Dressing up (a "450W" ATX supply) for Halloween -- YueLin YLP-013
  • needtoseepics
    • Oct 2013
    • 11
    So I have a supply labeled "ATX Switching power supply MODEL:450W",
    which is quite suspicious sounding already, but worse still is that
    there doesn't seem to be anyone claiming to be the manufacturer.
    There is a sticker labeled "200808023681 WP08023 Made In China",
    which seems to be the most informative indication on the outside.
    Input/output specs are listed however, and this is how they read:
    | VOLTAGE | CURRENT | FREQUENCY
    V~ | 115V~ | 7.5A | 60Hz
    INPUT | 230V~ | 4.5A | 50Hz
    ___ |+5V|+3.3V|+12V|-5V |-12V|+5VSB|PS-ON |POL | COM
    V--- |30A| 25A | 25A|0.6A|0.6A| 2A |REMOTE|P.G.|RETURN
    OUTPUT |
    | 180W |300W|3.0W|7.2W| 10W | -- | -- | --
    MAX | 430W | 20W |
    | 450W |
    (that looked so much better in vi...)
    Inside, what immediately catches my eye is the lack of input
    filtering, and the jumper wires where the coils should be, and the
    use of Canicon and Micon capacitors throughout. The PCB is labeled
    YUELIN YLP-013 PCB REV:2.2 070829.
    I got some Nichicon PL(M) and LQ(M) capacitors out of a blown Lucent
    623W supply (from an SGI Octane -- replaced with Cherokee 747W), so
    some of them are, how should we say, a bit overspec'ed for this
    supply, but, everybody seems to claim "you can go up in volts", and
    "you can go up in capacitance", so, that's what I'll be doing, and
    if I have to do point-to-point wiring off the PCB, then so be it.
    I salvaged a two 3300pF caps, and installed them in the spaces
    labeled "CX1" and "CX2" on the input stage, as well as replacing the
    jumper wires for "LF1", "J1", and "J2", with toroidal inductors
    which I'd obtained from a blown "Wan Nien/Enhance Electronics"
    supply. Also used was the input resistor, coded Green, Brown,
    Yellow, Gold, which looks up to mean 560k Ohm, 5% tolerance, which I
    fitted to the blank spot labeled "R1".
    Because the AC line input from the C14 inlet was using such flimsy
    wiring, I went ahead and upgraded the wiring, choosing to install
    two short leads to the L(ine) and N(eutral) inputs on the PCB and
    wire nut them to the wires from the inlet for easy removal (I would
    have used a connector, but the original of either ATX unit didn't
    have any, and sifting through my parts I only found a plug, and not
    the onboard header, so wire nuts it is -- no way am I gonna solder
    direct-to-the-plug like it was originally).
    The original bridge rectifier, labeled KBL406, is a 4Amp unit, which
    is right on the edge of the specifications: if you take 110V * 4Amp,
    you get 440Watts, so obviously they intend you to use 115V * 4Amp,
    in order to reach a "450Watts safe" 460Watt rated limit for that
    part. A scrap supply I had used a KBU605, slightly better at 6Amps
    rating, which, using their 115V * 6Amp formula yields 690Watts (660
    if using 110V), and backing off from it by 10Watts like they did,
    this particular component ought to be good for 680Watts of output
    (supposing I'm properly guessing the way to rate/use these things).
    Thankfully, it is a model with a hole in it, which allows for
    heatsinking, whereas the original 4Amp part wouldn't have been as
    easy to attach a heatsink to. I haven't tried this yet, but this
    part is next on my list to do. Question... the board I sourced this
    part in question from had burned/melted input switchers, as well as
    a charred area around the board nearby (which includes the bridge
    rectifier). Supposing this 6Amp bridge rectifier is still functioning,
    is it safe for extended use? It doesn't seem melted like the
    switchers, so I assume this is just like a "load test", and if it's
    still working, then no issue.
    I also have two international rectifier 30CPQ150's that I might use
    as an ouput-stage upgrade.
    At the moment, I've made a small perfboard-with-output-caps, and
    wired the Nichicons to the output holes using stranded speaker wire,
    which was thicker than any of the other wire I had available. I was
    thinking about putting some high-frequency ceramic or such capacitors
    on the output lines on the supply's PCB to "make up for" hosing the
    electrolytics off-board, but I haven't done this yet.
    The 8x output electrolytics I placed in have the values:
    3.3V: 1000uF, 2200uF
    12V: 1000uF, 2200uF
    5V: 1000uF, 6800uF
    -5V: 1000uF
    -12V: 1000uF
    Except for the 6800uF, which is rated 6.3V, the rest are rated for
    35V, and all are Nichicon PL(M) series with 105 degree C temp ratings.
    So right now, just the recap, input filter, fan connector, and AC-in
    wire thickness upgrade has been done. The fan that is in it does work,
    but I'll replace it before I'm finished. I tested with a single hard
    disk for load, and measured the following with a digital multimeter:
    3.3V output: 3.42-3.43
    5V output: 5.20-5.21
    12V output: 12.39-12.40
    -12V output: -11.69 - -11.70
    -5V output: -4.92 - -4.93
    I did take some pictures along the way to where I'm at now, so I'll
    be posting those soon, just have to sift out what's visible, and what's
    fuzzy, etc.
    Feel free to critisize this.. it's my first attempt at a power supply rebuild,
    and I'm especially interested in comments about off-boarding the output
    filter caps like I've done. I know it does increase their impedence, but..
    how much? Would some smaller caps right on the output lines overcome
    this? Also, I know the "X caps" that I've placed are rated as "Y caps",
    so not exactly correct.. but they are behind a fuse, so even if they fail
    closed (correct model fails open), that will just pop the fuse. Isn't this
    better than not having anything fitted there? I have a proper X2 cap,
    which I was thinking to mount on the power inlet, because it is too big
    to fit the PCB. If I do that, will it be a problem to leave the PCB as I've
    got it now, or should I remove the "Y caps" from the "X cap places" if
    I put a proper X cap on the power inlet.
    Note... it seems like a user here has done something similar before,
    so I'll leave this link here, so you know I'm aware of it (that's the thread
    that I "needed to see pics of" and why I created this account:
    https://www.badcaps.net/forum/showthread.php?t=7691
  • needtoseepics
    • Oct 2013
    • 11
    Re: Dressing up (a "450W" ATX supply) for Halloween -- YueLin YLP-013
    Here are the before pics, where I'd just taken the PCB out of the housing.
    3x from the top, and 3x from the side. I can't take anymore, 'cause it doesn't
    look like this anymore (more pics to follow):
    • needtoseepics
      • Oct 2013
      • 11
      I got the primary input filter from this Wan Nien Enhance Electronics supply,
      and this is where I also got that 6Amp bridge rectifier that I was considering
      installing in the YueLin. See what I mean about the pcb near the rectifier itself
      and the input coils all seemed fine, but obviously they've had quite the "burn in"
      load test (pun intended).
      The capacitors on this board are Jamicon and Teapo. The Jamicon seem fine,
      but the Teapo's were bulgy. Is Jamicon better or worse than Canicon? No idea.
      I removed all the Micon caps from the YueLin, but there are still some Canicons
      on it... would these Jamicons from this blown Wan Nien be better than Canicon?
      Last edited by needtoseepics; 10-15-2013, 08:26 AM. Reason: added paragraph about the capacitors
      • needtoseepics
        • Oct 2013
        • 11
        Here is the input filter stage installed on the YueLin, with also two "Y rated caps"
        in the X cap locations (is this a bad idea.. let me know if you think so). These Y caps
        came from the SGI's Lucent supply. Also you can see in this image the thicker
        input wires I've installed to the PCB.
        • needtoseepics
          • Oct 2013
          • 11
          Here you may see the recap itself, with one additional point.. when I took that
          decapped output stage pic, I had only removed the micons, and not yet the
          canicons from the output stage. I ended up putting those two "Low ESR"
          Canicon 470uF 16V to replace two Micon 470uF 10V on the 5VSB circuit.
          I don't know if that's a good idea or not, but I figured I'd remove Micon entirely,
          and so this leaves me with only Canicon (presumably not-the-best) and
          Nichicon (from a Lucent supply SGI used in 1997ish, so not the newsest,
          but also, not fake like apparently we have to worry about when buying "new"),
          with only Canicon on the main PCB, and all the Nichicon on the perfboard.
          (perhaps I'll replace the canicons with out-of-spec other parts, or eventually
          buy new (hopefully not fake) exact-spec replacements, but I'm not done yet,
          and just doing output stage here).
          So this is where I made it to, and it does power on, and that's how it looked
          when I measured the voltages above (in initial thread post). I think the next
          thing to do would be replacing the silicon parts (bridge rectifier 4A -> 6A first),
          but I was thinking I should probably hit it with a scope first to get an idea
          of what's going on. So might be awhile before I return to this thread.
          (hint: this is where the comments should happen)
          • needtoseepics
            • Oct 2013
            • 11
            Alright, this morning I did a scope test using an HP 16500B. For a light load,
            I attached one hard disk, and two additional fans to the YueLin YLP-013.
            I took two shots each with the scope set at 1ms, and 10ms. The A1 channel
            is connected to the 5V output line, and the A2 channel the 12V output line.
            Each channel is offset/centered by/on it's respective "correct" voltage,
            5V for A1, and 12V for A2, and the scale is set to 250mV, so each is slightly
            under "1 tick" away from center.
            I notice nobody's commented yet... perhaps this will spawn some replies.
            If there is another test you'd recommend I do to have more information, let
            me know, and if I can try it, I will. I think the results here look good, so I'll
            likely hook this up to a pc later today and see how it goes, then perhaps try
            my bridge rectifier upgrade to the 6Amp heatsinkable model. We'll see.
            • ben7
              Capaholic
              • Jan 2011
              • 4059
              You really should put in better secondary rectifier diodes, better primary side switching transistors, and possibly a better transformer.
              I personally don't like the idea of mounting the capacitors like that, but, whatever works for you ... xD
              Muh-soggy-knee
              • needtoseepics
                • Oct 2013
                • 11
                Alright, I've picked up a few more "bad" supplies, and I suspect I got
                lucky with regards to a compatible transformer, and NPN power
                transistors. Now, the YueLin's original primary switching parts are
                labeled:
                0107
                P13009-0
                And although I don't know who makes them, I notice that there is an
                FJP13009 "High Voltage Fast-Switching NPN Power Transistor" from
                Fairchild Semi. which these are likely similar to.
                The PCB has wider drilled spacing than what is used, so it will fit the
                TO3P package, but the stock part is the TO-220 package, so I ought to
                be able to use any NPN power transistor in either of those packages,
                and, presumably, if for no other reason than heat-dissipation due to
                increased external surface area, I should consider parts in the TO3P
                package to be "better". Now, one of the AT supplies I've got has two
                parts labeled SKC4138 in the TO3P package, which seems to be SanKen
                brand, and according to:
                https://cdn.badcaps-static.com/pdfs/...6fe02e2b70.pdf
                on page 5, it says C4138 is good for 500Vcbo, 400Vceo, and 10Amps.
                Of course, since I don't know the specs of the original part, I don't know
                if that's better or worse, but it is in the larger package size, so,..upgrade?
                Also, the SGI/Lucent supply contains 2x International Rectifier 541L2
                parts. I'm not actually sure if these are even compatible, or even if
                they are NPN transistors or not, but they are in a TO3P package, and
                it's I suppose possible they are compatible. I'm of course more
                likely to fit the SK parts since I at least have some information
                about them, and coming from an AT supply it's easy to tell they were
                being used for the same thing, whereas with the 3x pcb's inside this
                Lucent supply, it's hard to tell what's going on where, but it's possible
                that the IR parts are better than the SK and could be used. Any idea?
                With regard to the output stage, I think these International Rectifier
                30CPQ150 parts I harvested out of the Lucent supply (yeah, they used
                IR all over the place in it) will work as "Schottky Rectifier", I looked up the
                original parts, the MOSPEC F12C20, and ?PANJIT? SB3040ST, and
                it seemed like a match, with the IR part having an improve rating.
                Also... like I said, that Lucent supply has several PCB's and uses ribbon
                cable and thicker "power cable" connecting them, so I guess that's why
                I started thinking "maybe it's okay to mount capacitors on a different PCB
                from the control circuitry", since Lucent is reputable (though this particular
                supply is dead, hmm... red flag?), and this supply they did that way is
                rated for 623W, and the physical dimensions of it are massive compared
                to those of a standard "box" sized AT/ATX supply.
                @ben7 --
                Thanks for the response by the way, but now I feel the need to ask:
                Is your reason to dislike this mounting method technical in nature
                (because of increased impedence, or extra heat now that the transformer,
                heatsinks, and, additionally these caps, are all being in the same place),
                purely cosmetic, fire hazard or what?
                (now I'll have a lot of desoldering to do before I can continue, but it seems
                like I ought to have enough usable parts in this ever-increasing scrap heap
                of power supplies to go through a few more stages of this dress-up)
                • kc8adu
                  • Nov 2003
                  • 8832
                  • U.S.A!
                  those long thin leads totally defeat the purpose of low esr caps.
                  just get proper caps and install on the board.panasonic fm are a good choice.
                  • needtoseepics
                    • Oct 2013
                    • 11
                    So here are the parts I have harvested so far, and might be good candidates for
                    upgrade use. I don't really know how to tell if one transformer is better than another
                    one, but the size of the one I've extracted, labeled DVE E401201-014, is slightly
                    larger than the one that's installed by default in the YueLin. The pinout is at least
                    two pins on the input side, and six pins on the output side, with the "out-the-top"
                    center tap to ground. The spacing on the two input pins is a bit closer together
                    I thnk than the one that is installed by default, which itself is labeled El-33 4.5mH EL.
                    Supposing I can "make it fit", I'm guessing it's worth a try and would be considered
                    an upgrade?
                    The schottky rectifiers I collected are the (above mentioned) International Rectifier
                    30CPQ150 units. I also collected these International Rectifier 541L2 parts which
                    I couldn't find any information on. I'm hoping they are switching transistors, does
                    anybody know for certain? If not, I do have those SK C4138 which I still haven't
                    desoldered yet, so, I'll collect them if nobody knows what these 541L2 parts are.
                    And, as I mentioned before, I've got this KBU 605 6Amp bridge rectifier, which I'll
                    be testing first because I don't know if it's good or not (you can see where the
                    char marks are on its legs perhaps in the photo) since the supply I pulled it from
                    had a burned up input stage. I did pull the stock 4Amp bridge rectifier, A KBL406,
                    off of the YueLin by now, but I haven't put this one on (need to clean the pins off
                    a bit more I think and bend them about before I can get it to fit in)
                    So, that's where I'm at so far.
                    @kc8adu, how low is the resistance of low-esr caps, because if I calculated it right
                    (I used the "Wire Resistance Calculator" from cirrus, found here:
                    http://www.cirris.com/testing/resistance/wire.html
                    ) 22AWG wire for 1.2ft (7in there, 7in back) ought to be adding 0.019 Ohms of
                    resistance, in series obviously, to the capacitor. So, is this a considerable amount
                    for these components? Everything I notice seems to say "Low ESR", but.. low,
                    compared to what? I haven't seen any hard numbers like, for instance, is 1 Ohm
                    considered "High ESR", is 0.1 Ohm High ESR? If 0.019 Ohms is enough to push
                    a Low-ESR component out of that status, then I suppose that must be the case.
                    At what specific numerical value does a component become "Low ESR", and, if
                    I know that, then I could add 0.019 to the ESR of such a typical component,
                    and see if it is still low or not.
                    Not to say I wouldn't buy "the right" parts, but even if I do, with shipping and all
                    it kinda doesn't let me do anything for a while, and at least at the moment I can
                    test the other upgrades, and the supply seems usable, and within ATX spec.
                    From what I was reading, ATX spec says 50mV ripple is allowed/permitted, and
                    the voltages must be within 5% of the "correct value", so, looking at my scope
                    results, I'm very much "within spec", at least for this simple test (which I realize
                    is not representative of an actual computer system with activity), so I don't really
                    see a problem, yet. If you can propose a test that will demonstrate that my supply
                    has a technical issue caused by this layout, I'll do my best to carry it out, and
                    perhaps adapt things to work around the issue.
                    If not, I'm gonna assume you all just have an issue with the cosmetics of an
                    additional PCB
                    • kaboom
                      "Oh, Grouchy!"
                      • Jan 2011
                      • 2507
                      It's not going to filter nearly as well.
                      The common mode chokes need a low common mode impedance where those X caps normally go. The Y caps are only a few thousand picofarads at most = higher Xc = higher common mode impedance. The chokes then won't "dominate" with their Xl, resulting in more RFI leakage from the supply.
                      Use caps that are at least .1u, preferably .22u.
                      ^This.
                      You've got minimal load on there in your test. The output inductor (after secondary rectifiers, but before filter caps) and filter caps have minimal ripple- duty cycle is way down, ripple is (normally) relatively low.
                      Some large resistors, for 5 and 3, and a few bulbs for 12, will really test it. Don't push it too far, with those wimpy heatsinks. Too thin, and too little surface area, to adequately cool those devices.
                      "pokemon go... to hell!"
                      EOL it...
                      All style and no substance.
                      guilty,guilty,guilty,guilty!
                      guilty of being cheap-made!
                      • LLLlllou
                        • May 2011
                        • 201
                        That's analogous to asking "what tastes better, cat turds or dog turds?"
                        • cheapie
                          null
                          • Jul 2010
                          • 849
                          see a problem, yet.
                          You're actually being harder on it than the spec says. (Yes, I sat down one day and read the ATX12V (version 2.2) power supply specs. I was bored, okay?) The spec calls for some amount of capacitance at the load (10uF lytic and 0.1uF ceramic caps), and allows 120mVp-p ripple on +/-12V. See section 3.2.6.
                          • Pentium4
                            CapXon Be Gone
                            • Sep 2011
                            • 3741
                            To be fair, I'd rate Jamicon quite a bit higher than Canicon. But I still wouldn't use them in a power supply.
                            I'd install the filtering caps traditionally, but it looks like you have a nice project going here. Good job man!
                            • Th3_uN1Qu3
                              Believe in
                              • Jul 2010
                              • 6031
                              • Romania
                              Jamicon capacitors are decent for general purpose applications. They are completely unsuitable for use in a SMPS.
                              A working TV? How boring!
                              • kc8adu
                                • Nov 2003
                                • 8832
                                • U.S.A!
                                looks like several psu makers missed that memo!
                                im looking at you bestec!
                                • Pentium4
                                  CapXon Be Gone
                                  • Sep 2011
                                  • 3741
                                  Yeah, Bestec loved using GP Jamicon caps. How did they manage to keep the ripple low then with non low ESR caps? Maybe the PI coils did most of the work?
                                  • Wester547
                                    • Nov 2011
                                    • 1268
                                    • USA.
                                    ^ Well, for the +3.3V rail, Bestec seems to use a unique design that has one ferrite coil after the first (input) capacitor and another after the output capacitor, which seems to reduce the ripple voltage even more. That being said, the toroid for the +3.3V has a bit of a different method than the group regulated +5V/+12V toroid you see in most PSUs. It uses a saturated coil which means less squandered power and better crossloading. As for why Bestec get away with GP capacitors from Jamicon and Teapo and still manage to keep the ripple voltage low? Well, the toroidal coils and ferrite coils do a huge part of the filtering (unlike in flyback topology which lacks the toroids) so there isn't actually a tremendous difference between low ESR and GP capacitors in the output of a SMPS (not that I would recommend using GP capacitors at all, but extremely low ESR capacitors aren't necessarily a benefit either). The use of -52 material instead of -26 material seen in older PSUs also suggests a higher switching frequency which also means you can get away with smaller components and inductors. It also depends heavily on how well the feedback loop is designed (some designs are so bad that they have the feedback after the output filter...) and how well the PSU is cooled and ventilated, of course.
                                    • Pentium4
                                      CapXon Be Gone
                                      • Sep 2011
                                      • 3741
                                      Interesting. Too bad Bestec got such a bad reputation from the 250-12E because they make some pretty good power supplies. I may try an experiment recapping a Bestec with Rubycon PX to see if it can keep ripple maintained with the higher quality GP caps
                                      • needtoseepics
                                        • Oct 2013
                                        • 11
                                        I found two (different) X2 capacitors, one labed IEC384-14II
                                        55/100/56X2 250V~, then on the backside (frontside?) it's labeled
                                        OKAYA RE224 0.22uF K 275V~ iP, and black in colour. The other one
                                        is labeled MEX/TSC MKT GPF .1 M 250V~X2 40/85/21 211, then in the
                                        bottom left is 565-1, blank on back side, and orange in colour.
                                        Due to spacing, I've placed the shorter (orange) one by the bridge
                                        rectifier, and the slightly taller one (black) one closer to the
                                        input fuse. I assume these are a better match than the Y caps I had
                                        before.
                                        I've installed the 6Amp bridge rectifier, and replaced the Canicon
                                        470uF 200V 85'C caps with identically equivalent Rubycon capacitors,
                                        labeled: 200v470uF(M) NEG. +85'C USP. These are of course used, and
                                        harvested out of (yet another) junk supply. But presumably used
                                        Rubycon is better than used Canicon?
                                        Also I swapped out the primary-stage switching transistors with the
                                        SK C4138 parts, which are in the larger package size, and the
                                        secondary-stage schottky barriers with the international rectifier
                                        30CPQ150 parts, with two of the IR parts on the right, and the one
                                        original part that was the larger package size went to the far left,
                                        as was done in that other thread.
                                        I took pictures of this process, which I'll attach here, but then
                                        after taking the pictures, I did some more stuff that I don't yet
                                        have pictures of. Specifically, I also swapped the Y caps for some
                                        larger blue ones I pulled from another supply, and recased into a
                                        Sparkle Power box which has a much larger vent area, and used the
                                        fan from the Sparkle since it is in much better condition (The
                                        sparkle had a big burn mark on the pcb and apparently it got hot
                                        enough to desolder one of the large resistors which the burn
                                        is centered around, 'cause it just fell out and was bouncing around
                                        in the box, hmm.. no wonder it didn't work). I went ahead and soldered
                                        an A/C power input connector to the YueLin board that I harvested from
                                        (yet another) supply, and put the appropriate plug on the switch in
                                        the back, along with a Teapo X2 cap direct on the inlet, like I've
                                        seen posts on here with people doing that -- is this a problem since I
                                        now have 3x X2 caps? two on the board, and one on the inlet, or is
                                        this fine to do?
                                        I ran the reboxed YueLin overnight with a single harddisk load, and
                                        with no issue in the morning, I put it in a machine (Athlon X2 4200,
                                        4GB, Quadro FX 3500, Broadcom GigE), and ran it for a day in Linux
                                        with glxgears running. Didn't seem to have any trouble, so I suppose
                                        at this point I can think again about the transformer, or possibly
                                        ordering some correct output filter caps. That transformer I have
                                        harvested didn't have its input pins in the same spots as the holes
                                        are drilled on the YueLin pcb. I assume the two input pins must be
                                        "the same function" as the two holes on the pcb provide, so, if I
                                        mount it at an angle, with the output pins in the holes, and run
                                        some wiring to the transformer's input pins from the pcb holes,
                                        should I expect this would still be an improvement? At least if
                                        I do it now, I know it worked before, whereas if I had done it at
                                        the same time as all the other stuff and it hadn't have worked, I
                                        wouldn't have known how to assign blame.
                                        I might be ordering some "proper" output capacitors, but I guess
                                        I need to do more reading/pricing before I make a decision on what
                                        exactly to get. Whenever I open it again, I'll take pics of how it's
                                        looking now with the recase, etc.
                                        I was gonna try that whole "lamp" suggestion, but I'm just wondering
                                        are we talking 60Watt light bulbs on the 12Volts line, or what does
                                        "lamp" mean? Obviously I didn't try it yet, and just stuck it in a
                                        computer system for a load testing (and didn't scope it since the
                                        silicon upgrade yet).
                                          A Dead/DOA Gigabyte GP-P450b 450W ATX PSU
                                          by momaka
                                          More eBay specials: this time a Gigabyte GP-P450b 450W ATX PSU for $8 with free S&H. Listed “For part/not working”, as usual. Showed up in a beat up EVGA box without any padding… but that's understandable given the total price. PSU made it out fine without damage. It's a non-modular (yay!), pretty modern-day looking PSU with a black paint and all black cables (meh, not my favorite, but figure it's good to keep a few of these in case I do need one for a more modern PC build, just so my old gray boxes don't look too odd in there.)
                                          And a shot of the label:...
                                          01-26-2022, 03:26 AM
                                        • AdrenalinTL
                                          450W PSU capacitor Identification and where to get another one?
                                          by AdrenalinTL
                                          Howdy peeps.
                                          Can someone tell me where to get another one of these? google doesn't seem to have a 560uF and I am still quite a newbie to the whole world of electronics so trying to learn as I go. The one to the right of this pic is definitely swollen. PSU does random shut downs / no power on etc
                                          Also, can it be replaced with a higher uF cap ? Same voltage?
                                          Thanks...
                                          12-11-2024, 02:20 AM
                                        • harp
                                          Learning with ATX CHIEFTEC 450W PSU
                                          by harp
                                          I have a little question regardless a way to repair ATX power supply CHIEFTEC MODEL: GPS-450AA-101 A. It is 450w power for computer.
                                          The psu has initial short on +5VSB (purple), and I found and replace shorted diode SB340. After replace, unpowered measurement of all outputs seems fine. All other big double diode on heatsink are ok in secondary side. I connect psu with 60w bulb in serial, and bulb shortley flash as it would be. There is steady 5.2v on +5VSB and PS-ON, on any other is nothing, like 0.00v. So the primary is ok, right?
                                          But, when I PS-ON (green) connect, bulb...
                                          06-17-2022, 05:04 PM
                                        • ntoskrnl
                                          [FIXED] XFX Core Edition Pro 450W ticking noise
                                          by ntoskrnl
                                          My XFX Core Edition Pro 450W power supply started making a weird ticking noise some time ago. Eventually, I figured out what the issue was, and managed to fix it. I purchased this unit back in 2012, so it's an old model, but I'm still posting this in case someone finds it helpful or interesting.
                                          THE PROBLEM
                                          The power supply started making an irregular ticking noise when running, but would otherwise work perfectly fine. The issue would sometimes come and go. A Google search for "xfx core ticking" found multiple other people reporting the same issue, so I assume this...
                                          04-19-2022, 10:36 AM
                                        • Paar
                                          Corsair HX 450W shuts itself down after a while
                                          by Paar
                                          Hi. I have bought used Corasir HX 450W power supply and have an annoying problem. I have a PC case with PSU situated on the top of it. Whenever I install this PSU in it and close the cover the computer shuts down after a while. If I let the case open everything is fine. I'm using mATX AM3 motherboard with integrated graphics and am cooling the CPU with big Scythe cooler. With it the CPU idles at circa 40°C/104°F.
                                          I tried two different PSUs and none of them shows this symptom. The integrated GPU (in the northbridge) gets pretty hot but I'm sure it's not hot enough to overheat the...
                                          09-19-2020, 03:39 AM
