Crest CPX1500
  • 5inc
    • Dec 2016
    • 120
    • Spain
    #61
    Re: Crest CPX1500
    just got back from the shops where none had a single J174 (Q100,101) the closest I found was a J175 in a bigger enclosure...
    http://www.datasheetcatalog.com/data...1/2SJ175.shtml
    I'll have to order some J174s online...
    • Agent24
      I see dead caps
      • Oct 2007
      • 5097
      • New Zealand
      #62
      Re: Crest CPX1500
      The other end of R100 goes to GND, so you will hear nothing there. Don't worry about that.
      The signal should keep going through R101 and R102, but will be shunted to ground also if Q100\Q101 are switched on. They are controlled by the DELAYA signal. If the DELAYA signal is not working correctly, then Q101 and Q100 may never be getting turned off.
      You need to check what the DELAYA signal is doing and what voltage is on the gates of Q100 and Q101 before you replace them. They may not be faulty.
      Another problem could be if CR100 or CR101 are shorted\leaky. They are clamping diodes to protect the input of U100 but could cause a problem with the audio signal if they are faulty.
      The same goes for if U100 has a shorted input, or if U101 is faulty or being incorrectly driven.
      -David VanHorn
      • 5inc
        • Dec 2016
        • 120
        • Spain
        #63
        Re: Crest CPX1500
        I just checked voltage on gates of Q100/101 and have 13.9vdc. I then turned the tracer on and checked the both ends of R106 assuming that would be DELAYA, I hear nothing. Is that what you ment by "check what the DELAYA signal is doing"?
        CR100/101 are unlikely shorted or leaky, I allready shotgunned them... and the same goes for U100/101. Well, appart from the fact that I fiddled with them with the tracer and thats when I blew the fuses. Could they have gone bad by that?
        And when you say U101 could be incorrectly driven, are you talking about wrong voltage or wrong/no signal?
        • Agent24
          I see dead caps
          • Oct 2007
          • 5097
          • New Zealand
          #64
          Re: Crest CPX1500
          Q100\101 are P-Channel JFETs so that 13.9v on the gates should be turning them off, assuming they are not faulty (shorted).
          To troubleshoot further, you could try the following:
          Check R101 and R102 to be sure they are not open-circuit or have gone high-ohms.
          If R101 and R102 are OK, then temporarily remove Q100 and Q101 from the circuit and see if the signal then gets through to Pin 3 of U100.
          If it still does not, disconnect Pin 3 of U100 and see if the signal is now present at R102\CR101.
          If there is still no signal, disconnect Pin 6 & 1 of U101.
          If the signal is still not getting through R102, the only other possibility is that C103 has shorted and the signal is getting pulled to GND through it and R107.
          DELAYA is a control line, not an audio signal line. It controls the power-up sequence, likely to prevent switch-on thumping of the speakers.
          I meant for you to check what voltage it is at. Measure the voltage on DELAYA (at R106) while the amplifier starts up. When you first turn on the amp, DELAYA should be at 0v or close to it. After a second or two, the voltage should rise to that 13.9V.
          If the voltage on DELAYA goes high immediately after you turn the amplifier on, or if it stayed at 0v, it would indicate a fault with the DELAYA signal generation circuitry.
          Yes, it's possible you blew U100/101 again. I don't know what or how you probed where, but maybe you caused them to short out and that's what blew the fuses.
          Yes, U101 could be getting some incorrect signal or no signal from another part of the circuit. It could also be faulty.
          -David VanHorn
          • 5inc
            • Dec 2016
            • 120
            • Spain
            #65
            Re: Crest CPX1500
            Done, R101/102 are ok. Resistance Check reveals a short of 40 ohms, probably Q100/101.
            Took out Q100/101 and trace probed pin3 of U100 to hear fizz. (J174's have a 40ohm short from source to drain)
            Tried to unsolder pin3 of U100 but ended up taking the trace with me. It is now disconnected.
            Probed R102/CR101 and finally got signal.
            Does that mean U100 is faulty, appart from Q100/101?(these transistors and the ones on the other channel are equally shorted, I had seen it and thought it was a circuit matter)
            I'm ordering a pack of J174s, and will change U100 and fix the broken line when those arrive. Do you recomend using a couple of strands of copper wire to do so?
            • Agent24
              I see dead caps
              • Oct 2007
              • 5097
              • New Zealand
              #66
              Re: Crest CPX1500
              The transistors may still be OK, a JFET will show some conduction between drain\source because there is no PN-junction there, and being a P-channel there is no voltage on the gate to shut it off while you are testing it.
              Put the transistors back, one at a time, and with U100 still disconnected, do you still get signal?
              If U100 had a shorted input on Pin 3, it would cause the signal to be lost, but also, if there is some problem with the feedback signal going to U101, then U101 could have its output of Pin 6 pulled low and be sinking the signal to ground. Is the DDT LED for that channel constantly on? That could indicate such a problem.
              With U100's Pin 3 still disconnected, what resistance do you measure between Pin 3 and the power pins, 4 or 6? That could indicate a shorted input if you measure a low resistance.
              Yes, you can repair the trace with a piece of wire. Just be sure you don't make any solder bridges!
              -David VanHorn
              • 5inc
                • Dec 2016
                • 120
                • Spain
                #67
                Re: Crest CPX1500
                Thanks for the note on the JFET's
                DDT led on that channel IS constantly on. I didn't find any low resistance between pin 3 and 4,6 or 8 (something like 200-400K).
                I connected Q100 and Q101 one by one and still have signal on R102/CR101.
                Will now proceed to replace U100.
                • 5inc
                  • Dec 2016
                  • 120
                  • Spain
                  #68
                  Re: Crest CPX1500
                  I just replaced U100 and of course, it wasn't shorted. It still shows the same symptoms, so I have to go one step back.
                  -"Yes, U101 could be getting some incorrect signal or no signal from another part of the circuit. It could also be faulty."
                  I don't fully understand the feedback thing, so I followed BDFED (hoping it would stand for feedback) on P101 and ended up at the bridge/stereo switch. I saw it was set to bridged so switched it back to stereo, and fired the amp but everything is the same. I may be on the wrong foot but that's all I could figure out for now.
                  I have been checking the delay signal on R106 and it does rise after the delay startup time(I forgot to mention earlier) Should I understand that this indicates DELAYA signal is correct?
                  I also just checked voltage on R102/CR101 and it recieves voltage (-11,8VDC) from the second I turn on the device...
                  R105 does the same.
                  • Agent24
                    I see dead caps
                    • Oct 2007
                    • 5097
                    • New Zealand
                    #69
                    Re: Crest CPX1500
                    Yes it sounds like DELAYA signal is working properly. So that's good...
                    But, the DDT LED indicates that the audio signal is clipping (or it thinks it is) and that the amplifier is running in limiting mode.
                    Since I assume you are not overdriving anything during testing, then there seems to be a fault somewhere which is causing the limiter to want to drive the JFETs off, either because of a fault in the limiter itself or in the signal path before it.
                    If you toggle the GCL Enable\Defeat switch on the rear of the amp, can it turn the DDT LED off?
                    Also, have you checked or changed or done anything else with U102 yet?
                    -David VanHorn
                    • 5inc
                      • Dec 2016
                      • 120
                      • Spain
                      #70
                      Re: Crest CPX1500
                      Yay, DELAYA seems ok!
                      I think I'm not overdriving anything, there's no load on the outputs, but I have pulled the volume knob all the way up to get the signal through the amp for tracing.
                      I just toggled GCL enable switch while the amp was on, it made no difference.
                      About U102, no, I still havn't changed it. I have tested it on diode mode on the DMM but didn't find any shorts. I could give it a shot, after checking resistors around.
                      • 5inc
                        • Dec 2016
                        • 120
                        • Spain
                        #71
                        Re: Crest CPX1500
                        I just replaced U102. I must have done something really wrong in the process because now it blinks again(both DDT leds and Channel A yellow led).
                        Found a short between pins 3-4 of U101.
                        R194 has raised to 76ohms
                        Gonna replace them both tomorrow evening unless you think there might be something that caused this to short again (obviously appart from me)
                        • 5inc
                          • Dec 2016
                          • 120
                          • Spain
                          #72
                          Re: Crest CPX1500
                          Sorry the short on U101 wasn't on pins 3-4 but on 7-8, that both go to 15V+IA.
                          That, and having a burned R194 seems to point that U100 was bad again instead. Replaced both, and now everything looks stable:
                          I fire it up and first both DDT leds turn on, then they go off for a glimpse to turn back on together with green power leds for a few seconds. Finally, after another second or two, DDT leds go off and only power leds remain on.
                          After seeing all this I plugged in my testing speaker, saw it made no difference in fan speed, and tested the volume, but no luck, will try to trace and see where it leads.
                          • Agent24
                            I see dead caps
                            • Oct 2007
                            • 5097
                            • New Zealand
                            #73
                            Re: Crest CPX1500
                            Sounds like you're making some progress!
                            The fact that both channels do the same thing now is probably a good sign.
                            Though, hopefully the user manual could tell you if the LED sequence you're seeing is the correct one.
                            I assume you have tried the test speaker on both channels?
                            And just a thought, but maybe it needs two connected at the same time?
                            -David VanHorn
                            • 5inc
                              • Dec 2016
                              • 120
                              • Spain
                              #74
                              Re: Crest CPX1500
                              you see, channel B has been dissconnected at TX15/16/17 from almost the beggining.
                              Perhaps that has something to do with the fact the signal won't arrive to the output? I wanted to get channel A running before jumping on to B, for the sake of simplicity and a better learning experience.
                              And since everything looks stable now (no shorts) it could be the time to start laying hands on channel B...
                              But on the meantime, I started signal tracing, and could only hear up to R101/R102. Nothing around U100/101/102.
                              and about U101... I'm a bit confused with this one. I've been seeing different things that seem weird, but I'm not sure. Appart from the fact that I don't know nothing about feedback. And appart from the fact DDT led finally goes off, I have no other constance that that is ok (so I'm not certain yet).
                              Then, between pins 3-6 I have a 40 ohm short. It took me 2 pieces to realize the short wasn't on the IC but on the track.
                              Between pins 2-3 I have 100ohm (as expected), but pins 1-2 show 140ohms. Also on the track.
                              The ICs out of circuit read fine.
                              I also made one test with U101 out of circuit and signal arrived to the output! But as soon as I put it back on I lost all signal again to return to a constant pinkish noise.
                              Do you know how I can test feedback signal?
                              • 5inc
                                • Dec 2016
                                • 120
                                • Spain
                                #75
                                Re: Crest CPX1500
                                Just plugged TX15/16/17 in to start troubleshooting Channel B.
                                Turned it on and averything started fine, but channel B DDT red led stayed on.
                                I then tried replacing U200, U202 thinking that would do the trick but nothing, still have red DDT led staying on.
                                I haven't replaced U201 again because I don't think I damaged it(and even so, on channel A U101 is blocking the signal anyway). What I may try is replacing CR200/201, those are still the originals, even though they read ok..
                                I tried to follow the steps you gave me to troubleshoot channel A (pin3 of u200, pins 1/6 of U201, and measured C203 for short) with no success, channel B red DDT led remains on.
                                I even took out U101 (which allows the audio signal to arrive to channel A output) and U201 to see if that gave me any news. I still have channel B DDT led on.
                                Pin3 on u100 and r102 are in contact with CR100/101 on the track, pin3 on u200 and R202 are NOT in contact with CR200/201. I broke off a track under U200 so I had to run a wire from R202 to JMP112 to arrive to CR200/201. No luck.
                                Also, now, when I lift pin3 of U100, the whole amp goes into short and the bulb lights up.
                                Now I have a channel A (with a missing U101 and Q100/Q101) that sends output, and a channel B (missing U201, Q200/Q201, and pin3 of U200 lifted) with a red DDT led staying on.
                                • Agent24
                                  I see dead caps
                                  • Oct 2007
                                  • 5097
                                  • New Zealand
                                  #76
                                  Re: Crest CPX1500
                                  So when you say channel A sends output, there is audio all the way to the speaker? Is it at an OK volume or do you have to have the volume control all the way up for that as well?
                                  Can you get some photos of the board where DDTS2A/DDTS2B lines etc comes from?
                                  Also some photos of the areas we are working on: U101\U201 etc.
                                  Take voltage measurement at U202 pins 2,3,7 and compare with U100 on the same pins. What do you get?
                                  -David VanHorn
                                  • 5inc
                                    • Dec 2016
                                    • 120
                                    • Spain
                                    #77
                                    Re: Crest CPX1500
                                    Yes, I have audio all the way to the speaker. It seems to be at an ok volume on my testing speaker, havn't tried feeding it to a sub yet so I can't assure it is responding at 100% yet. It does sound pretty neat though.
                                    I took voltages at pins 2,3,7 on U202 to compare with U102
                                    pin-----2----------3---------7
                                    U202 -14,3v_____10,9v_____14,6v
                                    U102 14,4v_____14,5v____-14,2v
                                    Here are pictures of the back and front of the zone, DDTS2A is pointed and written DDTS2B is only pointed out. U102/202 also appear. Ask if you want more pictures
                                    • Agent24
                                      I see dead caps
                                      • Oct 2007
                                      • 5097
                                      • New Zealand
                                      #78
                                      Re: Crest CPX1500
                                      What is the type of IC used for U101 and U201? I was fooled by the schematic into thinking it was using an NJM13700M but I can see now it was not.
                                      Also I noticed some potential damage to a track in one of your photos. If that is a bad track, you should inspect the whole board for any more damage like that.
                                      Can you check voltages at U102, pins 5 and 6, as well as on the other side of R118 and R119? Do the same for U202 as well.
                                      -David VanHorn
                                      • 5inc
                                        • Dec 2016
                                        • 120
                                        • Spain
                                        #79
                                        Re: Crest CPX1500
                                        The IC is a NJM4560.
                                        What you saw is stain a of permanent marker i made when replacing them. The track is ok, and there aren't any other obvious bumbps, flaps or cracks on the board.
                                        Voltages at U102 pin5 has 1.6v and pin6 has 1.6v. This is ok
                                        R118/119 have 1.6v on both ends.
                                        U202 has 0.8v on pin5 and -9,9 on pin6. This is weird.
                                        R218 has -9,9v and R219 has 0,8v on both ends.
                                        • 5inc
                                          • Dec 2016
                                          • 120
                                          • Spain
                                          #80
                                          Re: Crest CPX1500
                                          i've been comparing voltages, but don't comprehend some things, maybe you can help me understand, or at least see something I can't
                                          THERMAL_A is 3.75v, THERMAL_B is -12.11v
                                          pin6 of U202 carries -9.8v
                                          Q104/105 carry 80.4v on their collectors, 1.4v on their emitters.
                                          Q205 had nothing on collector, and followed all the way up to -leg of bridge rectifier, whitch was apparently disconnected, reconnected it, and now when I start it it starts fine for two seconds, then blinks(along with the lightbulb) every second and a half. Soft speaker thump included.
                                          --THERMAL_B is now at 3.7v!
                                          --U202 pin6 is now at 1.5v!
                                          That reminds me of something an IC did when another IC controlling it had a fault. And considering I have no U101/201 or Q100/101... I just tried placing U101/201 and it starts as shorted. Ended up taking them out.
                                          • reessi
                                            Acer Alienware Nitro 5 (GH51G LA-L181P Rev: 1C) & Dell Alienware M15 (GDP50 LA-k452P Rev:1) - Both Shorted PCH (SRKMA)!
                                            by reessi
                                            Hello:
                                            I have both these laptops that appeared to have the same fault so I thought I would have a go at replacing the PCH as something to learn, they both are using the same PCH.
                                            I started on the ACER:
                                            Up on connecting power no signs of life but current is consumed from the bench power supply and the PCH is hot.
                                            All the caps on the top of the PCH are showing 0 ohms and the 3.3v rail is shorted.
                                            I followed the videos from Sorin, Electronics Repair School on Youtube heating from the bottom to install. I practised on around 10 scrap...
                                            10-29-2024, 11:31 AM
                                          • learnmore
                                            Help choosing replacement transformer for Levoit 600S SMPS
                                            by learnmore
                                            Hello Community,
                                            I'm attempting to repair the SMPS on a Levoit 600S air purifier. The transformer is likely open. It has six pins. 2 pairs on the primary side (one pair tested open) and 1pair of legs on the secondary side. Legs on the secondary side are spread wide apart (end-to-end). I've circled them in blue. Dimentons of the transformer are 22x23mm. Marking on the transformer are EE16 1.9mH-B. Fuse for the powerboard is fuse 250V 3.15A. I've been unable to find an exact match and I'm hoping for some help here to find a suitable replacment.
                                            Peace
                                            LM...
                                            11-04-2025, 09:35 AM
                                          • Yanleb
                                            Panasonic TC-P50S1 shorted TXNSC1EDUU (TNPA4782AB) SC Board
                                            by Yanleb
                                            Hi every one. I just want to share the successful repair of my shorted Panasonic TNPA4782AB SC board.
                                            I don't usually do it but so many parts were shorted that I decided to buy the ShopJimmy repair kit instead of sourcing each components individually. Here is the list of components the kit contained, with a remark indicating if I needed that part or not :
                                            D608, good
                                            D618 RF101L4S, good (See warning 1 below)
                                            D619 RF101L4S, good (See warning 1 below)
                                            D865 RF101L4S, shorted (See warning 1 below)
                                            IC464 Inverter HA MC74HC1G14, shorted...
                                            03-09-2020, 12:00 AM
                                          • modtool
                                            HP Victus 15-FA1025NL , board DAG3PAMB8D0 REV:D, shorted CPU DrMOS and n-mos
                                            by modtool
                                            Hi to all,
                                            the owner of this machine gave it to me after having opened it to take off the dust from the inside, telling me that after that it didn't start anymore.
                                            I found that one of the DrMOS (AOZ5016QI) on the CPUCORE was shorted between VIN and VSWH, and that one of the two n-mos connected to the power jack was shorted aswell.
                                            I unsoldered the two shorted components and verified that there were no other short circuits, then decided to power the mainboard, firstly with the lab power supply, then with its original HP 200w charger, but the behaviour is the same, the board starts...
                                            04-01-2025, 08:22 AM
                                          • ohren
                                            Original Xbox 1.0 PSU: Foxlink FTPS-0002 Rev. B. — 12 V low, 5 V high — transformer?
                                            by ohren
                                            Hello!
                                            I'm looking at an original Xbox PSU: Foxlink ftps-0002 rev. B. The 12 V output is too low, and the 5 V output is too (?) high. Unfortunately I don't have an oscilloscope but only a DMM for diagnosis.
                                            Measured voltages with no load:
                                            After rectifier diodes
                                            12V: 6.17 V
                                            5V: 5.67 V
                                            Secondary transformer pins AC measurement (really don't know if this says anything)
                                            12V: 2-2.5 V
                                            5V: 1.04 V
                                            I also have another (working) Xbox psu from Delta. The same measures there, in order, being: 10.79,...
                                            03-18-2023, 10:33 AM
