Fluke 189 series 1 DC voltage is off

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • tibimakai
    Badcaps Legend
    • Jan 2012
    • 3680
    • USA

    #1

    Fluke 189 series 1 DC voltage is off

    I have just realized that my meter is off, when I measure the car's battery, and I can't imagine how that happened.
    I also have a 87, and when I doubled checked the voltage, the 87 was showing 3.5V higher, than the 189. Most likely the 189 is off.
    Where should I start with the 189?
    Switched leads between the two meters, but nothing has changed.
    I have not opened it yet.
    Any help would be appreciated.

  • stj
    Great Sage 齊天大聖
    • Dec 2009
    • 31157
    • Albion

    #2
    is it under warranty?
    if not then start by checking the MOV because they are known to get leaky over time.
    it may also damage the big current-limiter resistor.

    Comment

    • tibimakai
      Badcaps Legend
      • Jan 2012
      • 3680
      • USA

      #3
      MOV?! What exactly is that?

      Comment

      • redwire
        Badcaps Legend
        • Dec 2010
        • 3912
        • Canada

        #4
        Take it apart.
        It could be the supercap is leaking and taking the piss on the board. Or a protection component (clamp transistors) are damaged.

        What does it read on Ohms with the leads not connected to anything?

        Typical pics of 189: http://mrmodemhead.com/blog/gallery/fluke-189-clean-up/
        Supercap thread on 289: https://www.badcaps.net/forum/troubl...repair?t=72162

        Comment

        • tibimakai
          Badcaps Legend
          • Jan 2012
          • 3680
          • USA

          #5
          I will check when I get home, but I remember that I have measure the Ohm scale with the 87, and it was reading 15.3Mohm instead of 10Mohm.

          Comment

          • tibimakai
            Badcaps Legend
            • Jan 2012
            • 3680
            • USA

            #6
            Opened it and found at least one resistor blown, with trace lifted as well. The one next to it, in series reads 1.152KOh. I'm not sure, if that OK, or not.Click image for larger version

Name:	20231128_183910.jpg
Views:	479
Size:	796.4 KB
ID:	3155019
            Attached Files

            Comment

            • redwire
              Badcaps Legend
              • Dec 2010
              • 3912
              • Canada

              #7
              Those appear to be the shunt resistors for the uA and mA ranges? The bridge rectifier and diode are supposed to clamp and protect the shunts and just blow the 440mA fuse... so I don't quite understand what went wrong here.
              One horizontal resistor going to the via I see in another pic is a blue precision 100Ω. There should be a 1Ω somewhere... the badly smoked one I cannot see clearly in any pics maybe 251x? or something written on it.

              I have a Fluke 189-II service manual 2008 with schematic- but it's different than your pic. They might have fixed an error or beefed them up.
              Current shunts:
              R167 100Ω 1% 1W
              R165 1Ω 5% 2.5W
              R5 0.010Ω 2% 1W WW big shunt for 10A range

              I can't see them causing problems for the Voltage and Ohms functions. Unless it's the leakage current from the shrapnel

              Comment

              • tibimakai
                Badcaps Legend
                • Jan 2012
                • 3680
                • USA

                #8
                The replacement supercap in the 289 thread is the same part number for the 189 as well? Mine is growing something as well. I think it would be time to replace that.

                Comment

                • stj
                  Great Sage 齊天大聖
                  • Dec 2009
                  • 31157
                  • Albion

                  #9
                  the blue parts are MOV's (varisters) and should meter open-circuit.
                  whats printed on the red one?

                  Comment

                  • redwire
                    Badcaps Legend
                    • Dec 2010
                    • 3912
                    • Canada

                    #10
                    Originally posted by tibimakai
                    The replacement supercap in the 289 thread is the same part number for the 189 as well? Mine is growing something as well. I think it would be time to replace that.
                    I believe it's the same supercap used in Fluke 189, 287, 289 - but not 100% sure.
                    07/2023 Elna is phasing out making coin supercaps, they are NRND but some production and stock available.
                    Elna 0.2F 3.3V DSK-3R3H204T614-H2L new p/n is RSCSK2043R3D01004T. Mouser 555-DSK3R3H204T614H2
                    Elna 0.33F 3.3V DSK-3R3H334T-HL new p/n is RSCSK3343R3D02008T. Mouser 555-DSK-3R3H334T-HL (thicker 2.1mm vs 1.4mm- fits)

                    I would first get the voltage channel working. The damaged current shunt resistors are on a different (current input) channel to the big IC so the two problems are not related that I can see, which is a bit puzzling.
                    The red MOV is an old GE part and must be a different voltage than the custom blue Epcos S05K575. For the old Fluke 87's it was V575LA10P or V910LSX1399 5mm red 910V 1mA almost same as the Epcos part.
                    I would guess either a MOV is leaky or the protection clamp is damaged, for the voltage channel - leakage would make it read lower V.

                    Comment

                    • tibimakai
                      Badcaps Legend
                      • Jan 2012
                      • 3680
                      • USA

                      #11
                      The red one measures 1.107Kohm.
                      I was measuring parasitic drain at my wife's CRV, maybe I did something stupid and I did not even realize?!
                      What should I look for at the voltage scale?
                      MOVs show OL with the Ohm meter, except the red one.
                      Which one is the protection clamp?
                      Attached Files
                      Last edited by tibimakai; 11-29-2023, 10:22 PM.

                      Comment

                      • stj
                        Great Sage 齊天大聖
                        • Dec 2009
                        • 31157
                        • Albion

                        #12
                        whats printed on the other side of the red part?
                        btw, all 3 are protection - that whole area below the switch is.
                        i suspect it's the same part as the blue ones - i hope anyway.
                        i have the schematic of the series2 here.
                        Last edited by stj; 11-30-2023, 07:15 AM.

                        Comment

                        • redwire
                          Badcaps Legend
                          • Dec 2010
                          • 3912
                          • Canada

                          #13
                          I found a Fluke 189-I board pic with the two burnt resistors "2611" for 2k61 1% and "1001" for 1k00 1%, in series with each other? and then PCB trace to the red part. I forgot the blue (foil?) ones, I thought a Vishay line but 1206 1% would be i.e. CRCW12062K61FKEA. It's probably better than 1% part though.
                          I don't see those resistors used in a Fluke 189-II. I wrongly thought they are the mA-range current shunts- but those are 1R (big black through-hole) and 100R 1W both located on the other side of the board.

                          For curiosity, can you tell us ALL markings on the red part? I did see "LF" for Littelfuse. Apparently it is not a MOV, but is a PTC (also used in 87-I) and it's normal to read ~1.1kΩ resistance across it - so it looks fine right now- but failed at its job so the 189-II ditched it lol. I think early Fluke DMM protection was a bit unrefined.
                          For reference data I have: Fluke p/n 1277360 THERMISTOR, POSITIVE, 1.1kΩ, +-20%, COATED, RADIAL Original G.E. YS3961, possible substitute Amphenol YS4020.

                          So anyway, back to the reading low V problem... I doubt it's the burnt resistors but not sure. I would replace them, clean up the smoke with IPA and a Q-tip, and see where that takes us? Red part seems OK.
                          Otherwise, it's the protection clamp and with no schematic a little hard to point those parts out.
                          Attached Files

                          Comment

                          • stj
                            Great Sage 齊天大聖
                            • Dec 2009
                            • 31157
                            • Albion

                            #14
                            the carbon that was the board could be conducting between traces

                            Comment

                            • tibimakai
                              Badcaps Legend
                              • Jan 2012
                              • 3680
                              • USA

                              #15
                              That is a great picture!! Thank you for that.
                              What wattage could those have, 0.25W?
                              I will post a picture with the red part, when I get home.
                              Thank you guys, it is much appreciated.

                              Comment

                              • redwire
                                Badcaps Legend
                                • Dec 2010
                                • 3912
                                • Canada

                                #16
                                They look like size 1206 [Metric 3216] 1/4W resistors, Fluke uses Vishay/Dale CRCW series so I would guess: CRCW12062K61 and CRCW12061K00.

                                To burn them up, say at 1W would be ~82V for them. 12V work could not have cooked them.

                                Comment

                                • tibimakai
                                  Badcaps Legend
                                  • Jan 2012
                                  • 3680
                                  • USA

                                  #17
                                  On the back of the red part is D44, not sure what that means. I did not se the LF logo.
                                  Thanks for the resistor info.
                                  What do you think I should do, order and replace those resistors and we go from there, or should we do more diagnosing, before I order parts?

                                  Comment

                                  • stj
                                    Great Sage 齊天大聖
                                    • Dec 2009
                                    • 31157
                                    • Albion

                                    #18
                                    clean it up with IPA - see if the board is damaged or it's just soot around the resistors

                                    Comment

                                    • tibimakai
                                      Badcaps Legend
                                      • Jan 2012
                                      • 3680
                                      • USA

                                      #19
                                      The PCB is black under that resistor(the resistor's shape), but it is not conductive. I also managed to pop up the PCB with the hot air station(yeahh).
                                      I have just realized that that one of the big fuses are open.
                                      I can swear that it was good when I started.
                                      I guess I will place an order for those two resistors, and a fuse, and we go from there.
                                      Will be back, when I have replaced those components.
                                      Since the trace is lifted and one side is gone, should/could I use through hole resistors?
                                      Last edited by tibimakai; 12-05-2023, 12:38 AM.

                                      Comment

                                      • VoicesFYC
                                        Member
                                        • Apr 2021
                                        • 26
                                        • United States

                                        #20
                                        not sure if this will help... but since i'm pretty good with fluke meters from time to time i went looking for a service manual.... welp good luck with that...i don't think fluke made one..or if they did it is in-house ONLY... but i did come across something that MIGHT help... just not totally sure so i will post it here... BE ADVISED it's Both Chinese and English...so don't shoot me over it....it's all i could find on the 187 & 189 fluke meters . hope this does some good . Daniel. https://www.eevblog.com/forum/repair...attach=1054516 click this to download the PDF they made.

                                        Comment

                                        Related Topics

                                        Collapse

                                        • Document Archive
                                          ASUS Pro Series D310MT-0G32500650 Pro Series G3250 Specification for Upgrade or Repair
                                          by Document Archive
                                          This specification for the ASUS Pro Series D310MT-0G32500650 can be useful for upgrading or repairing a desktop PC that is not working. As a community we are working through our specifications to add valuable data like the Pro Series D310MT-0G32500650 boardview and Pro Series D310MT-0G32500650 schematic. Our users have donated over 1 million documents which are being added to the site. This page will be updated soon with additional information. Alternatively you can request additional help from our users directly on the relevant badcaps forum. Please note that we offer no warranties that any specification,...
                                          09-12-2024, 03:28 PM
                                        • cheeky2
                                          Anyone know how to take apart a Fluke T100 Voltage Tester without damaging it?
                                          by cheeky2
                                          See attached picture. I have a Fluke T100 Voltage Tester that the resistance and diode measurement isn't working (it's not the batteries (as I have already checked them and put decent Duracell 1.5V AAA batteries inside it).
                                          I have looked on the web for any information and on youtube without any results for this model.
                                          There are no screws visible or anything obvious. So I am hoping someone here has the solution to take it apart?
                                          The only way I can see is underneath the yellow rubber handle area, however I don't wish to damage it if it isn't the case!...
                                          06-04-2024, 07:21 AM
                                        • Fernon@Repair
                                          Gigabyte Z490 Gaming X Motherboard No VCORE Voltage
                                          by Fernon@Repair
                                          i have Gigabyte Z490 Gaming X Motherboard .it has a power on no display fault .. no vcore voltage and debug led stuck CPU LED .

                                          VCCIO voltage ok
                                          VCCSA Voltage OK
                                          VCCPLL Voltage OK
                                          Ram Voltage ok
                                          RamVTT OK
                                          VPP 2.5 OK

                                          But No Core Voltage

                                          and i want to this vrm ic datasheet or pinout (ISL69269)
                                          anyone have this ic pinout voltages?...
                                          06-25-2025, 07:19 AM
                                        • electron_jj
                                          Panasonic TH-L42D25A checking the PSU output voltage
                                          by electron_jj
                                          Hi Everyone, I am new to this forum. I have a Panasonic LED TV which doesnt turn on and is blinking 4 times repeatedly when I turn on the switch. The TV does not respond to the remote control so I had to turn on using the physical switch on the side of the TV. When I turn on the switch the 4 x blinking starts and repeat itself until I turn off the TV. There is a black screen. I turned off the light in the room to check the panel and I think there is no backlight as far as I can tell.

                                          I have been inspecting and testing the capacitors and diodes in the power supply board to try to...
                                          08-29-2021, 08:13 AM
                                        • sam_sam_sam
                                          Trip light 2400 watt line conditioner bought used issue with low voltage oscillation going on between relays
                                          by sam_sam_sam
                                          I found two capacitors on the board and one of the leads on one capacitor had bad soldering joint and the other joint connections not much better and there were several soldering joints that way to much flux around the joint redid several of them

                                          The capacitors were a non brand that I did not recognize they got replaced and now it does not oscillating between relay but what concerns me a little bit is when the voltage is a little bit low the output voltage is around 130 volts and then drops down to 115 volts this happens around 105 to 108 volts if remember correctly

                                          Now...
                                          06-09-2024, 10:57 AM
                                        • Loading...
                                        • No more items.
                                        Working...