Kenwood CS-4035 Oscilloscope repair questions
  • Crystaleyes
    • Sep 2021
    • 493
    • Brazil
    Hola all,
    I recently picked up two dead scopes in a job lot. One had missing and blown components as well as a toasted trace on the PSU, whereas the other was complete and not burnt.
    Anyway, managed to get both PSU's working enough to switch the scopes on, with the LV rails all more less where they should be, and that is;
    +140v
    +8v
    +5v
    -8v
    Minus 8v is the adjustable reference voltage, which actually doesn't seem to adjust on either PSU. There is a KMA 01 hybrid IC with a couple of op-amps in it which the voltages pass through.
    The problems arise when the PSU is connected to the main board and immediately all the rails drop by around 20% or so.
    I'm just wondering how to go about locating the problem? I have a millivolt meter but no working scope.
    Would I be correct in thinking that considering how the fault is across all the rails and that all rails are dependent on the minus 8v, that this is where the problem should lie?
    There are 37 electrolytics on each main board. I'm planning on testing them first.
    Also enclosing the service manual, in case anyone wants a look...
    Would be great to get these working.
    • 31017
    Re: Kenwood CS-4035 Oscilloscope repair questions
    can you post a photo of the kma01 hybrids?
    i see a 33uf cap, wondering if it's tantalum or an electrolytic!
    • redwire
      • 3910
      The power supply's rails are all interdependent which makes troubleshooting really difficult.
      The -8V rail is the master and if it sags, it will upset the other rails. The +8V (thus +80V and +140V) regulators use the -8V as their reference.
      The module schematic does not look quite right, there should be a path to some reference for the top two op-amps (140V, 5V reg) but not a big deal.
      So I would watch the -8V rail for shifting, it's critical that rail does not sag when you plug in the other board.
      If the HV has problems it will overload the raw +/-10V rail which feeds the +/-8V regs. I'd think about pulling flyback transistor Q8 to stop that.
      With no load, many outputs i.e. the +80V and +140V will go too high due to the helper resistors R51, R1. This is normal. There's R5 helper as well for the +8V so it might too go higher with no load. You might have to make up some dummy loads to test the PSU stand alone. It's a big hassle.
      • Crystaleyes
        • Sep 2021
        • 493
        • Brazil
        Looks more like an smd electrolytic but everything is painted black
        • Crystaleyes
          • Sep 2021
          • 493
          • Brazil
          Hi Redwire
          Excellent news then?
          On the forum where I found the .jpg (and also a custom replacement board design) they were also discussing this design perhaps being incorrect
          Haven't been able to test the HV properly due to lack of probe, only measured on TP12 where it measured -353v which is within spec
          This is an area I have next to no experience. Would the dummy loads be resistors and how would I calculate the values if so?
          It's a big hassle.
          Possibly. At the moment, it is still a challenge which I'm up for attempting
            • 31017
            that cap on the hybrid probably needs to go
            • Crystaleyes
              • Sep 2021
              • 493
              • Brazil
              Cool. Thanks.
              Shall give that a go.
              Could it be that there is actually no problem on the main board and that the voltages are dropping due to the -8v rail failing under load?
              The way I had been thinking is that as something is pulling the voltages down then something is 'failing' as opposed to s dead short, so I was planning to test all the caps, followed by the semiconductors etc.
              And here is the IC off the board
              • Crystaleyes
                • Sep 2021
                • 493
                • Brazil
                Actually, upon closer inspection there appears to be corrosion on the cap and on one of the op amps.
                Doing a quick test shows that only IC pin 3 has disconnected, but that is seemingly unconnected anyway.
                The cap however, gives a reading of 1430nF, an ESR of 115Ω and 32% Vloss, so that is fooked, and will delicate removal
                • redwire
                  • 3910
                  It's a good find, all of my early (1990's) SMT electrolytics did not last for some reason it's like the PCB assembly heat was too hard on them.
                  I'm not sure what solvent would work to dissolve the epoxy coating, or if heat is enough. The pads are likely fragile if this is thick-film substrate with tinned metal on top. It's surely got fragile thick-film resistors printed on the board so scraping is pretty risky.
                  Can you try anything like acetone or xylene and a q-tip swab etc.
                  • Crystaleyes
                    • Sep 2021
                    • 493
                    • Brazil
                    Didn't need anything other than a quick blast of heat after running a craft knife along the edges
                    The pads look a bit oxidised but as I am in the humidity of Rio at the moment it's not so surprising on a 30+ year old pcb, and the 33uF cap now repeatedly measures at 43pF!!!
                    Just gonna connect up a standard electrolytic to test and I'll see about tidying things up properly if it works
                    • redwire
                      • 3910
                      The cap is across the 5.1V reference zener, so I would expect 5.1V if it's working. Note it's -ve wrt GND, a -5.1V ref. so (+) goes to GND.
                      • Crystaleyes
                        • Sep 2021
                        • 493
                        • Brazil
                        Thank you for that. I hadn't noticed that minor detail in the polarity
                        So on a more serious tip, why is the cap reversed polarity? What job is it performing being that was around?
                        • redwire
                          • 3910
                          The cap is part of the -5.1VDC reference for the -8V regulator. It then makes the master reference for the other rails. Just to save money, use one low drift zener reference instead of a bunch, like one per for each rail. It's also a single calibration at the factory.
                          I would use a variac to power it up after, just to make sure the -8V rail is working.
                          Back in the 80's these Japanese scopes were copied by the Koreans and made a lot cheaper.
                            • 31017
                            i would put an smd tantalum cap across the pads
                            • Crystaleyes
                              • Sep 2021
                              • 493
                              • Brazil
                              Unfortunately changing the cap didn't resolve the lack of -8v adjustment and therefore neither any of the voltages on the main board, although I suspect there is something else bringing the voltages down once that board is connected.
                              With that board disconnected the voltages are not too far out and the tube has a bright spot in the middle of the screen.
                              Anyway, whilst the Hybrid was off the board, I took the opportunity to take measurements based on the schematic, including the zener and all the resistance combinations which all measured fine.
                              Therefore, if this problem 'is' on the hybrid, I imagine it to be on the TO 072 op-amp, which judging by the capacitor, shouldn't be such a problem to replace. Just need to find one now...
                              Thank you so much though for the support and suggestions.
                              Any fresh ideas on where to head next will still be appreciated
                              • redwire
                                • 3910
                                Can you measure the DC voltages on the SIP module's pins, just a quick check to see if anything looks off. Probably better with the scope board connected to the PSU board.
                                • Crystaleyes
                                  • Sep 2021
                                  • 493
                                  • Brazil
                                  I'm presuming that the SIP module is this hybrid IC, and I think you meant to ask to see what 'isn't' off.
                                  Enclosing the schematic again for quick reference
                                  The voltages with the main board 'disconnected' are as follows...
                                  (Schematic listed voltages are in brackets)
                                  Pins;
                                  1: +158.8v (+140v)
                                  2: NC
                                  3: -0.3v (+0.6)
                                  4: +7.33v
                                  5; +3.3v
                                  6: 0.0v
                                  7: +8.6v (+8v)
                                  8: +8.09v (+0.33)
                                  9: +4.84v (-0.33)
                                  10: -9.47v (-8v)
                                  11: -0.65v
                                  12: 0.0v Gnd_
                                  Seems tto me that the op amp has failed internally. (Noticable above are pins 1 and 10, which were 142v and 8.2v at the last time of reading)
                                  Then with the main board 'connected'.. Things get curiouser and curiouser...
                                  1: +100v
                                  2: NC
                                  3: -0v
                                  4: +0.6v
                                  5; 3.25v
                                  6: -0.7v
                                  7: +2.33v
                                  8: +3.45v
                                  9: +3.15v
                                  10: -5.7v
                                  11: -1.26v
                                  12: Gnd_0.0v
                                  It's perhaps a good thing that I have two of these scopes...
                                  Speaking of which I pulled the KMA 01 from the 2nd scope and once again the cap measured bad. This one had 77% VLoss and 110Ω ESR
                                  • Crystaleyes
                                    • Sep 2021
                                    • 493
                                    • Brazil
                                    I'll bear that in mind for once these are fixed.
                                    One question, "would the failing electrolytic be enough to fry the op-amp?"
                                    • redwire
                                      • 3910
                                      Can you measure voltage across the new cap, things are acting like the 5.1V zener is pooched or the op-amp driving pin 9 is not working, or 10V zener D5 or Q5 has failed, connecting to pin 9.
                                      I'm guessing if the 5.1V zener reference is not there (shorted or open), the op-amp can look bad but it's really just making the wrong output based on a bad ref.
                                      What really doesn't make much sense is pin 11 (-8V feedback pot) reads more negative when pin 10 is more positive. That's either a bad measurement or that op-amp has a failed input.
                                        • 31017
                                        no idea
                                        but op-amps fail a lot anyway
                                        btw, TO 072 is probably supposed to be TL072
                                        • SarcasticOP
                                          General question about MacBook power rails for repair.
                                          by SarcasticOP
                                          I am looking at an A2337 and found this for the order of power rails and what they do. Can this be confirmed before I dive into hunting down power rails that do not exist or are not neccisary?
                                          • PPBUS_G3H (12.6V-12.8V)
                                            • This is the main power rail that supplies power to the entire system. Check this rail first, as it must be present for other power rails to work.
                                          • PP3V42_G3H (3.42V)
                                            • This power rail is derived from PPBUS_G3H and is used to power the SMC (System Management Controller) and other essential circuits. It should be present if PPBUS_G3H is working correctly.
                                          • PPVRTC_G3H (3.3V)
                                            • This
                                          12-12-2024, 05:50 PM
                                        • Cemal Odabasioglu
                                          MacBook Pro A1707 Motherboard Diagnostic Help – CD3215 heating and Fluctuating Rails
                                          by Cemal Odabasioglu
                                          I'm currently diagnosing a MacBook Pro A1707 motherboard that does not power on. I've gathered some data but am stuck and need expert advice.
                                          Note that i'm a beginner in electronics and i'm trying to learn while trying to repair this macbook that my work gave me. I'm really curious and love understanding how thing works and i already learned so much.
                                          Only one usb c port is taking 20v and have 12.5v on PPBUS_G3H. when trying other ports, 2 are not going to 20v and one is going to 20v but the CD3215 is getting really hot.
                                          Observations...
                                          01-14-2025, 03:54 PM
                                        • Tarot Superstars
                                          How not to Damage an Oscilloscope
                                          by Tarot Superstars
                                          Hello.
                                          I have jut got my first analogue oscilloscope. An Isotech 620 20MHz.
                                          I want to learn how not to damage it, and I found instructions on the internet obscure.
                                          They talk about ground, earth, floating ground, this, that....
                                          But they don't actually say anything anything useful so far.
                                          I want to know:
                                          Don't connect exactly this to exactly there or it blows up.
                                          For example, I am tracing a signal from a function generator in an audio amp, or testing for voltages in different parts of a circuit.
                                          I want to know what not to connect...
                                          08-29-2023, 02:27 PM
                                        • ferah831208
                                          help with AMD Ryzen 5 7535hs power rails
                                          by ferah831208
                                          Hello there, does anyone have info about AMD Ryzen 5 7535hs power rails or an schematics from a compatible CPU?. I have a quanta G3PR from a HP victus laptop, I suspect cpu is damaged but I would like to know what power rail is the one I guess is damaged. There are 4 coils aprox 4.2 ohm to gnd in parallel next to CPU (I guess vcore) ,one coil same size next to those 4 around 0.7ohms to gnd (damaged rail I guess), and other two smaller I guess associated to less current power rails for CPU logic purposes. Anyone have more specific info? I attached picture.
                                          03-09-2025, 01:59 AM
                                        • DrewPhillips
                                          SFX PSU - Need High Current on 3.3V & 5V Rails?
                                          by DrewPhillips
                                          I have an older system which was built in the Windows 7 era and which has developed, I suspect, a power supply problem. The system had the same problem, random crashes, a couple of years ago, which I subsequently remedied with a new power supply.
                                          The power supply which came with the machine is a no-name "Safe Power PE-300". It is SFX form factor and claims a high current capacity: +3.3V @ 22A and 5V @30A (the +12V rail claims a meager 16A). Problem is that SFX supplies (or even ATX supplies for that matter) having high current ratings on the 3.3V and 5V rails seem to be...
                                          02-18-2025, 01:03 PM
