Badcaps.net Forum
Go Back   Badcaps Forums > Electronics Theory and Troubleshooting > Troubleshooting Computer Displays
Thread Tools Display Modes
Old 08-27-2010, 03:26 PM   #41
PlainBill
PlainBill's Avatar
City & State: Phoenix, AZ
My Country: USA
Line Voltage: 120VAC 60Hz
I'm a: Hobbyist
Default Re: Acer x193w with Darfon B092-XXX inverter - assistance needed! [lotsa pics!]
Quote:
Originally Posted by HexInverter View Post
Okay Bill, so while getting all set up for this, I've run across an issue (maybe?) and just wanted to get some clarification before I plug anything into line voltage - As you can see form these shots, the inverter backplate design isn't going to let me test it in its chassis with everything hooked up the way it's supposed to be. The limiting factor appears to be the short, wider ribbon cable that goes from the logic board to the LCD.
So, for these tests - I think I either need to not have the LCD plugged into the logic board, or I need to unscrew the inverter/logic board from the chassis and put it on some insulating material for the test.
Which is preferred, or do you have some other idea judging by the pictures?
Thanks kindly
You've got a bad habit of taking close up shots rather than a single all - over shot. I've got a problem linking the various pictures to together. But the first picture on this page gives me all the information I need.
This is pretty simple. You MUST have the power supply / inverter hooked to both the CCFLs and the logic card. You also must have access to the back side of the power supply / inverter. You should have the front panel hooked to the logic card. It is desirable, but not absolutely necessary to have the LCD panel hooked to the logic card.
In situations like this I like to take a piece of the pink foam that usually comes with motherboards and use that as an insulator between the cards and the back of the LCD panel. I'll then hook the ground traces of the power supply to the back shell of the LCD panel. A primary consideration of this arrangement is you must avoid contacting the area of the power supply that is at 165 (or higher) volts. That would be uncomfortable.
PlainBill
__________________
PlainBill is offline   Reply With Quote
Old 08-27-2010, 03:45 PM   #42
HexInverter
CD40106
HexInverter's Avatar
City & State: Manitoba, Canada
PlainBill
Great, makes sense now - thanks! Sorry for all my silly questions by the way...I'm one of those people that can't stand being unsure about something...especially when it's line voltage I'm dealin' with!
Yeah...I'm real bad with those close ups. I don't have a tripod for my camera, and the image stabilization on it is AWFUL you can't take a clear shot unless it's on a tripod or your arms are rested up against something - hence why my shots are always up close (resting my forearms/wrists or elbows on the table).
Off to scope-n'!
HexInverter is offline   Reply With Quote
Old 08-27-2010, 04:19 PM   #43
HexInverter
CD40106
Okay, I've got the whole thing rigged up and ready to go - just one more thing before I proceed with the tests, if you don't mind.
When you say the oscilloscope ground goes to the inverters corner ground plane, do you mean the ground clip on the probe, or the oscilloscope's ground connector on the bottom right in the picture I attached? I've always used the probe's ground for low current/voltage circuits, without hooking up the ground on the oscilloscope, but I have no idea which one to use for a line powered device like this. If I had to guess, I'd think the oscilloscope's ground is there for when running on battery mode (and not properly earth grounded)?
It's a battery powered oscilloscope running off of the wall plug - it's connected to a properly earthed circuit with a 3 prong plug right now - I read somewhere you could create a ground loop, and that would be bad if you do it wrong, so figured I'd check before doing something potentially stupid
Old 08-27-2010, 05:28 PM   #44
PlainBill
PlainBill's Avatar
My Country: USA
I'm a: Hobbyist
Caution is good. The probe's ground would go to the ground for the power supply / inverter. If you visualize the monitor assembled the ground pin on the power cord, the DC ground on the power supply, and the metal shell on the VGA connector are all connected to the metal cover over the logic card and power supply.
PlainBill
Old 08-27-2010, 06:35 PM   #45
HexInverter
CD40106
Okay, well I just got back from trying these tests out...
I tested the scope just to make sure I can get a waveform on it with a simple square wave oscillator. No problemo, so I think it works. Onto the tests...
When testing over the transformer, I can't get a stable waveform - though there is one when the buzzing I described earlier happens.
This is what happens:
- I turn the monitor/power bar on, the blue power LED comes on
- A moment later, there is a buzzing sound coming from the board. At this point, the oscilloscope displays a waveform that is either noise, or I just can't get focused on it for the second that the transformer is on
- The buzzing stops a second later, and the oscilloscope displays nothing.
- The board waits another second, then buzzes again, then stops - same thing.
- The power LED turns orange.
After that, there is nothing displayed on the oscilloscope, and no buzzing. I assumed this is what you meant by not getting a waveform and went onto the next test, but I went to point V1, and tried numerous settings with the probe clipped on there. I cannot get a waveform to display for V1! Weird.
Out of curiosity, I went back and forth between the top and bottom transformer points. The top one displays a higher amplitude waveform than the bottom. Does that matter, or is the amplitude relative to how close I am placing the probe on the board? Whether the waves are the same or not in frequency or shape, I can't tell because it's just a blur of lines for the second it's on.
Does this mean the whole board is losing power (is that possible?) or am I just not using the scope right somehow?
Old 08-27-2010, 08:37 PM   #46
PlainBill
PlainBill's Avatar
My Country: USA
I'm a: Hobbyist
The buzzing is significant; that may have to be investigated.
Here is a quick summary of what happens when the monitor is powered up.
The power supply comes up. The processor(s) on the logic card initialize, and determine the last status. In this case, the status was 'On', so the inverter controller is switched on, and the processors look for a video signal.
The inverter controller drives the transformers (you actually have two transformers in one package) at full power while monitoring the voltage across and current through the CCFLs. If the voltage goes too high it will reduce the drive. As the CCFLs ignite the current rises and the controller reduces the drive. If the voltage across any CCFL is too high or the current through any CCFL is too low or too high the controller shuts down. This whole sequence takes a couple of seconds.
In the meantime the processor is displaying the video signal, if present. After 5-10 seconds without video signal the processor will switch the power LED to orange and turn the inverter off.
Now, back to troubleshooting. The waveform you should be seeing on V1 - V4 should be under 3 volts at about 100 Khz. The waveform at I1 - I4 should be over 3 volts and at the same frequency.
The fact that the 'non-contact' voltage is different at the two ends and the buzzing sound indicates a possible short. I'd suggest seeing if you can measure a voltage at V3 and V4, and I3 and I4.
PlainBill
Old 08-28-2010, 12:59 PM   #47
HexInverter
CD40106
PlainBill
Okay! Hopefully this means something I can only assume that this many zero voltages readings is a bad thing
I took my DMM and measured all of the points. Just like the oscilloscope, there's only ever a reading when the transformers buzz. I was pretty sure they are AC voltages, as someone mentioned near the beginning of the thread that everything's AC after the SMT FET, but figured I'd measure both AC and DC in case I was wrong.
Here are the results (I have put a "/" between results. The first result is the first time the transformers buzz, and the second is the second buzz of the transformer.)
AC Voltage Tests
V1: 14.7v / 12.6v
V2: 17.3v / 16.5v
V3: 8.6v / 8.5v
V4: 8.1v / 8.0v
I1: Zero / Zero
I2: Zero / Zero
I3: Very small number (.1v I think it was) / Zero!
I4: Zero / Zero
DC Voltage Tests
V1: 6.89v / 6.54v
V2: 8.22v / 7.84v
V3: 4.16v / 4.16v
V4: 3.4v / 3.4v
I1: Zero / Zero
I2: Zero / Zero
I3: .01v / Zero
I4: .05v / Zero
As for probing with the scope, I was having an impossible time trying to get a waveform on anything but the non contact transformer test. I think it would help if I had some way of securing the probe lead to the pad while playing with the scope settings. As it stands, I only have those IC leg clip on probes, so they're not very good for solder pads I did switch it on and off a bunch trying to get a reading as the transformers are powered on for those few seconds, but couldn't manage to get anything.
Old 08-28-2010, 04:25 PM   #48
PlainBill
PlainBill's Avatar
My Country: USA
I'm a: Hobbyist
Those numbers look very much like the lower side of the transformer has shorted turns in the secondary. At this point I would suggest either replacing it, or using a ring tester to verify shorted turns. No ring tester? Keep reading.
You had mentioned earlier that you had used a square wave to check the scope. Unsolder the transformer and hook one secondary across the output of the square wave generator. Hook the scope probe to the output of the square wave generator. You should see ringing like in the attached picture. Now try the other secondary. If they both show the same number of rings, the transformer is good. If only one secondary shows ringing, the transformer is bad. If neither shows ringing, I may be giving you bad advice.
PlainBill
Old 08-28-2010, 04:30 PM   #49
HexInverter
CD40106
PlainBill
Are we building ourselves a home made ring tester? Haha...I have no idea what one is, but this is cool!
Off to try that, I'll report back in a bit!
Old 08-28-2010, 05:59 PM   #50
HexInverter
CD40106
Okay - both of the waveforms appear to be the same. It doesn't look anything like the ring in your image. Is my oscilloscope broken? I don't know why it doesn't display vertical lines.
Also, the square wave generator I speak of is just a basic CMOS oscillator - not an extremely accurate function generator or anything, so the square wave has a very slight corner on one edge of the waveform. Is this going to render the test useless?
That would be a bummer if it was the oscilloscope being broken...it appears to work properly though - a square wave shows horizontal lines only.
So, have any ideas of what to try next?
Old 08-28-2010, 11:19 PM   #51
PlainBill
PlainBill's Avatar
My Country: USA
I'm a: Hobbyist
We are going to have to wait. I've got a similar transformer here; one secondary passes the ring test, the other fails. It might be a day or two before I can hook it up to the scope and square wave generator.
PlainBill
Old 08-29-2010, 03:00 AM   #52
Toasty
Toasty's Avatar
City & State: East Coast, USA
Line Voltage: 220-221
I'm a: Professional Tech
>>Are we building ourselves a home made ring tester? Haha...I have no idea what one is, but this is cool!<<
http://www.badcaps.net/forum/showthr...?t=8231&page=2
Toasty is offline   Reply With Quote
Old 08-29-2010, 06:16 AM   #53
PlainBill
PlainBill's Avatar
My Country: USA
I'm a: Hobbyist
Thanks, Toasty. I knew someone would step in with some help.
PlainBill
Old 08-29-2010, 06:39 AM   #54
Toasty
Toasty's Avatar
NP
They are so stupidly simple and yet an amazingly powerful tool.
I've found that some of these trafos have a tickler coil that is AFAIK open ended. It may be capacitively coupled internally, but I've been unable to confirm that. A cap wound into the trafo would be odd. It's used to pass a signal back to the SMPS chip. Ran across it in the Apple PSU's a while back. Confusing because ringing it gave no response and shorting it caused no loss in the other windings.
Old 08-29-2010, 06:51 AM   #55
PlainBill
PlainBill's Avatar
My Country: USA
I'm a: Hobbyist
NP
Yeah, the tertiary winding used to power the SMPS controller. The chances of a capacitor in the transformer is so close to zero as to not be worth mentioning. And that winding open WOULD cause some strange symptoms.
I've got an old Sencore ring tester from the '70's that comes in very handy in cases like that.
PlainBill
Old 09-15-2010, 03:57 PM   #56
HexInverter
CD40106
Alright!
Sorry for the age long reply...life kinda got in the way...it probably will again, but I must see to it that this puzzle is solved eventually!
I got some time today to revisit the ring test, as per the updated instructions. I had to isolate the signal with a capacitor as said, and BLAMO! - I got ringing to display! Unfortunately the transformer exhibits a very nice (and identical) ring on both secondaries...so no easy fix! Shucks!
So, any idea of what to try next?
PS: Thanks for getting me this far! I've learned so much already!
Old 10-23-2010, 12:13 PM   #57
lcdm
have you managed to resolve it? I am having the same problem with mine Acer x193w and Darfon B092-XXX. Unfortunately no bulged capacitors or burned elements, lamps go off after 1 or 2 sec
lcdm is offline   Reply With Quote
Old 10-23-2010, 01:51 PM   #58
HexInverter
CD40106
Quote:
That sounds exactly like my problem, and no - I still haven't resolved it
It's sitting in my room waiting to be fixed still :3
Old 11-04-2010, 02:03 PM   #59
lcdm
hey PlainBill, in your http://www.badcaps.net/forum/showpos...2&postcount=33 post you are saying "There are a few steps that we can take to TEMPORARILY bypass the protection so troubleshooting is easier.". Is there a way to make backlight always on regardless the video signal presence or lamp voltages/currents?
Old 05-30-2012, 02:26 PM   #60
jeffescortlx
I also have a X193w with the same problem, back light will only light for a few seconds.
But it looks like I have a different powersupply/inverter board then you other guys. P/N is E59670 ILPI-031.
The inverter is controlled by a 16pin SMD chip number MP1008ES. I cant find the data sheet for it. And would like to know what pin's are the enable/shut down.
The 5v and 12v supply runs like a champ (checked with scope). Just the darn inverter keep's trippen.
I get about 500ohms on the secondarys and .6 ohms on the primarys for both of the transformers.
jeffescortlx is offline   Reply With Quote
Reply
Thread Tools
Display Modes
Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts
BB code is On
Smilies are On
[IMG] code is On
HTML code is Off
Forum Jump
Badcaps.net Technical Forums 2003 - 2017
Powered by vBulletin ®
Copyright ©2000 - 2017, Jelsoft Enterprises Ltd.
