Lenovo Legion 5 Pro - NM-D601 REV1.0 Turns off with battery conected
  • judamex
    Active Member
    • Jan 2021
    • 183
    • Mexico
    Originally posted by mon2
    Can you confirm the gate voltage of PQ5407 WITH battery & power adapter? It does not look right.
    Click image for larger versionName:	gate_voltage.pngViews:	0Size:	50.3 KBID:	3306120
    A few other points:
    1) remove all power -> measure the resistance of the 4k resistor @ PR5405. Confirm this resistor matches the schematic value of 4k. Concern is that in your measurements, the CMSRC voltage to ground is lower than the power adapter voltage.
    2) double check the BATTERY only measurement for pins 1 & 2 on the charger IC.
    3) repeat the measurements WITH BATTERY & power adapter for pins 1 & 2 on the charger IC. Pin # 2 (ACP) should be a higher voltage than pin # 1 (ACN). In either case, the voltage measurements should be very close to each other. Note the voltage drop across ACP and ACN junction when using only the power adapter. Higher voltage difference when the battery is used or when both battery & power adapter is used.
    With no power to the board and no battery, confirm the resistance from pin # 2 (ACP) directly on the charger IC to the current sense resistor @ PR5401. Then repeat the same test with pin # 1 to the same current sense resistor. Both checks should be ~0 ohms. Touch your meter probes together and this is the same resistance measurement you should have for this current sense resistor path.
    Also, I think that for the BATTERY only, the pin # 19 measurement was missed in the above table. It should be the same voltage as pin # 20.
    Rectify the voltages on pin 23 gate of PQ5407 and it is still 0.5 with battery+charger, rectify the battery and charger alone and it is still 0V.
    Rectify measurement in PR5405 and it is 3.99Kohms
    Rectify measurement on pins 1 and 2 only with battery connected, they remain at 15.5 on both pins
    -battery+charger 20.4 on both pins 1 and 2
    measurement pin 1 and 2 with respect to PR5401 in both cases was 0.6 ohms
    I correct the measurement on pin 19 only with 15.74V battery pin 19 and 20​
    • judamex
      Active Member
      • Jan 2021
      • 183
      • Mexico
      A few other points:
      Update: bought a small oscilloscope to take the reading of pin 23 for PQ5407 attached a photo
      I did the readings with and without battery and the result was 0​
      • judamex
        Active Member
        • Jan 2021
        • 183
        • Mexico
        Update 2:
        Check the measurements regarding this video
        (https://www.youtube.com/watch?v=ODcc...LaV0BS&index=7) in diode mode and everything is fine except PIN 17 with a value in the video of 0.6622 and on my motherboard it is 0.501. It is the only difference I noticed, everything else is fine with respect to ground (RED ground, BLACK to the IC pins)
        Or what oscilloscope measurements can I take?​
        • mon2
          • Dec 2019
          • 14267
          As suggested by mcplslg123 , test with a known good battery. You can buy from Amazon and return it if not required. The issue may be with the battery itself (deep discharged state and is now locked or similar).
          • judamex
            Active Member
            • Jan 2021
            • 183
            • Mexico
            Thanks, that's what I'll do and order a new compatible replacement battery and I'll do the test. I will update the post with the result
            • judamex
              Active Member
              • Jan 2021
              • 183
              • Mexico
              #26
              Update: i Buy a replacement battery from amazon and still in the same error, I saw the BQ24780S get a little hot with the battery and charger connected.
              • mon2
                • Dec 2019
                • 14267
                Carefully , measure the voltage to ground on each pin on this charger IC. Post your full list of measurements.
                • judamex
                  Active Member
                  • Jan 2021
                  • 183
                  • Mexico
                  #28
                  I took the following measurements with the new battery connected just in case, I noticed several details that may be important:
                  1- the motherboard restarted after several seconds, it did not turn off completely, it just restarted
                  2- a strange oscillation in the ACDET, it only did it once and then showed stable voltage​
                  New Batery + Charger
                  20.36
                  20.25
                  20.17
                  25.26
                  25.27
                  2.98
                  0.31
                  0.011
                  0.113
                  3.32
                  3.3
                  3.3
                  GND
                  N/C
                  0.31
                  1.48
                  16.37
                  16.26
                  16.35
                  16.39
                  1.53
                  GND
                  0.58
                  6.06
                  21.33
                  20.24
                  16.49
                  19.55
                  • mon2
                    • Dec 2019
                    • 14267
                    Power down immediately. ACOK pin #5 should never be so high of a voltage. It appears to be shorted to the ACDRV pin #4.
                    Flux the ic well and reflow it and gently nudge the chip with tweezers to remove the short condition. This event can cause damage.
                    Double check your work with the meter to confirm the short is removed.
                    • judamex
                      Active Member
                      • Jan 2021
                      • 183
                      • Mexico
                      #30
                      I resoldered and resoldered again and it now shows me 2.98
                      • judamex
                        Active Member
                        • Jan 2021
                        • 183
                        • Mexico
                        #31
                        I resolder and took different measurements with a different charger, I realized that with the 12 amp charger the charging IC heats up to 48°C, so I decided to try with a charger with less amperage that is enough to keep it on, I put the measurements below.
                        I also attach measurements with only the battery connected and the laptop always on (I had to turn it on every 1 minute because it turns off)​
                        Click image for larger version  Name:	image.png Views:	0 Size:	5.9 KB ID:	3317603
                        • mon2
                          • Dec 2019
                          • 14267
                          Does the laptop power up correctly still with only the power adapter? The laptop will function ok till you power down if using only the adapter?
                          • judamex
                            Active Member
                            • Jan 2021
                            • 183
                            • Mexico
                            #33
                            That's correct, if I use the power adapter ONLY without a battery connected, I can even run benchmarks without a problem, but as soon as I use the battery alone or with the charging adapter connected, the laptop turns off. ONLY when the battery is connected does it turn off.
                            • mon2
                              • Dec 2019
                              • 14267
                              The above table of measurements should really have pin #s and also the same pin labels as used by the datasheet for the charger IC. It will make the review much easier to study.
                              Not sure what the GND followed by N/C means. Best to repost with the latest values.
                              What is the voltage to ground of pin # 15 = BATPRES# on the charger IC? This is a critical voltage that should be a logic low = '0' = 0 volts to denote that the battery is attached.
                              You may have other charger IC pins shorted but will review with a fresh table of measurements with pin #s and pin labels.
                              • judamex
                                Active Member
                                • Jan 2021
                                • 183
                                • Mexico
                                #35
                                Sorry for not indicating the number of pins, I update the table with the number of pins and their description based on the boardview of the motherboard, the charging IC corresponds to the BQ24780S
                                • mon2
                                  • Dec 2019
                                  • 14267
                                  Much cleaner. Thanks.
                                  Concerned about the gate voltage of the battery charge (PQ5404 & PQ5406) and also the battery mosfet (PQ5403). Each is a suspect.
                                  We know that the battery voltage is the voltage at pin #27 (CHG_LX) = 16 volts. For the battery to power the logic board, the battery mosfet gate voltage @ PQ5403 should be boosted to enable this N-channel mosfet. Yet, this gate voltage is the same voltage as on the source pins.
                                  Remove the power adapter. Connect only the battery. What is the voltage to ground of the drain pins on PQ5403 = main power rail to the logic board?
                                  In this condition, the drain pins should be ~16 volts and the laptop should function ok.
                                  To fully turn on PQ5403, the gate voltage needs to be boosted.
                                  Next, remove all power and disconnect the battery. Meter in resistance mode.
                                  Measure the resistance of PQ5403 as usual:
                                  source (1-2-3) & drain (5-6-7-8)
                                  source (1-2-3) & gate (4)
                                  gate (4) & drain (5-6-7-8)
                                  Post each measurement. The mosfets @ PQ5404 & PQ5406 are also suspects so does not hurt to review their resistance in the same manner.
                                  • judamex
                                    Active Member
                                    • Jan 2021
                                    • 183
                                    • Mexico
                                    #37
                                    The drain of PQ5403 is 16V in standby and 15.8 on
                                    The gate voltage of pq5403 is 22.06 in standby and 21.73 on,
                                    I monitored this voltage using the oscilloscope and noticed a very fast voltage drop and instantly the laptop turned off.
                                    The measurements of pq5403
                                    source (1-2-3) RED & drain (5-6-7-8) BLACK: 0.88 Mohms
                                    source (1-2-3)RED & gate (4) BLACK: 114 Kohms
                                    gate (4)RED & drain (5-6-7-8)BLACK: 1.66 Mohms
                                    The measurements in PQ5404
                                    source (1-2-3) RED & drain (5-6-7-8) BLACK: 0.80 Mohms
                                    source (1-2-3)RED & gate (4) BLACK: 8.18 Kohms
                                    gate (4)RED & drain (5-6-7-8)BLACK: 0.95 Mohms
                                    The measurements in PQ5406
                                    gate (4)RED & drain (5-6-7-8)BLACK: 0.95 Mohms​
                                    • mon2
                                      • Dec 2019
                                      • 14267
                                      The source to gate resistance of each mosfet appears to be low for a good mosfet.
                                      Consider to flux and carefully remove both PQ5404 and PQ5406 off the board. After you have done this, check if the source & gate resistance increases for PQ5403.
                                      Allow the removed mosfets to cool. Then measure their source to gate resistance for each while they are on your workbench. Each measurement should be a high resistance unless they are defective.
                                      • mcplslg123
                                        • Jun 2015
                                        • 7262
                                        • india
                                        #39
                                        pin15 rises to 3.3V when battery is dosconnected?
                                        Pin19 & 20 should read same voltage.
                                        • judamex
                                          Active Member
                                          • Jan 2021
                                          • 183
                                          • Mexico
                                          #40
                                          The resistance directly on the motherboard is 7.9 Kohms source RED, Gate BLACK.
                                          the mosfets on the outside mark O.L, they do not mark any measurement in Ohms
                                          I tested the mosfets in diode mode and everything seems to be in order​
                                            Lenovo Legion Laptop Gamer Notebook Legion 5 15ACH6H Notebook Specification for Upgrade or Repair
                                            This specification for the Lenovo Legion Laptop Gamer Notebook Legion 5 15ACH6H Notebook can be useful for upgrading or repairing a laptop that is not working. As a community we are working through our specifications to add valuable data like the Laptop Gamer Notebook Legion 5 15ACH6H boardview and Laptop Gamer Notebook Legion 5 15ACH6H schematic. Our users have donated over 1 million documents which are being added to the site. This page will be updated soon with additional information. Alternatively you can request additional help from our users directly on the relevant badcaps forum. Please...
                                            09-07-2024, 05:00 AM
                                            Lenovo Legion Legion 5 17ACH6H Notebook Specification for Upgrade or Repair
                                            This specification for the Lenovo Legion Legion 5 17ACH6H Notebook can be useful for upgrading or repairing a laptop that is not working. As a community we are working through our specifications to add valuable data like the Legion 5 17ACH6H boardview and Legion 5 17ACH6H schematic. Our users have donated over 1 million documents which are being added to the site. This page will be updated soon with additional information. Alternatively you can request additional help from our users directly on the relevant badcaps forum. Please note that we offer no warranties that any specification, datasheet,...
                                            Lenovo Legion Legion 5 15ITH6H Notebook Specification for Upgrade or Repair
                                            This specification for the Lenovo Legion Legion 5 15ITH6H Notebook can be useful for upgrading or repairing a laptop that is not working. As a community we are working through our specifications to add valuable data like the Legion 5 15ITH6H boardview and Legion 5 15ITH6H schematic. Our users have donated over 1 million documents which are being added to the site. This page will be updated soon with additional information. Alternatively you can request additional help from our users directly on the relevant badcaps forum. Please note that we offer no warranties that any specification, datasheet,...
                                            09-07-2024, 04:10 AM
                                          • twonk9098
                                            Lenovo Legion 5 high fan spin turns off after 34 seconds
                                            by twonk9098
                                            I have a Lenovo Legion 5 with a Ryzen 7 series CPU and RTX 3070 GPU. My issue is that it turns on and all seems fine except after a few seconds the fans spin up high and the laptop turns off after 34 seconds.
                                            I can get into bios and can even boot and login to Windows however after 34 seconds no matter if Im at the bios screen or booted into Windows it will just turn off.
                                            I have managed to run HWmonitor and the temperatures of the CPU and GPU are fine, I can also touch the heatsync which is warm but not hot.
                                            I have tried a few new bios but exactly...
                                            07-14-2024, 11:19 AM
                                            LENOVO Legion Y720 Legion i7-7700 Specification for Upgrade or Repair
                                            This specification for the LENOVO Legion Y720 can be useful for upgrading or repairing a desktop PC that is not working. As a community we are working through our specifications to add valuable data like the Legion Y720 boardview and Legion Y720 schematic. Our users have donated over 1 million documents which are being added to the site. This page will be updated soon with additional information. Alternatively you can request additional help from our users directly on the relevant badcaps forum. Please note that we offer no warranties that any specification, datasheet, or download for LENOVO Legion...
                                            09-12-2024, 02:30 PM
