    This is a Gallenkamp oven, used in a laboratory space, goes up to 220 Celcius, but is usually maintained at about 115 Celcius. Used for drying glassware, molecular sieves, and so on. Oven is in excess of 20 years old.

    It heats up fine, there is an orange light on the front that indicates that the unit is heating, and when it reaches the preset temp, the orange light goes out. All good so far.

    However, after the preset temp is reached, the temperature continues to rise, power to the heating elements isn't cut off.

    Temp continues to increase until the over-temp temp setting is reached, in which case the red light comes on, and the power to the elements is shut off, the over-temp switch is tripped, and the unit cools down.

    Pressing in the reset switch allows the unit to heat up the chamber again, and it will again over-temp and switch off, etc.

    I've attached a couple of photos of the front panel and electronics inside.

    Fuses are ok. Any area in particular to home in on for testing?
    Attached Files
  • redwire
    Badcaps Legend
    • Dec 2010
    • 3906
    • Canada

    #2
    Re: Drying Oven - Preset Temp Fail

    Check if the triac on the big heatsink is shorted between MT1 and MT2. Those are the fat wires on it. That would leave heat on all the time until the safety cutouts trip.
    Triac is TAG semiconductor T1513MJ 15A 600V TO-220 isolated, non-sensitive 50mA gate.

    Is the temperature readout correct? If it's stuck then I would look at the sensor for water ingress or corrosion etc.
    Last edited by redwire; 01-21-2021, 09:53 AM.

    Comment

    • tmcw
      Badcaps Veteran
      • Oct 2010
      • 382
      • Ireland

      #3
      Re: Drying Oven - Preset Temp Fail

      Thanks for the tests...

      With the multimeter on continuity/diode mode, I got a readout of 0.127 either way on the fat wires.

      When I disconnected one of the fat wires from the circuit, testing the triac, the meter read open.

      I'm pretty sure that the temperature readout is correct, at least a couple of degrees Celcius either way.

      We actually have 2 other ovens like this in the building, but I'm not sure if they work. I turned one of them on earlier, but the temperature display was a bit wonky, and the temp seemed to be fluctuating, so it may have different issues. I'll test the other one tomorrow.

      Comment

      • R_J
        Badcaps Legend
        • Jun 2012
        • 9535
        • Canada

        #4
        Re: Drying Oven - Preset Temp Fail

        What is the model number? maybe there is a manual/schematic available. That white/green wire connected to G2/G1 controls the triac, If this wire is disconnected does the oven heat up? if not the fault is likely on the board, maybe the optocoupler is bad.
        Is the heating element connected to triac? or is the heating element controlled by a relay somewhere?
        Last edited by R_J; 01-21-2021, 10:42 AM.

        Comment

        • tmcw
          Badcaps Veteran
          • Oct 2010
          • 382
          • Ireland

          #5
          Re: Drying Oven - Preset Temp Fail

          The model number is OVE.100.110S. A search throws up nothing. When it was supplied 20-odd years ago, it probably had all the schematics included, along with a hardware list.

          I'll try disconnecting the white/green wire tomorrow and see what happens.

          Yes, the heating element is connected directly to the triac, on one side of the element to the triac, the other side of the element to the door switch. When the door is opened, I would presume the heating element would normally switch off, and the circulation fan also switches off (which it does).

          Comment

          • tmcw
            Badcaps Veteran
            • Oct 2010
            • 382
            • Ireland

            #6
            Re: Drying Oven - Preset Temp Fail

            When I disconnected the white/green wire, the oven doesn't heat up at all. Turns on, and orange light comes on for heating, but no heat.

            On a more serious note, I just noticed that the door seal is completely perished away, probably like that for months/years. I don't think we'll try saving this one, maybe take some parts out of it for the others.

            Comment

            • stj
              Great Sage 齊天大聖
              • Dec 2009
              • 30998
              • Albion

              #7
              Re: Drying Oven - Preset Temp Fail

              try replacing the 2 chips,

              if you fix it first, you will know all the parts are good.

              Comment

              • R_J
                Badcaps Legend
                • Jun 2012
                • 9535
                • Canada

                #8
                Re: Drying Oven - Preset Temp Fail

                If disconnecting the triac gate does not turn on the oven, then it could be a bad optocoupler (IC4), can you work on the board when the unit is operating? if so check the voltage across pins 1 & 2 of the optocoupler when it is heating and again when it should be at temp and the heat element off.
                With no power applied, check the resistance between pins 4 & 6, maybe the internal triac is shorted
                Here is the datasheet for the ic.
                Attached Files
                Last edited by R_J; 01-22-2021, 10:59 AM.

                Comment

                • redwire
                  Badcaps Legend
                  • Dec 2010
                  • 3906
                  • Canada

                  #9
                  Re: Drying Oven - Preset Temp Fail

                  Mechanical failure - temp sensor beatup up, broke, got water in it, broken wires, corrosion I would also rule out. It might be a thermistor but the resistance could be compared between ovens.

                  The PC board has a +5V, +9V and +18V testpoints so I would check those.

                  The opto-coupler should have high resistance pin 4-6. Possible replacement MOC3063.

                  Comment

                  • kby202
                    New Member
                    • Dec 2022
                    • 2
                    • United Kingdom

                    #10
                    Drying Oven - Preset Temp Fail

                    Hi, hope someone can offer some advice. We have the same Gallenkamp drying oven. I tried to perform a Portable Appliance Test; it failed on the insulation test. I read if it's been unused in a long time there may be a build of moisture or if its dirty. It got cleaned and switched on to dry out any moisture. Initially it did pass the insulation test, but once the oven cooled it failed the insulation test. After some troubleshooting, the only way to pass the insulation test was to disconnect the element. As this oven is no longer supported does anyone have a pdf of the manual and recommend a generic replacement heating element?
                    Thank you in advance, kby

                    Comment

                    • stj
                      Great Sage 齊天大聖
                      • Dec 2009
                      • 30998
                      • Albion

                      #11
                      Re: Drying Oven - Preset Temp Fail

                      you dont insulation test devices that contain mains filters or switching psu's
                      not unless you want to repair them afterwards.
                      your putting 500-2000v into parts rated for under 400.

                      PAT testing is only required for PORTABLE equipment with fixed cables anyway - dont fall for the scammers who say you need to test everything.
                      https://www.hse.gov.uk/electricity/f...ce-testing.htm

                      Comment

                      • sam_sam_sam
                        Badcaps Legend
                        • Jul 2011
                        • 6037
                        • USA

                        #12
                        Re: Drying Oven - Preset Temp Fail

                        Can you please post some pictures of the heating elements

                        To add to what STJ said you want to ONLY test the heater element with a megohm meter no other part of the device or will probably be repairing more because of he said earlier
                        Last edited by sam_sam_sam; 12-03-2022, 06:32 AM.

                        Comment

                        • kby202
                          New Member
                          • Dec 2022
                          • 2
                          • United Kingdom

                          #13
                          Re: Drying Oven - Preset Temp Fail

                          Thank you for the information I've been reading a lot of conflicting information.

                          Comment

                          • stj
                            Great Sage 齊天大聖
                            • Dec 2009
                            • 30998
                            • Albion

                            #14
                            Re: Drying Oven - Preset Temp Fail

                            the whole idea behind pat testing is to check for cables that are damaged or coming loose inside a device.
                            thats why the "P" is for portable and you dont test stuff with removable cables - although you can inspect the cables and plugs.
                            the main things that actually should be inspected are power tools like drills and vac-cleaners/floor polishers etc.

                            Comment

                            • redwire
                              Badcaps Legend
                              • Dec 2010
                              • 3906
                              • Canada

                              #15
                              Re: Drying Oven - Preset Temp Fail

                              I think there's a myth out there that the heating elements absorb moisture and develop high ground-leakage currents. As if it's raining inside the oven, the heating element ends are outside the oven anyway and have silicone end seals. "Drying it out" there's really no way for moisture to get out of the heating element tube.
                              The MgO powder inside I think gets a thin spot or the element expands and just curls and shorts to the tube through it. So anything to shake it up a bit can temporarily make it work (no ground fault) sometimes, for a little while.

                              The PAT test looks like only 500V, the triac's and opto's insulation should take it.

                              Comment

                              • CapLeaker
                                Leaking Member
                                • Dec 2014
                                • 8133
                                • Canada

                                #16
                                Re: Drying Oven - Preset Temp Fail

                                Food for thought. About 20 years ago I had the top element in the oven fail like every 6 months or so. After a while it started popping the fuse too. Moisture one the element, because the climate around here?
                                Well, back then I had the old fuse panel which had clip in fuse blocks that contained two screw in fuses. Guess what?! Yep, the metal clips that connect the fuse block to the power bars in the fuse box were corroded. Cleaned the fuse block up with a wire wheel, put new fuses in and never had a problem with the stove again.

                                Comment

                                • stj
                                  Great Sage 齊天大聖
                                  • Dec 2009
                                  • 30998
                                  • Albion

                                  #17
                                  Re: Drying Oven - Preset Temp Fail

                                  Originally posted by redwire
                                  The PAT test looks like only 500V, the triac's and opto's insulation should take it.
                                  that depends on the device used,
                                  some can run upto a couple of KV and the really high end ones can dump several amps into a big crock clip to check for a solid earth connection between cable & chassis!

                                  Comment

                                  • redwire
                                    Badcaps Legend
                                    • Dec 2010
                                    • 3906
                                    • Canada

                                    #18
                                    Re: Drying Oven - Preset Temp Fail

                                    It makes no sense to do that PAT test in an earth-grounded appliance, a heating element short to chassis is a common fault and not a shock hazard. There's no cords to fail if it's mounted permanently.

                                    I think this thread found the moisture bake is a myth: https://www.badcaps.net/forum/showthread.php?t=84293
                                    Glowing red at 600°C/1,100°F expels moisture lol. I can see a GFCI/RCD tripping due to storage rust/corrosion and/or the heating wire coil expanding and touching the tube when it gets hot.

                                    Comment

                                    • CapLeaker
                                      Leaking Member
                                      • Dec 2014
                                      • 8133
                                      • Canada

                                      #19
                                      Re: Drying Oven - Preset Temp Fail

                                      I live in a high moisture area. I never had any success repairing or rejuvenating heating elements by driving the moisture out of them. Maybe it works for a very brief period, but to me a spent heating element is a spent heating element. Like I said before in my other post, the cause of my problem was corrosion in the fuse box destroying the heating element multiple times until I figured it out. A lot of older fuse boxes around here have this inside corrosion problem. Took me a few months and trips down the basement.

                                      Comment

                                      • redwire
                                        Badcaps Legend
                                        • Dec 2010
                                        • 3906
                                        • Canada

                                        #20
                                        Re: Drying Oven - Preset Temp Fail

                                        I don't get the fuse box affecting the heating element. Usually there is high temperature wire/spade connectors from the heating element to the control circuit, and then that goes to the fuse.
                                        Corrosion I find causes hot spots whichs melt things such as PVC wire or switches etc.

                                        Comment

                                        Related Topics

                                        Collapse

                                        • Fsufan1391
                                          LG Oven model LREL6325F
                                          by Fsufan1391
                                          Hello everyone my oven is less than 2 years old and all of a sudden recently when i run the oven for 15 minutes or so and i am done with it and shut it off. after about 10 minutes it starts chiming and randomly putting it in different modes like the touch screen is being touched but no one is around it. I have tried to reset it by unplugging it multiple times and no luck. I took it apart to see the main board and everything looks fine. I am wondering if i should replace the main board or if its in the control panel. the touch screen is very responsive and works like it should. its only acting...
                                          04-23-2025, 05:42 AM
                                        • schnauzer2021
                                          Chasis bond to neutral to overcome oven tripping problem
                                          by schnauzer2021
                                          Hi,

                                          I have a 6 years old Rowenta table top oven. The oven have both convection and grill function.
                                          Last Christmas I used it to roast a turkey with the convection mode for about 4 hours without any problem. When I switched to grill mode to deep browning the turkey for the final half hour, the power trips after a while.

                                          I have not been using the oven since then. Last week, when I try to use the oven again, both convection and grill mode trip the power supply a few minutes after turning it on.

                                          As I was too busy with my work, I leave it to the repairman...
                                          09-13-2021, 02:40 AM
                                        • bpjl
                                          SEF 551.10 - Error starting the reflow oven.
                                          by bpjl
                                          Hello,

                                          When starting the SEF 551.10 reflow oven, the following problem occurs:





                                          It seems to me that this is related to corrupted software. Does anyone have files that can fix this device? Or maybe someone uses such an oven and is able to make a sector copy of the memory card located in the computer controlling this oven?

                                          PS.: Of course, appropriate emails for help have been sent to the manufacturer and distributor in my country. No response yet, but I don't expect much. It's a fairly old device and from what I know, it's no longer...
                                          10-17-2024, 06:30 AM
                                        • foshland
                                          MSI MAG Z390 Tomahawk BIOS temp, allways at 0ºC and CPU goes protection 800Mhz
                                          by foshland
                                          Hi to all, my motherboard (bought second hand as faulty, but now working after cleaning the ram slots) have a strange failure that shows 0 Celsius degrees in BIOS, because of that, the CPU goes to protection mode and gets stuck at 800Mhz.

                                          Since its a 8-9 Intel generation, I know the CPU temp is the hottest temp of the cores (as i have read in intels documentation) so its not a diode or sensor on the board.

                                          External temperature (motherboard temp) is working OK, 27ºC more or less

                                          In windows, all the software I have tested shows correct the Core temps
                                          ...
                                          12-16-2023, 09:08 PM
                                        • mosfetratus
                                          concerning gtx675m vram temp
                                          by mosfetratus
                                          Is 98C vram temp are safe? Less demanding games goes 80ish.

                                          Funny thing, there were old non stock thermal pads very botched, so i replaced them. After, temperatures remained the same, but performance (fps) improved significantly. With both thermal pad application GPU core temp was similar.

                                          furmark goes about 104C, amperage keeps rising so do temps.
                                          01-30-2025, 08:41 AM
                                        • Loading...
                                        • No more items.
                                        Working...