Fixing a laboratory DSC

Collapse
X
 
  • Time
  • Show
Clear All
new posts

  • vrasp
    replied
    Re: Fixing a laboratory DSC

    Originally posted by sam_sam_sam
    Here is where you can find a cheaper fuse series

    https://www.amazon.com/Littelfuse-BL.../dp/B07DSPZPX7
    Thanks I ll have a look.

    Leave a comment:


  • sam_sam_sam
    replied
    Re: Fixing a laboratory DSC

    Here is where you can find a cheaper fuse series

    https://www.amazon.com/Littelfuse-BL.../dp/B07DSPZPX7

    Leave a comment:


  • vrasp
    replied
    Re: Fixing a laboratory DSC

    Thanks guys. It s much clearer. I think I ll be using that bulb method much more often from now on.

    Leave a comment:


  • sam_sam_sam
    replied
    Re: Fixing a laboratory DSC

    One note KTK-R-1 is a faster blowing than a KTK-1 fuse

    There are cheaper fuses if you do not need 600 volt voltage rating

    I once troubleshooted a machine that was blowing fuses on the controls circuit used the incandescent light
    ( 100 watt ) bulb instead of the the fuse to find the short in this machine if your light bulb is very bright and you only have a one to two amp fuse in this circuit then you definitely have an issue with the this circuit
    Last edited by sam_sam_sam; 07-15-2021, 05:03 PM.

    Leave a comment:


  • eccerr0r
    replied
    Re: Fixing a laboratory DSC

    Well the example did not indicate what the device under test (DUT) is represented by - it's just generic behavior of the circuit. The DUT could have been either the parallel or single bulb in series. So yes if you hook up two in parallel and treat the singleton one as the DUT you can see that the DUT will need to pass more current and is indicated by such as it'll burn brighter.

    That being said you don't want to allow 6A to go through continuously which is what will happen if you hook up a lot of bulbs in parallel. You only want only one or a few to start off to find "obvious" or critical shorts and then increase them as your confidence that the short has been solved.

    Leave a comment:


  • vrasp
    replied
    Re: Fixing a laboratory DSC

    I had to go back to reading on electricity basics but I think I understand the use of lamp bulbs as a current limiter.

    The part I have a hard time understanding is the number of bulbs to use. If I want to test the device that takes in 6A at 230v, wouldn't the one bulb at 1A be super bright? Then I wouldn't be able to tell if it's bright due to a short or if It's normal.

    Same with the other device that runs at 2A 230v. Wouldn't just one bulb be super bright even if there is no short?

    Leave a comment:


  • stj
    replied
    Re: Fixing a laboratory DSC

    lamps in parallel will allow more current.
    lamps in series are pointless unless they are the wrong voltage.
    i mentioned 2 in series because i thought you were going to do it with the 240v side and being canadian your lamps are 120v

    and dont give it 6a or a fault could cause a fire.
    if you disconnect the load from the power supplies then 1a should be more than enough to get them started.
    which is unlikely because we are searching for a fault!

    Leave a comment:


  • vrasp
    replied
    Re: Fixing a laboratory DSC

    It s more clear but there is still something I dont get.

    Will bulbs in parallel limit more current then in series?

    120v with 100w bulbs is about 0.83 amps. Should I put 7 of these in series to test the 6A Midec?

    Leave a comment:


  • eccerr0r
    replied
    Re: Fixing a laboratory DSC

    Think of it this way: one bulb straight connected to 120V... will not burn out
    Two bulbs connected parallel straight to 120V... will not burn out, and more the same... but more current is being drawn with each bulb added.

    Now connect two bulbs in series. Note that each bulb will be dimmer than before. Measure the current flowing through the two bulbs in series. It's necessarily lower than before! Double up on one of the bulbs. Note that the parallel bulbs are now dimmer than the single bulb...which will now be brighter! Now short out the parallel bulbs... The parallel bulbs will go out but you won't trip a fuse, and the single bulb will be full brightness.

    Don't know if that blurb above makes sense but that's how we're taking advantage of series bulbs to limit the amount of current going through a circuit.

    ---

    Hmm... that thing... now you might have to open it and look inside...

    Leave a comment:


  • vrasp
    replied
    Re: Fixing a laboratory DSC

    My mistake. Yes, the fuse is before the transformer so it s at 120v. I have more pictures of the setup but it s quite messy. Lots of cables everywhere. The Mitec is on the left.

    So if I have let s say, hypothetically, five 120w bulbs in parallel instead of the fuse, and there is a short to ground somewhere. This is what I understand: The current drawn will try to increase until something breaks and the circuit is open. Let s say each bulb only allow 1 amp. Will the rest be dissipated as heat? Or will the bulb break?
    Attached Files
    Last edited by vrasp; 07-15-2021, 08:55 AM.

    Leave a comment:


  • stj
    replied
    Re: Fixing a laboratory DSC

    the limit is the lamps - so in that instance, 100w
    at 240v that would be about 400mA

    your just trying to stat the power supply's and not run the whole unit at this time.

    Leave a comment:


  • eccerr0r
    replied
    Re: Fixing a laboratory DSC

    F**K those KTK fuses, that's those stupid fast blow midget fuses they use in multimeters and we had that other rant thread somewhere on BCN.

    The reason why light bulbs "limit" the current is that it's connected in series with the load. Now unless your device somehow has negative resistance (I'd love for someone to give me a resistor with negative resistance... things that are piecewise negative don't count) the amount of current going through the bulbs can only max out if your device was completely shorted out.

    Now you probably don't need a whole bunch of bulbs in parallel - yet. Unless you're sure the short is in the cooling/heating circuitry. However since this is a KTK fast blow fuse, it might be troubling and more pictures of how things are connected and what it's driving would be interesting. Normally you do not use KTK fuses on SMPS's because of inrush!

    As a rule of thumb, since 120V at 1A is 120W, a 100W bulb is a bit under 1 amp. You can make a rough guess at how many bulbs you need in parallel.

    So where is the fuse actually, is it in the 230V side or 120V side, I thought it was 120V wall plug - switch - fuse - step-up transformer 230V - that MITEC PSU that needs 230V.
    Last edited by eccerr0r; 07-15-2021, 08:23 AM.

    Leave a comment:


  • vrasp
    replied
    Re: Fixing a laboratory DSC

    Originally posted by stj
    no, your using it to limit the current - not allow it all
    a couple of 100w 120v lamps in series will do for now
    How do I know how much current I m limiting? Wouldn't I want enough bulbs that the current doesn't go over 5A so I dont damage anything? Also the voltage on the line is 230v.

    Thank you!

    Leave a comment:


  • stj
    replied
    Re: Fixing a laboratory DSC

    no, your using it to limit the current - not allow it all
    a couple of 100w 120v lamps in series will do for now

    Leave a comment:


  • vrasp
    replied
    Re: Fixing a laboratory DSC

    Originally posted by stj
    series lamp-bulb time.

    btw, can you show us this amazing $50 fuse??
    The fuse is a KTK-5 Bussmann. 5 amps 600VAC. It s actually $50 canadian.

    Can I really use the lamp bulb method in this case? The line is 230v. If I understand correctly, I would need a 230v*5A=1150w bulb? That sounds like a pretty powerful lamp bulb. Maybe as expensive if not more than the fuses. Correct me if I m wrong.

    Leave a comment:


  • stj
    replied
    Re: Fixing a laboratory DSC

    series lamp-bulb time.

    btw, can you show us this amazing $50 fuse??

    Leave a comment:


  • eccerr0r
    replied
    Re: Fixing a laboratory DSC

    just a hunch because they went out of business it seems (website is gone?!) and switchmode PSUs tend to have their own set of failure mechanisms that aren't shared with linear PSUs.

    But need to look at that PSU and confirm, it's only a speculation at this point.

    Leave a comment:


  • vrasp
    replied
    Re: Fixing a laboratory DSC

    Of course! Haha

    How do you know Midec is unreliable? Or are you saying this because of the switching power supply?

    Leave a comment:


  • eccerr0r
    replied
    Re: Fixing a laboratory DSC

    BadCaps.Net

    Leave a comment:


  • vrasp
    replied
    Re: Fixing a laboratory DSC

    Interesting. I ll take a closer look at that Midec. It does have a fan on the side which I suppose is next to a heatsink inside.

    By the way what's BCN?

    Leave a comment:

Related Topics

Collapse

  • magneh
    Transformer Hahn 230V 20v 6va
    by magneh
    Hello all.

    Not very familiar with transformers and did not find the datasheet for this one.


    Would these resistance values from Pimary and secondary normal? (check photo in red)
    Maybe someone with more experience can help me here
    Appreciate
    04-25-2025, 03:57 AM
  • ohren
    Original Xbox 1.0 PSU: Foxlink FTPS-0002 Rev. B. — 12 V low, 5 V high — transformer?
    by ohren
    Hello!

    I'm looking at an original Xbox PSU: Foxlink ftps-0002 rev. B. The 12 V output is too low, and the 5 V output is too (?) high. Unfortunately I don't have an oscilloscope but only a DMM for diagnosis.

    Measured voltages with no load:

    After rectifier diodes
    12V: 6.17 V
    5V: 5.67 V

    Secondary transformer pins AC measurement (really don't know if this says anything)
    12V: 2-2.5 V
    5V: 1.04 V


    I also have another (working) Xbox psu from Delta. The same measures there, in order, being: 10.79,...
    03-18-2023, 10:33 AM
  • mikey5791
    Is it possible for transformer substitution?
    by mikey5791
    Hi all,
    Got this local made DC power supply (12/14V) given free as the transformer has melted. There is no marking or indication to identify what type or rating of transformer used.
    Fyi,. I had a busted autogate mainboard with transformer with marking 13-0-13 . Is this the 13v dc type transformer?
    My questions are
    1. How do i test to see if the transformer is in good working condition?
    2. Can this transformer (marking 13-0-13) be used to replace the melted transformer on the DC power supply?
    Hope you guys understand what i was asking. My intention is...
    11-03-2022, 02:11 AM
  • dennis_elx33
    3PHASE STEP-DOWN Transformer in Check
    by dennis_elx33
    Hello All,

    i would like to post an interesting topic regarding 3-phase stepdown transformer which we suspect to cause an overcurrent (138A) which resulted to power tripped off of the HV power supply (80KV output).


    Problem statement: Power tripped Off

    Initial Problem: Found 2 IGBT and 2 rectifier diodes shorted (burnt).

    Action Done:
    1) Replaced the faulty parts with good spare.
    2) Measured impedance of the 3phase input lines and found that L2 to L3 has 23.3ohm while other pairs are 51Kohm. This pins will...
    12-02-2024, 06:52 AM
  • Document Archive
    ASUS Transformer T101HA - X5DHD Hybrid (2-in-1) Book Specification for Upgrade or Repair
    by Document Archive
    This specification for the ASUS Transformer T101HA - X5DHD Hybrid (2-in-1) can be useful for upgrading or repairing a laptop that is not working. As a community we are working through our specifications to add valuable data like the T101HA - X5DHD boardview and T101HA - X5DHD schematic. Our users have donated over 1 million documents which are being added to the site. This page will be updated soon with additional information. Alternatively you can request additional help from our users directly on the relevant badcaps forum. Please note that we offer no warranties that any specification, datasheet,...
    09-06-2024, 04:40 PM
  • Loading...
  • No more items.
Working...