now I did measure transistor Q1 even after replacment on psu, having a hard time, overheating up to 100 degrees Celsius with no load and 70 degrees Celsius with all loads attached.
            It was on "Q1" bf393 which is now replaced with mpsa42.
            Now Gould 4072-4074 psu units have if not the same circuits.

            So please can someone who has the 4070s manual tell me what does it say about this Q1 NPN transistor. Guideline for what to look for would be also very appreciated.
            Attached Files

            Comment

            Related Topics

            Collapse

            • dragometin
              Hitachi 40HYC42
              by dragometin
              TV Model: Hitachi 40HYC42
              Power Supply: 17IPS71
              MainBoard: 17MB95M
              Panel: VES400UNDS-2D-NO3

              When the set came in it was dead. Fuse was open primary side and the power mosfet was shorted. So I ordered a new power supply. Now it turns on, only displays "Hitachi" and then power off red and blue led flashing.

              I had to modify the 0ohm resistors in the led backlight of the new PS like the original power supply.
              Is there any variants on 17IPS71 power supply's?

              UART output pins 10 and 12 on scart connector:

              Code:
              12:24:04.595
              ...
              08-08-2024, 05:53 AM
            • Tarot Superstars
              How not to Damage an Oscilloscope
              by Tarot Superstars
              Hello.
              I have jut got my first analogue oscilloscope. An Isotech 620 20MHz.

              I want to learn how not to damage it, and I found instructions on the internet obscure.
              They talk about ground, earth, floating ground, this, that....
              But they don't actually say anything anything useful so far.

              I want to know:
              Don't connect exactly this to exactly there or it blows up.

              For example, I am tracing a signal from a function generator in an audio amp, or testing for voltages in different parts of a circuit.
              I want to know what not to connect...
              08-29-2023, 02:27 PM
            • MicroSMD Lab
              MSI MS-16W1 no power, EC replaced by other.
              by MicroSMD Lab
              Hi, I'm working on a MSI laptop, Motherboard number MS-16W11, GF-65 Thin 10SDR-459US. Went to another shop because the screen was flickering, but it was working well on an external monitor. The LCD was ordered and the laptop was returned to the customer and it no longer worked. It was then sent to another tech (by the shop who ordered the LCD) and the EC (ENE KB9028Q) was replaced with a donor from another MSI board with different board number and no programming was done. It looks like work was done on PU6 (TPS51225) as well as one of the GPU mosfets was burnt (I removed all three GPU mosfets)....
              03-27-2025, 07:35 AM
            • Evets
              Dynamco analogue oscilloscope
              by Evets
              I have recently acquired a Dynamco D7100 analogue oscilloscope without a user manual. As this is the first time I have owned or used an oscilloscope I would like to try and find an instruction book. The manufacturers no longer hold any information and a Google search only found one site but when I downloaded a 'service manual' all I got was gobbledegook. Any help would be gratefully appreciated or even pointers to a generic manual where I could learn more.
              08-30-2024, 06:54 AM
            • thesame
              Asus FX505DD REV 2.0 no power after someone replaced the cmos battery with a screwdriver.
              by thesame
              I have a dead FX505D, motherboard FX505DD REV 2.0 , the owner replaced the cmos using a screwdriver and he touched something on the board, since then the laptop its not starting, i have managed to see that U8800 (BQ24780SRUYR) charging IC is getting hot under a thermal camera , i have desoldered it but the laptop still doesn't power on, i also have no voltage at power connector on the board. I also replaced the U8800 but still no power. What should i check next?
              04-22-2024, 01:21 AM
            • Loading...
            • No more items.
            Working...