Metrix 803B Oscilloscope power supply issue
  • itsthatidiotagain
    • Dec 2014
    • 39
    I'm looking for a bit of advice or better yet a schematic for a Metrix 803B oscilloscope.
    The 40MHz animal here...
    .. I picked one up for very little on ebay, and it looks quite a useful thing to have around.
    The listing stated that it was dead, but in clean condition, so I took a gamble that it might be something simple.... needless to say it wasn't the fuse or the mains switch, and the next logical step is to check the power supply.
    Naturally I can find no schematic on line for it, so before I start to reverse engineer the thing, I wondered if anybody else had a schematic, or access to another 803B to compare.
    Everything looks clean and there are no obvious crispy bits or popped or bulging caps. When I power on, the mains gets from the power jack board, through the mains switch and over to the PSU board, and appears at the input side of the switch mode supply. Thats as far as I have probed so far, other things keep getting higher priority.
    I dont appear to get any audible high frequency squeak from the switch mode transformer or static on the tube, at start up suggesting it might be fairly simple to track down, but before I actually start to get my hands dirty I thought I would ask around in case the fault is a well known issue that someone here has enountered before.
    An incomplete set of pictures of the innards of the poor beast complete with optional dust are available here.... and yes, I know that the fuse is missing in some of those pictues, I had popped it out to test it. I did remember to re-fit it.
    • 31278
    Re: Metrix 803B Oscilloscope power supply issue
    upload the foto's, that google shit gives me a headache trying to see the foto's fullsize.
    and what is the part number on the 8pin chip near the big capacitor?
    • itsthatidiotagain
      • Dec 2014
      • 39
      I have every sympathy with your Google allergy. I posted there because I didn't want to clutter up this forum, and Google tends not to mash the original images up *too* much, or impose limits on their size.
      I plan on getting some better pictures of the PSU board in particular, this evening. I'll make sure that as many of the markings on the componets as possible are legible. If I have the time I might unsolder the mains input wires and pop the silkscreen side on my flatbed scanner too, in case I have to reverse engineer from the PCB.
      While I had it in bits, I spotted an un-connected single row of header pins, on the main PCB, which I suspect are for the "RS232 Serial" option (the case also has a punchout next to the 25 way D one, for what looks very like an IEEE connector, but there is no mention of that anywhere on the manufacturers site).
      If I manage to coax the thing back in to life, I will take a crack at seeing what the signals on that header connector look like.
      My gut geeling is that the serial option is probably a very expensive 25 'D' type on a bit of flat ribbon cable, with some equally expensive software to drive it.
      However first on the list of things to do is identify the 8 pin chip beside the 400V cap and the two other chips that look suspiciously like opto couplers nearby. I'll see if I can also identify the "transistor" on the black heatsink.
      Armed with that, and the values of some of the other eletrolytics, and passives you never know... it might not be such a poor investment after all.
        • 31278
        i suspect it's the startup capacitor,
        and it's only about 10p to find out.
        • itsthatidiotagain
          • Dec 2014
          • 39
          I hope you are correct, here is a slightly better crop from the previous images (I'm at work at the moment, so I cant bunk off and take some better pictures just yet). I presume you mean C5
            • 31278
            probably c5, but change c15 too.
            and use high temp, low esr stuff like panasonic FR
            i say probably because i dont know the chip number.
            psu chips need a startup current, most do it by charging a small cap.
            older ones used a high-voltage bleed using a couple of high value resistors in series to the mains smoothing cap.
            those often changed value but looked o.k. to the eye!!!
              • Jul 2011
              • 6070
              Originally posted by itsthatidiotagain
              However first on the list of things to do is identify the 8 pin chip beside the 400V cap and the two other chips that look suspicious.
              That would be the first thing I would do
              I just work on a power board that was very dead it was a 7 pin ic switcher
              • itsthatidiotagain
                • Dec 2014
                • 39
                Warning: Off topic post ahead.. and old f*rt alert.
                I must confess that it has been quite a while since I last did anything serious with switched mode PSUs.
                Many years ago I ran a small electonics repair workshop as a sideline to the software business I was with, so I cut my teeth on Wyse 50 terminals and large colour IBM CRT monitors. These days everything seems to be headed straight to the skip the moment something goes pop, which bothers me.
                Just for fun I googled Wyse 50 schematic, and now I feel really old... (c) 1986.... surely not... however on the plus side I still know one end of a capcitor from the other. If anybody wants to see a "classic" CRT/SMPSU circuit, take a look at the diagrams around page 158 of that manual.
                I don't have an ESR meter or even another O'scope, never mind one with an octopus component checker (these two devices probably kept me in bread and butter for quite a number of years), so this may take a little longer than would otherwise be the case.
                I was pleasantly surprised when I cracked open this particular scope and discovered a through hole board with a familiar layout. All of these ramblings are a prelude to a confession that I might not get a chance to take more pictures this evening, but rest assured, tomorrow at the latest...
                  • 31278
                  well fwiw i replace startup caps on sight usually, because so long as the kit isnt visibly burned it often fixes it without having to get any test gear out.
                  when i fixed sattelite boxes made by pace, i was buying startup caps in bags of 100.
                  • itsthatidiotagain
                    • Dec 2014
                    • 39
                    I popped the PSU board out from the 'scope and took a whole lot more pictures
                    available here..
                    The "mystery chip" is a TDA4605 and the two caps C5 = 100uF (250V?) exact voltage to be determined when I find my small dental mirror and poke it behind the cap to get a better look. C15 = 22uF 50V and the two other chips are as we suspected optocouplers - in this case 4N25 . Finally the switching transistor is an ST P4NB80FP
                    The TDA460 PDF is watermarked as "Obsolete Product", suggesting there will be a better alternative, does anybody here know what that might be?
                      • 31278
                      this one has an example circuit.
                      i would start by replacing those caps.
                      btw, did you check the fuse?
                      • itsthatidiotagain
                        • Dec 2014
                        • 39
                        Hi stj, yes I checked the fuse, and I even remembered to re-fit it afterwards but thanks for the reminder. Good advice... I even buzzed the fuse out in circuit, as it would not be the first time that grubby contacts on one of those cheap blade style fuse holders had caused things to stop working.
                        Looking at the application circuit, and comparing with the PSU, I would suggest that C5 is not likely to be 100uF 250V (as I previously suggested) since this cap is the chip supply cap (i.e. connected between pin 6 (V6) and ground and is shown as 100uF in the Seimens reference design).
                        Since the supply (V6) voltage to the TDA4605 is stated as having an absolute maximum rating of +20V, I would think that a 100uF 100V cap would do the trick, and since I have some of those "in stock" as it were, I will swap that.
                        For C15 I have just picked up a 22uF 100V from Maplin Electronics for the heady sum of 0.49p
                        Both caps are physically very similar in size to the originals so I think tomorrow might be time to break out the solder sucker and see if we are in luck. Before I do that though I think I might have a quick poke around on the low voltage side to see if I can spot any shorts on the regulators or anything else that might cause the TDA4650 to sulk.
                        Some quick notes from inspecting the board, in case anybody else is following along behind...
                        Metrix 803B PSU
                        12 pin power connector
                        Although the board has no obvious markings, assume pin 1 is the one with the black stripe.
                        Note: DC ground is tied to main earth, and that for safe operation of the scope it must be earthed.
                        As with most scopes, all scope input BNC connectors therefore have the Ground side tied to mains earth.
                        The following pins have been tentatively identified by visual inspection only.
                        Pin 1) -12V ?? Unverified
                        Pin 2) -5V ?? Unverified
                        Pin 3) +12V ?? Unverified
                        Pin 4)
                        Pin 5) +5V ?? Unverified
                        Pin 6) GND
                        Pin 7) GND
                        Pin 8)
                        Pin 9)
                        Pin 10)
                        Pin 11) GND
                        Pin 12)
                        NOTE: Mosfet may be Pin-to-Pin Replaceable with IRF IRFIBE30G , but the ST P4NB80FP looks to be available from several sources.
                        • itsthatidiotagain
                          • Dec 2014
                          • 39
                          Well, I finally got 5 minutes to swap those two capacitors, however I really should pay more attention when dealing with shop assistants. The 22uF 100V capacitor from Maplin turned out to be a 10uF 100V capacitor, so back to the store for a replacement in the next few days.
                          I did swap the 100uF 50V, for a nice shiney new 100uF 100V but it made no difference, and since I literally had 5 minutes, I didn't even get a chance to poke around to see if I could spot anything else that might be amiss. Well that isn't strictly true, I did carefully buzz through from the the live and neutral on the plug to the bridge rectifier, and did a quick diode test on the rectifier diodes and that all looks fine.
                          • CapLeaker
                            Leaking Member
                            • Dec 2014
                            • 8351
                            I would be surprised if the circuit would be that sensitive, that the 10uf 100V cap wouldn't work.
                              • 31278
                              hard to say with old 80's designs.
                              • itsthatidiotagain
                                • Dec 2014
                                • 39
                                A quick update for those following...
                                I have purchased the Mosfet, TDA4605 and the 4N25 optoisolators. I'll post the result of changing that lot.
                                My gut feeling is that I should be able to run the PSU board with the main board and HT connectors disconnected without having to add a dummy load, but that I may need some feedback to one of the optoisolators to make it stay up. One of them looks to be a voltage feedback, to control the SMPU the other looks to be some sort of standby circuit.
                                What is anybody elses take on this? Also what purpose do you think the two SOIC8 chips on the silkscreen side of the board serve, they look to be a TL082C and a TL071C - my guess was that they provide some sort of precision voltage references.
                                • itsthatidiotagain
                                  • Dec 2014
                                  • 39
                                  Parts haven't arrived yet, but I supect I might have spotted at least part of the problem.
                                  I checked most of the resistors the other day, but I thought I'd go over the circuit again and compare it with this similar one. and blow me if I didn't find a problem.
                                  R4 which I could have sworn I checked before (220K 1W 1%) looks to be open circuit. This appears to be connected to pin 3 on the TDA4605, (suggesting perhaps a short in the TDA4605, or a thermal stress crack in R4).
                                  I've nothing identical in my parts bins, so 5 of those are also now on order.
                                    Can you see how that resistor is connected to the other end in the circuit? It may give you the idea why the circuit is not working.
                                    Pin 3 is the Voltage monitoring pin, if it is too low then the SMPS will shut off, so that makes sense. You can temporary try another resistor in that resistance range to see if it will work.
                                    Page 3 of the IC spec sheet.
                                    Input for Primary Voltage Monitoring: In the normal operation V3 is moving
                                    between the thresholds V3H and V3L (V3H > V3 > V3L).
                                    V3 < V3L: SMPS is switched OFF (line voltage too low).
                                    V3 > V3H : Compensation of the overload point regulation (controlled by pin 2)
                                    starts at V3H : V3L = 1.7.
                                    • itsthatidiotagain
                                      • Dec 2014
                                      • 39
                                      For future reference, here are my notes from the examination of the patient to date (to assist Google and anybody else with a similar issue).
                                      Metrix 803B PSU - X00391C95A - 60286A
                                      12 pin power connector on green cable to main PCB.
                                      Pin 4) STD-BY> Unverified
                                      Pin 6) GND
                                      Pin 7) GND
                                      Pin 8)
                                      Pin 9)
                                      Pin 10)
                                      Pin 11) GND
                                      Pin 12)
                                      PSU Components to check.
                                      R1 - 12K 2W 1% (5 band Brown Red Black Red Brown)
                                      R2 - 12K 2W 1% (5 band Brown Red Black Red Brown)
                                      R28 - 1M2 1/2W 1% (5 band Brown Red Black yellow Brown)
                                      C5 - 100uF 50V - Checked - replaced with 100uF 100V
                                      C15 - 22uF 50V - http://www.ebay.co.uk/itm/ECA1HM220-...item2ed0cb867d
                                      Z1 - TDA4605 - Control IC for SMPSU
                                      https://cdn.badcaps-static.com/pdfs/...731590de00.pdf
                                      Z10 - 4N25 - Opto coupler - https://cdn.badcaps-static.com/pdfs/...9b513c4a42.pdf
                                      Z11 - 4N25 - Opto coupler - https://cdn.badcaps-static.com/pdfs/...9b513c4a42.pdf
                                      Q1? - P4NB80FP - N-CHANNEL 800V - 3Ω - 4A TO-220/TO-220FP PowerMESH MOSFET (Equiv = BUZ80FI?) -
                                      R3 - 470K 1/2W 1% (5 band Yellow Violet Black Orange Brown)
                                      R4 - 220K 1W? 1% (5 band Red Red Black Orange Brown)
                                      R10 - 100R 1/2W 1% (5 band Brown Black Black Black Brown)
                                      On silkscreen side of board. SOIC8 near voltage regulator - Marked 082C - TL082C smd
                                      or
                                      https://cdn.badcaps-static.com/pdfs/...967b60572a.pdf
                                      Possible source http://www.ebay.com/itm/TL082C-SMD-D...-/320734331798
                                      Similar SOIC8 at other end of voltage regulators - Appears to be TL071C -
                                      Most likely source http://www.ebay.co.uk/itm/TL071CD-ma...item4398feb620
                                        R4, at worst case the voltage will be around 330VDC, so that is about 1.5mA which means the power dissipation will be about 1/2W so I would use 1W.
                                        • Soulzink
                                          Samsug washing machine. Power supply DC41-00189a board issue
                                          by Soulzink
                                          Hi everyone,
                                          I am trying to fix a samsung washinmachine power supply main board.
                                          I already replaced it with another one to fix the machine but i want to try understand the issue with the old board.
                                          Therefore i hope to be able to get some help to explain what to check in order to troubleshoot.
                                          The power supply board reference is the following: DC41-00189a
                                          I have followed one thread (https://www.badcaps.net/forum/troubl...ingmachine-rel...
                                          04-19-2025, 06:26 AM
                                        • Tynan Dill
                                          by Tynan Dill
                                          Assuming...
                                          11-22-2024, 01:46 PM
                                        • KYBOSH
                                          Eufy Floodlight Cam 2 Pro Dead. Power Supply issue...
                                          by KYBOSH
                                          I have a floodlight-security camera that recently died on me.
                                          its hardwired to 115v po115v outlet. One evening it simply died. After some troubleshooting (ensuring it had power and ensuring the electronics were actually functional by supplying 5v via usb port) i realized the power supply must be the issue. The supply was noticably warm to the touch even though it was not powering anything.
                                          Trying to understand what i was up against i went looking for others with similar issues and anyone who might have attempted a repair. While a few ppl have had the...
                                          07-01-2024, 06:54 AM
                                        • Cookybaby
                                          Samsung UN60ES6500FXZA Power Supply Issue
                                          by Cookybaby
                                          Hello Everyone, this is my first post and I am excited to be a part of such a great community. I've been reading these forums for awhile and there really is some good friendly people here. No bashing and being rude like 90% of other forums Ive been to. Lets get to the good stuff.
                                          I have two years of college in Electronics Engineering Technology. Ive been tinkering with electronics sine about 1990. I know enough just to be dangerous, so be gentle.
                                          Make: Samsung
                                          Model: UN60ES6500FXZA
                                          Power Supply P/N: BN44-00525A
                                          Symptoms: The TV was...
                                          08-20-2023, 12:26 PM
                                        • lp20th
                                          Samsung Plasma TV Power Supply Issue
                                          by lp20th
                                          I have a vintage 2008 Samsung 50” Plasma TV, Model code PN50A550S1FXZA with a power supply part number BN44-00222A. I repaired the power supply board several years ago finding a couple of cold solder joints on the back side of the board and it's worked great up until now. Recently a fried sound occurred with the smell of burnt electronics and the TV shut down. Trying to turn it back on yielded the typical relay clicking with no TV boot up. Upon investigation, removing feeds to the other boards within the TV it was found that disconnecting the logic board from the power supply allowed the...
                                          11-28-2023, 10:28 AM