Fluke Scopemeter190

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • Domingi
    New Member
    • Mar 2009
    • 9

    #1

    Fluke Scopemeter190

    Hi folks,

    I have just come into a Fluke 190 Scopemeter that has sat idle and was put up for a few years. Everything looks to be in excellent physical condition including probes and leads. After reconditioning the battery, the unit came right up and seems to hold a good charge.
    The VOM functions all work fine, but the scope functions seem to flatline. An Attempt to trim the probes to the build in square wave generator was dysfunctional, even though I can read proper voltage there with an alternate VOM.

    I have been away from component level diagnostics for a few decades and even though I felt comfortable finding my way around this unit, I am not familiar with the particulars of this unit. My initial thought is that it is in need of repairs.

    So I have 2 questions. First...anyone here familiar enough with these to confirm or deny my thoughts on the condition of this unit. And secondly, any suggestions on a decent repair facility that might be able to get this thing back in the groove?

    Thanks much.
    Poultry is just another word for chickens.
  • Domingi
    New Member
    • Mar 2009
    • 9

    #2
    Re: Fluke Scopemeter190

    Ok,

    My n00bness is showing through, along with my age. The "Thumbs down" Icon registered in my brain as a "Question Mark" when I posted this. Working now to figure out how to edit that to something more appropriate. Bad eyes, bad brain, just one mark short of a trifecta here.
    Poultry is just another word for chickens.

    Comment

    • KeriJane
      Mac Enthusiast
      • Sep 2008
      • 681
      • USA

      #3
      Re: Fluke Scopemeter190

      Hi Dominigi

      Welcome to the Forum!

      A broken Fluke? Aren't those supposed to be like really good meters?

      Have you got the manuals for it? Maybe there's something you missed, like some goofy setting or other hidden in some obscure sub-menu.

      Fluke appears to have manuals for it here:
      http://us.fluke.com/usen/support/manuals/default.htm

      Good luck,
      Keri

      PS.
      Getting a scope to operate: Easy
      Figuring out what the darn thing is trying to say: AAAARRRRRGGGHHHH!
      The More You Learn The Less You Know!

      Comment

      • Toasty
        Badcaps Legend
        • Jul 2007
        • 4171

        #4
        Re: Fluke Scopemeter190

        Not to state the obvious, but -
        Is the probe good? Did you check it for continuity? Try another probe?

        Fluke offers repair and calibration services not only for their equipment but many others. There are also assorted repair and calibration services throughout the country. A Google search may be in order.

        Perhaps: http://www.byramlabs.com/
        veritas odium parit

        Comment

        • Domingi
          New Member
          • Mar 2009
          • 9

          #5
          Re: Fluke Scopemeter190

          Good morning folks,

          Thanks for the replies so far. I can see that I should have been more descriptive in my original post.

          This unit is in just mint condition as are all the leads, probes carrying case etc. which is all the more confusing to me. It doesn't look like this thing worked hard a day in its life. All of the original paperwork, manual, CD serial cable, 2 sets of probes and several adapters are present and in the same condition.

          I didn't think to originally, but I did go ahead and check the impedance of both probes and found then to be sitting at 9 M ohms and the grounds to 0 without any shorting between so they appear to be fine.

          When I go through the probe calibration routine, I read no wave on the screen. Understand that the lcd screen is functioning overall very nicely. This occurs on either probe in either (A/B) input. I went ahead and popped the case for a quick peak inside thinking that perhaps there was an internal fuse or 2. I saw no fuses, and what I did see was very pristine.

          Also just to clarify a bit , the unit is in the 190 series, but more specifically is a 192 Scopemeter 60 MHz 500MS/s V04.04 with calibration #1 dated 5/28/2001.

          I also tried setting the unit back to factory defaults just in case there was a "software/firmware" bug.

          Hmmm, I just found that a firmware update exists to take this to v4.08. Would hate to think that software was the issue here but I think I will download and installing that to see what happens.
          Poultry is just another word for chickens.

          Comment

          • Toasty
            Badcaps Legend
            • Jul 2007
            • 4171

            #6
            Re: Fluke Scopemeter190

            Okay. I mean no disrespect and I don't mean this in any way but good...

            Why are you screwing around with a $3,000 piece of equipment? If it's not working, get a hold of Fluke and get it serviced and calibrated. I'll bet they update the firmware while they're at it. Even if it cost a few hundred to do so... it's money in the bank.

            I've seen other folks "have a go" at fixing test equipment before, with some expensive lessons learned there...

            Just my 2¢ worth
            veritas odium parit

            Comment

            • Domingi
              New Member
              • Mar 2009
              • 9

              #7
              Re: Fluke Scopemeter190

              Thank you for the very succinct reply there Toasty, no disrespect taken here in the very least.

              I am screwing around with it simply because I am a man, and that is what we do. I've never been able to quite put my finger on it, but it's the same reason I screw around with my boats, bikes and cars.

              You needn't have any concern of me "having a go at fixing" this unit, I have no intent to perform any sort of component level repair on this unit. I was just looking to see if there was anything obvious. Perhaps things have changed in the last 30 years, but the first thing we were taught was to take a look and smell. Upgrading the firmware seems no more complicated than updating a bios in a computer, and is recommended as a user level "field upgrade" by Fluke.

              I have tried to get hold of Fluke's service department this morning without any success as of yet, but have found a few authorized Fluke repair firms that I have on my short list and should my efforts fail, it will be on the first bus out.

              I thought , perhaps there was a person or 2 here in these forums that might have first hand knowledge of this unit and any peculiarity or tip that might make for an easy fix. Thanks again for your positive and informative contributions to this thread.
              Poultry is just another word for chickens.

              Comment

              • Domingi
                New Member
                • Mar 2009
                • 9

                #8
                Re: Fluke Scopemeter190

                Ok, new information.

                The Firmware upgrade succeeded however the unit will still not read a waveform, so unless I happen across a "secret recipe" to fixing this, it will be off to the repair shop.

                I just got off the phone with the fine folks in the service area of Fluke out there in Everitt WA.

                They "flat rate" repairs as follows:

                $615.00 + freight to fix anything and everything that is wrong with the unit. Includes calibration.

                Calibration only is $198.00

                Their standard "target" turnaround after receipt of the unit in house is 10 days.

                Several of the repair centers I have looked at also go by the "flat rate" program, but the least expensive that I found was $695, leading me to believe that the take and reship the unit to Fluke and take a drag off the turnaround.

                I have found a few shops that charge a flat rate "bench testing" fee which would get me in the door for a bit less, but as hard as it might be to believe, none of them are down here in the Florida Keys and the thought of not being able to walk-in with this leaves me with a cause for concern.

                I'll stop back in once my final decision is made and let y'all know how it goes.
                Poultry is just another word for chickens.

                Comment

                • Toasty
                  Badcaps Legend
                  • Jul 2007
                  • 4171

                  #9
                  Re: Fluke Scopemeter190

                  >>I am screwing around with it simply because I am a man, and that is what we do.<<

                  Are you originally from Missouri? The "hold mah beer and watch this" driver state?



                  Anyway, the link that I put in my first reply, they have a facility in Orlando.

                  I might also inquire of them as to why a brand new unit of theirs has failed. *ahem* "Is this something I should expect from Fluke equipment?"

                  Toast
                  Last edited by Toasty; 03-06-2009, 01:10 PM. Reason: Mis-read -fixed
                  veritas odium parit

                  Comment

                  • Domingi
                    New Member
                    • Mar 2009
                    • 9

                    #10
                    Re: Fluke Scopemeter190

                    LoL

                    Nah, not from Missouri... but I did stay at a Holiday Inn Express last night.
                    Poultry is just another word for chickens.

                    Comment

                    • Toasty
                      Badcaps Legend
                      • Jul 2007
                      • 4171

                      #11
                      Re: Fluke Scopemeter190

                      Did you perform the unit Reset function? Maybe you did if you updated firmware...

                      It's in the Service Manual from Fluke.

                      4.4.1 Resetting the test tool
                      Proceed as follows to reset the test tool:
                      • Press [Power Button] to turn the test tool off.
                      • Press and hold [User Button].
                      • Press and release [Power Button] to turn the test tool on.
                      • Wait until the test tool has beeped twice, and then release [User Button]. When the test tool has beeped twice, the RESET was successful.

                      Cheers!
                      veritas odium parit

                      Comment

                      • Domingi
                        New Member
                        • Mar 2009
                        • 9

                        #12
                        Re: Fluke Scopemeter190

                        Hey Toasty,

                        Thanks for the info. Yes, I did perform the reset procedure, it is also included in the user's manual.

                        If you have access to a service manual, what does it say in the chapter titled "Darned thing looks perfect but won't grab a signal"?

                        Thanks again.
                        Poultry is just another word for chickens.

                        Comment

                        • Toasty
                          Badcaps Legend
                          • Jul 2007
                          • 4171

                          #13
                          Re: Fluke Scopemeter190

                          The Service Manual is available on Fluke's site.

                          But let me have a look for you...

                          Hmmm. Ahh! Here it is -

                          Open case and look for obvious signs of problem - broken connection, corroded wires or loose circuit board interconnects, rodent/insect infestation, gremlins, or Leprechauns abuse.

                          After that are some sub-sections, and the last section is "Other Uses":

                          a) How to use equipment as a hammer or other striking device
                          b) How to use equipment to gain entry to a locked vehicle
                          c) Proper method for shot-putting equipment
                          d) Proper method for drop-kicking and not causing self-injury
                          e) Use as bookcase, night table, foot rest, chair or table leveling device
                          f) Doorstop and wheel chock methods
                          g) Classroom/Educational uses on how not to treat test equipment
                          h) Small caliber impact resistance test
                          i) Large caliber impact resistance test
                          j) High voltage resistance testing and connection procedures
                          k) Acids and bases exposure resistance methods
                          l) Solvent resistance measurements
                          m) Proper method for flame resistance testing
                          n) Cooking and barbecuing with your meter
                          o) How to use as a beverage holder - non-alcoholic
                          p) I have to.... brb
                          q) How to perform the Urine Standard Analysis test
                          r) Proper display and use as a holder for alcoholic beverages
                          veritas odium parit

                          Comment

                          • Toasty
                            Badcaps Legend
                            • Jul 2007
                            • 4171

                            #14
                            Re: Fluke Scopemeter190

                            On a more serious note, I went digging in my records and found the company I used years ago. They repaired, certified, and calibrated my service monitors, signal gens and freq counters:

                            Auburn Electronic Labs
                            12345 Bowling Rd
                            ( Highway 68/80 )
                            Auburn, KY 42206

                            270-542-6000

                            I did a quick Google and it looks like they're still there.
                            veritas odium parit

                            Comment

                            • Domingi
                              New Member
                              • Mar 2009
                              • 9

                              #15
                              Re: Fluke Scopemeter190

                              Oh, nice find.

                              Looks like my Firefox + NoScript was keeping me a little too safe, wasn't able to get to those pull down menus before. Thanks much for the lead.

                              Looks like I have a little perusing to do this evening.
                              Poultry is just another word for chickens.

                              Comment

                              • Bob Parker
                                Technician
                                • Feb 2007
                                • 182
                                • Australia

                                #16
                                Re: Fluke Scopemeter190

                                As they say in the classics: "If all else fails, read the manual". Before you start concluding that there's something wrong with it (there probably isn't), be absolutely certain that you're driving it properly.

                                Go to the part of the manual about "Basic operation" and how to get it to display the probe compensation waveform. Maybe you've just got the inputs grounded or somethng really basic like that.

                                Good luck.
                                It is a good shrubbery. I like the laurels particularly...

                                Comment

                                • Toasty
                                  Badcaps Legend
                                  • Jul 2007
                                  • 4171

                                  #17
                                  Re: Fluke Scopemeter190

                                  Any word there Domingi?
                                  veritas odium parit

                                  Comment

                                  • joelfriend
                                    New Member
                                    • May 2013
                                    • 1
                                    • US

                                    #18
                                    Re: Fluke Scopemeter190

                                    Sorry to revive such an old post, but I haven't had success looking anywhere else.

                                    Domingi, if you're still an active member and are receiving messages, I am looking for the firmware upgrade file for the original 190 series like yours. Do you still happen to have that firmware lingering around on an old floppy somewhere?

                                    Comment

                                    Related Topics

                                    Collapse

                                    • ngml
                                      Repair of an LG split AC system indoor unit PCB
                                      by ngml
                                      I am trying to repair an LG split AC indoor unit, which does not start and gives a CH05 error code.
                                      This indicates an error in the communication between the outdoor and the indoor unit.
                                      The outdoor unit is fine, so the problem is the indoor PCB.
                                      The indoor unit is an LG PC09SQ NSJ (3SNM09JA2FA), and the PCB has the following type number: EAX35907219-1.2 (see photo1)
                                      I changed optocouplers ICO1X & ICO2X and the NPN transistor S9013 which drives optocoupler ICO2X (see photo2), but this did not remedy the fault.

                                      Does anybody have the schematics of this board,...
                                      02-01-2025, 12:55 PM
                                    • nobbnobb1
                                      Bose Companion 5 - Unit Powers On then Cuts Off Shortly After
                                      by nobbnobb1
                                      Hi guys,

                                      Trying to salvage a Bose Companion 5 computer speaker system that powers off shortly after powering on. It's got a primary USB input (built in DAC) and an aux 3.5mm input. The USB appears to be the primarily source that tells the unit to power on (control pod LED goes green) as the unit has no power switch.

                                      When I first connect power to the unit, plug in the USB, it will play audio nicely for a minute before it drops out and the unit appears to go into standby (control pod LED goes red). It stays in standby and I cannot seem to power it back on without physically...
                                      02-03-2022, 11:14 AM
                                    • Gjackson
                                      Air-conditioning unit power supply failing to power up
                                      by Gjackson
                                      Hi all, I managed to narrow down which PCB was causing my Haier AC to throw out an E7 error code (E7 indicates that the indoor unit and outdoor unit cannot communicate). I found that the outdoor unit power board which gives 5v and 15v to the outdoor unit module board fails to fully power up.. The power Led starts to fade in incredibly slowly and the 5v and 15v power buses jump around like crazy and never reach their required voltage. I unplugged the module board from the power board and the power board powered up fine! It had a steady 5v and 15v out and the power LED came on instantly and Bright....
                                      01-08-2022, 09:24 AM
                                    • Khoja
                                      Fluke 123 no image on the screen
                                      by Khoja
                                      Hello experts,

                                      I recently purchased a used Fluke 123 Industrial Scopemeter from an online Japanese auction site (jauce.com). It has now been delivered. Unfortunately, I overlooked the fact that the seller neither guarantees its working condition nor accepts returns.

                                      When I power it on, the backlight comes on, but there is no image on the screen. After opening the device, I noticed clear signs of previous soldering in several areas (photos attached). It seems someone attempted repairs before.

                                      At this point, I'll have to try and fix it myself. Has anyone here...
                                      04-10-2025, 09:24 AM
                                    • SolomonMan
                                      Vintage BK Precision 1403 Oscilloscope....
                                      by SolomonMan
                                      All,

                                      Some of you have helped me in the past and I am very appreciative of all the help! (I have fixed TVs/Stereos/ and other small appliances with your efforts...Thanks so much!!!



                                      Today I went to a local HAM Radio Festival. It was one of the smaller ones in our area. My stepfather is the HAM operator. I tinker in electronics but am a Software Programmer (27 years now).

                                      I was looking for a few odds and ends for some electronics work I have to do for the household/farm. Typically these type of shows have the grab bags or a electronics vendor...
                                      12-03-2023, 10:39 PM
                                    • Loading...
                                    • No more items.
                                    Working...