Sony Trinitron (KV-20FV12) Vertical Deflection Stopped
  • motorahead
    • Oct 2023
    • 17
    Hi everyone, I've been troubleshooting my 20" Sony Trinitron (KV-20FV12) CRT after it failed to show a picture. The standby LED is blinking 4 times, which according to the service manual is a failed vertical deflection via the self-diagnostic check during power up.
    After getting the TV checked out by a technician and doing research on my own. The most obvious first step was to replace the IC502. I've done so with an exact NOS replacement with no luck. I've read a few suggestions to make sure the 12V supply to pin 2, but that voltage check would seem misleading? It seems if the self diagnostic fails, the IC502 will get a supply voltage? I checked anyway and the multimeter reads no voltage.
    Throughout this process, I've been looking at the TV's schematics (attached the service manual) to try and better understand where the failure could be coming from, but with my limited understanding of circuit boards, it's incredibly difficult to follow; especially when the current splits in different directions. Near the IC502, I've spot checked a few of the diodes and resistors, to make sure they are reading per spec, but I've exhausted my skills and understanding of the issue.
    If someone is willing to take a look at the schematics and guide me through troubleshooting the vertical circuit, or suggest a fresh starting point, it would be hugely appreciated! Hopefully this troubleshooting process and resolution will be helpful to others, and I can keep this TV alive.
  • R_J
    • 9551
    Re: Sony Trinitron (KV-20FV12) Vertical Deflection Stopped
    You likely have an open R550 or R549 (0.47Ω) If the ic shorts, the two resistors can go open.
    • motorahead
      • Oct 2023
      • 17
      Wow, how did you nail that so quickly?!
      I checked and they are both open. The reading was first a little confusing so I checked R544 and it reads exactly 047.0; while both R550 and R549 read 001.0 (and the multimeter beeps). Not sure why the decimal is off, maybe a range setting on the multimeter?
      I'll order a couple, solder them, and update this thread.
      • petehall347
        • Jan 2015
        • 4427
        • United Kingdom
        you need to address the meter readings first . first thing it would be unusual for a resistor to measure exact value . but decimal place moved .. 47 ohms instead of 0.47 . i wouldn't even expect dead on 0.47 . your other reading says 1 ohm so that one might be a bit high or its meter error .. first set your meter to lowest ohms setting maybe 200 ohms then touch the probes together then pull out suspect resistors and check values again then deduct whatever it measured touching probes together .
        • motorahead
          • Oct 2023
          • 17
          Thanks for the reminder about touching the probes, I recall watching a video a while back and forgot about this.
          I followed your guidance and tested resistance with three multimeters (out of curiosity), but did not remove the resistors yet, but it does seem clear R549 / R550 are bad?
          Not sure why the decimals are showing like this.
          Klein CL390 (400 ohm)
          R544 - 046.7
          R549 / R550 - 000.4
          Innova 3340a (400 ohm)
          Did not record these numbers, but they were within a tenth.
          Allsun EM30 (200 ohm)
          Not a quality multimeter relative to the other two I have, but the reading seems to be within spec (5%)
          R544 - 45.9
          R549 / R550 - 000.5
          • petehall347
            • Jan 2015
            • 4427
            • United Kingdom
            R544 - 046.7 appears to be the bad one
            • motorahead
              • Oct 2023
              • 17
              Sorry for confusing things. I was reading the meter incorrectly based on a misunderstanding of R544.
              R544 1-260-312-11 CARBON 47 5% 1/2W
              R549 / R550 1-260-288-11 CARBON 0.47 5% 1/2W
              So as it turns out, all three are good...
              I'll be more careful moving forward. Next steps would be appreciated.
              • R_J
                • 9551
                Re check your voltages on IC501, you need +12v on pin2 and -15v on pin 4 Or it might be easier to check the voltages on the plug CN706, pins 1 (+12v) pin2 (-15v)
                R544 is for the +200v supply to the crt neck board
                • motorahead
                  • Oct 2023
                  • 17
                  Thanks, I'll check. Should the voltages still be present while the standby light is throwing the error code?
                  • motorahead
                    • Oct 2023
                    • 17
                    Plugged in the TV and checked the IC501as well as CN502 and CN706. I also re-tested IC502. No voltages.
                    Double checked the pins in the service manual and they are a little different than what you provided on IC501 (see attached). So I tested both ways.
                    • R_J
                      • 9551
                      I made a mistake, I meant IC502, the vertical output ic. Do you know if you have any high voltage to the crt?
                      Check if you have B+ on C654, should be around 130vdc
                      The +12 & -15 are supplied by T505 FBT
                      Check D504 and D505 to see if one is shorted
                      • motorahead
                        • Oct 2023
                        • 17
                        Thanks for being patient, I'm learning on the fly here and am a little embarrassed to admit my voltage checks have been incorrect. Sorry for steering you down the wrong path, but I am getting supply voltage at the IC502.
                        I was powering on the TV, which then powers off after 5 seconds (after the failed self check).So I was unknowingly checking for voltage when it was off...
                        Now I'm holding the probe at a contact with one hand, while powering on the unit with the other hand to get a reading. The negative lead is resting on the video input ground.
                        IC502 reading doesn't fully stabilize in time before the unit powers off.
                        Pin2: ~13.9v
                        Pin4: ~ -13.7v
                        CN502 / CN706
                        Pin1: ~13.9v
                        Pin2 ~ -13.7v
                        B+ on the C654: 136v
                        I'll test D504 and D505 in the morning.
                        • motorahead
                          • Oct 2023
                          • 17
                          Curious, I was tracing CN706 and am getting 1.4 ohm via R718 on the CB board. Is that correct? It seems out of spec unless resistance is split between R718 and R717 (no resistor here, but they look joined with solder)?
                          Per spec:
                          R718 1-216-375-00 METAL OXIDE 3.3 5% 2W
                          • R_J
                            • 9551
                            R718 is in the crt heater supply line, don't worry about it. You are measuring the crt heater resistance as well as the fbt heater winding which have one end connected to ground.
                            • motorahead
                              • Oct 2023
                              • 17
                              Understood, thanks. I checked the D504 and D505.
                              Used this article for reference on testing diodes: https://www.fluke.com/en-us/learn/bl...to-test-diodes
                              D504
                              Diode Test Setting
                              forward .468v
                              reverse 2.99 v
                              Resistance Test Setting
                              forward 1.41ohm
                              reverse 2.32ohm
                              D505
                              Diode Test Setting
                              forward .487v
                              reverse 2.98v
                              forward 1.45 ohm
                              reverse 2.27 ohm
                                • 12206
                                • U.S.
                                If your getting those readings out of circuit/best way to test, you have bad diode.
                                • motorahead
                                  • Oct 2023
                                  • 17
                                  I did not remove the diodes. Hopefully I tested them correctly. Forward bias was checked with the positive lead to the anode, negative to the cathode. Should I remove them to verify? Testing in reverse should read 0v?
                                    • 12206
                                    • U.S.
                                    Better let RJ guide you, I haven't touched one of these in prolly 12 years, unfortunately I don't remember anything about them.
                                    • motorahead
                                      • Oct 2023
                                      • 17
                                      Your feedback is certainly appreciated! And thank you RJ for sticking with me so far!
                                      • R_J
                                        • 9551
                                        The diodes are likely good, I remember if one would short it could shut down the tv, Sometimes IC702 (on cb board) can shut down the tv as well, Check if the 200v comes up on CN706 pin 6, If it seems low, disconnect CN706 from the cb board and see if the tv stays on. I have'nt worked on one of these in quite a few years.
                                        The tv checks if the +12v is missing and gives the 4 blink error, but if the high voltage shuts down, there will be no 12v so the tv can give 4 blinks unrelated to the vertical ic.
                                        • Returning Buick
                                          Viewsonic G810 vertical deflection problem
                                          by Returning Buick
                                          I have a G810 monitor i've had for years but it has developed a problem with vertical collapse. Is there anybody here who has worked on these and can help me find where the circuit even is for the vertical deflection?
                                          04-01-2025, 07:57 PM
                                        • gpo23
                                          Sony KDL-40V3000 - Vertical lines
                                          by gpo23
                                          i order a t-con because i had bad colors on my SONY KDL-40V3000 :
                                          With the new t-con the color are good but there are vertical lines on all the screen :
                                          I've tried the old t-con and it returned to the initial issue (bad colors but no vertical lines)
                                          Is there a way to fix the vertical lines on the new t-con ?
                                          Thank you for your help
                                          gpo23...
                                          10-04-2023, 09:11 AM
                                        • visionbluedemon
                                          Help. Sony kd-65x85dj vertical lines
                                          by visionbluedemon
                                          First. it has no picture at all. i solved it by covering the ckv lines using tape to to block the connections
                                          Second. now has picture but with vertical lines that follow the image. tcon board ic overheating
                                          tried blocking LC lines too, but same results
                                          is it lcd panel issue or tcon board issue?...
                                          07-25-2025, 04:56 AM
                                        • cmonkey
                                          Sony Trinitron KV32FV27 Suddenly Dimmed
                                          by cmonkey
                                          A couple of days ago I finally got my hands on a locally sourced Trinitron and after lugging all 175 LBS home I gave it a good cleaning inside and out. The guy said it had been in a shed for 2 years, but the outside wasn't too bad. The inside was another story and while there was no evidence of mouse, it had certainly been kept in a VERY dirty home.
                                          I cleaned it to an acceptable level of grime, at least clean enough to where nothing would heat up from the dust insulation. I then tried it out and it worked!
                                          I'd say I had it on for about 15 minutes or so and then lugged...
                                          04-28-2023, 06:25 PM
                                        • noisykid
                                          Sony XBR-85X950G Gray horizontal lines and vertical picture shift diagnosis?
                                          by noisykid
                                          5.5 year old 85 inch Sony Bravia X950G started showing gray horizontal lines across the bottom of the screen and the picture shifts slightly vertically when it happens. The lines sometimes get bigger/smaller vertically and bounce back. See the video for the problem. It seems to be happening after it warms up after an hour or so. Happens regardless of source. Right now there's nothing but power cable connected and it's happening. We did hook up a new Nintendo Switch to our HDMI switching receiver today, and played for a while with no issues, but this problem didn't occur til a few hours later while...
                                          12-30-2024, 07:15 PM
