    Re: Fixing a laboratory waterbath

    There is continuity between the output of the zener and ground. Now I have to find why that is.

    Comment

    • vrasp
      Senior Member
      • Nov 2019
      • 194
      • Canada

      #162
      Re: Fixing a laboratory waterbath

      I fully disconnected the display board and the short is still there.

      I noticed that Vcc of LM324 and OP07CP are also on this line and so there are shorted as well.

      Now this is the part that I don't understand. It looks as though they are all purposely grounded. If you look at the board on the attachment of this post, I drew a red line over the circuit I m talking about. This line ends up on the bottom right corner where it is grounded (I desoldered the capacitor there so we can see the line). I must be wrong as it wouldn't make sense. But do you see this as well?
      Attached Files
      Last edited by vrasp; 06-26-2021, 02:42 PM.

      Comment

      • petehall347
        Badcaps Legend
        • Jan 2015
        • 4440
        • United Kingdom

        #163
        Re: Fixing a laboratory waterbath

        you forgot to solder this ..
        Attached Files

        Comment

        • vrasp
          Senior Member
          • Nov 2019
          • 194
          • Canada

          #164
          Re: Fixing a laboratory waterbath

          Originally posted by petehall347
          you forgot to solder this ..
          I m gonna add some solder on that. Do you see that the Vcc seems to be connected to ground or is it just me?

          Comment

          • vrasp
            Senior Member
            • Nov 2019
            • 194
            • Canada

            #165
            Re: Fixing a laboratory waterbath

            Sorry but this is really bugging me. Is this pin circled on the picture a through hole? If it is then Vcc is directly connected to ground.
            Attached Files

            Comment

            • petehall347
              Badcaps Legend
              • Jan 2015
              • 4440
              • United Kingdom

              #166
              Re: Fixing a laboratory waterbath

              this needs looking at .
              Attached Files

              Comment

              • vrasp
                Senior Member
                • Nov 2019
                • 194
                • Canada

                #167
                Re: Fixing a laboratory waterbath

                Originally posted by petehall347
                this needs looking at .
                Thanks. I did check those connections and they are isolated from adjacent ones. I will still redo them.

                The waterbath is in the lab so that's where I work on it now when I have time during the week. It's good that it works because if that display issue can't be fixed it can still be used at a set constant temperature which is likely how it would have been used anyways. I just need to put a temperature probe in there and find the right temperature with the potentiometer knob. However, it would still be nicer with a display.

                I m thinking of a way I could see if that solder in my previous post is a through-hole or just on the surface. There is a solder pad on the other side of the board on that same spot which I thought meant it's a through-hole. Why would they run that line all the way to that same spot if it's not a through hole? But then if it is a though-hole, it means Vcc it connected to ground... and that wouldnt make sense.

                Comment

                • vrasp
                  Senior Member
                  • Nov 2019
                  • 194
                  • Canada

                  #168
                  Re: Fixing a laboratory waterbath

                  I redid those solder joints today.

                  I m still stuck on the same problem. The trace of Vcc of LM324 goes to the metal corner of the board. I cant figure out why.

                  Comment

                  • vrasp
                    Senior Member
                    • Nov 2019
                    • 194
                    • Canada

                    #169
                    Re: Fixing a laboratory waterbath

                    Continuing the waterbath saga.

                    I was able to resolve the issues I had. Everything worked fine until it didn't anymore. The problem is that I replaced the long pins with wires and I think there is a lot of strain on the solder joints. I keep having issues with wires popping off the solder joint and causing shorts as it touches other parts. Especially when I run the waterbath at high temperatures for long periods of time. I might not have the best soldering skills. There is also not much space so the wires end up being bent and twisted when I out everything back together. I m going to add solder to the weakest joints and add hot glue to all of them. If you have suggestions for this plz let me know.

                    The last time it happened it shorted something on the power supply card and the relay switch is now grounded. That short blew a trace so it probably killed the few components on the line. I think the wire that popped touched ground and a pin from the relay switch at the same time.

                    I m going to replace the relay switch and the other components around it. My problem is the same as the very first issue I had when starting to fix this waterbath: there is no info on the relay. I dont know what the specs are and so dont know what to replace it with. All I know is that the relay is connected to a 0.68 uF 250v capacitor getting 120v from the wall, and a 800v 15 amps triac. I believe I need a non latching relay since it has to turn off when power is off (for safety).

                    The relay is on the bottom left of the photo attached.

                    How can I figure out what relay to buy for this context?

                    Thank you!
                    Attached Files
                    Last edited by vrasp; 07-14-2021, 09:30 PM.

                    Comment

                    • sam_sam_sam
                      Badcaps Legend
                      • Jul 2011
                      • 6069
                      • USA

                      #170
                      Re: Fixing a laboratory waterbath

                      Originally posted by vrasp
                      Continuing the waterbath saga.

                      I was able to resolve the issues I had. Everything worked fine until it didn't anymore. The problem is that I replaced the long pins with wires and I think there is a lot of strain on the solder joints. I keep having issues with wires popping off the solder joint and causing shorts as it touches other parts. Especially when I run the waterbath at high temperatures for long periods of time. I might not have the best soldering skills. There is also not much space so the wires end up being bent and twisted when I out everything back together. I m going to add solder to the weakest joints and add hot glue to all of them. If you have suggestions for this plz let me know.

                      The last time it happened it shorted something on the power supply card and the relay switch is now grounded. That short blew a trace so it probably killed the few components on the line. I think the wire that popped touched ground and a pin from the relay switch at the same time.

                      I m going to replace the relay switch and the other components around it. My problem is the same as the very first issue I had when starting to fix this waterbath: there is no info on the relay. I dont know what the specs are and so dont know what to replace it with. All I know is that the relay is connected to a 0.68 uF 250v capacitor getting 120v from the wall, and a 800v 15 amps triac. I believe I need a non latching relay since it has to turn off when power is off (for safety).

                      The relay is on the bottom left of the photo attached.

                      How can I figure out what relay to buy for this context?

                      Thank you!
                      Do you have a temperature controlled soldering station
                      If you do then turn the heat up until you have it where when you can put iron on the terminal and it melts the solder with in a few second not the trace pad you want to heat this last

                      Because it sounds like you are not getting the terminal connection hot enough

                      Comment

                      • vrasp
                        Senior Member
                        • Nov 2019
                        • 194
                        • Canada

                        #171
                        Re: Fixing a laboratory waterbath

                        I have a hakko soldering station I use at 300°C. I melt enough solder to allow the whole tip of the wire to be fully "submerged" in solder (but not too much either). The solder stays on the pad but sometimes, when i move the board around to put everything back together, a wire will pop off while the solder stays on the pad. I ve been leaving the iron a bit longer and it seems more steady now.

                        Any idea about the specs of the relay switch?

                        Thank you!

                        Comment

                        • sam_sam_sam
                          Badcaps Legend
                          • Jul 2011
                          • 6069
                          • USA

                          #172
                          Re: Fixing a laboratory waterbath

                          Ok if the pads are ripping off the board then you are putting to much force on the connector wiring which should be avoided if possible even if this means you have to extend the wiring somewhat or solder the connectors when everything is mounted

                          Or you are not getting the wire connector hot enough like I said earlier heat the connector and touch the solder to the wire connector not on the iron tip or you need to use flux on the wire connector if the solder is not staying on the connector
                          Last edited by sam_sam_sam; 07-16-2021, 04:41 AM.

                          Comment

                          • sam_sam_sam
                            Badcaps Legend
                            • Jul 2011
                            • 6069
                            • USA

                            #173
                            Re: Fixing a laboratory waterbath

                            For the relay unless there is a part number on the relay it is going to very difficult to know what will fit on the board or use some type of relay that you can find the relay socket and wire them to the board

                            Comment

                            • vrasp
                              Senior Member
                              • Nov 2019
                              • 194
                              • Canada

                              #174
                              Re: Fixing a laboratory waterbath

                              Originally posted by sam_sam_sam
                              For the relay unless there is a part number on the relay it is going to very difficult to know what will fit on the board or use some type of relay that you can find the relay socket and wire them to the board
                              The relay is 24 x21 mm. I think as long as I know what specs I need then I ll try to find something that fits. The part I have trouble with is all the relay specs I dont understand. For example, it s getting 120VAC. Should the relay therefore be rated at 240v minimum? The 120VAC side is connected to the 800v 16A triac. Should the relay be rated at 16A or higher as well? I think both answers to these questions are yes, it should. However I know electronics can be more complicated than that. Same with the coil voltage and other specs. How do I know what to get?

                              Thank you!
                              Attached Files

                              Comment

                              • vrasp
                                Senior Member
                                • Nov 2019
                                • 194
                                • Canada

                                #175
                                Re: Fixing a laboratory waterbath

                                Or if you cant figure out the specs from the info I provided. What tests can I run to get that information. Should I measure the incoming activating voltage? Or is it the current that matters in this case?
                                Last edited by vrasp; 07-16-2021, 08:45 AM.

                                Comment

                                • eccerr0r
                                  Solder Sloth
                                  • Nov 2012
                                  • 8703
                                  • USA

                                  #176
                                  Re: Fixing a laboratory waterbath

                                  Why are you assuming it's a relay? What evidence do you have?

                                  I'm beginning to think it's not even a relay ... and have no clue what it is...

                                  Comment

                                  • sam_sam_sam
                                    Badcaps Legend
                                    • Jul 2011
                                    • 6069
                                    • USA

                                    #177
                                    Re: Fixing a laboratory waterbath

                                    Show a picture of the bottom of this device
                                    I am not sure what this device is it could be several things

                                    Comment

                                    • vrasp
                                      Senior Member
                                      • Nov 2019
                                      • 194
                                      • Canada

                                      #178
                                      Re: Fixing a laboratory waterbath

                                      The trace that goes from the 120VAC to the capacitor and relay was shorted and blew up. I m not 100% sure it's the relay, but the short happened where the relay was. Maybe it s the red capacitor next to it, or both. Without being 100% sure I was going to replace both the capacitor and relay. I have the specs for the capacitor but not for the relay.

                                      Also I dont know if this is normal but the relay has continuity on the common end (high current end). The circuit inside the relay, where the triac would be connected, is always closed even without current (it tested it off the PCB). I put it on a breadboard and connected 6v to the low current circuit to see if I can close or open the high current circuit at will but it doesnt change anything. I dont know if this is how a relay is tested.

                                      Comment

                                      • vrasp
                                        Senior Member
                                        • Nov 2019
                                        • 194
                                        • Canada

                                        #179
                                        Re: Fixing a laboratory waterbath

                                        Originally posted by sam_sam_sam
                                        Show a picture of the bottom of this device
                                        I am not sure what this device is it could be several things
                                        Originally posted by eccerr0r
                                        Why are you assuming it's a relay? What evidence do you have?

                                        I'm beginning to think it's not even a relay ... and have no clue what it is...
                                        I was told it s a relay. It has 4 pins on the bottom. I ll post a picture.

                                        Comment

                                        • R_J
                                          Badcaps Legend
                                          • Jun 2012
                                          • 9602
                                          • Canada

                                          #180
                                          Re: Fixing a laboratory waterbath

                                          You need to post a clear (in focus) high res, picture of the trace side of the board the relay is mounted on.
                                          Looking at the top suggests the small bridge rectifier and the 470µf capacitor are the supply for the relay, the to220 transistor controls the relay coil, The small glass diode next to the relay will be across the relay coil. I suspect the relay coil is either 12 or 24v based on the voltages marked on the transformer

                                          Comment

                                          Related Topics

                                          Collapse

                                          • ohren
                                            Original Xbox 1.0 PSU: Foxlink FTPS-0002 Rev. B. — 12 V low, 5 V high — transformer?
                                            by ohren
                                            Hello!

                                            I'm looking at an original Xbox PSU: Foxlink ftps-0002 rev. B. The 12 V output is too low, and the 5 V output is too (?) high. Unfortunately I don't have an oscilloscope but only a DMM for diagnosis.

                                            Measured voltages with no load:

                                            After rectifier diodes
                                            12V: 6.17 V
                                            5V: 5.67 V

                                            Secondary transformer pins AC measurement (really don't know if this says anything)
                                            12V: 2-2.5 V
                                            5V: 1.04 V


                                            I also have another (working) Xbox psu from Delta. The same measures there, in order, being: 10.79,...
                                            03-18-2023, 10:33 AM
                                          • mikey5791
                                            Is it possible for transformer substitution?
                                            by mikey5791
                                            Hi all,
                                            Got this local made DC power supply (12/14V) given free as the transformer has melted. There is no marking or indication to identify what type or rating of transformer used.
                                            Fyi,. I had a busted autogate mainboard with transformer with marking 13-0-13 . Is this the 13v dc type transformer?
                                            My questions are
                                            1. How do i test to see if the transformer is in good working condition?
                                            2. Can this transformer (marking 13-0-13) be used to replace the melted transformer on the DC power supply?
                                            Hope you guys understand what i was asking. My intention is...
                                            11-03-2022, 02:11 AM
                                          • dennis_elx33
                                            3PHASE STEP-DOWN Transformer in Check
                                            by dennis_elx33
                                            Hello All,

                                            i would like to post an interesting topic regarding 3-phase stepdown transformer which we suspect to cause an overcurrent (138A) which resulted to power tripped off of the HV power supply (80KV output).


                                            Problem statement: Power tripped Off

                                            Initial Problem: Found 2 IGBT and 2 rectifier diodes shorted (burnt).

                                            Action Done:
                                            1) Replaced the faulty parts with good spare.
                                            2) Measured impedance of the 3phase input lines and found that L2 to L3 has 23.3ohm while other pairs are 51Kohm. This pins will...
                                            12-02-2024, 06:52 AM
                                          • Document Archive
                                            ASUS Transformer T101HA - X5DHD Hybrid (2-in-1) Book Specification for Upgrade or Repair
                                            by Document Archive
                                            This specification for the ASUS Transformer T101HA - X5DHD Hybrid (2-in-1) can be useful for upgrading or repairing a laptop that is not working. As a community we are working through our specifications to add valuable data like the T101HA - X5DHD boardview and T101HA - X5DHD schematic. Our users have donated over 1 million documents which are being added to the site. This page will be updated soon with additional information. Alternatively you can request additional help from our users directly on the relevant badcaps forum. Please note that we offer no warranties that any specification, datasheet,...
                                            09-06-2024, 04:40 PM
                                          • Document Archive
                                            ASUS Transformer T304UA-P71SP-CA Hybrid (2-in-1) Pro Specification for Upgrade or Repair
                                            by Document Archive
                                            This specification for the ASUS Transformer T304UA-P71SP-CA Hybrid (2-in-1) can be useful for upgrading or repairing a laptop that is not working. As a community we are working through our specifications to add valuable data like the T304UA-P71SP-CA boardview and T304UA-P71SP-CA schematic. Our users have donated over 1 million documents which are being added to the site. This page will be updated soon with additional information. Alternatively you can request additional help from our users directly on the relevant badcaps forum. Please note that we offer no warranties that any specification, datasheet,...
                                            09-06-2024, 12:00 PM
                                          • Loading...
                                          • No more items.
                                          Working...