ESR Meter- MESR-100 feedback- did i do something wrong ?

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • herita
    Member
    • Aug 2012
    • 17
    • France

    #1

    ESR Meter- MESR-100 feedback- did i do something wrong ?

    Hi everybody,

    Recently, i have bought one ESR meter (MESR-100) from the ebay .

    link here :
    http://cgi.ebay.fr/Capacitor-Capacit...item5ae7d1210c

    But since my first try with one bad and good cap, the MESR-100 shows not so much difference betwen the good and the bad one. so i wish if there is someone can tell me if i did something wrong or the esr meter MESR is a brik can't be trusted ?

    Here is the new esr meter MESR-100


    I did Zero each time of test with each cap:


    Bad cap: 6.3v 3300µf


    With bad cap: - ok it shows : "-1.59 R" what means of "-" here ?
    [Img]http://img59.*************/img59/2815/img2403resize.jpg[/img]

    With good cap same value: 6.3v 3300µf : it shows "-1.70 R" not so much different with the bad one, i think



    any suggestion, pls

  • tom66
    EVs Rule
    • Apr 2011
    • 32560
    • UK

    #2
    Re: ESR Meter- MESR-100 feedback- did i do something wrong ?

    Looks like it is faulty, somehow not storing the zero measurement correctly, did you use a shorting link for zeroing though?
    Please do not PM me with questions! Questions via PM will not be answered. Post on the forums instead!
    For service manual, schematic, boardview (board view), datasheet, cad - use our search.

    Comment

    • herita
      Member
      • Aug 2012
      • 17
      • France

      #3
      Re: ESR Meter- MESR-100 feedback- did i do something wrong ?

      hi tom66, what did you mean use a shorting link for zeroing ?
      uhm, sure, i did zero the device, and have shorted caps with screw device already.

      Comment

      • tom66
        EVs Rule
        • Apr 2011
        • 32560
        • UK

        #4
        Re: ESR Meter- MESR-100 feedback- did i do something wrong ?

        To zero it, you may have to connect the plus and minus of the TESTER together, so it can measure any resistance of its contacts and eliminate them from the test.
        Please do not PM me with questions! Questions via PM will not be answered. Post on the forums instead!
        For service manual, schematic, boardview (board view), datasheet, cad - use our search.

        Comment

        • herita
          Member
          • Aug 2012
          • 17
          • France

          #5
          Re: ESR Meter- MESR-100 feedback- did i do something wrong ?

          ok, nice, but i think that i have done it already, because the MESR-100 has its own bouton Zero to zeroing, so before testing each cap, i did zero it down.

          Comment

          • mariushm
            Badcaps Legend
            • May 2011
            • 3799

            #6
            Re: ESR Meter- MESR-100 feedback- did i do something wrong ?

            You may have to connect the two probes together or use a wire to make the connection between the terminals BEFORE you press the zeroing button.

            The meter will have an internal resistance without the probes, when you put the probes in the sockets the resistance of the cables and the tips adds up and changes the measurements, so you have to zero it again.
            That's done by touching the probes (so shorting the circuit).

            If it still shows bad values, there should be some potentiometers on the back that can be used to adjust what you see.

            Comment

            • herita
              Member
              • Aug 2012
              • 17
              • France

              #7
              Re: ESR Meter- MESR-100 feedback- did i do something wrong ?

              humh, ok, so after you, there will be steps which i have to follow :

              1. with the probes already put in the meter, i connect them together and switch on the meter.
              2. alway with the two probes connect together, i press zero to zeroing the meter.
              3. after zeroing, i pull out the two probes and begin to test the cap.

              haha, i did the 3 steps above and after pulling out the two probes my meter says overflow.

              Comment

              • tom66
                EVs Rule
                • Apr 2011
                • 32560
                • UK

                #8
                Re: ESR Meter- MESR-100 feedback- did i do something wrong ?

                That is fine, now test with a cap and see if it still says overflow.
                Please do not PM me with questions! Questions via PM will not be answered. Post on the forums instead!
                For service manual, schematic, boardview (board view), datasheet, cad - use our search.

                Comment

                • severach
                  Badcaps Legend
                  • Aug 2007
                  • 1055
                  • USA

                  #9
                  Re: ESR Meter- MESR-100 feedback- did i do something wrong ?

                  Lead zero doesn't sound like the problem. Leads typically measure 0.10 ohms or less, not 1.5 ohms. I find it easier to subtract the value in my head rather than zeroing all the time. Meters often are not designed to notify you that the zero adjusted resistance is less than zero meaning the zeroing wasn't done accurately. Maybe that's what this meter is doing, indicated by the negative (-) value.

                  After zeroing, make sure that shorting the leads reads zero. Press the leads together hard and soft to see if varying pressure changes the zero value. Find some low ohm resistors to see if the meter shows the same resistance as marked.

                  While I agree that your blue cap (OST?) is bad I do not agree that your "good" Ltec cap is good. Can you try some known low ESR caps from Nichicon, Sanyo, or Panasonic that we can have a little more trust in?

                  Call me crazy but I trust a cheap meter before I trust a cheap cap.

                  I like the probe holes. The MESR-100 isn't cheaper enough than the Atlas ESR70 to be worth losing the microfarad measurement.
                  sig files are for morons

                  Comment

                  • herita
                    Member
                    • Aug 2012
                    • 17
                    • France

                    #10
                    Re: ESR Meter- MESR-100 feedback- did i do something wrong ?

                    hi everybody,

                    i think that everything is fine now, i did these steps before mesuring caps and the meter can finally tell me wich cap is good or bad.

                    Severach, i definitely agree with you, cheap cap can't be trusted. ok the green cap Ltec is a band new cap (i have got about 200pieces of these), but i have tested also with some others brand like Nichicon.

                    and with the MESR-100, these probe holes are ok, but i also have to use a tweezer in order to zeroing the meter before using.

                    i will try to make a video when i use this meter, in order to help others guys like me in the future

                    and yes, the Atlas ESR70 is expensive and you have to prepare battery for it, i think.

                    Comment

                    • int0x13
                      Member
                      • Nov 2012
                      • 42
                      • France

                      #11
                      Re: ESR Meter- MESR-100 feedback- did i do something wrong ?

                      Originally posted by herita
                      hi everybody,

                      i think that everything is fine now, i did these steps before mesuring caps and the meter can finally tell me wich cap is good or bad.

                      Severach, i definitely agree with you, cheap cap can't be trusted. ok the green cap Ltec is a band new cap (i have got about 200pieces of these), but i have tested also with some others brand like Nichicon.

                      and with the MESR-100, these probe holes are ok, but i also have to use a tweezer in order to zeroing the meter before using.

                      i will try to make a video when i use this meter, in order to help others guys like me in the future

                      and yes, the Atlas ESR70 is expensive and you have to prepare battery for it, i think.
                      Hello herita, i wanted to buy the same model on ebay, finally do you advice it or not?
                      Did you do the video to show how to use the meter?
                      Thanks

                      Comment

                      • MDOC
                        EngineeringTech
                        • Mar 2010
                        • 146
                        • USA

                        #12
                        Re: ESR Meter- MESR-100 feedback- did i do something wrong ?

                        herita,
                        Do a little test to find out if you really need to zero the meter when using the terminal block (the holes) in the meter:
                        1) Zero the meter by attaching the two probes together.
                        2) Remove the probes from the meter. Insert a known good cap in the terminal block. Write down the reading.
                        3) Without removing the cap, turn off the meter. Turn it back on again 5 seconds later.
                        4) Without zeroing again, measure the cap again. Note the meter reading.

                        If there is no appreciable difference in the reading, you don't need to do zeroing because you're not using the probe leads.

                        Comment

                        Related Topics

                        Collapse

                        • chth96
                          Is it possible to test small ceramic disc capacitor with analogue meter?
                          by chth96
                          I read EBOOK of Testing Electronic Component and found that

                          "In order to test the small blue resin coated ceramic disk capacitor which has 104 50v specification,it is possible to test this small ceramic capacitor by making use of analogue meter.
                          Set to times 10Kohm that has 12v output from the probe to check disorder of ceramic capacitor.
                          It is not uncommon that testing result of digital capacitance meter is okay,But It show shorted reading when you check it with analogue meter."

                          Is it really possible to test small ceramic disc capacitor with analogue...
                          03-09-2024, 10:03 PM
                        • marce002
                          UT210E clamp meter amps measurements wrong
                          by marce002
                          Hello all

                          I have an UT210E clamp meter that went wrong doing amps measurements in continuity DC, it shows an offset of around 4/5A when first set, (lever at DC 10A) at 2A DC shows fixed 0L always......,,,,,, and also does not allow me to (zero button) , and as i said it is even worse in the 2A !! setting which do nothing and shows "0L"... when you can measure DV in 20A after achiving a ZERO with many RESET ZERO times , you can measure with a lot of offset..... eg 1A current is measured at around 650 mamps, . Similarly, a 4A current is shown at around 3A. |O
                          ...
                          05-08-2020, 11:52 AM
                        • sam_sam_sam
                          Modification of a Fluke multi meter that uses three AA batteries to use a 14500 lithium ion battery
                          by sam_sam_sam
                          I have had with regular AA batteries that leak this meter is expensive new I already replaced the battery battery clips once and the same thing happened again and I am tried of so it is getting modified to accommodate lithium ion batteries that are the same size excluding the positive terminal and removing the battery clips I will have enough room for the BMS protection board

                          I going to put 3 lithium ion battery in parallel with the BMS protection board and a way to charge the battery externally and heat shrink the battery pack and use a JST connector so I easily remove the battery...
                          12-31-2023, 06:04 AM
                        • jorpec
                          Changing the scale of an analog meter
                          by jorpec
                          Hi

                          I want to change the scale of an analog meter, to monitor the charge of an li-ion single cell, so the needle at the end would point 4.2 volts and the 0 on the meter would point 2.5 volts or so, i tried to add diodes or a zener in series with the meter but the changes of the ambient temperature also changes the voltage value on the meter
                          I wonder how i can accomplish this so the meter would be precise at any temperature ?