    Re: electromagnetic field therapy

    Anybody?

    Thank you in advance

    Comment


      Re: electromagnetic field therapy

      The pulse transformer only passes... pulses. The transistor must give 16V AC blips in order to get the signal to the SCR. If the transistor was on all the time, you would not get anything on the pulse transformer output- it does not pass DC. The transistor would get very hot too. The SCR only needs 0.6V to turn on but it is only on for the EMF pulse, which is fast and hard to measure. Two things you can do:

      Pretend the SCR is actually getting triggered, and check parts downstream, like the diodes, red transformer, red led, coils and wiring etc. The SCR should have voltage on its anode. You can use a couple 12V lightbulbs (#194) in series to emulate the EMF coil.

      2. If the SCR (gate drive transistor) is getting nothing, then I would guess one of the signals on the AND gate inputs is not working. The signal on "pin10: this one acts crazy between 1-5-14-and OL, 3.747hZ", I would trace that back.

      I also noticed the 6 bigger resistors (like blue-gray-blue-silver?) which is 68Megohms. I think with the 10uF film (green) cap, it makes for a very long timer. This thing has a session timer, like 30 minutes? What does the G knob do?

      Comment


        Re: electromagnetic field therapy

        Originally posted by redwire View Post
        The pulse transformer only passes... pulses. The transistor must give 16V AC blips in order to get the signal to the SCR. If the transistor was on all the time, you would not get anything on the pulse transformer output- it does not pass DC. The transistor would get very hot too.
        Hi!
        correct me if i'm not understanding the purppose of the transitor NPN there,
        on the pulse transformer, the primary winding has voltage at pin 1-ground and voltage at pin2-ground (same at both 16V) and a frequency of 160Hz at the lowest frequency setting of the machine. collector of the transistor is also connected to pin1, emitter goes to ground and base goes to the AND gate, so i assume when the transistor is triggered pin1 is shoted to ground so there is a voltage between pin1 and pin2 that way the pulse is accomplished, because right now i dont get any voltage at pin1-2. it is that the way it works?
        Originally posted by redwire View Post

        The SCR should have voltage on its anode. You can use a couple 12V lightbulbs (#194) in series to emulate the EMF coil.
        SCR anode is conected to ground
        Originally posted by redwire View Post
        2. If the SCR (gate drive transistor) is getting nothing, then I would guess one of the signals on the AND gate inputs is not working. The signal on "pin10: this one acts crazy between 1-5-14-and OL, 3.747hZ", I would trace that back.
        yeah i think that pin10 it's a little wierd too, maybe that one it's keping the transistor off being triggered right??? that pin goes to the frequency knob and also goes to one of the cd4013 that i really have no idea how they work i got to check those
        Originally posted by redwire View Post
        I also noticed the 6 bigger resistors (like blue-gray-blue-silver?) which is 68Megohms. I think with the 10uF film (green) cap, it makes for a very long timer. This thing has a session timer, like 30 minutes? What does the G knob do?
        you're spot on regarding those resistors and the big cap, sesion timer is 30 minutes!
        the G knob is Gauss, intensity of the magnetic field from 40 to 100, but i dont know how it's that being regulated because i get 16V at pin1 and pin2 no matter where i set that knob
        i'm getting an oscilloscope that way i can troubleshoot better
        Thanks a lot for your help!

        Comment


          Re: electromagnetic field therapy

          SCR anode is conected to ground, are you sure?
          Never stop learning
          Basic LCD TV and Monitor troubleshooting guides.
          http://www.badcaps.net/forum/showthr...956#post305956

          Voltage Regulator (LDO) testing:
          http://www.badcaps.net/forum/showthr...999#post300999

          Inverter testing using old CFL:
          http://www.badcaps.net/forum/showthr...er+testing+cfl

          Tear down pictures : Hit the ">" Show Albums and stories" on the left side
          http://s807.photobucket.com/user/budm/library/

          TV Factory reset codes listing:
          http://www.badcaps.net/forum/showthread.php?t=24809

          Comment


            Re: electromagnetic field therapy

            Originally posted by budm View Post
            SCR anode is conected to ground, are you sure?
            Pretty sure, check this pic, the trace is coming from toroid transformer
            Attached Files

            Comment


              Re: electromagnetic field therapy

              If you look up the spec sheet, that pin on the bottom is Cathode (K), Middle pin is Anode, top pin is the Gate.
              http://pdf1.alldatasheet.com/datashe...LA/2N4444.html
              Never stop learning
              Basic LCD TV and Monitor troubleshooting guides.
              http://www.badcaps.net/forum/showthr...956#post305956

              Voltage Regulator (LDO) testing:
              http://www.badcaps.net/forum/showthr...999#post300999

              Inverter testing using old CFL:
              http://www.badcaps.net/forum/showthr...er+testing+cfl

              Tear down pictures : Hit the ">" Show Albums and stories" on the left side
              http://s807.photobucket.com/user/budm/library/

              TV Factory reset codes listing:
              http://www.badcaps.net/forum/showthread.php?t=24809

              Comment


                Re: electromagnetic field therapy

                Originally posted by budm View Post
                If you look up the spec sheet, that pin on the bottom is Cathode (K), Middle pin is Anode, top pin is the Gate.
                http://pdf1.alldatasheet.com/datashe...LA/2N4444.html
                You're right, my bad. so anode goes to choke

                Comment


                  Re: electromagnetic field therapy

                  Yes, the Gate has a 33 Ohms resistor to Cathode, and a cap in parallel, the Gate gets the drive signal through a diode, like I said before, you can manualy trigger the SCR by pulsing the Gate using resistor between Anode and Cathode by touching and un touching th resistor between Anode and Gate, you said it did not do anything last time, that because you did not know which pin is Anode or Catode. It is hard when we get wrong information.
                  See post 51 for the diagram by redwire.
                  Last edited by budm; 02-13-2013, 11:19 AM.
                  Never stop learning
                  Basic LCD TV and Monitor troubleshooting guides.
                  http://www.badcaps.net/forum/showthr...956#post305956

                  Voltage Regulator (LDO) testing:
                  http://www.badcaps.net/forum/showthr...999#post300999

                  Inverter testing using old CFL:
                  http://www.badcaps.net/forum/showthr...er+testing+cfl

                  Tear down pictures : Hit the ">" Show Albums and stories" on the left side
                  http://s807.photobucket.com/user/budm/library/

                  TV Factory reset codes listing:
                  http://www.badcaps.net/forum/showthread.php?t=24809

                  Comment


                    Re: electromagnetic field therapy

                    Originally posted by budm View Post
                    Yes, the Gate has a 33 Ohms resistor to Cathode, and a cap in parallel, the Gate gets the drive signal through a diode, like I said before, you can manualy trigger the SCR by pulsing the Gate using resistor between Anode and Cathode by touching and un touching th resistor between Anode and Gate, you said it did not do anything last time, that because you did not know which pin is Anode or Catode. It is hard when we get wrong information.
                    nothing happens. 330 ohm resistor between gate and anode (middle pin)

                    Comment


                      Re: electromagnetic field therapy

                      What dc and AC do you have at these two points?
                      By the way, this your drawing also show which pins are A, K, G for the SCR that you had marked.
                      I see why it does nothing, the power from the power transformer red wires flow through the transducer to supply the voltage to the SCR, so you have to have the transducer connected to the jacks.
                      Attached Files
                      Last edited by budm; 02-13-2013, 11:35 AM.
                      Never stop learning
                      Basic LCD TV and Monitor troubleshooting guides.
                      http://www.badcaps.net/forum/showthr...956#post305956

                      Voltage Regulator (LDO) testing:
                      http://www.badcaps.net/forum/showthr...999#post300999

                      Inverter testing using old CFL:
                      http://www.badcaps.net/forum/showthr...er+testing+cfl

                      Tear down pictures : Hit the ">" Show Albums and stories" on the left side
                      http://s807.photobucket.com/user/budm/library/

                      TV Factory reset codes listing:
                      http://www.badcaps.net/forum/showthread.php?t=24809

                      Comment


                        Re: electromagnetic field therapy

                        Originally posted by budm View Post
                        What dc and AC do you have at these two points?
                        By the way, this your drawing also show which pins are A, K, G for the SCR that you had marked.
                        yes, i did it right in that pic, but in the schematic is backward.
                        Do you mean between those two point or at each point and ground?
                        thank you

                        Comment


                          Re: electromagnetic field therapy

                          Do you have the transducer connected? Just read between the two points as shown.
                          Never stop learning
                          Basic LCD TV and Monitor troubleshooting guides.
                          http://www.badcaps.net/forum/showthr...956#post305956

                          Voltage Regulator (LDO) testing:
                          http://www.badcaps.net/forum/showthr...999#post300999

                          Inverter testing using old CFL:
                          http://www.badcaps.net/forum/showthr...er+testing+cfl

                          Tear down pictures : Hit the ">" Show Albums and stories" on the left side
                          http://s807.photobucket.com/user/budm/library/

                          TV Factory reset codes listing:
                          http://www.badcaps.net/forum/showthread.php?t=24809

                          Comment


                            Re: electromagnetic field therapy

                            Originally posted by budm View Post
                            Do you have the transducer connected? Just read between the two points as shown.
                            i tried with the transducer and without it. also in all settings. let me do that

                            Comment


                              Re: electromagnetic field therapy

                              Originally posted by budm View Post
                              Do you have the transducer connected? Just read between the two points as shown.
                              56 VAC pad connected, 8VAC no pad 60hZ, 350mV vDC. in the backplate machine says ouput 2x24V

                              Comment


                                Re: electromagnetic field therapy

                                "56 VAC pad connected" OK, that sound about right, now you should be able to pulse the Gate by connecting/diconnecting resistor between Gate and Anode with pad connected and see what happen.
                                By the way check the AC voltage between A and K, it should also have about 56VAC (no pulsing with the resistor).
                                Last edited by budm; 02-13-2013, 11:56 AM.
                                Never stop learning
                                Basic LCD TV and Monitor troubleshooting guides.
                                http://www.badcaps.net/forum/showthr...956#post305956

                                Voltage Regulator (LDO) testing:
                                http://www.badcaps.net/forum/showthr...999#post300999

                                Inverter testing using old CFL:
                                http://www.badcaps.net/forum/showthr...er+testing+cfl

                                Tear down pictures : Hit the ">" Show Albums and stories" on the left side
                                http://s807.photobucket.com/user/budm/library/

                                TV Factory reset codes listing:
                                http://www.badcaps.net/forum/showthread.php?t=24809

                                Comment


                                  Re: electromagnetic field therapy

                                  Originally posted by budm View Post
                                  "56 VAC pad connected" OK, that sound about right, now you should be able to pulse the Gate by connecting/diconnecting resistor between Gate and Anode with pad connected and see what happen.
                                  i already did like 10 minutes ago, with a 330 ohm resistor between gate and anode (middle pin) and nohing happened.
                                  i did test the SCR and it gets triggered, but not with the resistor

                                  Comment


                                    Re: electromagnetic field therapy

                                    with the pad connected? If that is the case, that SCR is not turning on. Also what AC do you get between A and K?
                                    Never stop learning
                                    Basic LCD TV and Monitor troubleshooting guides.
                                    http://www.badcaps.net/forum/showthr...956#post305956

                                    Voltage Regulator (LDO) testing:
                                    http://www.badcaps.net/forum/showthr...999#post300999

                                    Inverter testing using old CFL:
                                    http://www.badcaps.net/forum/showthr...er+testing+cfl

                                    Tear down pictures : Hit the ">" Show Albums and stories" on the left side
                                    http://s807.photobucket.com/user/budm/library/

                                    TV Factory reset codes listing:
                                    http://www.badcaps.net/forum/showthread.php?t=24809

                                    Comment


                                      Re: electromagnetic field therapy

                                      Originally posted by budm View Post
                                      "
                                      By the way check the AC voltage between A and K, it should also have about 56VAC (no pulsing with the resistor).
                                      nothing, .5 VDC and nothing AC. yes pad connected

                                      Comment


                                        Re: electromagnetic field therapy

                                        That means the choke has open circuit!
                                        Never stop learning
                                        Basic LCD TV and Monitor troubleshooting guides.
                                        http://www.badcaps.net/forum/showthr...956#post305956

                                        Voltage Regulator (LDO) testing:
                                        http://www.badcaps.net/forum/showthr...999#post300999

                                        Inverter testing using old CFL:
                                        http://www.badcaps.net/forum/showthr...er+testing+cfl

                                        Tear down pictures : Hit the ">" Show Albums and stories" on the left side
                                        http://s807.photobucket.com/user/budm/library/

                                        TV Factory reset codes listing:
                                        http://www.badcaps.net/forum/showthread.php?t=24809

                                        Comment


                                          Re: electromagnetic field therapy

                                          Originally posted by budm View Post
                                          That means the choke has open circuit!
                                          i think you got it!

                                          Comment