Fluke 23 Fusible Resistor
  • habanerocat
    • Nov 2014
    • 96
    • Ireland
    I have the above DMM with a blown 1K 2W Fusible Resistor on the input circuit.
    Maintenance manual says to replace with the exact part for safety reasons.
    But I can't find this part or get a substitute for my usual suppliers here in Ireland/UK.
    Any ideas?
    Thanks..............
  • habanerocat
    • Nov 2014
    • 96
    • Ireland
    Re: Fluke 23 Fusible Resistor
    Managed to get some 10% substitutes on ebay Germany.
    • redwire
      • 3910
      How did you blow that resistor? It takes over 1,000V sustained to kill it and the MOV's would certainly also be blown.
      Typically the part is a Ohmite OX102KE (1W 16.5mm) or OY102KE (2W 19mm) and has a safety function protecting the multimeter's input when measuring voltage. It's huge for the IEC61010 mains transient overvoltage spec.
      • retiredcaps
        • Apr 2010
        • 9271
        http://mrmodemhead.com/blog/fluke-87-fusible-resistor/
        --- begin sig file ---
        --- end sig file ---
        • habanerocat
          • Nov 2014
          • 96
          • Ireland
          I can't remember how I blew it. Think it was a 400V motor failtfind many years ago. Let's hope the MOVs aren't blown as I haven't checked them.
          Thanks for the link but getting parts from the US is very expensive.
          • habanerocat
            • Nov 2014
            • 96
            • Ireland
            Yea, came across that site in a search. Very informative.
            • retiredcaps
              • Apr 2010
              • 9271
              If you made an "oops" moment, you need to check all the input protection components for damage. While some may be obvious, others are not.
              1. All your MOVs should measure 0L in resistance mode.
              2. The PTC should measure around 1.1k ohm.
                • 31028
                the mov's are known to go leaky in old fluke meters.
                • habanerocat
                  • Nov 2014
                  • 96
                  • Ireland
                  I presume the PTC you are talking about is the Variable Resistor, R1?
                  Or the Thermister E1/DS1?
                  • habanerocat
                    • Nov 2014
                    • 96
                    • Ireland
                    Ok, there seems to be a lot of conflicting information out there in relation to the devices in the attached circuit and photos. People on here seem to be making the most sense thankfully, so thanks.
                    Replaceable fusible resistor is ordered as above and on the way.
                    The two red MOVs, RJ1 & RJ2 are reading open circuit in resistance mode.
                    The yellow thermistor (PTC) E1, is also reading open circuit, so I presume this is blown.
                    So if I'm correct above I just need to find a thermistor which is labelled as follows.
                    MDC
                    1-2 KV
                    9232
                    Any ideas where I might identify, and/or get this?
                    Thanks...........
                      • 31028
                      RT1 is the thermister,
                      RJ1&2 are mov's
                      E1 is a spark gap
                      • R_J
                        • 9535
                        I agree:
                        E1 is a spark gap and will read open.
                        RJ1 and Rj2, each protect at 910v, two in series give 1820v protection These will also read very high resistance.
                        As long as the RJ1 & RJ2 are not blown apart and not shorted they are likely ok.
                        The rt1 thermistor and will measure around 1k but the resistance will increase as it heats up
                        • habanerocat
                          • Nov 2014
                          • 96
                          • Ireland
                          RJ1&2 are mov's
                          E1 is a spark gap
                          Thanks, how do I test that spark gap?
                          • redwire
                            • 3910
                            Those MOV's are hard to find because they are high voltage yet small- so they are a specialty multimeter part. The modern TDK/Epcos part would be a S05K575 but I've never found them for sale.
                            You can get the same voltage rating, but twice the size in 10mm. Or, you can get the same size 5mm but slightly lower voltage 820V 1mA.
                            The old red MOV's are GE V910LSX1399 910V 1mA, 1500V/25A, 65J, 90pF, 2.5kA. I thought a 5mm size.
                            Similar part Voltlog #34 – Fluke 77 Teardown & Repair, Varistor Replacement
                            But- he used much fatter, bigger Panasonic ERZV10D911 10mm which would not fit here.
                            Littelfuse bought the GE line LA series MOV and offer only larger 10mm size only like V575LA10P.
                            TDK/Epcos Standard MOV lineup is bigger at 10mm B72210S0551K101 7.5mm lead spacing, or 5mm B72205S0461K101 but lower voltage 750V 1mA.
                            Bourns offers MOV-07D821K but lower voltage 820V 1mA and 7mm with 5mm lead spacing.
                            You'd have to see what you can fit on the pc board. Modern parts have 7.5mm lead spacing and differing thickness, I've never found the smaller original 5mm offerings for 910V.
                            In your locale, there are a few electronic component distributors. MOV's, PTC, 1k 2W Ohmite OY102KE resistor Digikey Ireland or Farnell etc. sell.
                            • habanerocat
                              • Nov 2014
                              • 96
                              • Ireland
                              Originally posted by R_J
                              I agree:
                              Thanks, I replied above before I refreshed the page.
                              So are Thermistors and Fusible Resistors the same thing?
                              That looks like Rt in the circuit but I think it's actually R1.
                              • retiredcaps
                                • Apr 2010
                                • 9271
                                No.
                                A spark gap should measure open circuit resistance.
                                • retiredcaps
                                  • Apr 2010
                                  • 9271
                                  No. See
                                  https://www.youtube.com/watch?v=zUhnGp5vh60
                                  • retiredcaps
                                    • Apr 2010
                                    • 9271
                                    RT1 is close by C1. See page 7-2 on Fluke 70 service manual.
                                    https://cdn.badcaps-static.com/pdfs/...ad75dcb902.pdf
                                    • redwire
                                      • 3910
                                      I think I'm confused what rev this multimeter is? It's a Series 23I? I gave older 87V part numbers, that Fluke 70 service manual bill-of-materials says it's a 430V MOV.
                                      It also gives the 1,000 ohm PTC thermistor TDK p/n 911P84E102YU13.
                                      • retiredcaps
                                        • Apr 2010
                                        • 9271
                                        The Fluke 70 series I is grey. As I understand it, Fluke took the 70 series I and added better probes and made the case yellow for the Fluke 20 series I.
                                        That's why I referenced the Fluke series I manual in the post above.
                                        • jinu_j
                                          by jinu_j
                                          Hi
                                          05-03-2025, 10:51 PM
                                        • gica69
                                          Need help to identify and to locate a 0.33 ohms fusible resistor for power supply
                                          by gica69
                                          I have a quad charger that is split into 4 separate sections and one section doesn’t charge anymore. I have located what I think is a thermistor and on the shrink tube it says 125 degrees C. I cut the shrink tube and it's a 33 ohms resistor. It measures a little less than 1/2 inch in length. I am including pictures of it. Please help me identify it. So, it should test at 0.4ohms and probably go up in resistance with the rise in temp. On the board it clearly says R for resistor. But it has to be a special one.
                                          Resistor shows orange range silver gold and green lines.
                                          Thanks in adv...
                                          07-21-2025, 09:49 PM
                                        • bluto
                                          Is this a fusible resistor?
                                          by bluto
                                          This is from the power supply of an LED shop light. The light went completely dead suddenly. I removed the SMPS board and it looked really good except for some bad solder joints. I touched all of them up but the light was still dead. I then found this open resistor that was between the neutral wire of the plug and the transformer. The color bands seem non-standard. Does it indicate 13 ohm, 30 ohm...? I am not sure. What should I replace this with?
                                          07-01-2025, 01:30 PM
                                        • mikey5791
                                          Need help to replace burned resistor in 350w atx psu
                                          by mikey5791
                                          Hi to all forum members,
                                          Got this faulty 350 watt atx psu dirt cheap from a local online seller as i am keen to use it as a learning tool to repair faulty power supply.
                                          The faulty part is quite obvious from the burn area near the 16pin KA7500 BD pwm. The fuse is still intact with good continuity.
                                          Upon power on ac, there is 5vdc standby power on cathode of diode D14 and about 10vdc on cathode of diode D13.(This should be the aux power to generate the 12v and 5v rail.)
                                          Initially i replace the two burned resistors near my marked red arrow to 200 ohm resistor...
                                          12-26-2024, 04:55 AM
                                        • MidnightRoller
                                          Can a resistor lose resistance after hot air?
                                          by MidnightRoller
                                          I am curious if a 0402 resistor could go from 43 ohm to 0. In electronics class and my experience a resistor can degrade and gain resistance all the way up to not being able to conduct at all. I have never heard of a resistor decreasing in resistance and becoming 0 ohm or a dead short… that's not the direction they fail… or so I thought.
                                          I was attempting to remove a wson8 chip on an iMac motherboard and it required a lot of heat, on one side of the chip were 3 0402 resistors that were very close. I popped the chip out unintentionally out of the tweezers and sent those resistors...
                                          02-02-2022, 11:46 PM
