Philips PUS6501 - no picture
  • betmen
    • May 2019
    • 90
    • Croatia
    I have a Philips PUS6501 TV that I got as defective. The TV has no picture. I first tested the power supply and made a short circuit during the measurement on Q8101 between D and S. Therefore, the Q8102 and resistor R8101 burned out, which I changed. After that I went to test the LED backlight and one LED was burned out on each strip. Now I doubt that the LED backlight was a major malfunction and that the power supply was good.
    After repairing the LED backlight, the TV did not work again. I managed to turn it on when I was measuring the voltages on the power supply with an instrument, but it turned on so that initially the LED backlight flashed for a few seconds and the TV turned on. Then everything worked, the TV could be turned on and off, everything worked. After that I unplugged it overnight and it didn't work again, it couldn't turn on.
    Ambilight lights up but there is no picture.
    On schematic I circled with red on page 4 which part I changed and the one I shorted.
    Does anyone know what else needs to be checked or what is wrong?
  • CapLeaker
    Leaking Member
    • 8116
    Re: Philips PUS6501 - no picture
    No blinking of the backlight, nothing? Turn the Tv on and measure each LED string with the LEDs connected and without. Make sure that all the turn on voltages are provided from the main board to the PSU, like PSon, BLon, Dimm…
    • betmen
      • May 2019
      • 90
      • Croatia
      I measured the voltages and when the TV is connected to the power supply, the voltages are: STBY-3.74V, 3D_ON-1.24V.
      After I turn on the TV with remote voltages are: STBY-3.7V, DIM-2.55V, BL_O / F-3.08V, PS_ON-3.4V, 3D_ON-0.01V, + 12V and + 12VA- 12.2V .
      The LED voltage connector is 0V and there is no picture.
      When I touched the leads G and D of the Q8101 transistor with the electrodes of the voltmeter, the LED backlight came on briefly, then I touched the G and D leads a few more times and the LED backlight started flashing for about half a minute and the TV turned on and everything worked.
      After that I turn off the TV for half an hour and it will not turn on again.
      It remained so "on" but without a picture and after 6-7 minutes the LED backlight started flashing for about half a minute and the TV turned on and everything works again:
      I turn it off again and turn it on right away and then the TV turns on and works.
      I turn it off again for a few minutes and it won't turn on right away but after a few minutes the LED backlight starts flashing and the TV turns on.
      As if some part needs to be heated for the TV to work, and when it cools down, the TV doesn't work or that part works a little, it doesn't work a little.
      When the TV was working, the voltages on the LED connector were 45V, then 44.6V, then 40.6V on the end connectors left and right in relation to ground.
      I suspect transistor Q8101 is not working properly, but I have to unsolder it and test it.
      Does anyone have any idea what it might be?
      • R_J
        • 9535
        Check the voltage ACROSS the main filter C9802. What is this voltage in standby and when the tv is turned ON.
        Q8101 is for the PFC circuit. You can also check the VCC voltage across C9809 and post the voltage. Also check the voltage across C9304. It could be that C9304, C9305 or CC9313 are weak.
        If the main B+ voltage is not boosted by the pfc voltage, the led voltage will also be low and likely not light the led's.
        I assume 230vac line, so B+ will be 325VDC in standby but should rise to around 395vdc when the tv is on and the pfc circuit is working.
        Note: the led ic U8101 (Vsen) senses the main B+ and likely will not turn on until the B+ voltage is at 395vdc.
        • CapLeaker
          Leaking Member
          • 8116
          Did you measure each led string vled+ to the corresponding vled- for each string?
          • betmen
            • May 2019
            • 90
            • Croatia
            I measured on the connector +VLED on the left and rigth side according to the grond(chasis).
            The rest of measurements I will do after the work today.
            I try to turn it on today but it did˙t, after some time like it˙s turn of by itself, ambilight turn off and red led for standby turn on.
            • betmen
              • May 2019
              • 90
              • Croatia
              Today TV turn on when I measured voltage od backlihgt LED connector, first flickering then stable LED light and everything working again.
              Measurements are:
              1.C9802-STBY-387V, STBY-after few minutes-322V, ON-screed not working-387V, ON-screed working-387V
              2.bridge rectifier BD9901-STBY-235V, STBY-after few minutes-322V, ON-screed not working-225V, ON-screed working-216V
              3.LED connector CN8101-pins +VLED and -VLED 36V, +VLED and chassis, not working - 12,5V, then start flickering, after that 40-43V
              Rest of capacitor I did`t measured because I must have access from behind, I did`t have tools to disassemble it near at hand, will tray it tomorrow I hope.
              Help.....
              • betmen
                • May 2019
                • 90
                • Croatia
                Rest of the mausurement are:
                C9809-0,52V
                C9304-14V
                C9305-13,45V
                C9131-0,52V
                I performed the measurements so that the connector from the motherboard and the LED backlight is disconnected, because the measurement lead can only be reached from the back of the PSU.
                Does anyone have any idea?
                • R_J
                  • 9535
                  I had made a mistake in post #4, Q8101 was NOT the pfc mosfet, but is one of the backlight driver mosfets.
                  If Q8101 was shorted, check the gate driver circuit, D8101, R8112 (47Ω) and R8113 (10Ω), Q8102 may have been damaged as well. Note: D8101 is in parallel with R8112.
                  I suspect C9131 is C9313, C9809 & C9313 are on the VCC supply for U9801 (pfc ic) This 0.5v would be normal since the main board was disconnected (no PS_ON) If you want to work on the power supply with the main board disconnected you need to connect about a 1K resistor between +3.5v/5v and PS_ON. This will turn on the power supply without the main board connected.
                  Voltages on post #7 suggest the pfc circuit is working ok.
                  • CapLeaker
                    Leaking Member
                    • 8116
                    Weird. Check the back light connector and the wiring? Maybe there is cracked solder / broken wire. LED flickering on and off could be also a flakey LED.
                    • betmen
                      • May 2019
                      • 90
                      • Croatia
                      Measurement from today:
                      - R8113- 10ohm 1,7V, after some time voltage droping
                      - R8112 47ohm, 2,6V after some time voltage droping
                      - D8101 - diode mode 0,029V, probably because of paralel resistor
                      - Q8101 - D-G 382,7V; D-S - 380,2V; G-S - 0V
                      - Q8102 - D-G 0,85V; D-S - 0,38,2V; G-S - 0,5V
                      Voltages are not staedy, after some time drops a litle and wher TV does`t turn on it`s turn of after some time, few minutes and STBY LED lights red.
                      After some measurement I measured voltage to LED backlight, it is about 35V, I connect connector and backlight lite up and TV powers up. When TV is working voltages on Q8101 and Q8102 are obout 2V, before was abour 382 on Q8101 and under 1V on Q8102.
                      Leds for backlight are OK I check them with backlight tester and they all worked fine without flickering.
                      I will change Q8101 when I will have time to do it (first I will test it when I desoldet it), Q8102 is new.
                      Any advice?
                      • betmen
                        • May 2019
                        • 90
                        • Croatia
                        I changed Q8101 today, old one was ok but I chaged it anyway.
                        All is the same, TV can not be turned on.
                        Anybody have idea what to do next?
                        • lotas
                          • Jan 2016
                          • 4554
                          • Russia
                          If Q8102 struck, then the U8101 SSC9522S PWM controller and its strapping could have suffered
                          • R_J
                            • 9535
                            After some measurement I measured voltage to LED backlight, it is about 35V, I connect connector and backlight lite up and TV powers up. When TV is working voltages on Q8101 and Q8102 are obout 2V , before was abour 382 on Q8101 and under 1V on Q8102.
                            Any advice?
                            This is impossible, the backlights can not operate if the voltage on Q8101 Drain is 2v. R8101 (0.1Ω) supplies 385vdc to Q8101 Drain
                            Like I said If Q8101 was shorted, Q8102 may have been damaged so try changing it.
                            • betmen
                              • May 2019
                              • 90
                              • Croatia
                              That voltage was D-G and D-S, not D to earth.
                              Backlight, when working, have voltage about 35V, when TV is on but backlight and TV not working voltage is low, about 6-7V, sometimes up to 12V it is not stable.
                              Q8102 was changed in the beginning with the R8101....
                              • R_J
                                • 9535
                                So just to be clear, you changed both Q8101 and Q8102, correct? Usually when there are internment backlights it is one of the led's or one of the connections in the panel that is at fault.
                                IC8101 monitors the led voltage as well as the led current, if the current through the string changes and the led voltage lumps up for a split second, this will trigger the ic to shut down the led voltage.
                                Sometimes C8109 gave problems on some power supplies, but the cap usually just went open.
                                • betmen
                                  • May 2019
                                  • 90
                                  • Croatia
                                  Yes I changed Q8101 and Q8102.... but I made one error at the beginning, I put the wrong scheme of the PSU.
                                  I put the scheme for bigger screen PSU
                                  I print for me correct scheme and put on my post wrong one.
                                  In this posts attachment is correct one...sorry for that.
                                  • betmen
                                    • May 2019
                                    • 90
                                    • Croatia
                                    I would love to fix this tv so if anyone wants to help let me know.
                                    I did the following, in LED circuit I took out C8116 and C8117 and check them ,both are ok, I measured most of the resistors, all are ok, secundary side of T8101 in LED circuit has no voltage on diodes or capacitors, I made measurements on U8101 according to datasheet, Vcc - 12,07V, FB pin voltage 2,95V, CSS pin 0,71V, VB-VS - 2,5V...it seems fine to me.
                                    There is stanby voltage of 3,5V and when TV is turned on there is 12V on 12V pins, DIM pin have 1,52V, BLon -2,92V, PSON - 3,25V, 3D_ON - 0V.
                                    C9802 and C9803 have 386V, so I think AC input, PFC and stanby circuit are fine, there is some fault in LED circuit.
                                    Correct scheme is in post 17.
                                    Anyone have some idea what to do?
                                      • 12116
                                      • U.S.
                                      If you don't have a backlight tester it might be time to tear it down and see whats going on with the leds.
                                      • betmen
                                        • May 2019
                                        • 90
                                        • Croatia
                                        That is first what I have done, few leds did`t work and i changed them, now all leds are working, tested with backlight tester.
                                        On the led connector CN8101, pin +VLED there is no voltage.
                                        • matthewmcconaughey
                                          I need help with my friend Sony GDM-FW900 24” CRT
                                          by matthewmcconaughey
                                          Hello, Badcaps is my last resort. I am the first owner of the monitor. I reached almost every repair store in my area, but no one even wanted to hear about troubleshooting an old CRT monitor. At this point, I have been trying to figure it out by myself, but I have no idea where to start or look for.
                                          I wish to know how to think and locate a failing component. I will put all my efforts to learn and contribute.
                                          1. It started with a bright and greenish picture upon starting a cold monitor.
                                          2. About 2 years ago, only on the cold boot, the screen started to occasionally flash with
                                          05-16-2023, 03:15 PM
                                        • theone111
                                          by theone111
                                          3. Problem: TV shuts down
                                          07-22-2024, 10:36 AM
                                        • reessi
                                          Acer Alienware Nitro 5 (GH51G LA-L181P Rev: 1C) & Dell Alienware M15 (GDP50 LA-k452P Rev:1) - Both Shorted PCH (SRKMA)!
                                          by reessi
                                          Hello:
                                          I have both these laptops that appeared to have the same fault so I thought I would have a go at replacing the PCH as something to learn, they both are using the same PCH.
                                          I started on the ACER:
                                          Up on connecting power no signs of life but current is consumed from the bench power supply and the PCH is hot.
                                          All the caps on the top of the PCH are showing 0 ohms and the 3.3v rail is shorted.
                                          I followed the videos from Sorin, Electronics Repair School on Youtube heating from the bottom to install. I practised on around 10 scrap...
                                          10-29-2024, 11:31 AM
                                        • madyoyo
                                          Philips 37pfl560: Yellow picture
                                          by madyoyo
                                          HI.
                                          I have a weird problem with an old Philips 37pfl560 TV i put out of storage: the picture have a noticeable yellow hue, and its not caused by the settings (did a factory reset). I've never seen this before. I will try to get a photo with a real camera, with my phone its hard to see because of automatic white balance, but in real live the whites are really yellow.
                                          Anyone have an idea of what can cause this? Can CCFL tubes get "warmer" with time? This tv have several analog inputs,so its nice for retrogaming....
                                          05-30-2025, 10:16 AM
                                        • maneatingmonkey
                                          Missing LCD Panel Voltages VGH, VGL, VCOM (TCL 55P6US Sound & Backlight but NO Picture)
                                          by maneatingmonkey
                                          I have a TCL 55P6US with sound and backlight but no picture. This model does not have a T-CON board, and the right (rear perspective) panel board is fed from the left panel board. I have not been able to induce any change by disconnecting the secondary panel board or the tape cut-off method on the primary board.
                                          There are a bunch of voltages missing on the panel boards. I can't find any shorted caps of the panel boards, and all of the missing voltage rails read in the kiloohms, which leads me to suspect the issue may lie on the circuit board, rather than a short in the panel.
                                          03-27-2024, 07:17 AM
