inverter problem in LG 1934S LCD flatron
  • PlainBill
    replied
    Re: inverter problem in LG 1934S LCD flatron
    Originally posted by stevenps
    The test winding and the old transformer both light all LEDs. I tried using the inverter transformers in-circuit and found that the high-voltage side lights the 3 red LEDs but not the yellow or green ones. I'm having trouble getting a good reading on the low-voltage side though - I usually get either no LEDs, one red LED, or up through one of the green LEDs depending on which of the many pins I pick. What should I see on the shorted turns tester here? Can I get an accurate reading in-circuit, or should I desolder the transformers?
    Desolder the transformer. The ring tester I have uses an analog meter so it's possible to say something like 2.5 rings. With the LEDs it's a little more difficult.
    Do label them so you know where they came from.
    PlainBill
  • stevenps
    replied
    I built the kit and it seems to be working. I'm winding up the sample transformer to test it with now. I have a board from an old power supply I can steal transformers from to test with also.
  • stevenps
    replied
    By all means build up the shorted turns tester, I the problem is a bad transformer, but I ALWAYS suspect shorted turns.
    Definitely, a second monitor, even one with the same problem will be a big help.
    I brought the other monitor home from work. I'll hook up the inverter board I've been working with to the lamps from the other monitor and see if shorting the same resistor keeps the backlight on.
  • PlainBill
    replied
    I checked in-circuit, and R524 = 469 ohm, R527 = 467 ohm (both within 5% of 470 ohm).
    If it would help, I have a second monitor of this same model with the same symptoms. I only brought one home, but I can grab the other set of parts to see if swapping CCFLs changes anything. I also have a kit for a Dick Smith shorted turns tester that I could build - I just found all the parts that disappeared after my last move, except for the top cover. Let me know if either of these things would be helpful in diagnosing this problem and I can try to take care of it tomorrow.
    The light just went on. Let's assume a simplified version of this inverter circuit. It has multiple transformers, CCFLs, etc. Each transformer has one end of the secondary grounded. The other end goes to a voltage divider whick allows measuring the output voltage and to a CCFL. The other end of the CCFL goes to ground through a resistor, which allows measuring output current. We will further assume that the inverter system is shutting down due to a fault.
    Now if we compare voltages and currents and we find one transformer has a higher output voltage than the other, and the current through one CCFL is higher than through the other, we can come to several conclusions.
    1. If the transformer which has higher output voltage has a higher current it is probably operating normally.
    2. If the transformer which has a lower output voltage has a lower current, the transformer is probably defective.
    3. If the transformer which has a lower output voltage has a higher current, the CCFL is probably aging and drawing a higher current.
    4. If the transformer which has a higher output voltage has a lower current, the CCFL wiring may be bad.
    Of course, there are several problems - we don't know 'normal' voltage or current, and this assumes all parts of the measuring system are equivalent. A further complication is the fact that one transformer drives two CCFLs. Swapping the two CCFL leads on a single transformer doesn't really change anything - the CCFLs are in series.
    PlainBill
  • stevenps
    replied
    Compare the resistances across R524 and R527; they should match to within 5%.
    I'm going to have to rewrite the data in a table form. Right now I'm like three phase AC - jumping between suspecting a bad CCFL, a bad diode, or a bad transformer. THAT, however is going to have to wait for a day or two.
    PlainBill
  • PlainBill
    replied
    These numbers seem to confirm what I suspected, D506 (OP1) and D510 (OP2) should have approximately the same voltages. Compare the resistances across R524 and R527; they should match to within 5%.
    PlainBill
  • stevenps
    replied
    Great! I'll be looking at them later.
    Here are the results:
    • D501: Pk-Pk = 1.760v, Freq = 45.72KHz, RMS = 4.064v
    • D506: Pk-Pk = 2.400v, Freq = 45.56KHz, RMS = 663.3mv
    • D510: Pk-Pk = 7.440v, Freq = 45.68KHz, RMS = 2.675v
    • IC501 pin 11: Pk-Pk = 480.0mv, Freq = 91.41KHz, RMS = 780.0mv
    Again, this time with R506 shorted:
    • D501: Pk-Pk = 1.880v, Freq = 130.0KHz, RMS = 4.182v
    • D506: Pk-Pk = 2.240v, Freq = 66.25KHz, RMS = 700.9mv
    • D510: Pk-Pk = 7.920v, Freq = 45.63KHz, RMS = 2.889v
    • IC501 pin 11: Pk-Pk = 700.0mv, Freq = 134.2KHz, RMS = 794.2mv
    Images are again attached for reference. The view settings are not all the same, if you're looking at the waveforms; I think the offset ones are only offset by 1 div. I'm not sure what to make of the differences in frequency measurements between having the end of resistor R506 shorted to ground, and leaving it alone. The scope seems to have trouble measuring the frequency of the short pulses the backlight stays on for when the resistor is not shorted, but it's possible (maybe?) that the frequencies actually are different during that very short time right after the inverter turns on.
  • PlainBill
    replied
    The scope is an Owon PDS5022S (http://www.owon.com.cn/eng/pdsSeries.asp). The probe is a P6060 with both scope and probe set at x10. I just switched to DC coupling. I was attaching the ground clip to the case of the monitor, as the center screw hole is attached to cold ground and the board is screwed in to the case. I can try using another location, though it may be a little tricky.
    I took new readings but I don't have time to post the values right now; I'll get that done later this afternoon. I moved my ground position to the negative leg of C501 and used DC coupling. I have two sets of measurements, one with and one without shorting the end of R506 to ground to keep the inverter running long enough to properly adjust the scope and let it get a good reading.
    Great! I'll be looking at them later. Today was supposed to be my goof-off day, after running errands. It's now 10:20 AM and it'll be at least an hour before I can start on the errands. So much for a goof-off day.
    PlainBill
  • stevenps
    replied
    So to start at the beginning, what model Owon oscilloscope do you have? What kind of probe are you using, and where are you connecting the ground lead for the probe? And it appears you are using AC coupling, switch to DC coupling.
  • PlainBill
    replied
    I re-measured the connectors with my scope tonight, and I don't quite believe the readings. Hopefully this is due to my less-than-stellar job of re-attaching R510, but everything except CN501 measures really high. I had 32mv, 2.3v, 58mv, and 600mv for the 4 connectors last night. Tonight (after removing and replacing R510):
    • CN501: Pk-Pk = 2.280v, RMS = 687.0mv
    • CN502: Pk-Pk = 21.80v, RMS = 7.960v
    • CN503: Pk-Pk = 7.600v, RMS = 2.698v
    • CN504: Pk-Pk = 10.48v, RMS = 3.878v
    I'm attaching pictures of the scope screen for each, which provides a little extra information.
    Yes, it does, and it provides plenty of questions, too. I hope you don't mind, but we are going back to the basics on this for a few minutes. The most significant thing I see is the frequency on CN502 is shown as 13Khz, it's about 45Khz on the other connectors. From what I think I know about inverters, 45Khz is possible, 13Khz is flat out wrong.
    Next, I would like to see the traces at OP1, OP2, OP3. Probe the cathode ends of D506, D510, D501. Also, I'd like to see a trace for pin 11 of IC501. Clip the scope ground as close as reasonably possible to that area. Soldering a short piece of wire (resistor lead) to a ground point is a good idea.
    Hmm, another problem with SMD technology. No ready source of short lengths of pre-tinned wire.
    I had to go back to school to figure this out, but the impedance of C532 at 45Khz is 540 ohms. With the parallel resistance of R535 I'd expect the voltage at CN502 to be roughly double the voltage at CN501.
    By the way, the purpose of the waveform at pin 11 of IC501 is to get a reference.
  • stevenps
    replied
    Now, what does all of this mean? I'm not sure, but I have at least one more test for you to try. What are the RMS and P-P at the L end of CN501 - CN504. I'm assuming the readings you gave earlier were DC voltages.
    I'm beginning to wonder is Shakespeare was right. "The fault lies not in the stars, but in ourselves." Or something like that.
    That could absolutely be true. Don't discount the possibility of a case of operator error somewhere in here - I haven't had any training. I do this stuff as a hobby and most of the monitors I fix (Samsung) just require capacitor replacement and they're fine. I've got two models giving me trouble right now (this one and a Samsung Syncmaster 940T), and I've been dragging my feet about getting them fixed because I had no idea how to go about troubleshooting them.
  • PlainBill
    replied
    I measured voltage on all pins of the inverter IC when I first started gathering data. I measured 0.24v on pin 7 (OVPT) when the power was off, and 0.71v when the power was on. If something doesn't add up, I can re-measure tonight.
    DuH!!!! The light dawns!!! I was looking at this backwards. This is REALLY a screwy design!!!
    Look at CN501 and CN502. The H pins are outputting sine waves 180° apart - ends of a transformer winding, right? These numbers are approximate, by the way. When CN501H is at +800V CN502H is at -800V. Since the CCFL draws about 6 ma, the voltage at OP1 should be about 5.6V P-P, centered around 0 volts. On the positive peaks D506 is reverse biased, but on the negative peaks D506 conducts and pulls the voltage at the anode of D506, D510, D501, D505 down.
    Now looking at CN502, things are way different. When CN502H is at the negative peak D504 1-3 is forward biased, and there will be about -3.5V at CN502L. Because D504 3-2 is reverse biased, the voltage at OP2 will be being pulled toward 0V at a rate determined by the time constant of R535 and C532. On the positive half cycle D504 3-2 is forward biased, so C532 is being charged by R523 and being discharged by R535.
    CN503 behaves pretty much like CN501. The negative peaks will pull OP2 down in proportion to the current flowing through the CCFL and R527.
    CN504 is a whole 'nother matter. The negative peaks go through R520 and D507 1-3 to ground. The positive peaks go through R520 and D507 3-2 to the parallel network of C524 R528 and R533 and to the ISSEN input on the inverter controller. This provides the feedback to regulate the drive current.
    As an aside, troubleshooting the 'two seconds to black' problems has always been difficult. The bulk of my electronics training was nearly 40 years ago, and I had a hard time grasping what was happening without a schematic. Even with a schematic, the guy I was helping usually had only a DMM. About all I could do is trace circuits and look for components with unusual readings. Now I've got a schematic, a well documented controller, and someone with both a DMM and a scope. And I'm STILL struggling!!! I'm beginning to wonder is Shakespeare was right. "The fault lies not in the stars, but in ourselves." Or something like that.
    PlainBill
  • stevenps
    replied
    I'm wondering if the voltage at the OVPT pin is lower than normal. Could you measure that? Maybe that will provide a clue.
  • PlainBill
    replied
    I re-measured the voltages using my oscilloscope:
    • Left end of R506 = 706.5mv RMS, 2.24v pk-pk
    • Bottom end of C509 = 454.3mv RMS, 1.92v pk-pk
    • Left end of R525 = 589.2mv RMS, 1.96v pk-pk
    • Bottom end of C528 = 478.3mv RMS, 2.32v pk-pk
    I checked R506 in-circuit and it measures 2.96MΩ. For reference, R525 measures 3.005MΩ and the parts list says this should be 3MΩ 3KV.
    I desoldered R510 and it measures 3.37KΩ. The parts list says it should be 3.4KΩ and it is labeled "3401" which I think means it's supposed to be a precision resistor. If that's the case, then this one is close to the bottom end of 1% tolerance.
    I don't think I have appropriate tools to desolder D502 without damaging something. I used hot air from a butane torch to get R510 off, but the diode is a bit bigger and I don't want to risk melting anything. Looks like I'll have to order some Chip Quik (and probably just get another BAV99 for the replacement anyway).
    Since I already have to order parts, I might as well replace the resistors as they should be pretty cheap. What type of resistor would be best for R506? It's tan colored and should be 1/2W. How about R510? Looks like I can get thin film, thick film, carbon film, wirewound, metal film... I have no idea which to choose.
    Note that R506/R510 form a voltage divider. If the resistance of R506 is higher than spec or of R510 is lower than spec, the voltage at the junction of pin 3 of D502 will be lower. I don't think it is necessary to replace either of the resistors. The peak voltage at pin 3 is below the normal threshold for triggering shutdown, well below if you consider the .5 - .7 volt drop across the diode.
    PlainBill
  • stevenps
    replied
    That would indicate that either R506, R510, or D502 is the problem. There is about 100Khz AC on the points I had you measure; an oscilloscope would be the best way to measure the voltage. You can check the resistance of R510 in circuit; lift one end of R506 to measure it. About the only wan to reliably test D502 is by replacement. The diode in question is called D502 on the board, D602 on the schematic.
    PlainBill
  • PlainBill
    replied
    Thanks for the suggestions. The backlight comes on and goes off again twice after I push the power button. The first time the enable signal to the inverter IC goes low and the backlight turns off. A second later, the backlight turns on again for about half a second then turns off; the enable signal to the inverter IC stays on this time (this is where I think the problem is). Also I did find a surface-mount fuse near the connector to the display board (CN801). It's F802 (4A, 32v) and it's fine. I checked U501 and U502, and found that none of the pins are shorted.
    Thanks PlainBill. I think I sound like I know more about this than I actually do. I went back and took some more readings and found that the voltage from the power supply is about 13.7v when the inverter is running, which is pretty close to the 14v mentioned in the service manual. I also re-measured the voltage at the CCFL returns, and found that two are higher than the other two.
    • CN501 = 0.032v
    • CN502 = 2.3v
    • CN503 = 0.058v
    • CN504 = 0.6v
    Hopefully this makes some sense based on what's in the service manual on page 47.
    I took the measurements you requested:
    • Left end of R506 = 0.2v
    • Bottom end of C509 = 0.45v
    • Left end of R525 = 0.15v
    • Bottom end of C528 = 0.47v
    I also tried shorting each point to ground, and found that the only point that made the inverter stay on was the left end of R506. I'm attaching another picture of the whole bottom of the board. Lighting in my house is not particularly good for this, so I have to hold a flashlight or use the flash on my camera to get a reasonably clear picture.
    PlainBill
  • stevenps
    replied
    • CN501 = 0.032v
    • CN502 = 2.3v
    • CN503 = 0.058v
    • CN504 = 0.6v
    Last edited by stevenps; 09-23-2010, 12:45 AM. Reason: Add result of checking U501 and U502
  • PlainBill
    replied
    The service manual (including schematic) for a LG W1934S is available on www.elektrotanya.com.
    This employs a rather unusual design. An inverter is a constant-current AC source with over-voltage protection. Normally the return currents on the four CCFLs are summed and used to to regulate the inverter output. Voltage sense is used to shut down the inverter if a CCFL goes open. The individual return currents are used to shut down the inverter if the current through a CCFL goes too high. (As a CCFL nears 'end of life' the current increases). Also, the four current sense resistors have three different values!
    PlainBill
  • PlainBill
    replied
    U501 and U502 are Niko-Sem P6006HV(dual N-channel FET). I'll check them tonight. I didn't see any surface-mount fuses, though I'd expect the inverter would not work at all if a fuse was blown. The only fuse I see (aside from the line fuse, F801) is FB501 which is provides 12v to the inverter, and it's fine. I believe the 12v output from the power supply is actually a little over 13v, which is hopefully within tolerance.
    You've done an excellent job of troubleshooting . You've established the problem lies with the inverter, not the CCFLs. You are also correct, it's not due to a problem with a fuse.
    Breaking the problem down, there are two possibilities. The drive voltage to one set of CCFLs is wrong, or the circuitry sensing the voltage / current has a defect.
    Your troubleshooting of the inverter controller was good, but you did not go far enough. If you look at figure 1, the protection logic control turns off the FET, allowing the cap on Timer (pin 3) to charge, shutting off the inverter. If you look at the application drawing (figure 2) in the OZ9938 datasheet you will see that the voltage at Pin 6 comes from the circuits feeding D2 and D4. C22 filters the pulses, R25 provides pull down. Unfortunately, while the picture of the bottom of the board is excellent, I can't quite trace the actual circuit. However, there is another approach. Note that D2 and D4 in the app note form an OR circuit. The voltage at pin 6 will exceed 3 volts only if one (or more) of the voltages at the anodes of D2 and D4 exceeds about 3.5 volts.
    Referring to the picture of the bottom of the board, with the CCFLs operating, measure the voltage at the following points:
    Left end of R506
    Bottom end of C509
    Left end of R525
    Bottom end of C528? (blue cap, lower right hand corner of the picture).
    If you are unable to measure the voltages reliably, instead ONE AT A TIME short each of these points to ground. The one which results in the backlights staying on is the culprit. Report the results, we'll carry on from there.
    PlainBill
  • retiredcaps
    replied
    The monitor is a LG Flatron W1934S and its backlight won't stay on.
    1) Just to clarify, does the backlight initially come on and then go to black? If yes, how long does the backlight stay on?
    2) Or does the backlight NOT come on at all?
  • Askic
    Problem with monitor DS-D5022QE-B
    by Askic
    Hello to all,
    I have a problem with monitor that is delivered together with CCTV system.
    The monitor appears to be completely dead, there is no LED power diode and it doesn't react to pressed power button.
    I have disconnect display and check many components on the power supply part of the board. There is no anything obvious and monitor is about 3 years old. There are no bulged caps. When measuring voltages on the output connector to VGA board I can measure stable 14.5 V on 12V pins and stable 5.07V on 5V pins. So I assumed that this part of the board is OK....
    05-31-2023, 03:14 AM
  • CptChronic
    Samsung LU32J590UQUXEN Monitor display problem
    by CptChronic
    I have an intermittent display problem with my monitor, when I plug in the HDMI cable to my (or any) laptop, you see the connection screen shrink on the laptop as it usually would before displaying on the monitor but 95% of the time the monitor display stays black.
    Before it cut out all together, tapping the monitor control button a fair few times would make the screen display, and everything work fine.
    But now it never does it. Not sure what's going on, hoping someone can help. I'm capable of stripping the thing down and soldering in new parts if needed, or programming...
    04-24-2023, 04:33 PM
  • valvashon
    Insignia NS-LCD19-09 (PS OK, Inverter problem?)
    by valvashon
    For some reason I'm just not going to let this 19" TV defeat me. I guess because I have a schematic to work from I feel like I'm oh so close.
    History- while adding a turn-on circuit I let a small screw roll under the power board. Worked through all of that circuitry, probably replacing a MOSFET and PWM controller that were not bad but I was following the service manual. Finally figured out that the power supply is in working order when the inverter section (for the backlights) is disconnected. When the +12 supply to the inverter is disconnected the +12 and +5 will come up...
    01-04-2024, 01:44 PM
    by momaka
    03-15-2023, 10:17 PM
  • tmhobadcap
    LG 24MP56HQ-P Monitor Backlight problem?
    by tmhobadcap
    I have a few successful experience in fixing tv backlight problem. But for monitor, I have no experience. However, I think they are more or less the same. Am I correct?
    Recently, I got the captioned monitor and the owner said it could not be turned on. I open it up and found that one of the SMT capacitors seemed to be bad. I replaced it as well as the other one. The monitor still could not be turned on.
    I found the power indicator is on. When I shine a flashlight on the screen, I can see the image. Therefore, I think it is with backlight problem. I took out the led strip....
    10-13-2021, 01:52 PM
